740 Wrist strap +shoe tester

Operating Instructions MO 0688-0201

February, 2001

II. Declaration of Conformity

The 740 Wrist Strap and Shoe Tester is in conformity with the regulations of the EMC-directive 89/336/EWG. The following standards have been taken into account while designing and manufacturing the instrument: EN 60204-1/85,EN 60204-1/91, EN 61010 (Safety),EN 50082-1/92 and EN 50082-2/92.

I. General

Personnel grounding devices such as wrist straps and conductive footwear are the primary method to minimise charge generation on the human body. They must be able to drain this charge as rapidly as it is generated.

For this reason, it is important to test wrist straps and conductive footwear on a regular basis to ensure they are functioning properly and that the test results can be recorded

Ground cord 4 mm stud

Contact plate for wrist strap test

e.g. - data transfer to PC, electrical door opening system, acoustic signal, etc.

selection for wrist strap and shoe test

shoe electrode On / Off switch

R

max III. Description

The 740 wrist strap and shoe tester is an electronic test instrument which is easy to operate and designed to be wall mounted. An AC/DC -transformer is supplied. The 740 switches off automatically after non-use for approximately 30minutes.

The unit consists of two measurement circuits with an open circuit test voltage of 20 V for wrist strap and shoe test applications. For wrist strap and footwear testing, a minimum threshold

resistance of 750 k ? is set internally. The desired maximum allowable resistance level can be selected separately for each measurement. If the resistance of the tested wrist strap or footwear is within the selected range, the green "o.k."-LED will illuminate. The red "<" -LED alerts the

operator that the resistance of the tested device is

under 750 k ?. The red ">"-LED indicates that the

selected maximum resistance is exceeded.

The 740 can be connected to external devices such as computers (for data recording), electrical door opening systems, counter or other evaluating components.

The 740 wrist strap and shoe tester

consists of:

-3.5" - diskette containing instruction manual and 740 daily log sheet - Base unit

-Data output connector plug

-Wall mounting kit, 3M Dual Lock ? fastening system and template

-Cover for wrist strap plug-in jack -AC/DC - transformer -Ground cord

Accessories:

-Shoe electrodes type 741 (single electrode) or 741D (dual electrode) (to be ordered separately).

IV. Operation

Connect the AC/DC - transformer (DC 24-30 V,150 mA) to the 740 tester.

Note: If you notice a delay in the response time of the LED illumination please connect the ground cord supplied to the ground socket of the 740.Please ensure proper earth ground connection!

Switch the unit on. All LEDs will flash for

approximately 1 s to check their function. Select

the desired upper maximum resistance level for

both the wrist strap and shoe test by using a small

screw driver.The 741/741D shoe electrode must be connected if footwear testing is required.

V. Wall Mounting

The enclosed 3M Dual Lock? fastening discs

must be screwed to the wall using the attached

template approximately 1500 mm (5 feet) above

the floor. Drill three holes of 5mm (0.2 in)

diameter at marked locations. Make sure the area

on the 740 in which the Dual Lock? fastening

strips are placed is free of dust and dirt. Attach the

3M Dual Lock? fastening strips to the rear of the

740 tester to the bottom and top in parallel; see

drawing for locations.

For securing on the wall, press the 740 firmly

against the discs. For removal of the 740, lift the

top corners with both hands.

VI. Wrist Strap Test

Put the wrist band on with ground cord attached

and insert the ground cord into the wrist strap

plug-in jack. Depress the metal contact plate and

hold it. One of the indicator-LEDs will illuminate.

An illuminated green LED indicates that the wrist

strap performs within the resistance range of

750k? to the upper selected value of 2, 5, 10 or

35 M?. If a red "<"-LED is on, the current

limiting resistor in the ground cord is bypassed

and the cord has to be replaced. The red ">"-LED

indicates that the selected resistance range is

exceeded. Check to see if the high resistance is in

the cord, the wrist band or in the contact between

wrist band and the operator’s skin. To check the

resistance of the ground cord, leave the plug end

of the ground cord attached to the tester and

disconnect the snap end from the wrist band.

Connect ground cord snap to the stud in the hand

symbol or press it onto the contact plate and

depress the plate until one of the LEDs is

illuminated.

Template

If the green LED illuminates

now, the ground cord can be

used.

If the red ">"-LED is lit,

replace the ground cord.

In some cases, high contact resistance between the

skin and wrist band will cause the tester to show a

red ">" - LED condition. This resistance may be

caused by dry skin or the presence of hair in the

wrist area. The use of a skin lotion is

recommended to solve this problem.

If a red ">"- LED condition still exists, replace the

wrist band.

VII. Shoe test

Either the 741/741D shoe electrode must be connected to the 740 shoe jack.

a) when the 741 single shoe electrode is used:

If the red ">"-LED is illuminated, clean the soles of the shoes or check shoe grounding system (heel ground or toe ground strap) and retest. If, after cleaning, the red ">"-LED remains lit, replace it by an appropriate device. If the red "<"-LED is illuminated, the minimum resistance of the shoes or shoe grounding system is under 750 k ?.

An electrical hazard to personnel may exist at the work place.

b) when the 741D dual shoe electrode is used:

The dual shoe electrode allows differentiation between the right and left shoe; giving the right shoe a preference in indication. This means, if both shoes or just the right one are out of range,you will get identical indications as it would be with the single shoe electrode. If just the left shoe is out of range, the corresponding red LED will flash. Consequently, in case of a steadily lit red LED, you shall clean or remove right shoe first and retest. If the right shoe is in the "ok"-range then, you will obtain information about

performance of the left shoe and you can react accordingly.

The shoe test does not influence the wrist strap test, therefore the shoes can be tested while the wrist strap is still connected to the 740 or the wrist strap can be tested while the operator is standing on the 741 shoe electrode. Both metal plates shall not be pressed at the same time.

Switch the 740 on. Stand on the applied shoe electrode. If you are using shoe straps with a 1M ? - resistor on both feet to be tested with the 741 shoe electrode, you have to test the straps one after the other, to avoid a red "too low" -indication. Care must be taken not to put the non-tested foot on ESD - protective flooring to avoid a bypass to ground. Depress the metal contact plate for shoe test until one of the indicator - LEDs lights up. The green LED indicates that the

resistance of the person through the footwear is in the range between 750 k ? and the desired upper maximum level (10, 35, 50 or 100 M ?).

VIII. Data Output Connector

If you are going to use the data output connector you must follow a test sequence. It must be wrist strap test first and then the shoe test. If this sequence is not strictly kept, the output at the pins described on page 7 will be mixed up .

