GUIDE TO INSPECTIONS OF PHARMACEUTICAL QUALITY CONTROL LABORATORIES

Note: This document is reference material for investigators and other FDA

personnel. The document does not bind FDA, and does no confer any rights,

privileges, benefits, or immunities for or on any person(s).

1. INTRODUCTION

The pharmaceutical quality control laboratory serves one of the most

important functions in pharmaceutical production and control. A significant

portion of the CGMP regulations (21 CFR 211) pertain to the quality control

laboratory and product testing. Similar concepts apply to bulk drugs.

This inspection guide supplements other inspectional information contained

in other agency inspectional guidance documents. For example, Compliance

Program 7346.832 requiring pre-approval NDA/ANDA inspections contains

general instructions to conduct product specific NDA/ANDA inspection audits

to measure compliance with the applications and CGMP requirements. This

includes pharmaceutical laboratories used for in-process and finished

product testing.

2. OBJECTIVE

The specific objective will be spelled out prior to the inspection. The

laboratory inspection may be limited to specific issues, or the inspection

may encompass a comprehensive evaluation of the laboratory's compliance with

CGMP's. As a minimum, each pharmaceutical quality control laboratory should

receive a comprehensive GMP evaluation each two years as part of the

statutory inspection obligation.

In general these inspections may include

-- the specific methodology which will be used to test a new product

-- a complete assessment of laboratory's conformance with GMP's

-- a specific aspect of laboratory operations

3. INSPECTION PREPARATION

FDA Inspection Guides are based on the team inspection approach and our

inspection of a laboratory is consistent with this concept. As part of our

effort to achieve uniformity and consistency in laboratory inspections, we

expect that complex, highly technical and specialized testing equipment,

procedures and data manipulations, as well as scientific laboratory

operations will be evaluated by an experienced laboratory analyst with

specialized knowledge in such matters.

District management makes the final decision regarding the assignment of

personnel to inspections. Nevertheless, we expect investigators, analysts

and others to work as teams and to advise management when additional

expertise is required to complete a meaningful inspection.

Team members participating in a pre-approval inspection must read and be

familiar with Compliance Program 7346.832, Pre-Approval

Inspections/Investigations. Relevant sections of the NDA or ANDA should be reviewed prior to the inspection; but if the application is not available

from any other source, this review will have to be conducted using the

company's copy of the application.

Team members should meet, if possible, prior to the inspection to discuss

the approach to the inspection, to define the roles of the team members, and to establish goals for completion of the assignment. Responsibilities for

development of all reports should also be established prior to the

inspection. This includes the preparation of the FDA 483.

The Center for Drug Evaluation and Research (CDER) may have issued

deficiency letters listing problems that the sponsor must correct prior to the approval of NDA/ANDA's and supplements. The inspection team is expected to review such letters on file at the district office, and they are expected to ask the plant for access to such letters. The team should evaluate the

replies to these letters to assure that the data are accurate and authentic. Complete the inspection even though there has been no response to these

letters or when the response is judged inadequate.

4. INSPECTION APPROACH

A. General

In addition to the general approach utilized in a drug CGMP inspection, the inspection of a laboratory requires the use of observations of the

laboratory in operation and of the raw laboratory data to evaluate

compliance with CGMP's and to specifically carry out the commitments in an application or DMF. When conducting a comprehensive inspection of a

laboratory, all aspects of the laboratory operations will be evaluated.

Laboratory records and logs represent a vital source of information that

allows a complete overview of the technical ability of the staff and of

overall quality control procedures. SOPs should be complete and adequate and the operations of the laboratories should conform to the written procedures. Specifications and analytical procedures should be suitable and, as

applicable, in conformance with application commitments and compendial

requirements.

Evaluate raw laboratory data, laboratory procedures and methods, laboratory equipment,including maintenance and calibration, and methods validation data to determine the overall quality of the laboratory operation and the ability to comply with CGMP regulations.

Examine chromatograms and spectra for evidence of impurities, poor

technique, or lack of instrument calibration.

s use systems that provide for the investigation of

laboratory test failures. These are generally recorded in some type of log. Ask to see results of analyses for lots of product that have failed to meet specifications and review the analysis of lots that have been retested,

rejected, or reworked. Evaluate the decision to release lots of product when the laboratory results indicate that the lot failed to meet specifications and determine who released them.

B. Pre-Approval

Documents relating to the formulation of the product, synthesis of the bulk drug substance, product specifications, analysis of the product, and others are examined during the review process in headquarters. However, these

reviews and evaluations depend on accurate and authentic data that truly

represents the product.

Pre-approval inspections are designed to determine if the data submitted in an application are authentic and accurate and if the procedures listed in

the application were actually used to produce the data contained in the

application. Additionally, they are designed to confirm that plants

(including the quality control laboratory) are in compliance with CGMP

regulations.

The analytical sections of drug applications usually contain only test

results and the methods used to obtain them. Sponsors are not required to

file all the test data because such action would require voluminous

submissions and would often result in filing redundant information. Sponsors may deliberately or unintentionally select and report data showing that a

drug is safe and effective and deserves to be approved. The inspection team must decide if there is valid and scientific justification for the failure to report data which demonstrates the product failed to meet its

predetermined specifications.

Coordination between headquarters and the field is essential for a complete review of the application and the plant. Experienced investigators and

analysts may contact the review chemist (with appropriate supervisory

concurrence) when questions concerning specifications and standards arise.

Inspections should compare the results of analyses submitted with results of analysis of other batches that may have been produced. Evaluate the methods and note any exceptions to the procedures or equipment actually used from

those listed in the application and confirm that it is the same method

listed in the application. The analyst is expected to evaluate raw

laboratory data for tests performed on the test batches (biobatches and

clinical batches) and to compare this raw data to the data filed in the

application.

5. FAILURE (OUT-OF-SPECIFICATION) LABORATORY RESULTS

Evaluate the company's system to investigate laboratory test failures. These investigations represent a key issue in deciding whether a product may be

released or rejected and form the basis for retesting, and resampling.

In a recent court decision the judge used the term "out-of-specification"

(OOS) laboratory result rather than the term "product failure" which is more

common to FDA investigators and analysts. He ruled that an OOS result

identified as a laboratory error by a failure investigation or an outlier

test. The court provided explicit limitations on the use of outlier tests

and these are discussed in a later segment of this document., or overcome by retesting. The court ruled on the use of retesting which is covered in a

later segment of this document. is not a product failure. OOS results fall into three categories:

-- laboratory error

-- non-process related or operator error

-- process related or manufacturing process error

A. LABORATORY ERRORS

Laboratory errors occur when analysts make mistakes in following the method of analysis, use incorrect standards, and/or simply miscalculate the data. Laboratory errors must be determined through a failure investigation to

identify the cause of the OOS. Once the nature of the OOS result has been

identified it can be classified into one of the three categories above. The inquiry may vary with the object under investigation.

B. LABORATORY INVESTIGATIONS

The exact cause of analyst error or mistake can be difficult to determine

specifically and it is unrealistic to expect that analyst error will always be determined and documented. Nevertheless, a laboratory investigation

consists of more than a retest. The inability to identify an error's cause with confidence affects retesting procedures, not the investigation inquiry required for the initial OOS result.

The firm's analyst should follow a written procedure, checking off each step as it is completed during the analytical procedure. We expect laboratory

test data to be recorded directly in notebooks; use of scrap paper and loose paper must be avoided. These common sense measures enhance the accuracy and integrity of data.

Review and evaluate the laboratory SOP for product failure investigations. Specific procedures must be followed when single and multiple OOS results

are investigated. For the single OOS result the investigation should include the following steps and these inquiries must be conducted before there is a retest of the sample:

o the analyst conducting the test should report the OOS result to the

supervisor

o the analyst and the supervisor should conduct an informal laboratory

investigation which addresses the following areas:

1. discuss the testing procedure

2. discuss the calculation

3. examine the instruments

4. review the notebooks containing the OOS result

An alternative means to invalidate an initial OOS result, provided the

failure investigation proves inconclusive, is the "outlier" test. However, specific restrictions must be placed on the use of this test.

1. Firms cannot frequently reject results on this basis.

2. The USP standards govern its use in specific cases only.

3. The test cannot be used for chemical testing results. An initial content uniformity test was OOS followed by a passing retest. The initial OOS result was claimed the result of analyst error based on a statistical evaluation of the data. The court ruled that the use of an outlier test is inappropriate in this case..

