e.bloxx A6-2CF Instruction Manual

e.bloxx A6-2CF

HB_EBLOXX-A62CF_E_V19.doc

1

Gantner Instruments Test & Measurement GmbH

? Copyright 2004 by GANTNER INSTRUMENTS Test & Measurement GMBH, Schruns (Austria).

Copyrights: Operating instructions, manuals and software are protected by copyright ?. All rights are reserved.

Copying, duplication, translation, installation in any electronic medium or machine-readable form in whole or in part is prohibited. The sole exception is represented by creation of a back-up copy of software for own use as a safeguard, so far as this is technically possible and recommended by us. Any infringement will render the party committing such infringement liable to compensation payment.

Liability: Any claims against the manufacturer based on the hardware or software products described in this manual shall depend exclusively on the conditions of the guarantee. Any further-reaching claims are excluded, and in particular the manufacturer accepts no liability for the completeness or accuracy of the contents of this manual. The right is reserved to make alterations, and alterations may be made at any time without prior notice being given.

Trade marks: Attention is drawn at this point to markings and registered trade marks used in this manual, in particular to those of Microsoft Corporation, International Business Machines Corporation and Intel Corporation.

Important: before commencing installation, commissioning, putting into service and before any

maintenance work is carried out, it is essential that the relevant warning and safety instructions in this manual are read!

!

2HB_EBLOXX-A62CF_E_V19.doc

Gantner Instruments Test & Measurement GmbH

e.bloxx A6-2CF

HB_EBLOXX-A62CF_E_V19.doc

3

Gantner Instruments Test & Measurement GmbH

General warning and safety instructions:

Dear customer, We congratulate you on having selected a product of Gantner Instruments Test & Measurement GmbH. So that our

product functions in your installation with safety and to your complete satisfaction, we take this opportunity to familiarize you with the following ground rules:

1. Installation, commissioning, operation and maintenance of the product purchased must be carried out in accordance with instructions, i.e. in accordance with the technical conditions of operation, as described in the corresponding product documentation.

2. Before installation, commissioning, operation or maintenance it is therefore essential that you read the corresponding chapter of this manual and observe its instructions.

3. If there are still some points on which you are not entirely clear, please do not take a chance, but ask the customer adviser responsible for you, or ring the Gantner Instruments Test & Measurement GmbH hot line.

4. Where not otherwise specifically laid down, appropriate installation, commissioning, operation and maintenance of the appliance is the customer’s responsibility.

5. Directly on receipt of the goods, inspect both the packaging and the appliance itself for any signs of damage. Also check that the delivery is complete (-> accessories, documentation, auxiliary devices, etc.).

6. If the packaging has been damaged in transport or should you suspect that it has been damaged or that it may have a fault, the appliance must not be put into service. In this case, contact your customer advisor. He will make every effort to resolve the problem as quickly as possible.

7. Installation, commissioning and servicing of our appliances must only be carried out by suitably trained personnel. In particular, correspondingly qualified specialists may only make electrical connections. Here, the appropriate installation provisions in accordance with the relative national Electrical Engineers construction regulations (e.g. ?VE, [Austrian] VDE, [German]...) must be observed.

8. Where not otherwise stated, installation and maintenance work on our appliances is exclusively to be carried out when disconnected from the power supply. This applies in particular to appliances which are normally supplied by low-tension current.

9. It is prohibited to make alterations to the appliances or to remove protective shields and covers.

10. Do not attempt yourself to repair an appliance after a defect, failure or damage, or to put it back into operation again. In such cases, it is essential you contact either your customer adviser or the Gantner Instruments Test & Measurement GmbH hot line. We will make every effort to resolve the problem as quickly as possible.

11. Gantner Instruments Test & Measurement GmbH accepts no responsibility for any injuries or damage caused as a result of improper use.

!

e.bloxx A6-2CF

4

HB_EBLOXX-A62CF_E_V19.doc

Gantner Instruments Test & Measurement GmbH

12. Although every care is taken and we are continuously aiming for improvement, we cannot exclude completely the possibility of errors appearing in our documentation. Gantner Instruments Test & Measurement GmbH therefore accepts no responsibility for the completeness or the accuracy of this manual. The right is reserved to make alterations, and we may carry out alterations at any time without giving prior notice.