The 740 will provide digital signals (high/low) while testing the personnel grounding devices. These digital signals can be used for data recording (Example 1) and authorisation control equipment such as electrical door opening systems (Example 2). The output of the 740 will give you permanently +5 V at pin 3 and DC-return at pin 6. The remaining pins 1, 2, 4 and 5 (open collector) will be high

(max.+30 V/20 mA have to be provided externally) or low (DC-return) while pressing on the contact plates either for wrist strap or shoe tests. The chart on page 7 shows all of the possible test results and the corresponding output levels.

Example 1: Output connection for use with a computer Example 2: Output connection for use with a controlling device

Computer input Computer ground

Data output connector pin status when testing wrist strap/shoe

The sequence for testing must be wrist strap test first and then the shoe test.

Shaded pin indicators will read "hi" if the sequence is reversed or only the shoe tests are performed.

Data Output Connector

pin 1, 2, 4, 5-high or low pin 3-+5 V pin 6

-

DC-return

IX. Specifications

Dimension of base unit :138 mm x 190 mm x 53 mm (5.25 x 7.5 x 2) inches Weight :450 g (15.75 oz)

Power supply :

external AC/DC - transformer,DC 24-30 V/150 mA

Accuracy :

± 10% of 2, 5, 10, 35 and 50 M ?-ranges ± 20% of 100 M ?-range Output :

open collector at pin 1,2, 4, 5(max. +30 V/20 mA)

DC-return at pin 6, +5 V at pin 3Measurement voltage :DC (20±1) V (open circuit)

750 k ?

?

?

?X. Calibration Procedure

The following procedure can be used to determine if the 740 operates within its specifications.Please note: The tester has no adjustable components.

Equipment needed:

1.Resistors

750 k ? - 120 M ?, tolerance ± 1%2. 2 leads

as required to connect the reference resistor

a) calibration of wrist strap test circuit:

Connect the reference resistor to the 740 as shown in the figure below. Switch the 740 unit on. Select the 2 M ?-range and use the reference resistors as indicated in the test table and press contact plate A.

b) calibration of shoe test circuits:

The 740 Tester can either operate with the single or dual shoe electrode. They are identified by the wiring in the 3-pin plug according to DIN 41 524,type HIRSCHMANN MAS 3100, that connects the shoe electrode to the 740. For the single shoe electrode, the cord is connected to pin 1 while pins 2 and 3 are shunted and grounded.The dual shoe electrode uses the same plug but has got the following connections:pin 1: right shoe electrode pin 2: open

pin 3: left shoe electrode.

Reference Resistance LED -resistor range settings Indication *750 k ? 2 M ? - range

Red 900 k ?Green 1.8 M ?Green >2.2 M ?Red 4.5 M ? 5 M ? - range Green >5.5 M ?Red 9.0 M ?10 M ? - range Green >11.0 M ?Red 31.5 M ?35 M ? - range

Green >38.5 M ?

Red

*This example is used above

The LEDs will indicate as shown below if the 740performs within specifications. Repeat this procedure for 5, 10 and 35 M ?-ranges.

11.1 M?II) calibration of the dual shoe test circuits: The 740 Tester measures the resistance of both shoes almost at the same time. The test result for the circuit of the right shoe has got priority. The calibration can be either performed with two reference resistors for both circuits to be checked at the same time or -corresponding to the single shoe test - with one reference resistor for one circuit to be checked and an appropriate bridge resistor (1M? e.g.) for the other circuit. For completion of calibration, the test has to be repeated with connections of the resistors vice versa. Follow the instructions for the single electrode calibration test accordingly.

For the operation with two reference resistors proceed as follows: Connect one reference resistor between contact 1 of the socket and the stud and the other one between contacts 2 and 3. Select the 10 M?-range and set the reference resistors as indicated in the test table and press contact

plate B. The LEDs shall light as shown below. Repeat this procedure for 35, 50 and 100 M?-ranges.

Right shoe Left shoe

electrode electrode

circuit circuit Signal

OK OK green steadily high OK>red steadily

low OK

OK high>red flashing

OK lowred steadily

low highred steadily

low low

I) calibration of the single shoe test circuit: Connect a reference resistor to the 740 between contact 1 of the socket and the stud as shown below. Contacts 2 and 3 of the socket shall be bridged via an appropriate resistor (1M? e.g.) Select the 10 M?-range and set the reference resistors as indicated in the test table and press contact plate B.

The LEDs shall light as shown below. Repeat this procedure for 35, 50 and 100M?-ranges.

Shoe electrode

circuit Signal

OK green steadily

high>red steadily

low

3M 740 Wrist Strap + Shoe Tester Logbook

Distributed by:

SJM Eurostat UK Ltd.Countess Avenue

Stanley Green Trading Estate

Cheadle Hulme, Cheshire SK8 6QS United Kingdom

Tel: +44 (0) 161 485 5002Fax: +44 (0) 161 485 4678

3M France

Electronic Handling & Protection Boulevard de l ’Oise

95006 Cergy Pontoise Cedex T él.: +33 (0) 130316809Fax: +33 (0) 130316181

SA au capital de 52500000francs RC Pontoise b 542078555, ape 246 C

3

Magnab Eurostat

Skvasta Industriby 16BN 61192 Nyk ?ping Sweden

Tel: +46 (0) 155 20 26 80Fax: +46 (0) 155 26 98 34

Important notice to purchaser

"Warranty-3M expressly warrants that for a period of twelve (12) months from the date of purchase,3M Static Control Products will be free of defects in materials (parts) and workmanship (labor).Defects occurring during the warranty period will be repaired or products will be replaced at 3M ’s option and expense, if 3M receives notice during the warranty period. Defective products must be returned to 3M with proof of purchase date.Warranty exclusions: the foregoing express warranty is made in lieu of all other product

warranties, express and implied, including fitness and merchantability. The express warranty will not apply to defects or damage due to accidents,neglect, misuse, alterations, operator error, or failure to properly maintain, clean, or repair products.

Limit of liability-in no event will 3M or seller be responsible or liable for special, incidental, or consequential losses or damages, whether based in tort or contract. Fulfillment of 3M ’s warranty obligations will be customer ’s exclusive remedy and 3M ’s and seller ’s limit of liability for any breach of warrant or otherwise."