4. It is never appropriate to utilize outlier tests for a statistically

based test, i.e., content uniformity and dissolution.

Determine if the firm uses an outlier test and evaluate the SOP.

Determine that a full scale inquiry has been made for multiple OOS results. This inquiry involves quality control and quality assurance personnel in

addition to laboratory workers to identify exact process or non process

related errors.

When the laboratory investigation is inconclusive (reason for the error is not identified) the firm:

1. Cannot conduct 2 retests and base release on average of three tests

2. Cannot use outlier test in chemical tests

3. Cannot use a re-sample to assume a sampling or preparation error

4. Can conduct a retest of different tablets from the same sample when a

retest is considered appropriate (see criteria elsewhere)

C. FORMAL INVESTIGATIONS

Formal investigations extending beyond the laboratory must follow an outline with particular attention to corrective action. The company must:

1. State the reason for the investigation

2. Provide summation of the process sequences that may have caused the

problem

3. Outline corrective actions necessary to save the batch and prevent

similar recurrence

4. List other batches and products possibly affected, the results of

investigation of these batches and products, and any corrective action.

Specifically:

o examine other batches of product made by the errant employee or machine

o examine other products produced by the errant process or operation

5. Preserve the comments and signatures of all production and quality

control personnel who conducted the investigation and approved any

reprocessed material after additional testing

D. INVESTIGATION DOCUMENTATION

Analyst's mistakes, such as undetected calculation errors, should be

specified with particularity and supported by evidence. Investigations along with conclusions reached must be preserved with written documentation that enumerates each step of the investigation. The evaluation, conclusion and

corrective action, if any, should be preserved in an investigation or

failure report and placed into a central file.

E. INVESTIGATION TIME FRAMES

All failure investigations should be performed within 20 business days of

the problem's occurrence and recorded and written into a failure or

investigation report.

6. PRODUCT FAILURES

An OOS laboratory result can be overcome (invalidated) when laboratory error has been documented. However, non-process and process related errors

resulting from operators making mistakes, equipment (other than laboratory equipment) malfunctions, or a manufacturing process that is fundamentally

deficient, such as an improper mixing time, represent product failures.

Examine the results of investigations using the guidance in section 5 above and evaluate the decision to release, retest, or rework products.

7. RETESTING

Evaluate the company's retesting SOP for compliance with scientifically

sound and appropriate procedures. A very important ruling in one recent

court decision sets forth a procedure to govern the retesting program. This district court ruling provides an excellent guide to use in evaluating some aspects of a pharmaceutical laboratory, but should not be considered as law, regulation or binding legal precedent. The court ruled that a firm should

have a predetermined testing procedure and it should consider a point at

which testing ends and the product is evaluated. If results are not

satisfactory, the product is rejected.

Additionally, the company should consider all retest results in the context of the overall record of the product. This includes the history of the

product. The court ordered a recall of one batch of product on the basis of an initial content uniformity failure and no basis to invalidate the test

result and on a history of content uniformity problems with the product.,

type of test performed, and in-process test results. Failing assay results cannot be disregarded simply on the basis of acceptable content uniformity results.

The number of retests performed before a firm concludes that an unexplained OOS result is invalid or that a product is unacceptable is a matter of

scientific judgment. The goal of retesting is to isolate OOS results but

retesting cannot continue ad infinitum.

In the case of nonprocess and process-related errors, retesting is suspect. Because the initial tests are genuine, in these circumstances, additional

testing alone cannot contribute to product quality. The court acknowledged that some retesting may precede a finding of nonprocess or process-based

errors. Once this determination is made, however, additional retesting for purposes of testing a product into compliance is not acceptable.

For example, in the case of content uniformity testing designed to detect

variability in the blend or tablets, failing and non-failing results are not inherently inconsistent and passing results on limited retesting do not rule out the possibility that the batch is not uniform. As part of the

investigation firms should consider the record of previous batches, since

similar or related failures on different batches would be a cause of

concern.

Retesting following an OOS result is ruled appropriate only after the

failure investigation is underway and the failure investigation determines in part whether retesting is appropriate. It is appropriate when analyst

error is documented or the review of analyst's work is "inconclusive" , but it is not appropriate for known and undisputed non-process or process

related errors.

The court ruled that retesting:

o must be done on the same, not a different sample

o may be done on a second aliquot from the same portion of the sample that was the source of the first aliquot

o may be done on a portion of the same larger sample previously collected

for laboratory purposes

8. RESAMPLING

Firms cannot rely on resampling. The court ordered the recall of one batch of product after having concluded that a successful resample result alone

cannot invalidate an initial OOS result. to release a product that has

failed testing and retesting unless the failure investigation discloses

evidence that the original sample is not representative or was improperly

prepared.

Evaluate each resampling activity for compliance with this guidance.

9. AVERAGING RESULTS OF ANALYSIS

Averaging can be a rational and valid approach when the object under

consideration is total product assay, but as a general rule this practice

should be avoided. The court ruled that the firm must recall a batch that

was released for content uniformity on the basis of averaged test results. because averages hide the variability among individual test results. This

phenomenon is particularly troubling if testing generates both OOS and

passing individual results which when averaged are within specification.

Here, relying on the average figure without examining and explaining the

individual OOS results is highly misleading and unacceptable.

Content uniformity and dissolution results never should be averaged to

obtain a passing value.

In the case of microbiological turbidimetric and plate assays an average is preferred by the USP. In this case, it is good practice to include OOS

results in the average unless an outlier test (microbiological assays)

suggests the OOS is an anomaly.

10. BLEND SAMPLING AND TESTING

The laboratory serves a vital function in blend testing which is necessary to increase the likelihood of detecting inferior batches. Blend uniformity testing cannot be waived in favor of total reliance on finished product

testing because finished product testing is limited.

One court has ruled that sample size influences ultimate blend test results and that the sample size should resemble the dosage size. Any other practice would blur differences in portions of the blend and defeat the object of the test. If a sample larger than the unit must be taken initially, aliquots

which resemble the dosage size should be carefully removed for the test,

retests, and reserve samples. Obviously, the initial larger sample should

not be subjected to any additional mixing or manipulation prior to removing test aliquots as this may obscure non-homogeneity.

Multiple individual blend uniformity samples taken from different areas

cannot be composited. However when variation testing is not the object of

assay testing, compositing is permitted.

If firms sample product from sites other than the blender, they must

demonstrate through validation that their sampling technique is

representative of all portions and concentrations of the blend. This means that the samples must be representative of those sites that might be

problems; e.g. weak or hot spots in the blend.

11. MICROBIOLOGICAL

The review of microbiological data on applicable dosage forms is best

performed by the microbiologist (analyst). Data that should be reviewed

include preservative effectiveness testing, bioburden data, and product

specific microbiological testing and methods.

Review bioburden (before filtration and/or sterilization) from both an

endotoxin and sterility perspective. For drug substance labs evaluate

methods validation and raw data for sterility, endotoxin testing,

environmental monitoring, and filter and filtration validation. Also,

evaluate the methods used to test and establish bioburdens.

Refer to the Microbiological Inspection Guide for additional information

concerning the inspection of microbiological laboratories.

12. SAMPLING

Samples will be collected on pre-approval inspections. Follow the sampling guidelines in CP 7346.832, Part III, pages 5 and 6.

13. LABORATORY RECORDS AND DOCUMENTATION

Review personal analytical notebooks kept by the analysts in the laboratory and compare them with the worksheets and general lab notebooks and records. Be prepared to examine all records and worksheets for accuracy and

authenticity and to verify that raw data are retained to support the

conclusions found in laboratory results.

Review laboratory logs for the sequence of analysis versus the sequence of manufacturing dates. Test dates should correspond to the dates when the

sample should have been in the laboratory. If there is a computer data base, determine the protocols for making changes to the data. There should be an audit trail for changes to data.

We expect raw laboratory data to be maintained in bound, (not loose or scrap sheets of paper), books or on analytical sheets for which there is

accountability, such as prenumbered sheets. For most of those manufacturers which had duplicate sets of records or "raw data", non-numbered loose sheets of paper were employed. Some companies use discs or tapes as raw data and

for the storage of data. Such systems have also been accepted provided they have been defined (with raw data identified) and validated.

