微波消解测食品中的铅的方法

微波消解测食品中的铅的方法
微波消解测食品中的铅的方法

(一)、仪器:

微波消解罐8个,塑料勺2个,三角瓶1个,容量瓶(100ml)5个,塑料试剂瓶3个,15ml锥形离心管8个天平(0.0001g),移液器(1ml-5ml、100μl-1000μl),微波消解仪,智能样品处理器,耶拿AA700

(二)、试剂:

1、准备试剂:

硝酸(优级纯),过氧化氢(30%),超纯水,铅标准液(100μg/mL),基体改进剂(硝酸钯)

2、配制溶液

铅标准液(10μg/ mL,3%硝酸):10mL铅标准液(100μg/mL)+3mL硝酸(优级纯)+87mL超纯水

铅标准液(30μg/L,0.5%硝酸):0.3mL铅标准液(10μg/ mL,3%硝酸)+0.5mL硝酸(优级纯)+99.2mL超纯水铅标准液(20μg/L,0.5%硝酸):4mL铅标准液(30μg/L,0.5%硝酸)+2mL超纯水

(三)、步骤:

1、超纯水冲洗仪器,烘干备用,盖薄膜防污染。

2、称样:0.3g于微波消解罐中。

3、高温室通风橱中,加硝酸7ml,过氧化氢1ml,安装微波消解外罐,放入微波消解仪中,使用“ 3”程序消解,时间约1小时40分钟。放置过夜。

4、高温室通风橱中,取出消解管,超纯水冲洗消解罐盖子3-4次于消解罐中,放置智能样品处理器上170℃(首页选第二个run----,点enter,输入170,点R/S)赶酸(约1小时30分钟),液体剩余绿豆大小即可,期间配备溶液。

5、冲洗消解罐4-5次倒入10ml塑料定容管中定容备用。

标准曲线法:Pb-food--bzqx

标准加入法:Pb-food--bzjr

测定食品中铅含量及实验条件研究

毕业论文(设计) 题目:几种毕节特色食品中铅的测定 及实验条件研究 学号:0713******* 姓名:苏志卫 教学院:化学与化学工程学院 专业班级:2008级应用化学本科班 指导教师:陈燕芹 完成时间:2012年5 月10 日 毕节学院教务处制

目录 摘要 (1) Abstract (1) 第一章绪论 (2) 1.1 研究背景 (2) 1.2 研究现状 (2) 1.3 研究内容和方法 (3) 1.4 实验仪器、试剂 (3) 第二章氢化物发生—原子荧光光谱法 (4) 2.1 原理 (4) 2.2 仪器装置 (4) 2.3 特点 (5) 2.4 应用 (5) 第三章实验条件研究 (7) 3.1 灯电流选择 (7) 3.2 负高压选择 (7) 3.3 原子化器高度选择 (7) 3.4 载气流量选择 (8) 3.5 屏蔽气流量选择 (8) 3.6 实验条件研究结果 (9) 第四章几种毕节特色食品中铅的测定 (10) 4.1 标准曲线和检出限 (10) 4.2 样品处理 (11) 4.3 测量 (11) 第五章结果与讨论 (12) 5.1 结果分析 (12) 5.2 讨论 (12) 5.2.1 介质酸浓度对实验的影响 (12) 5.2.2 硼氢化钾对实验的影响 (13) 5.2.3 铁氰化钾对实验的影响 (13) 5.3 问题与展望 (14) 参考文献 (15) 致谢 (16)

