5. Technical Safety Inspection and Preventive Maintenance 4/20025

贝朗血透机维修手册1

-1MT-MD-DE08C M.KAY DialogAd_sm_Chapter 5_4-2002.doc/pdf <280602> ddmmyy D i a l o g /D i a l o g A d v a n c e d B. Braun Medizintechnologie GmbH Copyright

This technical documentation contains confidential information of B. Braun Medizintechnologie GmbH and may not be copied, partly or whole, nor made accessible to third parties without our written consent and remains with all rights the property of B. Braun Medizintechnologie GmbH.Service

Only competent individuals that are trained in the repair,maintenance and software installation of the Dialog/Dialog Advanced should attempt to service the machine!The servicing of the Dialog/Dialog Advanced in accordance with this service manual should never be carried out in the absence of proper tools, calibration equipment and the most recent edition of this service manual, which must be clearly and thoroughly understood!Prevent Electrical Shock Hazard

Switch off the Dialog/Dialog Advanced and disconnect unit from mains before opening the machine!Do not touch any exposed wiring or conductive surfaces while the Dialog/Dialog Advanced is open. The voltages present when electrical power is connected to the Dialog/Dialog Advanced can cause serious injury or death!If necessary open the machine for maintenance and repair ESD Recommendations

Pay attention to ESD requirements, because electronic components are sensitive to electrostatic discharges (see appendix)!Fuses

If fuses are replaced they must exactly match the type and rating specified by the manufacturer. Where applicable: Fuses must be approved by UL/CSA!Calibration

Only perform a calibration after the Dialog/Dialog Advanced has reached working temperature!