Carefully examine and evaluate laboratory logs, worksheets and other records containing the raw data such as weighings, dilutions, the condition of

instruments, and calculations. Note whether raw data are missing, if records have been rewritten, or if correction fluid has been used to conceal errors. Results should not be changed without explanation. Cross reference the data that has been corrected to authenticate it. Products cannot be "tested into compliance" by arbitrarily labeling out-of-specification lab results as

"laboratory errors" without an investigation resulting in scientifically

valid criteria.

Test results should not have been transcribed without retention of the

original records, nor should test results be recorded selectively. For

example, investigations have uncovered the use of loose sheets of paper with subsequent selective transcriptions of good data to analyst worksheets

and/or workbooks. Absorbance values and calculations have even been found on desk calendars.

Cut charts with injections missing, deletion of files in direct data entry systems, indirect data entry without verification, and changes to

computerized programs to override program features should be carefully

examined. These practices raise questions about the overall quality of data.

The firm should have a written explanation when injections, particularly

from a series are missing from the official work-sheets or from files and

are included among the raw data. Multiple injections recorded should be in consecutive files with consecutive injection times recorded. Expect to see written justification for the deletion of all files.

Determine the adequacy of the firm's procedures to ensure that all valid

laboratory data are considered by the firm in their determination of

acceptability of components, in-process, finished product, and retained

stability samples. Laboratory logs and documents when cross referenced may show that data has been discarded by company officials who decided to

release the product without a satisfactory explanation of the results

showing the product fails to meet the specifications. Evaluate the

justification for disregarding test results that show the product failed to meet specifications.

14. LABORATORY STANDARD SOLUTIONS

Ascertain that suitable standards are being used (i.e. in-date, stored

properly). Check for the reuse of stock solutions without assuring their

stability. Stock solutions are frequently stored in the laboratory

refrigerator. Examine the laboratory refrigerators for these solutions and when found check for appropriate identification. Review records of standard solution preparation to assure complete and accurate documentation. It is

highly unlikely that a firm can "accurately and consistently weigh" to the same microgram. Therefore data showing this level of standardization or

pattern is suspect and should be carefully investigated.

15. METHODS VALIDATION

Information regarding the validation of methods should be carefully

evaluated for completeness, accuracy and reliability. In particular, if a

compendial method exists, but the firm chooses to use an alternate method

instead, they must compare the two and demonstrate that the in-house method is equivalent or superior to the official procedure. For compendial methods firms must demonstrate that the method works under the actual conditions of use.

Methods can be validated in a number of ways. Methods appearing in the USP are considered validated and they are considered validated if part of an

approved ANDA. Also a company can conduct a validation study on their

method. System suitability data alone is insufficient for and does not

constitute method validation.

In the review of method validation data, it is expected that data for

repetitive testing be consistent and that the varying concentrations of test solutions provide linear results. Many assay and impurity tests are now

HPLC, and it is expected that the precision of these assays be equal or less than the RSD's for system suitability testing. The analytical performance

parameters listed in the USP XXII, <1225>, under the heading of Validation of Compendial Methods, can be used as a guide for determining the analytical parameters (e.g., accuracy, precision, linearity, ruggedness, etc.) needed to validate the method.

16. EQUIPMENT

Laboratory equipment usage, maintenance, calibration logs, repair records, and maintenance SOPs also should be examined. The existence of the equipment specified in the analytical methods should be confirmed and its condition

noted. Verify that the equipment was present and in good working order at

the time the batches were analyzed. Determine whether equipment is being

used properly.

In addition, verify that the equipment in any application was in good

working order when it was listed as used to produce clinical or biobatches. One would have to suspect the data that are generated from a piece of

equipment that is known to be defective. Therefore, continuing to use and

release product on the basis of such equipment represents a serious

violation of CGMP's.

17. RAW MATERIAL TESTING

Some inspections include the coverage of the manufacturer of the drug

substance. The safety and efficacy of the finished dosage form is largely

dependent on the purity and quality of the bulk active drug substance.

Examine the raw data reflecting the analysis of the drug substance including purity tests, charts, etc.

Check the impurity profiles of the BPC used in the biobatch and clinical

production batches to determine if it is the same as that being used to

manufacture full scale production batches. Determine if the manufacturer has a program to audit the certificate of analysis of the BPC, and, if so, check the results of these tests. Report findings where there is substantial

difference in impurity profiles and other test results.

Some older compendial methods may not be capable of detecting impurities as necessary to enable the control of the manufacturing process, and newer

methods have been developed to test these products. Such methods must be

validated to ensure that they are adequate for analytical purposes in the

control and validation of the BPC manufacturing process. The drug substance manufacturer must have complete knowledge of the manufacturing process and the potential impurities that may appear in the drug substance. These

impurities cannot be evaluated without a suitable method and one that has

been validated.

Physical tests such as particle size for raw materials, adhesion tests for patches, and extrusion tests for syringes are essential tests to assure

consistent operation of the production and control system and to assure

quality and efficacy. Some of these tests are filed in applications and

others may be established by the protocols used to manufacture the product. The validation of methods for such tests are as important as the test for

chemical attributes.

Physical properties tests often require the use of unique equipment and

protocols. These tests may not be reproducible in other laboratories,

therefore, on site evaluation is essential.

18. IN PROCESS CONTROLS AND SPECIFICATIONS

Evaluate the test results from in-process tests performed in the production areas or laboratory for conformance with established sampling and testing

protocols, analytical methods, and specifications. For example, evaluate the tests for weight variation, hardness, and friability. These tests may be

performed every fifteen or thirty minutes during tableting or encapsulating procedures. All testing must comply with CGMP's.

The drug application may contain some of the in-process testing plan,

including methods and specifications. The inspection must confirm that the in-process tests were done, as described in the plan, and ascertain that the results were within specifications. The laboratory work for the lengthier

tests should also be reviewed.

The methods used for in-process testing may differ from those used for

release testings. Usually, whether the methods are the same or different,

the specifications may be tighter for the in-process tests. A product with a 90.0%-110.0% assay release specification may have a limit of 95.%-105.0% for the in-process blend. Some of the tests done may differ from those done at release. For example, a firm may perform disintegration testing as an

in-process test but dissolution testing as a release test.

Expect to see consistent in-process test results within batches and between batches of the same formulation/process (including development or exhibit

batches). If this is not the case, expect to see scientific data to justify the variation.

19. STABILITY

A stability-indicating method must be used to test the samples of the batch. If there is no stability-indicating assay additional assay procedures such as TLC should be used to supplement the general assay method. Evidence that the method is stability indicating must be presented, even for compendial

methods.

Manufacturers may be required to accelerate or force degradation of a

product to demonstrate that the test is stability indicating. In some cases the sponsor of ANDA's may be able to search the literature and find

background data for the specificity of a particular method. This information may also be obtained from the supplier of the drug substance. Validation

would then be relatively straightforward, with the typical parameters listed in the USP in chapter <1225> on validation of compendial methods addressed as applicable.

Evaluate the manufacturer's validation report for their stability testing. Again, review the raw laboratory data and the results of testing at the

various stations to determine if the data actually reported matches the data found in on site records.

Evaluate the raw data used to generate the data filed documenting that the method is stability indicating and the level of impurities.

20. COMPUTERIZED LABORATORY DATA ACQUISITION SYSTEMS

The use of computerized laboratory data acquisition systems is not new and is addressed in the following CGMP guidance documents:

o Compliance Policy Guide 7132a.07 Computerized Drug Processing:

Input/Output Checking.

o Compliance Policy Guide 7132a.08 Computerized Drug Processing:

Identification of "Persons" on Batch Production and Control Records.

o Compliance Policy Guide 7132a.11 Computerized Drug Processing: CGMP

Applicability to Hardware and Software

o Compliance Policy Guide 7132a.12 Computerized Drug Processing: Vendor

Responsibility

o Compliance Policy Guide 7132a.15 Computerized Drug Processing: Source Code for Process Control Application Programs

o Guide to Inspection of Computerized Systems in Drug Processing.

It is important, for computerized and non computerized systems, to define

the universe of data that will be collected, the procedures to collect it, and the means to verify its accuracy. Equally important are the procedure to audit data and programs and the process for correcting errors. Several

issues must be addressed when evaluating computerized laboratory systems.

These include data collection, processing, data integrity, and security.

