Rev. 1 5193154-800

POSITIONER ROADRUNNER

CVC CPU BOARD INSTALLATION

APPLIES TO

Advantx-E Single Plane (LC+, LCV+, LCA+) and Biplane (LC/LP+, LCN+, LCA/LP+)

PURPOSE/SYMPTOM

Initial Software Download of a blank Positioner CPU RR CVC board

PROCEDURE

1. TOOLS

Standard Service Engineer’s toolkit

2. SPECIAL SAFETY PRECAUTIONS

GE X光机装机资料 SN_04684_5193154-800_R1

3. PREREQUISITES

1. Verify that a Positioner Database Backup exists.

If it does not, you will need to perform a Reset to Defaults, then perform a full system calibration

2. Close TP6 for download

Jumper description

Setting

ADVANTX E INNOVA 2000/2100-IQ/3100/4100

TP1, TP2 Not populated Not populated

GE X光机装机资料 SN_04684_5193154-800_R1

GE X光机装机资料 SN_04684_5193154-800_R1

GE X光机装机资料 SN_04684_5193154-800_R1

present Present (ARCNET)

TP5 Not

GE X光机装机资料 SN_04684_5193154-800_R1

TP6 Force

Force download

download

GE X光机装机资料 SN_04684_5193154-800_R1

bypass Watchdog bypass

TP7 Watchdog

GE X光机装机资料 SN_04684_5193154-800_R1

TP8, TP9 Not present Present (CAN 1)

GE X光机装机资料 SN_04684_5193154-800_R1

Rev. 1 5193154-800

Illustration 1: CPU CVC Board

GE X光机装机资料 SN_04684_5193154-800_R1

3. Connect the laptop with the cable 2348769 to J17 of RR CVC

4. Launch PST

5. Turn On the system

This message appears:

GE X光机装机资料 SN_04684_5193154-800_R1

Click “OK”

Rev. 1 5193154-800

This message appears:

C

Cancel

? For all systems EXCEPT AdvantX-E & Innova 2000

Click “OK”

? For AdvantX-E & Innova 2000, if the TITAN power button (near the reset button on the console) is “ON”

Switch OFF the console & Click “OK”

6. Select files on the accompanying Applications CD-Rom, or the system Applications CD-Rom.

The file begin by “RR” & finish by “.txt”

Example AdvantX-E : RR-2.02.txt, RR_advantxe_1.0H.txt,

RR_heroic5.2AD-2.txt,

:

Innova

2000

Innova 2100, 31200, 4100 : RR_POS_LBY_1.3.6C_2.txt,

GE X光机装机资料 SN_04684_5193154-800_R1

7. Click “OPEN”

The download start. Click “Finish” on the last screen when the download is completed.

Rev. 1 5193154-800

8. Switch “OFF” the positioner cabinet

9. Open the jumper TP6 on RR CVC

10. Switch “ON” the system

GE X光机装机资料 SN_04684_5193154-800_R1

11. Restore the Positioner Database Backup

If it was available, using PST - Configuration - Load, as shown below

GE X光机装机资料 SN_04684_5193154-800_R1

GE X光机装机资料 SN_04684_5193154-800_R1

If the Pos DB Backup was not available, Reset the DB to Defaults, using PST - Configuration - Reset

to Defaults, as shown below.

Rev. 1 5193154-800

GE X光机装机资料 SN_04684_5193154-800_R1

GE X光机装机资料 SN_04684_5193154-800_R1

Also perform a full Positioner calibration (AP and LAT gantries, if available), table calibrations , and

Smart Handle/Smart Box/JoyStick calibrations, if applicable

Rev. 1

5193154-800

On-Line Centers:

OLC–Americas OLC–Europe OLC–Japan OLC–ANZ 1–800–321–7937 (USA/CDN) +33 (0)1 30 83 13 00 81–426–48-2940 61-3-33235000

1(262) 896-2890 (Elsewhere)

Author:

Jean-Louis VAISSIERE GEMS

Headquarters Information: Are parts in stock? Yes __ No _ _ Available On Demand _ _ N/A _ √_

(If No, Service Note is not to be distributed)

This Service Note answers PQR 13071904

REVISION HISTORY

REV DATE REASON FOR CHANGE

1 02-oct-06 Initial release