安规综合测试仪安全操作规程

常州有则合众光电有限公司 T/HZ-03-57 支持性文件 (A 版) 文件名称程控安规综合测试仪安全操作规程 受控状态 发放部门 2016年 2 月 25 日发布 2016年 3 月1日实施

常州合众光电有限公司T/HZ-03-57 共4页第 1页支持性文件:程控安规综合测试仪安全操作规程第A版第0次修改 程控安规综合测试仪安全操作规程 1.目的 为使程控安规综合测试仪得到有效的维护,合理的使用,以达到保证机台精度及应有的使用寿命,确保产品品质、产能和维护、操作人员安全。 2.适用范围 适用设备:组件车间程控安规综合测试仪。 适用对象:组件车间设备技术员及指定的经培训过的设备操作人员。 3.定义 程控安规综合测试仪:包含接地电阻测试、耐压、绝缘测试三位一体的测试仪器。 本规程适用程控安规综合测试仪。 4.引用文件 程控安规综合测试仪说明书。 5.职责 5.1生产操作人员:按规程要求操作设备,日常维护。 5.2设备技术员:执行设备维修及保养作业。 5.3工程师:督导技术员对设备进行维修保养,对设备存在的隐患进行改善预防。 6.开机前检查 6.1检查周围地面清洁,无与生产不相关的杂物; 6.2检查耐压仪接线完好,线缆无破损; 6.3检查探针接线良好,螺丝紧固,如有异常通知设备处理;

常州合众光电有限公司T/HZ-03-57 共4页第 2页支持性文件:程控安规综合测试仪安全操作规程第A版第0次修改 6.4检查接地完好; 6.5按《程控安规综合测试仪PM作业指导书》要求执行日常点检; 7. 操作步骤 7.1打开电脑电源和仪器电源,电脑启动完成后打开测试软件,点击连接设备; 7.2显示连接成功后,即生产准备工作已完成,可投入生产。 8安全注意事项 8.1安全基本注意事项 1)安装后开机前确保电源电压输出符合要求; 2)不要触摸控制盘、电气箱等带有标示牌的机器内部电路。否则可能碰触高压电路,有触电死亡危险; 3)机器运转中,不可触摸机器和其他可动部分,如升降机构、规正等。不要将身体的任何一部分置于可能被机构夹住的地方; 4)确实熟记紧急停止按钮的位置,以便随时按下。不要在卸下门、盖、护罩的状态下接通电

静电测试仪使用说明书

Hand-held static sensor locates and meas-ures static voltages, tests air ionizers.New from 3M,the 718 Static Sensor can help companies competing in the global high-tech marketplace prevent cost-ly losses due to electro static discharge (ESD) damage by playing a vital and valuable role in their own ESD control program. Easy to use,the hand-held 3M ?718 Static Sensor is designed to measure static voltages on objects and sur-faces arising from electrostatic charge buildups,and can help identify ESD trouble-spots — ensuring product relia-bility and customer satisfaction which translates into com-pany profits. As a bonus,when used in conjunction with the 3M TM Model 718A Air Ionizer Test Kit,the 718 can also be used to verify the operation of air ionizers.718 Static Sensor Features ? Small-size,lightweight,conductive plastic housing ? Membrane switches for Power,Range/Zero,and Hold functions. ? Digital,LCD (liquid-crystal) display is easy to read and updates quickly. ? Ranging system assists user in making quick and easy measurements ? Measurements accurate to 5% ? Output jack available for continuous measurements Convenient Size/Low Power Requirements The 718 is small enough to be carried in a pocket and weighs less than 5 oz. (142 g),including battery. The light-weight plastic housing is conductive,allowing a properly-grounded user to dissipate all electrostatic charges from the surface of the meter.Meter Functions The meter is equipped with three membrane switches which control different functions. The POWER switch turns the instrument on and off. The RANGE/ZERO button performs two functions; when pressed momentarily it switches between the two measurement ranges of 0-2,000 volts and 0-20,000 volts,and if held for longer than 3 seconds,it resets the voltage display to 0 volts. The HOLD button allows the user to freeze a measurement on the LCD for later review.Ranging System Included with the 3M 718 Static Sensor is a ranging system consisting of two light-emitting diodes (LEDs) which each emit a circular red light onto the surface being measured for static. When the two lights intersect and form a single focused light,the measurement distance is the prescribed 1 inch (2.54 cm). Accuracy The Model 718 Static Sensor is accurate to within ±5% of the displayed measurement,at a distance of one inch (2.54 cm) from the target. Accuracy will vary as the dis-tance between measured object and instrument changes from the one inch (2.54 cm) specification.Analog Output Jack The analog output jack located in the front of the unit pro-vides a convenient hook-up,via a 3/32 inch (2.5 mm)monophone jack,to a recorder/data acquisition console. The 3M 718 Static Sensor may then be used for remote monitoring or permanent recording of electrostatic voltage readings. 3M 718 Static Sensor Specifications Dimensions 0.85" (H) x 2.4" (W) x 4.15" (L) 2.2 cm (H) x 6.1 cm (W) x 10.5 cm (L)Weight 4.5 oz. (128 g) with battery Power Requirements One 9-volt alkaline battery Measurement Ranges 0 – 2 kV Low Range 0 - 20 kV High Range V oltage Display 3) digit liquid crystal display V oltage Output 1/1000 of measured voltage @ low range 1/10,000 of measured voltage @ high range Distance Indicator LED targets. Aligned targets indicate 1 in. (2.54 cm) measurement distance Measurement Accuracy Within 5% of actual voltage Certifications UL,C-UL,CE,CB-scheme,NOM 3 718 Static Sensor 718A Air Ionizer Test Kit 718 Range Finder Unfocused 718 Range Finder at 1" away 3M 718 Static Sensor 1 2 3

静电放电抗扰度测试仪作业指导书

静电放电抗扰度测试仪作业指导书 1. 目的 本规范规定了ESD-2005静电放电抗扰度测试仪的操作方法、维护保养及安全规范,指导操作人员正确作业。 2. 适用范围 测试中心静电放电抗扰度测试仪(ESD-2005) 3. 检测单位 质控部 4. 设备操作方法 4.1静电放电抗扰度测试仪测试面板说明: ○1放电电压极性切换按钮 ○2显示窗 ○3电压设定选择按钮 ○4防电间隔时间设定选择按钮 ○5放电次数设定选择按钮