13. Should you discover any fault with the product or in its accompanying documentation, or have any suggestions for improvement, you may confidently approach either your customer adviser or Gantner Instruments Test & Measurement GmbH directly.

14. However, even if you just want to tell us that everything has functioned perfectly, we still look forward to hearing from you.

We wish you a successful application of our appliances. We will be pleased to welcome you as a customer again soon.

Contact address / manufacturer:

Gantner Instruments Test & Measurement GmbH Montafonerstrasse 8 A - 6780 Schruns/Austria Tel.: +43 5556 73784 - 410 Fax: +43 5556 73784 - 419

E-Mail: office@https://www.wendangku.net/doc/8a17709316.html, Web: https://www.wendangku.net/doc/8a17709316.html,

Gantner Instruments Test & Measurement GmbH Industriestra?e 12 D-64297 Darmstadt Tel.: +49 6151 95136 - 0 Fax: +49 6151 95136 - 26

E-Mail: testing@https://www.wendangku.net/doc/8a17709316.html, Web: https://www.wendangku.net/doc/8a17709316.html,

e.bloxx A6-2CF TABLE OF CONTENTS

HB_EBLOXX-A62CF_E_V19.doc

5

Gantner Instruments Test & Measurement GmbH

TABLE OF CONTENTS

1. ABOUT THIS MANUAL ....................................................................................................................................7

2. M ODULE DESCRIPTION ..................................................................................................................................8 2.1. System Overview .................................................................................................................................................8 2.2. Types of Modules .................................................................................................................................................9 2.

3. Module Parts ......................................................................................................................................................10 2.

4. Functional Overview ...........................................................................................................................................11 2.

5. Front-LED ..........................................................................................................................................................11 2.

6. DC-Isolation .......................................................................................................................................................11 2.

7. Functional Diagram ............................................................................................................................................12 3. MOUNTING E.BLOXX A6-2CF AND CONNECTING WIRES ..............................................................................13 3.1. Environmental Conditions ....................................................................................................................................13 3.2. Connection Technique ........................................................................................................................................13 3.3. Power Supply .....................................................................................................................................................13 3.4. Bus Connection ..................................................................................................................................................14 3.5. Shielding ............................................................................................................................................................19 3.6. Rapid Bus Link Plug ...........................................................................................................................................19 4. MEASUREMENTS .........................................................................................................................................21 4.1. General .............................................................................................................................................................21 4.2. Analog Input .......................................................................................................................................................24 4.2.1. Strain Gauge and Inductive Full Bridge (4-Lead Connection)..................................................................................25 4.2.2. Strain Gauge and Inductive Full Bridge, Piezo-Resistive Transducer (6-Lead Connection)..........................................25 4.2.3. Strain Gauge and Inductive Half Bridge (3-Lead Connection)..................................................................................26 4.2.4 Strain Gauge and Inductive Half Bridge (5-Lead Connection)..................................................................................26 4.2.5. Single strain gauge with external completion resistor (3-Lead Connection)..............................................................27 4.2.6 Single strain gauge with external completion resistor (5-Lead Connection)..............................................................27 4.2.7. Strain Gauge Quarter Bridge (3-Lead Connection).................................................................................................27 4.2.

8. LVDT Transducers (Linear Variable Differential Transducer)...................................................................................28 4.2.

