Agilent 16900-Series Logic Analysis System Quick Start Guide for 16700-Series Users

Notices

? Agilent Technologies, Inc. 2003-2004 No part of this manual may be reproduced in any form or by any means (including electronic storage and retrieval or transla-tion into a foreign language) without prior agreement and written consent from Agi-lent Technologies, Inc. as governed by United States and international copyright laws.Manual Part Number

16900-92001

Edition

First edition, March 2004

Printed in USA

Agilent Technologies, Inc.

1900 Garden of the Gods Rd.

Colorado Springs, CO 80907 USA

Warranty

The material contained in this docu-

ment is provided “as is,” and is sub-

ject to being changed, without notice,

in future editions. Further, to the max-

imum extent permitted by applicable

law, Agilent disclaims all warran-

ties, either express or implied, with

regard to this manual and any infor-

mation contained herein, including

but not limited to the implied warran-

ties of merchantability and fitness for

a particular purpose. Agilent shall not

be liable for errors or for incidental or

consequential damages in connection

with the furnishing, use, or perfor-

mance of this document or of any

information contained herein. Should

Agilent and the user have a separate

written agreement with warranty

terms covering the material in this

document that conflict with these

terms, the warranty terms in the sep-

arate agreement shall control.

Technology Licenses

The hardware and/or software described in

this document are furnished under a

license and may be used or copied only in

accordance with the terms of such license.

Restricted Rights Legend

If software is for use in the performance of

a U.S. Government prime contract or sub-

contract, Software is delivered and

licensed as “Commercial computer soft-

ware” as defined in DFAR 252.227-7014

(June 1995), or as a “commercial item” as

defined in FAR 2.101(a) or as “Restricted

computer software” as defined in FAR

52.227-19 (June 1987) or any equivalent

agency regulation or contract clause. Use,

duplication or disclosure of Software is

subject to Agilent Technologies’ standard

commercial license terms, and non-DOD

Departments and Agencies of the U.S. Gov-

ernment will receive no greater than

Restricted Rights as defined in FAR

52.227-19(c)(1-2) (June 1987). U.S. Govern-

ment users will receive no greater than

Limited Rights as defined in FAR 52.227-14

(June 1987) or DFAR 252.227-7015 (b)(2)

(November 1995), as applicable in any

technical data.

Safety Notices

CAUTION

A CAUTION notice denotes a haz-

ard. It calls attention to an operat-

ing procedure, practice, or the like

that, if not correctly performed or

adhered to, could result in damage

to the product or loss of important

data. Do not proceed beyond a

CAUTION notice until the indicated

conditions are fully understood and

met.

WARNING

A WARNING notice denotes a

hazard. It calls attention to an

operating procedure, practice, or

the like that, if not correctly per-

formed or adhered to, could result

in personal injury or death. Do not

proceed beyond a WARNING

notice until the indicated condi-

tions are fully understood and met.

Agilent 16900-Series Logic Analysis System

Quick Start Guide for 16700-Series Users

Getting Started with the 16900-Series

Logic Analysis System

Main Menu Navigation4

Simple and Quick Triggers5

Advanced Triggers6

How to Quickly Move Between Windows7

Using Markers in Display Windows7

Tips for Using the Touch Screen9

Using the Overview Window10

Making Simple Inter-Module Measurements12

If you are familiar with the 16700-series logic analysis system, you will notice many similarities in the 16900-series logic analysis system; however, there are lots of improvements too.

By describing the key differences between the two logic analysis systems, this guide will help you quickly become productive with the new 16900-series logic analysis system.

About the Agilent Logic Analyzer Application

The 16900-series logic analysis system is controlled by the Agilent Logic Analyzer application, a Windows-based user interface that also controls the 1680/90-series logic analyzers and can be used for offline analysis of data captured by 16900-series, 1680/90-series, and 16700-series logic analyzers. The Agilent Logic Analyzer application can run on Windows XP/2000 computers and still connect to and control 16900-series logic analysis systems. This is called the hosted power mode, and it lets you improve the performance of the application when desired.

The Agilent Logic Analyzer application can control multiple 16900A or 16902A logic analysis system frames connected in a multiframe configuration and present all their modules as if they were in a single frame.

Getting Started with the 16900-Series Logic Analysis System

Main Menu Navigation

In the 16700-series logic analysis system:?The primary method of navigation is through the Window menu.?The main menu bar provides access to a limited number of the system features.?The Workspace window is used primarily by advanced users.?There are lots of multi-tabbed dialogs.In the Agilent Logic Analyzer application (which controls the 16900-series logic analysis system):?You can access just about everything from the main menu.?The Overview window is the main navigation aid for all users.?There are separate dialogs for setup and advanced trigger, and there are relatively few multi-tabbed dialogs.Table 1Agilent Logic Analyzer Menus to Know Command To Edit>Find...Search for bus/signal values.Setup>Bus/Signal...Set up buses and signals.Setup>Timing/State (Sampling)...Choose the acquisition mode.Tools>New Filter/Colorize...Add a Filter/Colorize tool.Markers>New...Add new markers.Markers>Go To...Go to a marker.Run/Stop>Run

Run the logic analyzer.