○6 IEC标准放电电压指示灯 ○7 IEC标准放电电压设定按钮 ○8移位按钮 ○9数字加1按钮 10 数字减1按钮 ○ 11‘DISCHARGE MODE’选择按钮 ○ 12 气隙放电模式指示灯 ○ 13 接触放电模式指示灯 ○ 14 电源开关 ○ 15 高压输出指示灯 ○ 16‘TRIGGER MODE’选择按钮 ○ 17‘STOP’按钮 ○ 18‘STOP’按钮指示灯 ○ 19‘START’按钮 ○ 20‘START’按钮指示灯 ○ 21‘20pps’触发模式指示灯 ○ 22‘COUNT’触发模式指示灯 ○ 23‘SINGLE’触发模式指示灯 ○ 24 高压插座 ○ 4.2操作指南 4.2.1 放电电压设置 首先点按钮○3进入放电电压设置状态,此时放电电压显示窗口十位数以闪烁方式 10可以改变该位数字,而后点按钮○8移动输显示,表示该位处于编辑状态。此时点按钮○9、○ 入位到个位数,此时放电电压设定值的十位数已输入仪器,放电电压显示窗口个位数以闪烁方式显示,表示该位处于编辑状态,点按纽○3退出放电电压设置状态。 4.2.2 放电电压极性设置 点按纽○1即可实现切换电压极性的操作。 4.2.3 放电间隔时间设置 首先点按纽○4进入放电间隔时间设置状态(注意:触发模式应为COUNT),此时放

安全综合测试仪操作规程示范文本

安全综合测试仪操作规程 示范文本 In The Actual Work Production Management, In Order To Ensure The Smooth Progress Of The Process, And Consider The Relationship Between Each Link, The Specific Requirements Of Each Link To Achieve Risk Control And Planning 某某管理中心 XX年XX月

安全综合测试仪操作规程示范文本使用指引:此操作规程资料应用在实际工作生产管理中为了保障过程顺利推进,同时考虑各个环节之间的关系,每个环节实现的具体要求而进行的风险控制与规划,并将危害降低到最小,文档经过下载可进行自定义修改,请根据实际需求进行调整与使用。 1. 安全性能综合测试仪及被测产品所使用的电源插座 均应为完好无损的固定插座。禁止使用规格不符的插座。 插座破损应及时更换。 2. 安全性能综合测试仪及被测产品所使用的电源插座 均应为完好无损的固定插座。禁止使用规格不符的插座。 插座破损应及时更换。 3. 测试过程中,严禁操作人员身体及仪器带电部位和 被测负载壳体,谨防触电。 4. 启动测试仪:按下面板电源开关,测试仪随即启 动。延时5秒后,显示自检界面。 5. 自检:确认接线盒上未连接待检设备后再按确认 键,仪器自检。自检无误,自动进入主菜单。

6. 选择测试后,根据不同规格型号的产品设立几个测试组,在主菜单下,按确认键,首先设置密码,然后在每一行的行首按上移键或下移键来选择绝缘、接地、耐压、泄漏、启动等测试项,根据测试要求设置各项目测试条件、报警值、测试时间,用左右移动键设定增减的参数值。 测试项目设置值时间 接地电阻25A,≤0.1Ω1秒 耐压试验220V工作电压1800V,10 mA 1秒 115V工作电压1500V,10mA 泄漏电流220V工作电压233V,≤1.5mA 2秒 115V工作电压115V,≤1.5mA 7. 将被测产品的电源插头插在接线盒,测试夹夹住被测产品金属外壳。 8.禁止重复关机,每次开关机间隔10秒以上,当测

检测仪使用说明书

检测仪使用说明书 一.概述 核酸蛋白检测仪、紫外检测仪是液湘色谱仪中的一种紫外检测装置,核酸蛋白检测仪、紫外检测仪是根据生命科学的发展对于现代色谱仪器的要求而改进设计的一种新型紫外检测仪。该仪器在创新方面的主要特点为: 1.该仪器除配有输出10mV记录仪信号外,还配有输出适合计算机积分仪的输口,这 样很方便构成色谱工作站系统。(可同时进行计算机和记录仪信号输出,亦可省去记录仪) 2.该仪器的数字显示设计为固定光吸收,A显示计算机用和可变量程光吸收A显示记 录仪用两种可选模式,这样可方便于规范化读数(特别是可应用于药品生产的GMP 工艺规范化管理),同时亦可根据科研需要进行可变量程的高灵敏度读数,这样可方便于对低浓度样品检测。 3.该仪器采用新型进口IP28光电倍增管和改进型电路结构,使仪器工作更为稳定可 靠。 该仪器配有上层析柱、恒流泵、部分收集器等等,即组成一套完整的液色湘色谱分离分析系统。它可应用于现代生物学研究,药物测定、农业科研、化工、食品及医疗单位对具有紫外吸收的样品作定量分析。本仪器主要元器件均采用进口,仪器全部采用LED数字显示,使用方便。 二.主要技术性能 (1)核酸蛋白检测仪提供波长:254nm、280nm。 (2)紫外检测仪提供波长:220nm、254nm、280nm、340nm。 (3)量程范围:0~100%T、0~2A、0~1A、0~0.5A、0~0.2A、0~0.1A、0~0.05A。 (4)流式样品池:容积100微升、光程3毫米。 微量样品池:容积30微升、光程10毫米。 (5)记录仪输出:10mV (6)积分仪输出:0.1A/mV (7)数显模式:固定A量程读数(0~2.0A);可变A量程读数(0~2.0A、0~1.0A、0~0.5A、0~0.2A、0~0.1A、0~0.05A)。 (8)量程在0.05A档时:噪音≦0.002A。 (9)工作环境温度:0℃~35℃。 (10)仪器可连续工作。 (11)电源:220VAC±10%50HZ。 (12)单体外形尺寸:280×180×158(mm)。 (13)主机重量:5㎏。 三.工作原理 从光源发出的光经狭缝,滤色器聚焦到样品池上,此单色光通过样品池射到光电倍增管的光阴极面上,使光束由于样品浓度不同所引起透光强度的变化转换成光电流变化,此光电流经放大器放大,并输入到对数转换器、使透光率T转换成光吸收A输出即A=lgT/1=ε·CL式中ε为待测样品的摩尔消光系数,C为样品浓度,采用摩尔/升单位,L为光程,用厘米作单位。根据上式只要测出了A、L和ε就可求出样品浓度C。若从放大器直接输入到记录仪,则在记录仪上绘出的是样品透光率T变化的图谱,若从对数转换器输入到记录仪上,在记录仪上绘出的是样品光吸收变化的图谱。 四.仪器结构 核酸蛋白检测仪、紫外检测仪是单光路结构,由紫外检测器、和记录仪部分组成现将其构造分别说明如下: 1.紫外检测仪: 它由光源、干涉滤色片、样品池、光电倍增管、放大和对数板、低压板和高压板等组成。面板上有四氟塑料管的进样口和出样口,A调零以及调节“光量”大小旋(光

FMX-003静电场测试仪中文使用说明(1)