9. Potentiometer Transducers..................................................................................................................................29 4.3. Analog Output ....................................................................................................................................................29 4.3.1. Analog Voltage Output........................................................................................................................................29 4.4. Digital Input ........................................................................................................................................................30 4.4.1. Status................................................................................................................................................................30 4.4.2. Frequency Measurement.....................................................................................................................................30 4.4.3. Counter..............................................................................................................................................................31 4.4.4. Up/down counter................................................................................................................................................32 4.4.5. Quadrature Counter............................................................................................................................................33 4.5. Digital Output .....................................................................................................................................................34 4.5.1. Digital Status Output - Host-Controlled.................................................................................................................34 4.5.2. Digital Status Output - Process-Controlled.............................................................................................................34 5. CONFIGURATION ..........................................................................................................................................36 5.1. General Information about Configuration Software ICP 100....................................................................................36 5.2. Setting Address and Baud Rate of an e.bloxx A6-2CF ...........................................................................................36 5.3. Module Settings .................................................................................................................................................37 5.4. Definition of Channels .........................................................................................................................................38 5.4.1

Sensor Adaptation, Calibration (39)

e.bloxx A6-2CF TABLE OF CONTENTS

6

HB_EBLOXX-A62CF_E_V19.doc

Gantner Instruments Test & Measurement GmbH

5.4.2 Signal Conditioning..............................................................................................................................................51 5.4.3 Arithmetic...........................................................................................................................................................58 5.4.4 Analog Output.....................................................................................................................................................63 5.4.5 Alarm.................................................................................................................................................................65 5.4.5 Digital Output......................................................................................................................................................66 5.4.7 Digital Input........................................................................................................................................................68 6. SYNCHRONIZATION OF THE E.BLOXX A6-2CF ..............................................................................................69 7. SPECIFICATIONS ..........................................................................................................................................70 7.1. Analog Inputs ......................................................................................................................................................70 7.2. Signal Conditioning .............................................................................................................................................70 7.3. Analog Outputs ...................................................................................................................................................71 7.4. Digital Inputs .......................................................................................................................................................72 7.5. Digital Outputs ....................................................................................................................................................72 7.6. Communication Interface .....................................................................................................................................72 7.7. Power Supply .....................................................................................................................................................73 7.8. Mechanical .........................................................................................................................................................73 7.9. Connection .........................................................................................................................................................73 7.10. Environmental Conditions ....................................................................................................................................73 8. DECLERATION OF CONFORMITY (74)

e.bloxx A6-2CF ABOUT THIS MANUAL

HB_EBLOXX-A62CF_E_V19.doc

7

Gantner Instruments Test & Measurement GmbH

1. ABOUT THIS MANUAL

This manual describes the installation and set-up of the e.bloxx A6-2CF.

The following information can be found in this manual:

- Description of the e.bloxx system with detailed information on the hardware and module features.

- Installation description of the e.bloxx A6-2CF and how it is connected to the power supply and bus lines. - Description of the different types of measurement.

- A short introduction on how the e.bloxx A6-2CF is configured with the CONFIGURATION SOFTWARE ICP 100. This software has an integrated help that includes detailed description on the configuration process. - Possible errors and solutions.

- Technical specifications of the e.bloxx A6-2CF.

e.bloxx A6-2CF MODULE DESCRIPTION

8

HB_EBLOXX-A62CF_E_V19.doc

Gantner Instruments Test & Measurement GmbH

2. M ODULE DESCRIPTION

2.1. System Overview

The e.bloxx modules have been developed for the industrial and experimental testing technology, especially for the multi-channel measurement of electrical signals of thermal or mechanical quantities at test beds and test sites.

Picture 2.1 - e.bloxx A6-2CF

The e.bloxx A6-2CF, which is described in this manual, is a modular measuring system for the automation of industrial

test stands as well as for monitoring of assembly and production processes. This module is designed for the measurement of quantities such as strain, force, displacement, torque, weight and pressure simple and safe. The module is part of a whole product line of different e.bloxx.

Due to the modular design, flexible definition of solutions and the standardized interface, the module provides maximum freedom, thus leaving enough space for further testing concepts.

Standardized interfaces guarantee the integration of up to 127 modules into a single network.

With the e.gate module very high data rates via Profibus-DP and Ethernet can be realized. The customer-specific signal processing supplements the standard conditioning of the single e.bloxx modules.

e.bloxx A6-2CF

MODULE DESCRIPTION

HB_EBLOXX-A62CF_E_V19.doc

9

Gantner Instruments Test & Measurement GmbH

2.2. Types of Modules

There are several types of e.bloxx that differ in their number and type of analog and digital inputs and outputs.