Procedures should only be judged adequate when data are secure, raw data are not accidentally lost, and data cannot be tampered with. The system must

assure that raw data are stored and actually processed.

The agency has provided some basic guidance on security and authenticity

issues for computerized systems:

o Provision must be made so that only authorized individuals can make data entries.

o Data entries may not be deleted. Changes must be made in the form of

amendments.

o The data base must be made as tamperproof as possible.

o The Standard Operating Procedures must describe the procedures for

ensuring the validity of the data.

One basic aspect of validation of laboratory computerized data acquisition requires a comparison of data from the specific instrument with that same

data electronically transmitted through the system and emanating on a

printer. Periodic data comparisons would be sufficient only when such

comparisons have been made over a sufficient period of time to assure that the computerized system produces consistent and valid results.

21. LABORATORY MANAGEMENT

Overall management of the laboratory work, its staff, and the evaluation of the results of analysis are important elements in the evaluation of a

control laboratory. Span of supervisory control, personnel qualifications, turnover of analysts, and scope of the laboratory's responsibility are

important issues to examine when determining the quality of overall

management and supervision of work. Individually or collectively, these

factors are the basis for an objection only when they are shown to result in inadequate performance of responsibilities required by the CGMPs.

Review laboratory logs for the sequence of analysis and the sequence of

manufacturing dates. Examine laboratory records and logs for vital

information about the technical competence of the staff and the quality

control procedures used in the laboratory.

Observe analysts performing the operations described in the application.

There is no substitute for actually seeing the work performed and noting

whether good technique is used. You should not stand over the analysts, but watch from a distance and evaluate their actions.

Sometimes the company's employees have insufficient training or time to

recognize situations that require further investigation and explanation.

Instead they accept unexplained peaks in chromatograms with no effort to

identify them. They may accept stability test results showing an apparent

increase in the assay of the drug with the passage of time with no apparent question about the result. Also, diminishing reproducibility in HPLC

chromatograms appearing several hours after system suitability is

established is accepted without question.

Good manufacturing practice regulations require an active training program and the documented evaluation of the training of analysts.

The authority to delete files and override computer systems should be

thoroughly examined. Evaluate the history of changes to programs used for

calculations. Certain changes may require management to re-examine the data for products already released.

ENDNOTES

There are no references from this document.

实验室电子数据管理规程

一、目的: 建立化验室电子数据管理规程,实行电子数据管理标准化,确保检验结果准确无误,避免遗失和损坏,保障电子数据具有可追溯性。 二、范围: 适用于化验室电子设备(仪器)采集、打印与编制的数据管理。 三、责任者:系统管理员、检验人员、化验室负责人、QA。 四、内容: 1、定义 1.1电子数据:也称数据电文,是指以电子、光学、磁或者类似手段生成、发送、接收或者储存的信息。 1.2电子签名:是指电子数据中以电子形式所含、所附用于识别签名人身份并表明签名人认可其中内容的数据。 1.3数据审计跟踪:是一系列有关计算机操作系统、应用程序及用户操作等事件的记录,用以帮助从原始数据追踪到有关的记录、报告或事件,或从记录、报告、事件追溯到原始数据。 1.4数据完整性:是指数据的准确性和可靠性,用于描述存储的所有数据值均处于客观真实的状态。 2、授权管理 2、授权管理 2.1化验室的电子数据一般来源于电子系统,计算机操作系统作为工作站的运行平台应进行账户管理,应用于数据采集的计算机应设置管理员权限、标准用户权限、受限制用户权限 2.2工作站软件是数据采集和处理的工具,直接产生检测数据,因此需由品质部授权相应的

系统管理员、实验室管理员、分析员、QA完成数据采集、复核、修改、备份、恢复和管理;为确保电子数据的真实、有效,针对不同的人员,授权操作,设置相应的访问权限。 2.2.1系统管理员:一级管理员,经公司授权的人员,一般为IT人员,这些人员一般不超过两人,他们具有系统的所有访问权限。有权建立实验室管理员、QA账户、分析员、操作员帐户,如电脑系统时间、系统日志、操作员权限、增减登录帐户和初始登录帐户密码等,对所有使用的应用软件进行原名称安装、修复、备份和卸载,必须使用和验证应用软件为仪器供应商提供的正版软件,保留证书,重装和更换电脑、系统升级应进行风险评估,通过质量管理部门批准;对系统和应用软件采集的电子数据进行备份和恢复,对下级人员没有权限的内容设置成灰色不能使用,下级人员没有删除和修改、复制和粘贴、剪切和转移、备份和重命名权限。定期校验,并对重要操作进行复核和审核。 2.2.2化验室负责人(QC经理):二级管理员,需经品质部负责人和系统管理员授权。 2.2.3分析员:三级管理员,一般由分管各仪器组长或具备1年以上仪器使用经验的人员担任,有权登录系统,进入分析员帐户,在授权后可进行分析方法修改。 2.2.4操作员:四级管理员,为实验室仪器组QC人员,有权进行常规的仪器分析操作。 2.2.5仪器管理员:四级管理员,一般由分析员兼任,有权登录系统,进入仪器管理员帐户,使用应用软件程序,对仪器设备进行硬件配置,更改仪器网络位置。 2.2.6 QA:四级管理员,进入系统后,登入QA账户,可查看软件的审计追踪日志,查阅分析员分析所用的方法,所得到的数据,并对方法、数据进行复核,可进行文件备份。 2.2.7 系统管理员、实验室管理员、仪器管理员、QA、分析员和操作员如果离开本岗位或不

大学英汉对照课程翻译

大学英汉对照课程翻译 高等代数 Elementary Algebra 数学分析 Mathematical Analysis 中共党史 History of the Chinese Communist Party 算法语言 Algorithmic Language 体育 Physical Education 英语 English Language 力学实验 Mechanics-Practical 德育 Moral Education PASCAL语言 PASCAL Language 政治经济学 Political Economics 电学实验 Electrical Experiment 数字逻辑 Mathematical Logic 普通物理 General Physics 计算方法 Computing Method 离散数学 Discrete Mathematics

汇编原理 Principles of Assembly 概率与统计 Probability & Statistics 数据结构 Data Structure 哲学 Philosophy 微机原理 Principles of Microcomputer 编译方法 Compilation Method 系统结构 System Structure 操作系统原理 Principles of Operating System 文献检索 Documentation Retri 数据库概论 Introduction to Database 网络原理 Principles of Network 人工智能 Artificial Intelligence 算法分析 Algorithm Analysis 毕业论文 Graduation Thesis

公司理财(英文版)题库2

CHAPTER 2 Financial Statements & Cash Flow Multiple Choice Questions: I. DEFINITIONS BALANCE SHEET b 1. The financial statement showing a firm’s accounting value on a particular date is the: a. income statement. b. balance sheet. c. statement of cash flows. d. tax reconciliation statement. e. shareholders’ equity sheet. Difficulty level: Easy CURRENT ASSETS c 2. A current asset is: a. an item currently owned by the firm. b. an item that the firm expects to own within the next year. c. an item currently owned by the firm that will convert to cash within the next 12 months. d. the amount of cash on hand the firm currently shows on its balance sheet. e. the market value of all items currently owned by the firm. Difficulty level: Easy LONG-TERM DEBT b 3. The long-term debts of a firm are liabilities: a. that come due within the next 12 months. b. that do not come due for at least 12 months. c. owed to the firm’s suppliers. d. owed to the firm’s shareholders. e. the firm expects to incur within the next 12 months. Difficulty level: Easy NET WORKING CAPITAL e 4. Net working capital is defined as: a. total liabilities minus shareholders’ equity. b. current liabilities minus shareholders’ equity. c. fixed assets minus long-term liabilities. d. total assets minus total liabilities. e. current assets minus current liabilities. Difficulty level: Easy LIQUID ASSETS d 5. A(n) ____ asset is on e which can be quickly converted into cash without significant loss in value.