Simco便携式数字静电测量仪 型号FMX-003 指导书 安装/操作/维护 注意 在你试图安装或者操作之前仔细阅读和理解这些指导是非常重要的。没有这么做可能会造成严重的个人受伤及[或] 损坏仪器. 在这一本手册最后部分提供有一手写的保证证书。请小心地保存起来. 使用者注意事项 注意 此仪器没有设计在机密的(危险的) 环境下使用. 它不能够被暴露用于易起火的或腐蚀的材料和气体中. 注意 此仪器专门用于静电的处理,需要远离水,油,溶剂和其他具有传导性的污染物. 对暴露在如此的污染物中将会引起产品的电气绝缘故障. 特别小心确定环境没有露珠形成. 此仪器应该在一个湿度小于60% 的环境下保存和使用. 不要插入任何的物体于用于固定感应器的口子里;不要有任何异物进入感应器的开口. 这个仪器不能在有腐蚀性的酸/碱烟或腐蚀性的气体如氯之类的环境中使用操作. 这个仪器是用电池操作的,不要连接任何其他电线. 正常的操作条件在它的名牌上有列出. 当不使用的时候,请关闭仪器. 为了仪器精确的测量请将他正确的接地. 如果跌落,此仪器很容易损坏。如果发生此类事情,它应该交由授权的技术人员做仔细的检查和必要的修理. 此仪器为电子设备,其包含的传感器可能对机械的震动和冲击非常敏感. 当然它也包含有一个微型处理芯片和电子电路,所以它不能在有许多电子噪音的环境中使用. 当电压读数超出测定范围之外的时候会中止测量. 如果超过量程使用,有可能损坏感应器. 在离子化的空气中使用该产品是可能的。但是在这种情况下正描述的10% 精度不

能够被保证. 不要在LCD 屏幕的上方施加任何压力. 此仪器在25 ± 0.5 毫米的测定距离已经被校正. 对于超出此距离校正是无效的. 不要干预感应器位臵和LED方向,否则可能改变测定距离. 该仪器在Simco 日本公司装配和检查. 请不要试图拆解或改变它的构造. 如果你不清楚有关它的操作事项和维护,请联系你所在区域的Simco 日本公司的授权代理人. 第1页 谢谢你购买Simco 产品. 当仪器被正确的安装和维护后,将会迎合你的要求并提供安全的操作. 仪器清单 请小心地将仪器从硬纸盒移出并且检查. 注意任何可能在运输期间发生的损坏. 倒空硬纸盒确定没有小部份被丢弃. 如果在运输期间发生任何的损害,请立刻通知当地的运输员. 损坏报告应该立即寄到SIMCO 日本公司. 地址和其他的相关信息在背面封页上有写. 包装项目和配件 (1) FMX-003 静电场测试……………………………………………………… 1件 (2) 离子平衡盘…………………………………………………………………….1件 (3) 1 m 长接地线…………………………………………………………………1件 (4) 软皮包………………………………………………………………………….1件 (5) 电池(类型6 F22,9 V)…………………………………………………….. 1件 (6) 操作手册/证书(这一本书) ……………………………………………….. 1件请检查是否有部件不见或没有满意的完成. 如果这样的情况发生,请连络我们或我们的代理人. 第2页 内容 使用者注意 (1) 收到仪器 (2) 包装项目和配件 (2) 目录 (3) 第一节一般的描述 (4) 第二节特征 (5) 第三节规格 (6) 第四节外观 (7) 第五节电源 (8)

通风机综合测试仪操作规程

TF-3通用机综合测试仪操作规程 1、概述 TF-3通风机综合测试仪是新一代嵌入式系统测试仪,自动测试通风机风速、风量、相对静压、电机输入功率、转速、温度、湿度、环境大气压、风机全压与静压的差压,能同便携微机实现数据通讯,完成测试,绘制通风机运行特性曲线,并能实现测试数据处理,自动生成测试报告。 2、主要技术指标: (1)风速:0.1~3.5m/s 精度:±2% (2)启动风速:V≤0.6m/s 分辨率:0.1m/s (3)相对静压:0~10KPa 精度:±0.5% (4)大气压:0~200KPa 精度:±1% (5)湿度:0~100 精度:±5% (6)温度:0~50℃误差:±1℃ (7)转度:0~3000r/min 精度:±0.2% (8)电机输入功率:输入电压范围交流0~500V 电流范围0~500A 功率测试范围0~999KW 精度:±1% (9)差压:0~1000Pa 精度:±1% 3、仪器使用条件 (1)电源电压:交流220±22V 频率50Hz±1Hz (2)使用温度:风速探头:(-10℃~50℃) 主机:0~40℃ 微机:室温

(3)湿度:≤85% (4)功耗:<35VA 4、现场检测(使用方法、适用范围及注意事项) (1)风机测定方案的确定 1)不停产测定 不停产测定风机必备以下条件:①工作风机和待测风机能有效地隔离②电网满足两台风机同时工作,不致因过载而跳闸;③被测风机能形成独立风流回路且便于调节;④便于安装传感器,特别是测风和测压传感器。 2)停产测定 与不停产测定相比,停产测定能为测定工作提供较佳的条件,特别是在风流稳定方法,从而使测定结果更接近实际。 (2)工况调节 在进行风量调节时,应备足方木板,并且要具备足够的强度,并准备2~3块较窄的木板以备工况微调用。 轴流风机的风量调节通常是由大到小调节,离心风机反之。 在入风口调节工况的人员应佩带好安全带,其他人员不得进入调风口,以确保人身安全。 根据现场具体条件,可以灵活采用各种方式调节工况。 (3)测量截面位置的选择 选择测量截面的位置应位于风道的缓变流处,在一般情况下,要保证其截面为缓度变流,要求在测量截面位置的上游和下游有2倍风硐直径的直线距离。(4)安装测量仪器 1)先仔细检查电源电压是否符合仪器使用电压的工作范围,确认无误后方

继电保护测试仪说明书

微机继电保护测试仪 使 用 说 明 书

目录 目录 (1) 第一部分微机继电保护测试仪使用说明 (3) 第一章装置特点与技术参数 (4) 第二章装置硬件结构 (6) 第三章单机操作模块功能说明 (8) 第四章外接PC机操作说明 (21) 第二部分继保软件操作说明 (21) 第五章软件操作方法简介 (22) 第六章交流试验 (24) 第七章直流试验 (32) 第八章状态系列 (34) 第九章谐波叠加试验 (38) 第十章频率及高低周试验 (41) 第十一章功率方向及阻抗试验 (45) 第十二章同期试验 (49) 第十三章整组试验Ⅰ和Ⅱ (54) 第十四章距离和零序保护 (59) 第十五章线路保护 (64) 第十六章阻抗特性 (70) 第十七章差动保护 (73) 第十八章 6-35KV微机线路保护综合测试 (80) 附录1:外接电脑串行通信口的设置 (85) 附录2:插接U盘等设备时设备驱动安装方法 (87) 附录3:各种继电器的试验方法 (87)