Table 2.1. - Characteristics of the e.bloxx modules

e.bloxx A6-2CF MODULE DESCRIPTION

10

HB_EBLOXX-A62CF_E_V19.doc

Gantner Instruments Test & Measurement GmbH

2.3. Module Parts

1 …Pluggable Screw-Type Terminal Strip for Connection of RS 485 Bus, Power Supply,

Synchronization and Grounding 2 … Power/Error-LED (red/green)

3 … Pluggable Screw-Type Terminal Strips for Sensor Connection and In-/Outputs

4 … Rapid Bus Link Plugs

Picture 2.2. - Parts of the e-bloxx A6-2CF

Terminal Strip for RS 485, Power Supply, Synchronization and Grounding

Table 2.2.

- Description of Terminal Strip for RS 485 Bus and Power Supply

Terminal Strip for Sensor Connection

Table 2.3. - Description of Terminal Strips for Sensors and In-/Outputs

Terminal

Description

UOut1 Voltage Output 1 GND Ground Out2 Voltage Output 2 GND Ground UExc+ Bridge Excitation Voltage (+) UExc- Bridge Excitation Voltage (-) USen+ Sense Leads (+) USen- Sense Leads (-) USig+ Measuring Signal (+) USig- Measuring Signal (-) DO 1 .. 4 Digital Output 1 up to 4 DI 1 .. 6 Digital Input 1 up to 6

1

2 3

4

e.bloxx A6-2CF MODULE DESCRIPTION

HB_EBLOXX-A62CF_E_V19.doc

11

Gantner Instruments Test & Measurement GmbH

2.4. Functional Overview

This manual describes the e.bloxx module A6-2CF, which is a 16-channel (real plus virtual variables) module.

The first channel always defines the analog input which is a bridge sensor input for 3-, 4-, 5- or 6-wire bridges, each one with 1V, 2.5V or 5V supply. The other channels can be used for digital I/O’s, to process the digital input which can be used for example as a trigger, to make arithmetical calculations or to process the measured value (scaling, run/hold, minimum/maximum envelope curve). The channels are defined in the configuration table of the Configuration Software ICP 100.

The values of each channel can be read out via the RS 485 bus.

2.5. Front-LED

The LEDs at the front of the e.bloxx module provides the following information:

Right LED LED green Module works well, no signal overflow, no communication error… LED red general error like signal overflow, broken sense leads LED red + short off period general error like signal overflow, broken sense leads + communication timeout LED green + short red flash Signal ok + communication timeout LED red fast flashing global error, no suitable firmware Left LED (DSP function) LED green No DSP failure LED red DSP failure Notice: The right LED will get red when the signal leaves the selected range and the error checking is activated

(see ICP 100 column Range/Error).

2.6. DC-Isolation

The following groups are galvanic isolated from each other: - power supply, digital inputs and digital outputs - bus interface - analog circuit

e.bloxx A6-2CF MODULE DESCRIPTION

12

HB_EBLOXX-A62CF_E_V19.doc

Gantner Instruments Test & Measurement GmbH

2.7. Functional Diagram

The following picture can describe the e.bloxx A6-2CF.

Picture 2.3. - Functional Diagram of the e.bloxx A6-2CF

e.bloxx A6-2CF

MOUNTING E.BLOXX AND CONNECTING WIRES

HB_EBLOXX-A62CF_E_V19.doc

13

Gantner Instruments Test & Measurement GmbH

3. MOUNTING e.bloxx A6-2CF AND CONNECTING WIRES

3.1. Environmental Conditions

The e.bloxx A6-2CF is protected against water and dirt according IP 20. If required by the conditions of the operating site the modules have to be installed accordingly, e.g. in a water-resistant or waterproof case, compliant with the regulations of electrical engineering.

For the allowed ambient temperatures for the e.bloxx A6-2CF please see the technical specifications at the end of this manual.

3.2. Connection Technique

The wires are connected to the modules via screw-type terminals. The captive terminal screws are part of the terminal strips. All terminal strips are of plug-in type and can be detached from the modules.