毕节学院本科毕业论文 几种毕节特色食品中铅的测定及实验条件研究 摘要 铅是对人体毒性最强的重金属之一,并广泛存在于自然界,食物链是人体铅的主要来源,所以对食品中铅含量进行测定极为重要。应用氢化物发生—原子荧光光谱法测定铅含量,具有操作简便快捷、灵敏度高、干扰少、节省试剂等优点。采用微波消解法消解样品,由于样品的消解在高压密封的消解罐中进行,所需的试剂用量较少、铅损失小、消化时间短、消解效果好。 本实验先对实验条件进行研究,根据所设定的实验条件做出标准曲线,然后对样品进行处理,最后测出的几种毕节特色食品(荞饭、酸菜、臭豆干、土豆)中铅含量均低于食品中铅允许限量。 关键词:铅;氢化物发生—原子荧光光谱法;微波消解;毕节特色食品 Abstract Lead is one of the body of the most toxic heavy metals, and widely distributed in nature, the food chain is the main source of human lead, the lead content in food were extremely important. The application of hydride generation - atomic fluorescence spectrometric determination of lead content, with easy operation, high sensitivity, less interference, reagent saving advantages.The samples were digested by microwave digestion method required less reagent and sample digestion in the digestion of high-pressure sealed cans, the lead loss, digestion time, digestion effect. In this study, experimental conditions and to make a standard curve, according to the set of experimental conditions, and then the samples were last measured several the Bijie characteristics of food (buckwheat rice, sauerkraut, smelly bean curd, potatoes) in the lead levels were lower than the allowable levels of lead in food. Key words: lead; hydride generation - atomic fluorescence spectrometry; Microwave Digestion;The specialty food in Bijie

消解所用酸的种类和用量

消解所用酸的种类和用量 消解试样使用最广泛的酸是HN03、HCl、HF、HCl04、H202等都是良好的微波吸收体,它们在微波炉中的稳定性、沸点和蒸汽压以及与试样的反应。 (1)HNO3(比重1.42,70%水溶液,w/w) 在常压下的沸点为120℃。在0.5MPa下,温度可达176℃,它的氧化电位显著增大,氧化性增强。能对无机物及有机物进行氧化作用。金属和合金可用硝酸氧化为相应的硝酸盐,这些硝酸盐通常易溶于水。部分金属元素,如Au、Pt、Nb、Ta、Zr不被溶解。Al 和Cr不易被溶解。硝酸可溶解大部分的硫化物。 硝酸消化的主要有机样品有:饮料、植物、废水、聚合物等。 (2)HCl(1.19,37%),沸点110℃ HCl不属于氧化剂,通常不用来消解有机物。HCl在高压与较高温度下可与许多硅酸盐及一些难溶氧化物、硫酸盐、氟化物作用,生成可溶性盐。许多碳酸盐、氢氧化物、磷酸盐、硼酸盐和各种硫化物都能被盐酸溶解。 (3)HCl04(72%),沸点130℃ 是一种强氧化剂,能彻底分解有机物。但高氯酸直接与有机物接触会发生爆炸,因此,通常都与硝酸组合使用。或先加入HN03反应一段时间后再加入HC104。HC104大都在常压下的预处理时使用,较少用于密闭消解中,要慎重使用。 (4)HF(38.3%),沸点112℃ 在密闭容器中达180℃,会产生约0.8Mpa 的分压,能有效地使硅酸盐变成可挥发的SiF4,而留下其他要测量的元素。少量HF与其他酸相结合使用,可有效地防止样品中待测元素形成硅酸盐。 (5)H2O2(30%),沸点107℃ 过氧化氢的氧化能力随介质的酸度增加而增加。H202 分解产生的高能态活性氧对有机物质的破坏特别有利。使用时通常先加HN03预处理后再加入H2O2 。 (6)H2S04(1.84,98.3%),沸点338℃ 硫酸是许多有机组织、无机氧化物、合金、金属及矿石等的有效溶剂。它几乎可以破坏所有的有机物。但在密闭消解时要严格监控反应温度,因为浓H2S04在达到沸点温度时可能熔化聚四氟乙烯内罐,浓H2S04的沸点是338℃,而聚四氟乙烯的使用温度不能超过240 ℃。所以,一般不单独用H2S04,而是与HN03一起组合使用。 (7)HP04(1.71,85%),沸点158℃ 磷酸有较低的蒸气压,在0.8MPa时温度可达240 ℃。热HP04 适用于消解那些用HCl消解时会使某些特定痕量组分挥发损失的铁基合金,磷酸还可溶解铬矿、氧化铁矿、铝炉渣等。在微波消解时,常常采用两种或两种以上的酸组合(混合)使用,消解效果更好。常使用的混合酸有以下几种。 (1)水,HCl:HNO3=3:1 v/v 王水需现配现用。王水可用来溶解许多金属和合金,其中包括钢、高温合金钢、铝合金、锑、铬和铂族金属等。植物体与废水也常使用它来进行消化。王水可从硅酸盐基质中酸洗出 部分金属,但无法有效的加以完全溶解。除王水外,硝酸和盐酸还常以另外的比例混合在一起使用, 所谓的勒福特(Lefort)王水,也叫逆王水,是三份硝酸与一份盐酸的混合物。 可用来溶解氧化硫和黄铁矿。 (2) HN03:H2S04, 常用的比例为1:1(v/) 这种混酸的最高温度仅比单纯HN03时的最高温度高10℃左右。高温条件下,易于形成硫酸盐络合物,还具有脱水和氧化的性质。通常在完成最初的消化后,可加入H2O2 以完成

