Agilent 16900-Series Logic Analysis System Quick Start Guide for 16700-Series Users
Notices
? Agilent Technologies, Inc. 2003-2004 No part of this manual may be reproduced in any form or by any means (including electronic storage and retrieval or transla-tion into a foreign language) without prior agreement and written consent from Agi-lent Technologies, Inc. as governed by United States and international copyright laws.Manual Part Number
16900-92001
Edition
First edition, March 2004
Printed in USA
Agilent Technologies, Inc.
1900 Garden of the Gods Rd.
Colorado Springs, CO 80907 USA
Warranty
The material contained in this docu-
ment is provided “as is,” and is sub-
ject to being changed, without notice,
in future editions. Further, to the max-
imum extent permitted by applicable
law, Agilent disclaims all warran-
ties, either express or implied, with
regard to this manual and any infor-
mation contained herein, including
but not limited to the implied warran-
ties of merchantability and fitness for
a particular purpose. Agilent shall not
be liable for errors or for incidental or
consequential damages in connection
with the furnishing, use, or perfor-
mance of this document or of any
information contained herein. Should
Agilent and the user have a separate
written agreement with warranty
terms covering the material in this
document that conflict with these
terms, the warranty terms in the sep-
arate agreement shall control.
Technology Licenses
The hardware and/or software described in
this document are furnished under a
license and may be used or copied only in
accordance with the terms of such license.
Restricted Rights Legend
If software is for use in the performance of
a U.S. Government prime contract or sub-
contract, Software is delivered and
licensed as “Commercial computer soft-
ware” as defined in DFAR 252.227-7014
(June 1995), or as a “commercial item” as
defined in FAR 2.101(a) or as “Restricted
computer software” as defined in FAR
52.227-19 (June 1987) or any equivalent
agency regulation or contract clause. Use,
duplication or disclosure of Software is
subject to Agilent Technologies’ standard
commercial license terms, and non-DOD
Departments and Agencies of the U.S. Gov-
ernment will receive no greater than
Restricted Rights as defined in FAR
52.227-19(c)(1-2) (June 1987). U.S. Govern-
ment users will receive no greater than
Limited Rights as defined in FAR 52.227-14
(June 1987) or DFAR 252.227-7015 (b)(2)
(November 1995), as applicable in any
technical data.
Safety Notices
CAUTION
A CAUTION notice denotes a haz-
ard. It calls attention to an operat-
ing procedure, practice, or the like
that, if not correctly performed or
adhered to, could result in damage
to the product or loss of important
data. Do not proceed beyond a
CAUTION notice until the indicated
conditions are fully understood and
met.
WARNING
A WARNING notice denotes a
hazard. It calls attention to an
operating procedure, practice, or
the like that, if not correctly per-
formed or adhered to, could result
in personal injury or death. Do not
proceed beyond a WARNING
notice until the indicated condi-
tions are fully understood and met.
Agilent 16900-Series Logic Analysis System
Quick Start Guide for 16700-Series Users
Getting Started with the 16900-Series
Logic Analysis System
Main Menu Navigation4
Simple and Quick Triggers5
Advanced Triggers6
How to Quickly Move Between Windows7
Using Markers in Display Windows7
Tips for Using the Touch Screen9
Using the Overview Window10
Making Simple Inter-Module Measurements12
If you are familiar with the 16700-series logic analysis system, you will notice many similarities in the 16900-series logic analysis system; however, there are lots of improvements too.
By describing the key differences between the two logic analysis systems, this guide will help you quickly become productive with the new 16900-series logic analysis system.
About the Agilent Logic Analyzer Application
The 16900-series logic analysis system is controlled by the Agilent Logic Analyzer application, a Windows-based user interface that also controls the 1680/90-series logic analyzers and can be used for offline analysis of data captured by 16900-series, 1680/90-series, and 16700-series logic analyzers. The Agilent Logic Analyzer application can run on Windows XP/2000 computers and still connect to and control 16900-series logic analysis systems. This is called the hosted power mode, and it lets you improve the performance of the application when desired.
The Agilent Logic Analyzer application can control multiple 16900A or 16902A logic analysis system frames connected in a multiframe configuration and present all their modules as if they were in a single frame.
Getting Started with the 16900-Series Logic Analysis System
Main Menu Navigation
In the 16700-series logic analysis system:?The primary method of navigation is through the Window menu.?The main menu bar provides access to a limited number of the system features.?The Workspace window is used primarily by advanced users.?There are lots of multi-tabbed dialogs.In the Agilent Logic Analyzer application (which controls the 16900-series logic analysis system):?You can access just about everything from the main menu.?The Overview window is the main navigation aid for all users.?There are separate dialogs for setup and advanced trigger, and there are relatively few multi-tabbed dialogs.Table 1Agilent Logic Analyzer Menus to Know Command To Edit>Find...Search for bus/signal values.Setup>Bus/Signal...Set up buses and signals.Setup>Timing/State (Sampling)...Choose the acquisition mode.Tools>New Filter/Colorize...Add a Filter/Colorize tool.Markers>New...Add new markers.Markers>Go To...Go to a marker.Run/Stop>Run
Run the logic analyzer.