Not more than 2 leads should be connected with one clamp. In this case both leads should have the same conductor cross section. For the precise clamping of stranded wire we recommend the use of wire-end ferrules.

Notice: Connecting wires respectively the plugging-in and –out of the terminal strip is only allowed with modules in

power-off status.

In order to prevent interference with sensors, signals and modules, shielded cables have to be used for the power supply, bus connection and signal lines.

We strongly recommend using a single screened cable each input signal. To use more signals in one cable could generate interacting influences.

Notice: For optimal performance the e.bloxx modules must be grounded properly. This is achieved by utilizing the

Ground/Earth screw on the back of each e.bloxx module. The screen of the sensor cable has to be grounded at the same potential.

3.3. Power Supply

Non-regulated DC voltage between +10 and +30 VDC is sufficient for the power supply of the modules. The input is protected against excess voltage and polarity connecting error. The power consumption remains approximately constant over the total voltage range, due to the integrated switching regulator.

Due to their low current consumption (max. 180 mA at 24 VDC) the modules can also be remotely supplied via longer lines. Several modules can be supplied in parallel within the permissible voltage range and drop in the lines. If required, the supply lines together with the bus line may be incorporated in one cable.

In order not to overload the module power supply needlessly and to avoid unnecessary line troubles, a separate power supply is recommended for sensors with a large current drain.

The distribution voltage for the e.bloxx A6-2CF has to be protected by a fuse with maximum 1 A (inert). The modules have an internal fuse (reversible) for protection against excess voltage, excess current and wrong polarity.

e.bloxx A6-2CF

MOUNTING E.BLOXX AND CONNECTING WIRES

14

HB_EBLOXX-A62CF_E_V19.doc

Gantner Instruments Test & Measurement GmbH

Notice: It depends on the power supply unit and its noise and internal grounding issues whether it is helpful to

connect earth of the power supply unit with ground/earth of the e.bloxx module.

Picture 3.1. - Power Supply of the e.bloxx A6-2CF

3.4. Bus Connection

Only the connection of the e.bloxx A6-2CF to the bus is described in here. A detailed description of the bus and the communication of the modules can be found in the Communication Guide of the e.bloxx modules.

The e.bloxx A6-2CF have an RS 485 bus interface for connection to the serial fieldbus. The bus has to be terminated on both sides with characteristic impedance. The maximum line length depends on the transmission speed (refer to the Communication Guide for details) and can never be higher than 1.2 km per bus segment or 4.8 km via a physical bus string by using 3 repeaters. A maximum of 32 devices are possible at each bus segment and up to 127 devices via a physical bus string.

Wiring In general, the e.bloxx A6-2CF is connected to the bus by connecting both signal leads A and B of the incoming bus

cable and A' and B' of the outgoing bus cable together to one terminal on the module (Picture 3.2). Alternatively, the bus can also be connected by a "stub cable" (Picture 3.3). This guarantees that the bus connection to other modules remains in place, even if one module has to be exchanged, due to the removable terminal strip.

e.bloxx A6-2CF

MOUNTING E.BLOXX AND CONNECTING WIRES

HB_EBLOXX-A62CF_E_V19.doc

15

Gantner Instruments Test & Measurement GmbH

Picture 3.2. - Bus Connection of an e.bloxx A6-2CF to the RS 485 Fieldbus with Derivation

Picture 3.3. - Bus Connection of an e.bloxx A6-2CF to the RS 485 fieldbus via Stub Cable

The stub-cable should be as short as possible, not longer than 30 cm.

Notice: The terminal designations A and B of all modules of the e.bloxx series are exchanged compared with the

PROFIBUS-definitions. Consequently, in multi-vendor systems the bus lines A and B have to be exchanged when connecting them to the e.bloxx modules.