第一部分 继保使用说明

第一章装置特点与技术参数 第一节主要特点 ◆标准的4相电压3相电流输出具有4相电压3相电流输出,可方便地进行各种组合输出进行各种 类型保护试验。每相电压可输出120V,电流三并可输出120A,第4相电压Ux为多功能电压项,可设为4种3U0或检同期电压,或任意某一电压值的情况输出。 ◆单机操作方便单机由方便灵活的旋转鼠标通过大屏幕液晶显示屏进行操作,全部中文显示。可 完成现场大多数试验检定工作,可对各种继电器及微机保护进行检定,并可模拟各种复杂的瞬时性、永久性、转换性故障进行整组试验。开机即可使用,操作方便快捷。 ◆双操作方式,联接电脑运行通过Windows平台上的全套中文操作软件,可进行各种大型复杂 及自动化程度更高的校验工作,可方便地测试及扫描各种保护定值,可实时存贮测试数据,显示矢量图,绘制故障波形,联机打印报表等。 ◆软件功能强大可完成各种自动化程度高的大型复杂校验工作,如三相差动试验、厂用电快切、 备自投试验、线路保护检同期重合闸等,能方便地测试及扫描各种保护定值,进行故障回放,实时存储测试数据,显示矢量图,联机打印报告等。 ◆开关量接点丰富7路接点输入和2对空接点输出。输入接点为空接点和0~250V电位接点兼容, 可智能自动识别。输入、输出接点可根据用户需要扩展。 ◆大屏幕LCD显示屏本机采用320×240点阵大屏幕高分辨率图形液晶显示屏,全部操作过程均在 显示屏上设定,操作界面和试验结果均汉化显示,显示直观清晰。 ◆自我保护采用合理设计的散热结构,并具有可靠完善的多种保护措施及电源软启动,和一定的 故障自诊断及闭锁功能。 ◆具有独立专用直流电源输出装置设有一路110V 及220V专用可调直流电源输出。 ◆性价比高属于跨专业联合设计产品,综合了多专业的先进科技成果。兼具大型测试仪的性能, 和小型测试仪的价位,具有很高的性能价格比。

【精品】静电放电测试仪操作规程

静电放电测试仪(三基)操作规程 为正确、安全、规范的使用SKS-0230,以评定样机在经受静电放电干扰时的性能,特制定本操作规程。 一、【操作程序】 1、按SKS-0230使用说明书接好放电枪,并接通电源。 2、给被试样机施加额定电压。 3、打开静电放电发生器电源开关,将开关拨动到上方CONTACT位置。按下HV ON 开关,接通直流高压电源,按顺时针方向调节VOLTS ADJUST调节旋钮,使表头显示的电压值为8kV。 4、持枪,将放电枪垂直于被试品表面,按下REDAY键,再按下“枪机”对样机进行放电试验。按下STOP键并结束该极性放电试验。 5、换极性(+←→-),按3、4重复试验,注意应确保电压调节旋钮逆时针旋转到底且高压开关处于OFF状态方可转换电压极性。 6、试验结束后,应将电压调节旋钮逆时针调到底,使高压开关处于OFF状态,按下“枪机”对着铝板放完电。然后断开工作电源,拔掉电源插头,撤去电表工作电源,清理好现场。 二、【测试条件】 1、环境温度:15℃~35℃;相对湿度 2、相对湿度:30%~60%(如果湿度过大,必须经过除湿处理再进行试验); 三、【注意事项】 1、试验人员必须经培训后才能进行设备操作,操作前应阅读设备使用说明书;

2、仪器的F.G端子要良好接地; 3、仪器通电后,不得用手触摸放电电极,以防电击; 4、仪器使用过程中,不得随意切换量程,如需切换量程,务必先将电压调节旋钮逆时针旋转到底,使高压开关处于OFF状态。以免在量程切换过程中电压突变损坏高压电源。 5、关机时也必须先将电压调节旋钮逆时针旋转到底,使高压开关处于OFF状态,然后才能切断主机的工作电源,否则关机瞬间电压的突变可能会损坏高压电源。 6、非有关人员严禁操作本仪器。

气体检测仪操作规程

便携式四合一气体检测仪操作规程 1 范围: 本规程规定了设备启动前,对罐内气体实施检测,四合一气体检测仪的检查准备,检测操作步骤及安全注意事项。 本规程适用于机械清洗项目对各类储油罐清洗前的气体检测操作。 2规范性引用文件 SY6503-2000 可燃气体检测报警器使用规范 3 四合一气体检测仪检测前的检查准备 3.1检查电池电量是否充足(3.3V以上),不充足及时充电; 3.2检查进气口气滤有无杂物堵住,堵住需清理干净或更换; 4 操作步骤: 4.1开机操作: 4.1.1按[MODE]键并保持1秒,LCD显示“on”,LED亮,峰鸣器响一声,仪器开机; 4.1.2LCD显示版本号,同时进行预热和自检。 4.1.3预热和自检完成致10秒倒计时结束,仪器进入检测模式,确认仪器运行正常。 4.1.4确认确实在抽新鲜空气,确认氧气指示计的指示值确实为20.9%。 4.1.5将取样管端部插入测试点中,待测试值变化稳定后,读数并记

录。 4.1.6从测试点中拿出取样管,置于空气中,待LED显示值回复到空气中状态后,再进行下一测试点测试。 4.2关机操作: 4.2.1按住按键不放,LCD显示5秒倒计时,倒计时结束后LCD显示“off”,随后仪器无显示,仪器关机。 5注意事项: 5.1仪器更换电池或简单维修时应在安全场所进行。 5.2传感器和仪器要注意防水和杂质。 5.3仪器长期不工作时,应关机,置于干燥、无尘、符合储存温度的环境中。 5.4调整好的仪器不要随便打开盖。

硫化氢气体检测报警仪操作规程 1 范围: 本标准规定了设备启动前,对罐内气体实施检测,硫化氢检测仪的检查准备,检测操作步骤及安全注意事项。 本标准适用于机械清洗项目对各类储油罐清洗前的气体检测操作。 2规范性引用文件 《COWS施工手册》 3 硫化氢检测仪检测前的检查准备 3.1检查电池电量是否充足,不充足更换; 3.2检查进气口不被杂物堵住,堵住清理干净; 4 操作步骤: 4.1开机操作 4.1.1确认电池已经装入仪器,按住按键3秒,LCD显示“on”,红色LED亮,蜂鸣器响一声,振动器振动,仪器开机。 4.1.2LCD显示版本号,同时进行预热和自检。 4.1.310秒倒计时预热和自检完成后,仪器进入检测模式,显示实时读数。 4.2关机操作 按住按键不放,LCD显示5秒倒计时,倒计时结束后LCD显示“off”,随后仪器无显示,仪器关机。 4.3检测模式说明:

IXIA测试仪使用手册

IXIA测试仪使用手册 一、设备开机、关机 (一)开机 打开IXIA测试仪电源,等待设备启动完成,需将测试客户端IP设置为192.168.1.200,测试仪IP地址为192.168.1.100,开IxNetwork或IxLoad可连接测试仪表明设备完成开机。 (二)关机 在运行中输入mstsc进入远程桌面,连接到计算机192.168.1.100,在运行中输入shutdown.exe -s -t 3让设备在3秒内关闭,让系统自动关闭。 二、二三层测试配置(IxNetwork) 使用IxNetwork配置测试基本流程如下图所示。 (一)添加测试端口 点击标题栏中的 或Overview表页中的

连接192.168.1.100测试板卡 添加测试用端口

然后点击OK键,完成测试端口添加。 (二)配置端口、协议 启用测试端口Ping及ARP,如下图所示。 添加测试端口IP地址、网关,并使能端口,如下图所示,添加的网关地址需是实际存在的,可以是交换机的网关地址或测试端口对端IP地址。

(三)配置流量 选择配置流量。 Type of Traffic选项可选择Raw、Ethernet/vlan、IPV4分别对应原始报文流(需手动编辑,用来打广播包流)、二层流、三层流(需配置IP地址及网关),Bi-Directional表示流是双向流。 1、IPV4(三层流)

在设置好Type of Traffic、Traffic Mesh以及选择好端口后,点击添加Endpoint。点击NEXT,在Packet/QoS、Flow Group Setup、Frame Setup、Rate Setup保持默认配置。点击NEXT,进入Flow Tracking,选择Traffic Item选项。点击NEXT,在后续Protocol Behaviors、 Preview以及Validate中保持默认选项,同时检查配置的有效性。

SCS 718静电场测试仪操作规程

文件编号 发行日期2019年10月23 静电场测试仪 操作规程 版本更改类型生效日期更改内容 A0 / 2019-10-23 首次制定 会签部门: 品质:工程: 采购:业务: PIE:体系: 生产:PMC: 人力资源及行政:财务:

文件编号 版本 A0 静电场测试仪操作规程 页码 第1页,共3页 发行日期 2019年10月23日 1.目的: 为规范静电场测试仪使用,确保仪器能准确检测,使生产品质能够得到有效管控, 特制定此作业指导书。 2.范围: 本文件适用于公司内SCS 718静电场测试仪的使用。 3.仪器操作界面简述 3.1仪器图示 4、测试步骤说明: 4.1按POWER 键一次开机。 4.2 开机后,按一次设置键选择量程(低量程(0~±1.999KV )/高量程0~±19.99KV ) 制定部门 品质部 3 编写/日期 2 审核/日期 1 正电荷选择键 负电荷选择键 电源开关(POWER ) 测试值显示区域 输出端 电极板 设置键RANGE ZERO 电源开关(HOLD ) 信号孔 高压产生器 静电场测试仪主机 高量程 (小数点后两位数) 低量程 (小数点后三位数)

发行日期 2019年10月23日 4.3 归零 将量程切换至低量程,将测试仪电极板与接地导电体接触,按住“设置键”直至显“.000”。 4.4离子风扇测量方法 4.4.1 取高压产生器与归零好的718A 静电场测试仪电极板相接触,按”+”电荷按钮,显示为静电压+1KV 以上. 4.4.2将718A 静电场测试仪电极板对准离子风扇吹风口,保持有效距离在30-60cm,在3S 内充电时的电 压降低至50V 以下. 4.4.3 取高压产生器与归零好的718A 静电场测试仪电极板相接触,按”-”电荷按钮,显示为静电压-1KV 以上. 4.4.4将718A 静电场测试仪电极板对准离子风扇吹风口,保持有效距离在30-60cm,在3S 内充电时的电 压降低至50V 以下. 制定部门 品质部 3 编写/日期 2 审核/日期 1 “+”电荷充电 消除静电 “-”电荷充电

防静电检测方法

防静电检测方法 1 2020年4月19日

防静电检测方法 2 2020年4月19日

前言 本规范由公司ESD项目组提出。 本规范主要起草和解释部门:公司ESD项目组 本规范主要起草人:姜延平 本规范主要审核人:林建平,陈迎曦 本规范批准人:林建平,鞠英年,陈迎曦 3 2020年4月19日

1范围 本规范制定了UT公司防静电材料、用品、工具、防静电接地等防静电技术指标的检测方法和采用的仪器。目的在于规范防静电技术指标的检测工作,指导现场操作。 本规范适用于UT公司防静电系统各要素(包括地面、接地系统、工作台、工作椅、工位器具、物流传递工具、包装材料、人员、腕带、服装、离子风机、防静电工具等)的防静电性能指标检测。防静电专用检测仪器的检测按照仪器仪表的检测标准进行,本规范不涉及。 2规范性引用文件 下列文件中的条款经过本规范的引用而成为本规范的条款。凡是注日期的引用文件,其随后所有的修改单(不包括勘误的内容)或修订版均不适用于本规范,然而,鼓励根据本规范达成协议的各方研究是否可使用这些文件的最新版本。凡是不注日期的引用文件,其最新版本适用于本规范。 3 表面电阻(surface resistance):两个特定的放置于材料同一面的电极 4 2020年4月19日

之间的电压与它们经过电流的比值。 体电阻(volume resistance):单位厚度上的直流电压,与经过材料的单位面积电流的比值,测试电极放置于材料相对面的对应位置。 接地电阻(Earth Resistance):被接地体与地下零电位面之间的接地引线电阻、接地器电阻、接地器与土壤之间的过渡电阻和土壤的溢流电阻之和。 摩擦起电(Triboelectrification):用摩擦的方法使两物体分别带有等值异号电荷的过程。 衰减时间(decay time):静电电压从峰值电压降低到给定比例的时间。例如:在15%相对湿度的情况下,静电电压从 V降低到100V的衰减时间小于等于1秒。 屏蔽泄漏电压:因屏蔽体外部的高电场而使屏蔽体内部获得的感应电场电压或外部高电场穿透屏蔽体衰减的残留电场电压,又称屏蔽残余电压。 静电中和(Electrostatic Neutralization):带电体上的电荷与其内部和外部相反符号的电荷(电子或离子)的复合而使所带静电部分或全部消失的现象。 4防静电参数测试方法 4.1防静电主要参数及测试仪器 5 2020年4月19日