COD微波消解法

微波消解法测定化学需氧量 微波消解技术加热快,效率高,与标准的回流加热方法相比消解时间短,水样及各种试剂的消耗量小。实验仪器:微波炉,聚四氟乙烯消解罐,酸式滴定管,移液管,锥形瓶。 一、测试方法 1、所用试剂均采用基准或分析纯。 2、含Hg+消解液:称取经120℃烘干2小时的基准纯重铬酸钾9.806g溶于600ml 水中,再加入硫酸汞25.0g(如所测的水样氯离子浓度<100mg/l,可免加硫酸汞),边搅拌边慢慢加入浓硫酸250ml,冷却后移入1000ml容量瓶中,并稀释至刻度线,该溶液重铬酸钾浓度为0.20mol/l。 3、试亚铁灵指示液:称取邻菲罗啉1.458g、硫酸亚铁(FeSO4·7H2O)0.695g溶于水 中,稀释至1000ml,贮于棕色瓶中。 4、硫酸亚铁铵标准溶液:称取(NH4)2Fe(SO4)2·6H2O16.6g溶于水中,边搅拌边 缓慢加入浓硫酸20ml,冷却后移入1000ml容量瓶中,加水稀释至刻度,摇匀。其浓度约为0.042mol/l,临用前用重铬酸钾标准溶液标定。 标定方法:准确吸取5.00ml重铬酸钾标准溶液于150ml锥形瓶中,加水稀释至约30ml左右,缓慢加入浓硫酸5ml,混匀,冷却后加入2滴试亚铁灵指示剂(约0.10ml)用硫酸亚铁铵溶液滴定,溶液的颜色由黄色经蓝绿色至褐色即为终点。 C[(NH4)2Fe(SO4)2]·6H2O=(0.2000×5.00)/V 式中:C—硫酸亚铁铵标准溶液的浓度(mol/l) V—硫酸亚铁铵标准溶液的滴定用量(ml) 5、硫酸—硫酸银催化剂:于1000ml浓硫酸中加入10g硫酸银,放置1-3天,不时摇动使其溶解。 6、硫酸汞:结晶或粉末。 7、邻苯二甲酸氢钾标准溶液:C(KC6H5O4)=2.0824mmol/l,称取150℃时干燥2h的邻 苯二甲酸氢钾0.4251g,溶于水,并稀释至500ml,混匀。以重铬酸钾为氧化剂,将邻苯二甲酸氢钾完全氧化的COD值为1.176(指1g邻苯二甲酸氢钾耗氧1.176g)该标准溶液的理论COD值为1000mg/l。 二、测试步骤:

微波消解前处理测定食品中总汞方法论文

微波消解前处理测定食品中总汞方法的研究【摘要】目的应用微波消解前处理测定食品中总汞的含量。方法用微波消解系统对食品进行消解,样品用原子荧光仪测定其含量。结果本方法测定汞的检出限为0.012μg/kg,精密度在 2.4-9.5%回收率为93—109%。结论该方法简便、快速、灵敏度、精密度准确度较好,适合食品的快速消解以及微量金属污染的测定。 【关键词】微波消解食品总汞氢化物原子荧光法 中图分类号:r155文献标识码:a文章编号:1005-0515(2011)2-014-02 determination of microwave digestion pre-treatment method of total mercury o lihua liang zhijian liwenting (center for disease control and prevention, kunming, yunnan, kunming, yunnan 650228, china) 【abstract】objective determination total mercury in food by microwave digestion pre-treatment methods.methods mercury in food were determined by microwave digest-hydride generation atom fluorescence spectrometry.results the limit of detection mercury was 0.012μg/kg; the rsd was 2.4%-9.5%; the recovery rates of mercury were 93%-109%.conclusion the methods is simple,rapid,sensitive and of accurate results