Bus Structure The bus structure is a line structure where each bus segment will be terminated with characteristic impedance on both

ends. Branches can be set up by means of a bi-directional signal amplifier, so-called repeaters. Other types of branches are not permitted (no tree topology). The max. stub-length to a user must not exceed 30 cm. The following pictures show a few examples of possible bus topology structures. The meanings of the symbols are as

follows:

e.bloxx A6-2CF

MOUNTING E.BLOXX AND CONNECTING WIRES

16

HB_EBLOXX-A62CF_E_V19.doc

Gantner Instruments Test & Measurement GmbH

Picture 3.4. - Simple Line Structure

Picture 3.5. – Extended Line Structure

Picture 3.6. - Line Structure with Branches

e.bloxx A6-2CF

MOUNTING E.BLOXX AND CONNECTING WIRES

HB_EBLOXX-A62CF_E_V19.doc

17

Gantner Instruments Test & Measurement GmbH

Picture 3.7. - e.bloxx A6-2CF connected to e.gate

* ... If e.bloxx A6-2CF are used together with an e.gate, which is used to collect data of all connected e.bloxx modules

and processes them for fast transmission via the further network, a bus termination must be connected to the last e.bloxx in each bus line.

Bus Connection to PC

The bus interface of the e.bloxx A6-2CF is based on the RS 485 standard. Since most hosts are "only" equipped with RS 232 interfaces, an interface converter or a plug-in board with RS 485 drivers is required for conversion purposes.

Gantner Instruments Test & Measurement GMBH offers a compact interface converter, called ISK 200, with an inte-grated power supply and automatic baud rate detection. The power supply, bus connection and a separate 24 VDC-output are DC-isolated. Therefore, the interface converter ISK 200 is also applicable as a power supply for remote

applications. Additionally, the interface converter ISK 200 features the option of connecting the required bus termination via a switch. The converter is designed to be used as a desk device.

Another module IRK 100 from Gantner Instruments Test & Measurement GMBH is available which may be used as an RS 485 repeater or RS 485/RS 232 converter. The baud rate can be adjusted at the IRK 100. Also, for this module the required bus termination may be connected with a switch. The Repeater/ Converter IRK 100 has a snap-on mounting mechanism for the installation on standard profile rails (DIN rail) 35 mm according to DIN EN 50022.

Interface Converter ISK 200

Repeater/Converter IRK 100

Interface Converter ISK 101

Picture 3.8.

- Interface Converters ISK 200, IRK 100 and ISK 101

e.bloxx A6-2CF

MOUNTING E.BLOXX AND CONNECTING WIRES

18

HB_EBLOXX-A62CF_E_V19.doc

Gantner Instruments Test & Measurement GmbH

Bus Connection to Profibus-DP

For the connection of the bus cable and bus interface to a PC 9-channel D-subminiature plugs and sockets can be used. The pin assignment for the RS 485 connection according to PROFIBUS is given in Table 3.1.

Table 3.1. - Pin Assignment D-Subminiature Plug According to PROFIBUS

The signal leads A and B (and Shield) are mandatory for a (shielded) connection. Additional signal leads may be installed if required.

Notice: Due to the fact that the RS 485 interface is used for different protocols, in case of using Profibus-DP the

leads A and B has to be crossed.

Bus Termination at the e.bloxx A6-2CF

In order to avoid signal reflections on the bus, each bus segment has to be terminated at its physical beginning and at its end with the characteristic impedance. A terminating resistor is installed between the bus leads A and B for this

purpose. In addition, the bus lead A is connected via a pull-up resistor to potential (VP) and the bus lead B is connected via a pull-down resistor to ground (DataGround). These resistors provide a defined quiescent potential in case there is no data transmission on the bus. This quiescent potential is level high.

The e.bloxx A6-2CF has these bus termination resistors built in. They can be connect to the bus by plugging the Bus Termination Plug IBT 100, which is available as accessory, into the rapid bus link plug on the front side of the module. Instead of the bus termination plug IBT 100, also separate jumpers may be used for the bus termination. In this case, it is mandatory that the jumper clips are installed as indicated below, and that the bus leads or the bus termination are not short-circuited by mistake.

e.bloxx

Bus Termination

e.bloxx with connected Bus Termination

Picture 3.9. - Bus Termination at the e.bloxx A6-2CF