实验室记录控制培训试卷及答案

实验室记录控制培训试卷及答案 记录控制培训试卷 姓名 : 班组: 考核时间: 分数: 一、填空题(每空2分) 1 《检测和实验室能力认可准则》中4.13.2.2中规定:观察结果、数据和计算应在产生的当时予以记录,并能按照特定任务分类标识。 2《实验室记录控制程序》中规定:记录应实时记录,不允许事前预记或事后补记。记录应字迹清楚、端正、内容真实,数据完整。必须有书写人、复核人签名,必要时,要有审核人签名。 《检测和实验室能力认可准则》中4.13.2.3和《实验室记录控制程序》和《实验室记录控3 制程序》中规定:当记录中出现错误时,每一错误应划改,不可涂擦掉,以免字迹模糊或消失,并将正确值写在旁边。对记录的所有改动应有改动人的签名或签名缩写。对电子存储的数据也应采取等同措施,以避免原始数据的丢失或改动。 4《检测和实验室能力认可准则》中4.13.1.3中规定:所有记录应予以安全保护和保密。 5 实验室记录包括:质量记录和技术记录。质量记录包括:内部审核报告和管理评审报告以及纠正措施和预防措施的记录。 6 每项检测的记录应包含充分的信息,以便在在可能是时识别不确定因素,并确保该检测在尽可能接近原条件的情况下能够重复。 二、判断题(每题5分) 1 为保证记录的整洁完整,现场检测记录,可以在检测结束后统一整理后再填写( × ) 2 所有质量记录和技术记录均应归档并保存三年(×) 3 原始记录填写错了,复核人可以修改( × ) 4技术记录是进行检测或校准活动的记录,包括抽样、检测的原始记录、观察记录、人员培训记录、环境条件控制记录 ( ? ) 5 若

试验过程中人手少,未来得及记录。可以在试验结束后补记或追记。(×) 6 实验室无需规定记录的保存期限。( ×) 7 每次检测记录应包含足够的信息以保证能够再现。 ( ? ) 8 对电子存储的记录应该采取有效措施,避免原始信息或数据的丢失或改动 ( ? ) 下面是赠送的办公室6S管理资料,下载了不需要可以编辑删除的, 谢谢选择我们, 祝您工作顺利,生活愉快~ 办公室6S管理制度规定 1、彻底落实办公室个人责任区制度。办公室个人责任区是指个人的桌面、抽屉、电脑、文件架以及个人办公室的地面、桌洞、墙面、天花板、窗台以及相关办公设备和设施。每个人均有责任做好个人责任区的6S工作;落实“个人责任区负责”制度。 2、办公桌桌面除文件、电脑、口杯、电话、文具外,不允许放其他物品;文具必须竖放。办公用品一般的常用品:笔、订书机、即时帖、便条纸、橡皮、计算器,可以集中放在办公桌的一定区域内,电脑线、网线、电话线均需扎起来有序放置;人离开半小时以上应将桌面收拾干净。 3、个人抽屉(柜)管理:每星期整理一次,对三个月内用不到的东西,应从抽屉里清理走,最下面一个抽屉用于存放私人物品;抽屉内物品要分类好。第一层放个人的参考资料(文件夹、辞典、手册、商品目录等)及不常用文具;第二层放文件卷宗(各种信息、草拟的文件材料等);第三层放私人物品。同一方位(方向)按同一位置摆放。以上三层物品均需分类摆放整齐并贴上统一的标签。 4、文件整理:(1)文件不用均要放入文件夹,且要摆放整齐。文件夹太多而无法放下,应将文件夹整齐有序的排放在办公桌上。文件架(柜)要每半个月整理一次,对舍弃的文件、资料,应遵照安全、保密的原则进行销毁。必要文件送档案室

高等代数与解析几何教学大纲

附件1 高等代数与解析几何教学大纲 课程编号: 课程英文名:Advanced Algebra and Analytic Geometry 课程性质:学科基础课 课程类别:必修课 先修课程:高中数学 学分:4+4 总学时数:72+72 周学时数:4+4 适用专业:统计学 适用学生类别:内招生 开课单位:信息科学技术学院数学系 一、教学目标及教学要求 1.本课程是统计学专业的一门重要基础课。它不仅是学习后继课程及在各个学科领域进行理论研究和实际应用的必要基础,同时还为培养学生的独立工作能力提供必要的训练。学生学好这门课程的基本内容和方法,对今后的提高和发展有着深远的影响。 2.通过本课程的学习,要使学生了解高等代数与解析几何的概貌、各部分内容的结构和知识的内在联系;学会代数与几何方法,培养学生抽象思维能力、逻辑推理能力、想象能力、运算能力和综合应用能力。 3.要求学生熟练掌握本课程的基本概念、基本理论、基本运算及方法。通过课堂教学及进行大量的习题训练等各个教学环节,使得学生做到概念清晰、推理严密、运算准确,并且学会应用这些基本理论及方法去处理实际问题。 二、本课程的重点和难点 (略。由课任教师自行掌握) 三、主要实践性教学环节及要求

精讲、细读、自学相结合方法,加强课内外训练为手段。 四、教材与主要参考文献 教材:《高等代数与解析几何》(上、下)(第二版),孟道骥编著,科学出版社,2004年。 参考书: 1.《高等代数与解析几何》,陈志杰编著,高等教育出版社, 2000年; 2.《数论基础》,张君达主编,北京科学技术出版社,2002年。 五、考核形式与成绩计算 考核形式:闭卷考试。 成绩计算:平时成绩(包括平时作业、小测验、考勤等)占30%, 期末考试占70%。 六、基本教学内容 第二学期 第一周—第二周:(8课时) 第一章:向量代数与解析几何基础 1.代数与几何发展概述。 2. 向量的线性运算及几何意义:定义与性质、向量的共线、共面与线 性关系 3. 坐标系:标架、向量和点的坐标、n维向量空间。 4. 向量的线性关系与线性方程组。 5. 三维空间中向量的乘积运算:内积、外积、混合积、三重外积。 6. 方程及几何意义: (1)二元方程及几何意义:平面曲线的表示(非参数式、极坐标、 参数式、向量式); (2)三元方程及几何意义:直线与平面方程、曲线与曲面方程(非 参数式、参数式、向量式)。 第三周—第五周:(12课时)

实验室检测基础知识试题(含答案)

实验室检测基础知识试题(含答案) 1.光学金相试样制备要经过的步骤是:( )、镶嵌、( )、( )、( )。 取样磨光抛光显示 2.金相试样的取样必须具备( )性和( )性。 代表有针对 3.通常作为金相组织腐蚀剂的化学药品不外乎有四类,它们是( )和 ( )、( )、( )、( )。 各种有机酸无机酸各种碱各种盐类溶剂 4.金属中常见的晶格有三种类型,它们是( )、( )和( )。 体心立方晶格面心立方晶格密排立方晶格 5.铁素体的金相特征为( )的( )晶粒。 明亮多边形 6.奥氏体的金相特征为( )晶粒,晶界较铁素体( ),晶粒内常出现 ( )。 多边形平直孪晶 7.金属是具有金属光泽、( )和有良好( )、( )的物质。 可锻性导电性导热性 8.金属和合金的( )是不同的,这主要( )它们各自的结构和( )。性能取决于组织 9.金属在外力作用下,都会发生一定的变形,一般有两种形式,即( )变形和( )变形。 弹性塑性 10.热应力是金属在( )时( )所引起的应力。 加热内外温差 11.金属结晶主要受三个因素的影响:金属的( );( );金属结晶时的状态。 化学成份冷却速度 12.金属能够结晶,与液态金属的( )有密切的关系。 结构特点 13.纯金属是由( )元素组成的。 单一金属 14.合金则由( )或两种以上的金属、或金属与( )组成的具有金属特性的物质。 两种非金属 15.金属的原子在空间总是严格按照一定的( )而( )地排列。 规则周期 16.金属晶体是由原子通过( )结合而成的。 金属键 17.一般的合金钢在退火、正火状态下,具有( )+( )组织。 铁素体珠光体

高等代数与解析几何同济答案

高等代数与解析几何同济答案 【篇一:大学所有课程课后答案】 资料打开方法:按住 ctrl键,在你需要的资料上用鼠标左键单击 资料搜索方法:ctrl+f 输入关键词查找你要的资料 【数学】 o o o o o o o o o o o o o o o o o