Getting Started with the 16900-Series Logic Analysis System

Simple and Quick Triggers

In the 16700-series logic analysis system, the easiest way to set up triggers is

by using the predefined trigger functions.

In the Agilent Logic Analyzer application (which controls the 16900-series

logic analysis system):

?You can set up simple triggers from within the Waveform or Listing

windows without having to open a trigger setup dialog.

?You can set up quick triggers by drawing a box around data in the

Waveform or Listing window.

Simple and quick triggers let you set up a single trigger event defined as a

combination of patterns, edges, etc., across any signals and buses.

Simple Triggers

Waveform Display

Listing Display

Getting Started with the 16900-Series Logic Analysis System

Quick Triggers

Advanced Triggers

When you need to set up triggers that are more complex than just finding

particular bus/signal values (for example, when you need to trigger on a

sequence of events in the device under test), you can set up advanced triggers.

The advanced trigger dialog is similar to the Trigger tab in the 16700-series

logic analysis system, where you can choose from predefined trigger functions.

With advanced triggers, you can customize a trigger using trigger functions as

individual trigger events or as building blocks for complex scenarios.

To use a trigger function, drag-and-drop it into the Trigger Sequence area.

Getting Started with the 16900-Series Logic Analysis System

How to Quickly Move Between Windows

In the Agilent Logic Analyzer application, you can quickly move between

windows by clicking on the tabs at the bottom of the display.

Using Markers in Display Windows

In the 16700-series logic analysis system:

?You have two markers, x and o, that are local to each display tool.

?You also have two global markers, G1 and G2, which are used in

time-correlated measurements.

In the Agilent Logic Analyzer application (which controls the 16900-series

logic analysis system):

?You can have up to 1,024 markers.

?All markers are global.

?You can give markers long and short names, specify their color, and add comments (which appear in marker tool tips).

Getting Started with the 16900-Series Logic Analysis System

?On the front panel of the 16902A or 16903A, you can position markers by pressing the Choose button to select a marker and then by using the marker knob to position it.?In the Agilent Logic Analyzer application, you can position markers by dragging-and-dropping them, placing them at the mouse cursor, placing them at the center of the display, or by changing the position property in the Marker Properties tab.?When moving a marker in the Waveform display window, if the mouse cursor is over a waveform and the marker's Snap to Edge property is enabled, the cursor changes to a green "direction arrow" indicating the direction of the next valid edge. A yellow "cross hair" target appears on the edge at which the marker will be placed if you release the mouse button.If you don't want the marker to snap to an edge, move the mouse cursor so that it is not over any waveforms before releasing the mouse button.?You can set up time interval , sample interval , and value at marker measurements.?When searching for bus/signal data values with Find, you can place

markers where data values are found.

Getting Started with the 16900-Series Logic Analysis System

Tips for Using the Touch Screen

The 16902A and 16903A logic analysis system frames have a touch screen.

Here are some tips for using the touch screen:

?Use firm, even pressure on the touch screen.

?You may prefer to use a stylus.

?Use the front panel marker knobs to place markers. (Any marker can be

selected using the choose button.)

?For trees, the touchable area around the +/- buttons is expanded.

?For option selections, both the option and the caption are active.

?There is a Keyboard button on the front panel to open a keyboard dialog.

?Inside the Agilent Logic Analyzer application, most right mouse button

behavior is accessible just by touching the screen. Outside the Agilent Logic

Analyzer application, you can press down on the touch screen until a full

circle is drawn around your finger; then, a right-click occurs.

Getting Started with the 16900-Series Logic Analysis System

Using the Overview Window

The Overview window in the Agilent Logic Analyzer application looks like the

Workspace window in the 16700-series logic analysis system, but it’s different.

In the 16700-series Workspace window

Drawn lines grant

permission to use

the data flow

By default, all modules are not on workspace

Drag-and-drop from here to workspace

In the 16700-series Workspace window:

?Instruments, analysis tools, and displays are dragged onto the Workspace from the tool box.

?You draw lines between instruments, analysis tools, and displays to route data .

?Drawn lines give an analysis tool or window permission to use the data.

Getting Started with the 16900-Series Logic Analysis System In the Agilent Logic Analyzer application’s Overview window

By default, all modules are

always in the Overview.

Menus

Shortcut buttons

In the Agilent Logic Analyzer application’s Overview window:

?All modules are always displayed.

?Tools and display windows are added via menu selections.

?Lines are automatically drawn between modules, tools, and windows to reflect how you are using data.

Getting Started with the 16900-Series Logic Analysis System

Making Simple Inter-Module Measurements

In the 16700-series logic analysis system:

?There is an Inter-Module triggering dialog box.

?For analyzers split into two machines, there is a separate arming tree

dialog.

Getting Started with the 16900-Series Logic Analysis System In the Agilent Logic Analyzer application (which controls the 16900-series

logic analysis system):

?Inter-module triggers are set up just like any other event. (There is no special dialog for inter-module triggering.)

Notice that triggers from split analyzers are selected in the same place as external triggers from other instruments.

?The Trigger In and Trigger Out BNC connectors are set up through the

External Trigger dialog.

Getting Started with the 16900-Series Logic Analysis System