食品中铅的测定方法

食品中铅的测定方法 1.1 原理 试样经灰化或酸消解后,注入原子吸收分光光度计石墨炉中,电热原子化后吸收283.3nm共振线,在一定浓度范围,其吸收值与铅含量成正比,与标准系列比较定量。 1.2 试剂 1.2.1硝酸:优级纯。 1.2.2高氯酸:优级纯。 1.2.3硝酸(0.5mol/L):取3.2ml 硝酸加入50ml水中,稀释至100ml。 1.2.4硝酸(1mol/L):取6.4ml硝酸加入50ml水中,稀释至100ml。 1.2.5磷酸二氢铵溶液(20g/L):称取2.0g磷酸二氢铵,以水溶解稀释至100ml。 1.2.6混合酸:硝酸+高氯酸(4+1)。取4份硝酸与1份高氯酸混合。 1.2.7铅标准储备液:由国家标准物质研究中心提供。 1.2.8铅标准使用液:每次吸取铅标准储备液1.0ml于100ml容量瓶中,加硝酸(0.5mol/L)或硝酸(1mol/L)至刻度。如此经多次稀释成每毫升含10.0,20.0,40.0,60.0,80.0ng铅的标准使用液(可根据样品所含浓度进行配制)。 1.3仪器 所用玻璃仪器均需以硝酸(1+5)浸泡过液,用水反复冲洗,最后用去离子水冲洗干净。 1.3.1原子吸收分光光度计(附石墨炉及铅空心阴极灯)。 1.3.2消化装置 1.3.3可调式电热饭、可调式电炉。 1.4 操作 1.4.1 试样预处理 1.4.1.1 在采样和制备过程中,应注意不使试样污染。 1.4.1.2 粮食、豆类去杂物后,磨碎,过20目筛,储于塑料瓶中,保存备用。 1.4.1.3 蔬菜、水果、鱼类、肉类及蛋类等水分含量高的鲜样,用食品加工机或匀浆机打成匀浆,储于塑料瓶中,保存备用。 1.4.2 试样消化 湿式消解法:称取试样1.00g~5.00g 于锥形瓶或高脚烧杯中,放数粒玻璃珠,加10ml混合酸,加盖浸泡过夜,加一小漏斗电炉上消解,若变棕黑色,再加混合酸,直至冒白烟,消化液呈无色透明或略带黄色,放冷用滴管将试样消化液洗入或过滤入(视消化后试样的盐分而定)10ml~25ml容量瓶中,用水少量多次洗涤锥形瓶或高脚烧杯,洗液合并于容量瓶中并定至刻度,混匀备用;同时作试剂空白。 1.4.3 测定 1.4.3.1 仪器条件:根据各自仪器性能调至最佳状态。参考条件为波长283.3nm,狭缝0.2nm~1.0nm,灯电流5mA~7mA,干燥温度120℃,20s;灰化温度450℃,持续15s~20s,原子化温度1700℃~2300℃,持续4s~5s,背景校正为氘灯或塞曼效应。 1.4.3.2 标准曲线绘制:吸取上面配制的铅标准使用液10.0,20.0,40.0,60.0,80.0ng/ml(或μl)各10μL,注入石墨炉,测得其吸光值并求得吸光值与浓度有关系的一元线性回归方程。 1.4.3.3 试样测定:分别吸取样液和试剂空白液各10μl,注入石墨炉,测得其吸光值,代入标准系列的一元线性回归方程中求得样液中铅含量。 1.4.3.4 基体改进剂的使用:对于干扰试样,则注入适量的基体改进剂磷酸二氢铵溶液(20g/L)一般为5μl或与试样同量消除干扰。绘制铅标准曲线时也要加入与试样测定时等量的基体改进剂磷酸二氢铵溶液。