习题答案 o o o o o o o o o o o o o o o o o o o o 【计算机/网络/信息】 o

o o o 【经济/金融/营销/管理/电子商务】 o o o o o o o o o o o o 【 o o o o o

【篇二:各门课程课后答案】 式]《会计学原理》同步练习题答案 [word格式]《成本会计》习题及答案(自学推荐,23页) [word格式]《成本会计》配套习题集参考答案 [word格式]《实用成本会计》习题答案 [word格式]《会计电算化》教材习题答案(09年) [jpg格式]会计从业《基础会计》课后答案 [word格式]《现代西方经济学(微观经济学)》笔记与课后习题详解(第3版,宋承先) [word格式]《宏观经济学》习题答案(第七版,多恩布什) [word格式]《国际贸易》课后习题答案(海闻 p.林德特王新奎) [pdf格式]《西方经济学》习题答案(第三版,高鸿业)可直接打印 [word格式]《金融工程》课后题答案(郑振龙版) [word格式]《宏观经济学》课后答案(布兰查德版) [jpg格式]《投资学》课后习题答案(英文版,牛逼版) [pdf格式]《投资学》课后习题答案(博迪,第四版) [word格式]《微观经济学》课后答案(高鸿业版) [word格式]《公司理财》课后答案(英文版,第六版) [word格式]《国际经济学》教师手册及课后习题答案(克鲁格曼,第六版)

实验室测试题和答案解析

1、[判断题] 实验室内出现火情,若被困在室内时,应迅速打开水龙头,将所有可盛水的容器装满水,并把毛巾打湿。用湿毛巾捂嘴,可以遮住部分浓烟不被吸入。(分值1.0) 你的答案:正确 2、[判断题] 测试数据应进行异地备份。(分值1.0) 你的答案:正确 3、[判断题] 接触化学危险品、剧毒以及致病微生物等的仪器设备和器皿,必须有明确醒目的标记。使用后及时清洁,特别是维修保养或移至到其他场地前,必须进行彻底的净化。(分值1.0) 你的答案:正确 4、[判断题] 早晨、中饭、晚上等几个人员稀少的时间段,要特别注意随手关门,确保实验室财产和个人物品的安全。(分值1.0) 你的答案:正确 5、[判断题] 使用过的实验服脱下后,不得与日常衣服放在一起,也不得放在洁净区域。(分值1.0) 你的答案:正确 6、[判断题] 针头、玻璃、一次性手术刀等利器应在使用后放在耐扎容器中,尖利物容器应在内容物达到三分之二前进行置换处置。(分值1.0) 你的答案:正确 7、[判断题] 正在进行实验时,可带着防护手套接听电话。(分值1.0) 你未作答标准答案:错误 8、[判断题] 要保持实验室环境整洁,做到地面、桌面、设备三整洁,减少安全隐患。(分值1.0) 你的答案:正确 9、[判断题] 触电紧急救护时,首先应使触电者脱离电源,然后立即进行人工呼吸、心脏按压。(分值1.0) 你的答案:正确 10、[判断题] 在触电现场,若触电者已经没有呼吸或脉搏,此时可以判定触电者已经死亡,可以放弃抢救。(分值1.0) 你的答案:正确标准答案:错误 11、[判断题] 当有人发生触电事故时,应马上直接将其拉开。(分值1.0) 你未作答标准答案:错误 12、[判断题] 实验室人员发生触电时,应迅速切断电源,将触电者上衣解开,取出口中异物,然后进性人工呼吸。当患者恢复呼吸时,应立即送医院。(分值1.0) 你未作答标准答案:正确 13、[判断题] 溅入口中已下咽的强碱,先饮用大量水,再服用乙酸果汁,鸡蛋清。(分值1.0) 你未作答标准答案:正确 14、[判断题] 实验中遇到一般割伤,应立即取出伤口内的异物,保持伤口干净,并用酒精棉清除伤口周围的污物,涂上外伤膏或消炎粉。(分值1.0) 你未作答标准答案:正确 15、[判断题] 火或热水等引起的大面积烧伤、烫伤,必须用湿毛巾、湿布、湿棉被覆盖,然后送医院进行处理。(分值1.0) 你未作答标准答案:正确 16、[判断题] 加强计算机的安全管理,建立病毒防护系统并不断加以更新,重要的数据资

实验室管理规程

1.目的 规范实验室工作管理,确保检测的科学合理进行,保证检测结果的准确可靠。 2.范围 适用于实验室的管理。 3.职责 3.1.质保部负责本规程的执行。 3.2.其他部门协作质保部执行本规程。 4.规程 4.1.人员 4.1.1.微生物实验室应设置质量负责人、技术管理者、检验人员、生物安全负责人、生物安全责任人、生物安全监督员、菌种管理员及相关设备和材料管理员岗,可以设置一人多岗设置。 4.1.2.从事药品微生物试验工作的人员应具备微生物学或相近专业知识的教育背景。 4.1.3.检验人员必须熟悉相关检测方法、程序、检测目的和结果评价。微生物实验室的管理者其专业技能和经验水平应与他们的职责范围相符,如:管理技能、实验室安全、试验安排、预算、实验研究、实验结果的评估和数据偏差的调查、技术报告书写等。 4.1.4.实验人员应依据所在岗位和职责接受相应的培训,在确认他们可以承担某一试验前,他们不能独立从事该项微生物试验。应保证所有人员在上岗前接受胜任工作所必需的设备操作、微生物检验技术等方面的培训,如无菌操作、培养基制备、消毒、灭菌、注平板、菌落计数、菌种的转种、传代和保藏、微生物检查方法和鉴定基本技术等,经考核合格后方可上岗。 4.1. 5.实验人员应经过实验室生物安全方面的培训,熟悉生物安全操作知识和消毒灭菌知识,保证自身安全,防止微生物在实验室内部污染。 4.1.6.实验室应制定所有级别实验人员的继续教育计划,保证知识与技能不断地更新。 4.1.7.实验室应确定人员具备承担实验室活动的能力,以及评估偏离影响程度的能力。可通过参加内部质量控制、能力验证或实验室间比对等方式客观评估检验人员的能力,并授权从事相应的实验室活动,必要时对其进行再培训并重新评估。当使用一种非经常使用的方法或

高等代数教学大纲

中国海洋大学本科生课程大纲 课程属性:学科基础 课程性质:必修 一、课程介绍 1.课程描述: 高等代数是数学科学学院各专业的重要专业必修基础课,是学习其它数学课程的主要先修课之一。高等代数的内容主要包含两个模块:第一模块,方程和方程组的求解问题,主要内容有:多项式、行列式、线性方程组、矩阵、二次型;第二模块,线性空间相关理论,主要内容有:线性空间、线性变换、λ-矩阵、欧几里得空间。高等代数内容包含理工科所开设的线性代数的主要内容。 2.设计思路: 开设高等代数课程的目的是:一方面,使数学院本科生在中学所学初等代数的基础上继续学习更加高深的代数学知识,使其掌握系统的经典代数内容,为学习其它数学课程(如数值代数、近世代数、计算方法等等)提供坚实的代数基础知识;另一方面,通过本课程的学习,逐步培养学生的数值计算能力、逻辑分析能力和抽象思维能力,提高学生在数学思想、数学方法方面的修养。 19世纪以前的代数研究内容主要是解方程和方程组以及由此产生的相关理论,称为经典代数;19世纪以后的代数主要研究一些抽象代数结构,如群、环、域、模等,称为抽象代数或近世代数。高等代数课程的内容主要是经典代数内容,涵盖学习其它数学课程所要求的基本的代数基础知识。 - 2 -

高等代数的内容基本按照经典代数的发展编排的,主要有两条主线:第一,方程和方程组求解问题,第二,线性空间相关理论。第一条主线的主要内容有:多项式理论——对应高次方程,其求解需要降次,需研究多项式的因式分解;行列式理论——求解线性方程组的主要工具之一;线性方程组理论——解的判定与求法;矩阵理论——解线性方程组时用到的矩阵运算与性质;二次型理论——二次齐次方程的化简与对称矩阵。第二条主线的主要内容多是解析几何中内容的推广,主要有:线性空间——几何空间的推广与抽象;线性变换——线性空间中点的运动的描述;λ-矩阵——证明线性变换的矩阵与其标准形相似;欧几里得空间——带有长度、夹角与距离等度量性质的线性空间,是几何空间的推广。 3.课程与其他课程的关系: 先修课程:无; 并行课程:数学分析、空间解析几何; 后置课程:近世代数。高等代数与近世代数内容恰好实现对接,完整体现了代数学的基本内容,联系密切。 二、课程目标 本课程目标是:一方面使学生系统地掌握经典代数的内容,包括多项式、线性方程组、矩阵、二次型、线性空间、线性变换、欧几里得空间等,为学习其它数学课程打下坚实的代数知识基础;另一方面,通过本课程的学习,培养学生的数值计算能力、逻辑分析能力和抽象思维能力,提高学生运用数学思想、数学方法分析问题、解决问题的能力。 到课程结束时,学生应达到以下几方面要求: (1)知识掌握良好。会判断多项式的可约性,能计算两多项式的最大公因式;会计算行列式;会判定线性方程组是否可解,掌握线性方程组解的结构;熟练掌握矩阵的各种运算;可将二次型化为标准形;掌握线性空间基底理论以及子空间的运算;会写线性变换的矩阵,会判定矩阵是否对角化、准对角化,并能求出其相应对角形与准 - 2 -