Getting Started with the 16900-Series Logic Analysis System
Simple and Quick Triggers
In the 16700-series logic analysis system, the easiest way to set up triggers is
by using the predefined trigger functions.
In the Agilent Logic Analyzer application (which controls the 16900-series
logic analysis system):
?You can set up simple triggers from within the Waveform or Listing
windows without having to open a trigger setup dialog.
?You can set up quick triggers by drawing a box around data in the
Waveform or Listing window.
Simple and quick triggers let you set up a single trigger event defined as a
combination of patterns, edges, etc., across any signals and buses.
Simple Triggers
Waveform Display
Listing Display
Getting Started with the 16900-Series Logic Analysis System
Quick Triggers
Advanced Triggers
When you need to set up triggers that are more complex than just finding
particular bus/signal values (for example, when you need to trigger on a
sequence of events in the device under test), you can set up advanced triggers.
The advanced trigger dialog is similar to the Trigger tab in the 16700-series
logic analysis system, where you can choose from predefined trigger functions.
With advanced triggers, you can customize a trigger using trigger functions as
individual trigger events or as building blocks for complex scenarios.
To use a trigger function, drag-and-drop it into the Trigger Sequence area.
Getting Started with the 16900-Series Logic Analysis System
How to Quickly Move Between Windows
In the Agilent Logic Analyzer application, you can quickly move between
windows by clicking on the tabs at the bottom of the display.
Using Markers in Display Windows
In the 16700-series logic analysis system:
?You have two markers, x and o, that are local to each display tool.
?You also have two global markers, G1 and G2, which are used in
time-correlated measurements.
In the Agilent Logic Analyzer application (which controls the 16900-series
logic analysis system):
?You can have up to 1,024 markers.
?All markers are global.
?You can give markers long and short names, specify their color, and add comments (which appear in marker tool tips).
Getting Started with the 16900-Series Logic Analysis System
?On the front panel of the 16902A or 16903A, you can position markers by pressing the Choose button to select a marker and then by using the marker knob to position it.?In the Agilent Logic Analyzer application, you can position markers by dragging-and-dropping them, placing them at the mouse cursor, placing them at the center of the display, or by changing the position property in the Marker Properties tab.?When moving a marker in the Waveform display window, if the mouse cursor is over a waveform and the marker's Snap to Edge property is enabled, the cursor changes to a green "direction arrow" indicating the direction of the next valid edge. A yellow "cross hair" target appears on the edge at which the marker will be placed if you release the mouse button.If you don't want the marker to snap to an edge, move the mouse cursor so that it is not over any waveforms before releasing the mouse button.?You can set up time interval , sample interval , and value at marker measurements.?When searching for bus/signal data values with Find, you can place
markers where data values are found.
Getting Started with the 16900-Series Logic Analysis System
Tips for Using the Touch Screen
The 16902A and 16903A logic analysis system frames have a touch screen.
Here are some tips for using the touch screen:
?Use firm, even pressure on the touch screen.
?You may prefer to use a stylus.
?Use the front panel marker knobs to place markers. (Any marker can be
selected using the choose button.)
?For trees, the touchable area around the +/- buttons is expanded.
?For option selections, both the option and the caption are active.
?There is a Keyboard button on the front panel to open a keyboard dialog.
?Inside the Agilent Logic Analyzer application, most right mouse button
behavior is accessible just by touching the screen. Outside the Agilent Logic
Analyzer application, you can press down on the touch screen until a full
circle is drawn around your finger; then, a right-click occurs.
Getting Started with the 16900-Series Logic Analysis System
Using the Overview Window
The Overview window in the Agilent Logic Analyzer application looks like the
Workspace window in the 16700-series logic analysis system, but it’s different.
In the 16700-series Workspace window
Drawn lines grant
permission to use
the data flow
By default, all modules are not on workspace
Drag-and-drop from here to workspace
In the 16700-series Workspace window:
?Instruments, analysis tools, and displays are dragged onto the Workspace from the tool box.
?You draw lines between instruments, analysis tools, and displays to route data .
?Drawn lines give an analysis tool or window permission to use the data.
Getting Started with the 16900-Series Logic Analysis System In the Agilent Logic Analyzer application’s Overview window
By default, all modules are
always in the Overview.
Menus
Shortcut buttons
In the Agilent Logic Analyzer application’s Overview window:
?All modules are always displayed.
?Tools and display windows are added via menu selections.
?Lines are automatically drawn between modules, tools, and windows to reflect how you are using data.
Getting Started with the 16900-Series Logic Analysis System
Making Simple Inter-Module Measurements
In the 16700-series logic analysis system:
?There is an Inter-Module triggering dialog box.
?For analyzers split into two machines, there is a separate arming tree
dialog.
Getting Started with the 16900-Series Logic Analysis System In the Agilent Logic Analyzer application (which controls the 16900-series
logic analysis system):
?Inter-module triggers are set up just like any other event. (There is no special dialog for inter-module triggering.)
Notice that triggers from split analyzers are selected in the same place as external triggers from other instruments.
?The Trigger In and Trigger Out BNC connectors are set up through the
External Trigger dialog.
Getting Started with the 16900-Series Logic Analysis System