综合检测仪作业指导书

ZY-06-2012 JMZX -3006综合测试仪操作规程 一、概述 JMZX-3006综合测试仪是一种便携式、多功能、智能读数仪。该系列仪器均能对钢弦传感器、电感调频类传感器、半导体温度传感器进行测量。该系列仪器有三种型号:JMZX-3001综合测试仪(单弦);JMZX-3003综合测试仪(三弦);JMZX-3006综合测试仪(六弦)。其中三弦、六弦测试仪为多通道综合测试仪,可测量多弦传感器(如压力传感器)。 JMZX-3006综合测试仪具有检测速度快、精度高、使用简单方便等特点。仪器体积小、重量轻,采用可充电电池供电,使用携带极为方便。 该仪器配合本公司生产的智能型传感器使用。能在传感器内自动记录传感器编号、系数,自动计算应变、自动检测温度的结果并作温度修正,保存记录测试结果,供以后查阅或送计算机处理。 二、使用方法 (1)传感器接线连接 ①振弦式应变传感器或压力盒连接:传感器导线为四芯屏蔽线,如传感器配有九芯插头(连接方式为1:红、2:黄或黑、3:兰或白、4:绿),则直接插入仪器面板之INP插口即可,如传感器未接插头,可配本公司之转换插头再按红-红、黄-黄或黑、兰-兰或白、绿-绿接线即可。 ②三弦穿心或压力传感器连接:上述传感器导线为六芯屏蔽线,如传感器配有九芯插头(连接方式为1:红、2:黄、3:兰、4:绿、5:橙、6:紫),则直接插入仪器面板之INP插口即可,如传感器未接插头,可配本公司之转换插头再按红-红、黄-黄、兰-兰、绿-绿、橙-橙、紫-紫接线即可。 ③六弦穿心或压力传感器连接:上述传感器导线为九芯屏蔽线,该传感器均配为九芯插头,直接插入仪器面板之INP插口即可。 ④温度传感器:温度传感器导线为三芯屏蔽线,配专用转换插头后接线,其中红色接传感器红线,兰色接传感器兰线,绿色接传感器黄线。 (2)开机:按“ON/OFF”键 (3)仪器显示时钟及功能提示 测量、查阅、通讯、校时如果连接了温度传感器或温度型应变计(如:JMZX -215AT),仪器还应显示温度值。

ESD测试操作规程(强力推荐)

静电放电(ESD)测试指引 一、目的 规范手机静电测试方法,达到测试的准确性和可靠性 二、测试仪器设备 静电放电测试仪、静电试验台、垂直耦合板、水平耦合板、静电枪 三、测试样机的状态 静电测试前,首先检查测试样机的按键,LCD显示,背光灯、通话等各项功能都正常,保证样机是良品。 四、测试条件 测试环境:温度:15℃~35℃,湿度:30%RH~60%RH 五、测试等级 空气放电:+2KV、+4KV、+6KV、+8KV、+10KV(开发阶段:±10KV;量产阶段:±8KV)接触放电:+2KV、+4KV、+6KV(开发阶段:±6KV;量产阶段±4KV) 垂直耦合放电:+2KV、+4KV、+6KV、+8KV、+10KV 水平耦合放电:+2KV、+4KV、+6KV、+8KV、+10KV 六、测试注意事项 1、在进行放电操作时不要将枪头接触到人体,以免对人体产生危险(注意:高压危险)。 2、在信息收集时要详细,描述要清晰。 3、空气放电时不允许放电头接触到机壳本身(必须距离机壳5-10mm)。 4、翻盖机在测试时必须打开翻盖,然后再进行测试,滑盖机在测试时必须滑开方可进 行全方位测试。 5、每次静电测试后,用接地线对枪头放电的部位放电。然后检查放电部位的功能是否 正常,再继续放电。

七、静电放电(ESD)测试流程 八、测试判定标准 样机出现功能暂时性(失效时间大于3秒)或永久性失效(故障不恢复)判为不合格。 出现屏闪或3秒内可自行恢复之故障判为合格。样机出现严重外观不良(如镜片及装饰件严重掉漆),应判定为不合格。试验样品无功能降级故障(如通话质量下降,MP3音质降低,摄相图片画面降级),存储的信息不可丢失。 九、记录 放电后进行功能测试,将测试中与到的异常情况详细清楚的记录。 附件一

甲醛检测仪使用说明书

美国INTERASCAN 4160型甲醛检测仪 使用说明书

美国INTERASCAN 4160型甲醛检测仪简介美国INTERASCAN 4160型甲醛检测仪是专门用于检测空气中甲醛含量的直读式仪器,仪器有内置采样泵,检测元件是长寿命、高可靠性的电化学传感器。 操作时,样气进入传感器,气体浓度以ppm为单位直接显示。 仪器的基本部件是:传感器、电池、电路。电路包括:零点补偿、传感器偏置、放大器、采样泵和液晶显示器等。 传感器原理 美国INTERASCAN 4160型甲醛检测仪采用三电极传感器有一系列优点,INTERASCAN传感器电解质是不活动的类似于闪光灯和镍镉电池中的电解质。所以不需要考虑电池损坏或酸对仪器的损坏。游离的电解质减少将清除不希望的传感器噪音干扰,特别是低浓度测量需要高放大倍数时,传感器有一个密封的储气室,这不仅使传感器寿命更长,而且消除了参比电极污染的可能性,不象二电极传感器需要空气来操作,INTERASCAN 传感器可用于厌氧环境。 INTERASCAN传感器是高灵敏度的检测器,根据气体的类型,其灵敏度大于扩散性传感器的50~200倍。可以测量非常低的值,这对于低浓度气体测量是重要的条件。 操作原理 美国INTERASCAN 4160型甲醛检测仪采用电化学电压型传感器,是一种电化学气体检测器,它是在控制扩散的条件下运行的。样气的气体分子被吸收到电化学敏感电极,经过扩散介质后,在适当的敏感电极电位下气体分子发生电化学反应,这一反应产生一个与气体浓度成正比的电流,这一电流转换为电压值并送给仪表读数或记录仪记录。 技术参数: 显示:液晶数字显示 测量范围:0~19.99ppm 最小读数:0.01ppm 不确定度:0.03ppm (0~0.5ppm) 体积:178×102×225mm3 重量:2Kg

相关文档