完整word版公司理财英文版题库8

CHAPTER 8 Making Capital Investment Decisions I. DEFINITIONS INCREMENTAL CASH FLOWS a 1. The changes in a firm's future cash flows that are a direct consequence of accepting a project are called _____ cash flows. a. incremental b. stand-alone c. after-tax d. net present value e. erosion Difficulty level: Easy EQUIVALENT ANNUAL COST e 2. The annual annuity stream o f payments with the same present value as a project's costs is called the project's _____ cost. a. incremental b. sunk c. opportunity d. erosion e. equivalent annual Difficulty level: Easy SUNK COSTS c 3. A cost that has already been paid, or the liability to pay has already been incurred, is a(n): a. salvage value expense. b. net working capital expense. c. sunk cost. d. opportunity cost. e. erosion cost. Difficulty level: Easy OPPORTUNITY COSTS d 4. Th e most valuable investment given up i f an alternative investment is chosen is a(n): a. salvage value expense. b. net working capital expense.

实验室检查试题及答案

实验室检查 1.简要病史患者女性,33岁,头痛、全身不适半个月 化验报告脑脊液检查:外观微混,蛋白1.0g/L,葡萄糖1.7mmol/L,氯化物81mmol/L,白细胞35×106/L,脑脊液放置12小时后形成薄膜 参考化验报告作出初步临床诊断结核性脑膜炎 2.简要病史患者7岁,急性发热、头痛3天 化验报告脑脊液检查:外观微混,蛋白1.7g/L,葡萄糖0.8mmol/L,氯化物95mmol/L,白细胞98×106/L,多形核细胞占80%,脑脊液放置2小时后出现凝块参考化验报告作出初步临床诊断化脓性脑膜炎 3.简要病史患者男性,28岁,低热,左胸痛半月 化验报告胸水检查:草黄色,比重1.020,蛋白定量35g/L,有核细胞计数1200×106/L,多形核细胞占80%。 参考化验报告作出初步临床诊断细菌感染或其它疾病引起的渗出液 4.简要病史患者女性,47岁,十年前曾患乙型肝炎,因体检发现脾大、腹水就诊化验报告腹水检查:外观淡黄透明,比重1.010,蛋白定量10g/L,有核细胞计数20×106/L,单个核细胞占70% 参考化验报告作出初步临床诊断肝硬化所致漏出液 5.简要病史患者女性,32岁,因“感冒”后一周出现全身浮肿、腰痛就诊化验报告尿常规检查:尿蛋白“+++”,尿糖“-”,红细胞3~5/高倍,白细胞3~5/高倍,24小时尿蛋白定量5g。血浆白蛋白22.6g/L 参考化验报告作出初步临床诊断肾病综合征 6.简要病史患者男性,45岁,因头晕、乏力、腰痛、浮肿1年余就诊。查BP160/100mmHg 化验报告尿常规检查:比重1.010,蛋白“+”,红细胞15~20/高倍,白细胞1~2/高倍,颗粒管型1~2/高倍,蜡样管型0~2/高倍 参考化验报告作出初步临床诊断慢性肾小球肾炎 7.简要病史患者男性,60岁,体检时尿常规检查发现异常 化验报告尿比重1.030,尿蛋白“-”,尿糖“++”。血浆生化检查:空腹血糖7.8mmol/L,餐后2小时血糖12.3mmol/L 参考化验报告作出初步临床诊断糖尿病 8.简要病史患者女性,16岁,因感冒就诊,尿常规发现异常 化验报告尿常规检查:比重 1.025,尿蛋白“-”,尿糖“++”。复查空腹时血糖 4.6mmol/L,尿糖“-”;餐后2小时血糖7.24mmol/L,尿糖“++” 参考化验报告作出初步临床诊断肾性糖尿 9.简要病史患者男性,56岁,因恶心、呕吐、嗜睡、呼吸深快就诊 化验报告尿常规检查:比重1.030,尿蛋白“±”,尿糖“++++”,酮体“+++”,红细胞5~7/高倍,白细胞2~3/高倍,颗粒管型1~3/高倍 参考化验报告作出初步临床诊断糖尿病酮症酸中毒 10.简要病史患者男性,68岁,患十二指肠溃疡,因进食后频繁呕吐、呼吸困难就诊化验报告血气和电解质检查:pH 7.55,PaCO2 57mmHg,PaO2 63.9 mmHg,HCO3ˉ52.6mmol/L,Na+ 141 mmol/L,K+ 2.5mmol/L,Clˉ72 mmol/L

51、电子数据管理规程

文件名称中心化验室电子数据管理规程 编码版本AP-SMP-QC-T-051-01 第1版 制定部门质量监督管理部 审核人批准人 制定日期年月日 起草人审核日期年月日批准日期年月日起草日期年月日颁发部门质量监督管理部生效日期年月日分发部门质量监督管理部、中心化验室颁发2份 中心化验室电子数据管理规程 一、目的:建立中心化验室电子数据管理规程,实行电子数据管理标准化,确保检验结果准确无误,避免遗失和损坏,保障电子数据具有可追溯性。 二、范围:适用于中心化验室电子设备(仪器)采集、打印与编制的数据管理。 三、责任者:系统管理员(QA)、中心化验室检验人员(QC)、中心化验室负责人。 四、内容: 1、定义 1.1电子数据:也称数据电文,是指以电子、光学、磁或者类似手段生成、发送、接收或者储存的信息。 1.2电子签名:是指电子数据中以电子形式所含、所附用于识别签名人身份并表明签名人认可其中内容的数据。 1.3数据审计跟踪:是一系列有关计算机操作系统、应用程序及用户操作等事件的记录,用以帮助从原始数据追踪到有关的记录、报告或事件,或从记录、报告、事件追溯到原始数据。 1.4数据完整性:是指数据的准确性和可靠性,用于描述存储的所有数据值均处于客观真实的状态。 2、授权管理 计算机分为系统软件和应用软件,对各软件的操作分别授权,可以使用的人员由质量管理部设置。 2.1中心化验室的电子数据一般来源于电子系统,数据采集、修改、备份、恢复和管理均由质量监督管理部授权的系统管理员(由QA担任)、化验室主任和操作员完成;为确保电子数据的真实、有效,针对不同的人员,授权操作,设置相应的访问权限:

《解析几何与高等代数(I)》课程教学大纲

《解析几何与高等代数(I)》课程教学大纲 一、课程基本信息 1、课程代码:MA113 2、课程名称(中/英文): 解析几何与高等代数(I)/Analytic Geometry and Higher Algebra (I) 3、学时/学分:144/8 4、先修课程:无 5、面向对象:数学系数学专业一年级本科生 6、开课院(系)、教研室:数学系代数教研室 7、教材、教学参考书: 《高等代数与解析几何》上册,孟道骥编著,科学出版社,2004年第二版 二、课程性质和任务 本课程是数学系最根本的两门基础课之一(另一门是数学分析),又是数学系新生一入学就要面对的,因此是从中学数学走向大学数学的第一道门槛,跨不过去就无法继续前行。所以本课程从大的方面说主要是在学习方法和数学思想上初步完成从中学数学走向大学数学的适应与过渡;从学习内容来说,则主要是一些基础性知识,包括线性代数、多项式、空间解析几何。这些知识介于中学代数与近世代数之间,既有承上启下作用又有广泛应用性。 三、教学内容和基本要求 0.1 引言(关于内容、教与学的说明) 0.2 预备知识(连加号、连乘号;数学推理与数学归纳法)两次共6学时

1.行列式(24学时) 1.1 数域2学时 1.2 矩阵及其初等变换4学时 1.3 行列式归纳定义2学时 1.4 行列式的性质 1.5 行列式的展开式两次共6学时 1.6 Cramer法则2学时 1.7 行列式算例4学时 2.矩阵(14学时) 2.1矩阵的运算2学时 2.2可逆矩阵2学时 2.3矩阵的分块2学时 2.4矩阵的初等变换与初等方阵2学时 2.5矩阵与线性方程组2学时 2.6矩阵的性质举例(包括Bimet-Cauchy公式)4学时 3.线性空间(26学时) 3.1 空间向量(包括坐标系、平面与直线方程)4学时 3.2 数域上线性空间的概念2学时 3.3 线性相关 3.4 秩、维数与基两次共6学时 3.5 矩阵的秩与矩阵的相抵4学时 3.6 线性方程组理论2学时

英文版罗斯公司理财习题答案

CHAPTER 8 MAKING CAPITAL INVESTMENT DECISIONS Answers to Concepts Review and Critical Thinking Questions 1. In this context, an opportunity cost refers to the value of an asset or other input that will be used in a project. The relevant cost is what the asset or input is actually worth today, not, for example, what it cost to acquire. 2. a.Yes, the reduction in the sales of the company’s other products, referred to as erosion, and should be treated as an incremental cash flow. These lost sales are included because they are a cost (a revenue reduction) that the firm must bear if it chooses to produce the new product. b. Yes, expenditures on plant and equipment should be treated as incremental cash flows. These are costs of the new product line. However, if these expenditures have already occurred, they are sunk costs and are not included as incremental cash flows. c. No, the research and development costs should not be treated as incremental cash flows. The costs of research and development undertaken on the product during the past 3 years are sunk costs and should not be included in the evaluation of the project. Decisions made and costs incurred in the past cannot be changed. They should not affect the decision to accept or reject the project. d. Yes, the annual depreciation expense should be treated as an incremental cash flow. Depreciation expense must be taken into account when calculating the cash flows related to a given project. While depreciation is not a cash expense that directly affects c ash flow, it decreases a firm’s net

实验室认可资质认定内审员培训试题 含答案

实验室认可/资质认定内审员培训试题 一、填空题 1、国际标准化(ISO)和国际电工委员会(IEC)于2005年共同发布了有关实验室能 力认可第二版国际标准,文件名称是 ,代号是 2、中国合格评定国际认可委员会等同采用上述国际标准作为我国实验室认可的依 据,文件名称是,代号是 3、实验室资质认定评审的依据文件名称是 4、实验室或实验室作为其一部分的组织须是能承担的实体。 5、如果实验室要作为第三方实验室获得认可,应能证明其。 6、实验室应有政策和程序确保客户的和得到保护。 7、术语“管理体系”一词是指控制实验室运作的、和体系。 8、实验室应建立与其工作范围的管理体系,并维持管理体系 的。 9、凡发给实验室人员使用的管理体系文件,都应是版本。 10、在合同中对包括所用方法在内的要求应予,形成文件,并易于理解。实 验室有和满足客户要求。 11、实验室需要将工作分包时,应分包给的分包方,将分包方安排以形 式通知客户,并在适当时得到客户的准许。 12、实验室应确保所购买的影响检测和校准质量的供应品,在经检查证明有关 要求之后才投入使用。应保存所采取的符合性检查活动的。 13、实验室应积极与客户或其代表,并应收集客户的。 14、实验室须有处理投诉的。须保存所有与投诉处理有关的。 15、实验室当发现不符合工作时应进行纠正。 16、实验室应通过实施和、应用、、、和以 及来持续改进管理体系的有效性。 17、如果不合格可能再次发生,实验室须按规定的政策和程序采取,此程序须 从调查确定问题的入手。 18、预防是事先主动确定主动行为,消除可能发生不合格的原因。 19、每份记录应包含信息,观察结果、数据和计算须在工作予以记 录,出现错误应该,不许,以免难以辨认。 20、实验室的人员培训计划应与实验室和的任务相适应。实验室应 评价这些培训活动的。 21、实验室的设施须有利于检测或校准的正确实施,环境条件不会使无效或 对要求的产生不良影响。当环境条件危及到结果时,应检测或校准。 22、实验室应采用能满足需要并且的检测或校准方法。应优先使用

实验室计算机化系统管理规程

1.目的:规范我公司实验室计算机化系统管理,确保其准确、稳定、可靠运行。 2.范围:适用于实验室计算机化系统 3. 责任:QC负责对本规程的编写、修订、审核、培训和执行;QA负责对本规程的执 行进行监督;质量部负责人负责对本规程的批准。 4. 内容: 计算机化系统定义:计算机化系统由一系列硬件和软件组成,利用一系列指令(软件)执行自动功能的设备(硬件)。 实验室计算机化系统分类 a 没有测量能力或没有常规校验要求的检验设备;没有用户界面,不产生原始数据及测试结果。如氢气发生器、冰箱、磁力搅拌器、微波消解仪、离心机、菌落计数器等。 b 提供测量值的标准仪器和设备;仪器可通过需要校准的物理参数(如温度、压力、或流量)进行控制;仪器提供的功能和操作限制与用户要求一致;没有用户界面,产生原始数据及测试结果但不存储及处理。如硬度仪、崩解仪、折光仪、旋光仪、电子天平、滴定仪、恒温恒湿箱、PH计等 c 用户对于功能、运行和操作的限度要求与实际的分析应用相关;配置参数可存储并可再使用;没有用户界面,产生原始数据及测试结果但不存储及处理,如键盘控制仪器。 d 仪器结构复杂,包括仪器和与之相关的计算机化系统;有单一用户界面,配置参数可存储及再使用,但软件不能配置;产生原始数据及测试结果及处理。如HPLC、GC、UV、IR、AAS等。 实验室计算机化系统管理原则 计算机化系统代替人工操作时,不会对产品的质量、过程控制和其质量保证水平造成负面影响,不增加总体风险。 风险管理应当贯穿计算机化系统生命周期全过程,应当根据书面的风险评估结果确定验证和数据完整性控制的程度。 制定实验室计算机化系统供应商管理规程,供应商提供产品和服务时,企业应当与供应商签订正式协议,明确双方责任。 根据风险评估结果提供与供应商质量体系和审计相关的文件。 计算机化系统权限管理 实验室计算机化系统实行三级管理权限:管理员权限(IT管理员)、标准用户权限(实验室负责人、QA经理)、受限制用户权限(仪器分析员)

健康评估实验室检查习题和答案

一、单选题 1.空腹采血一般是指采血的时间在空腹(B) A.6小时以后B.8小时以后C.10小时以后D.12小时以后 2.正常男性的血红蛋白含量是(C) A、110-150g/L B、120-160g/L C、131-172g/L D、170-200g/L E、90-110g/L 3.支气管哮喘、荨麻疹、血清病等变态反应性疾病者,外周血(A) A.嗜酸性粒细胞增多B.中性粒细胞增多C.单核细胞增多D.淋巴细胞增多 4.网织红细胞减少主要见于(E) A、缺铁性贫血 B、溶血性贫血 C、失血性贫血 D、巨幼细胞性贫血 E、再生障碍性贫血 5.下列哪种疾病可引起中性粒细胞减少(C) A、尿毒症 B、慢性粒细胞性白血病 C、脾功能亢进 D、肺吸虫病 E、骨髓纤维化症 6.镜下血尿指高倍视野(Hp)红细胞数为(C) A.0-1个B.>2个C.≥3个D.>5个 7.多尿是指24h尿量大于(D) A、1000ml B、1500ml C、2000ml D、2500ml 8.无尿是指24h尿量少于(A) A、100ml B、200ml C、50ml D、250ml E、400ml 9.蛋白尿是指尿蛋白测定为(E) A、0~40mg/24h尿 B、0~80mg/24h尿 C、20~80mg/24h尿 D、100mg/24h尿 E、>150mg/24h尿 10.急性肾小球肾炎患者尿中可出现(A) A、颗粒管型 B、红细胞管型 C、肾衰竭管型 D、脂肪管型 E、蜡样管型 11.脑脊液标本采集后应立即送检,一般不超过(A) A、1小时 B、1.5小时 C、2小时 D、2.5小时 12.正常人尿中可出现(C) A、细胞管型 B、颗粒管型 C、透明惯性 D、蜡样管型 E、脂肪管型

相关文档