微波消解知识

转一篇文章: 密闭微波样品消解原理及常识 刘伟阎军武刚 The Knowledge of Hermetic Microwave Digestion 4样品制备目标和原则 现在,在生物有机物样品及矿物岩石、矿石、矿渣和玻璃等痕量元素分析中,常常是从待分析物的溶样开始,绝大多数都要进行化学预处理,要把固体制成溶液,需要分解和破坏样品基体。一般使用原子吸收或发射光谱分析元素时,若样品为水溶液时,通常都可以达到极佳的分析结果。非水溶液也可以分析,但是样品溶液中,高浓度的有机化合物对某元素会引起严重的分析干扰问题。绝大部分的样品都无法直接进行固体分析,因此必须先转变成溶液型态。在分析前先经过消化处理,大多情况下都可产生较精确的分析结果。以下为样品消解前处理所应考虑的前提: 4.1确定样品消解是否必要 1).没有样品消解能否分析? 2).通过消解样品能否改善分析? 4.2理想样品制备的进行步骤 1).将固体和液体样品转化成液体溶液,以避免在测定中阻塞仪器液体传输及雾化系统。 2).破坏所有的有机物质使它不会干扰火焰燃烧或增加背景讯号。 3).把感兴趣的分析物以可检测的浓度保留在溶液中。 4).不加任何元素或化合物干扰离子。 5).调整样品的粘度和颗粒百分比到分析的最佳条件。 4.3采用微波消解手段的十大理由 1)与加热板消化比较时,消化可快4-100倍完成。 2)通常采用的2450MHz的微波,只能导致分子(粒子)运动,不引起分子结构变化,从而不会改变消解反应的方向。 3)微波直接向样品释放能量(热是副产物),避免了传统方式(热传导、热对流)中能量的损失,提高了能量的使用效率。 4)大多数传统试剂在微波消解中仍然可以使用,因此对大多数的反应操作者无须改变试剂的种类。且用于消化的酸类不会因为其活性成分的蒸发而降低或失去强度。 5)样品的消解可以进行的更精确、彻底。在许多消化程序中可避免过氯酸的使用,如HNO 3在微波消化期间,基于消化瓶内压力的缘故,会产生较高的温度而得到较好的消化结果,以取代过氯酸的使用。 6)密闭微波消解可通过提高温度/压力协助反应,使反应物在特定温度下发生快速分解,减少分解所需的时间,提高工作效率,对传统方法这是不可能的。 7)挥发元素如:As,Hg等可被保留在消化溶液中,防止挥发造成结果的偏差和对环境的污染。同时也使操作人员避免接触酸雾和有害的气体,如氢氟酸。 8)由于微波消解试剂用量少且密闭,可消除由于空气传播的微粒或渗出现象而导致的样品污染。因此有较低的空白值。 9)最先进的微波消解仪器能够通过磁控管的自动调节,定量的控制微波能量的传递,以此控制分解条件并实现对反应的自动控制。避免了人为操作产生的错误和误差。 10)通过温压控制可以保证消解的质量,保证反应一致的平行性和重复性。 5酸与样品分解 消解的目的是希望酸能分解样品基体同所感兴趣的金属离子形成可溶盐。硝酸及过氯酸等氧化性的酸类,是最常用来破坏有机物质与分解金属化合物的试剂。有机物质氧化成

微波消解_原子吸收法测定食品中的钙含量

第7卷 第8期 食品安全质量检测学报 Vol. 7 No. 8 2016年8月 Journal of Food Safety and Quality Aug. , 2016 *通讯作者: 李卫群, 高级工程师, 主要研究方向为光谱分析。E-mail: lwq@https://www.wendangku.net/doc/9b10998337.html, *Corresponding author: LI Wei-Qun, Senior Engineer, Hangzhou Wahaha Group Co., Ltd., Hangzhou 310018, China. E-mail: lwq@https://www.wendangku.net/doc/9b10998337.html, 微波消解-原子吸收法测定食品中的钙含量 李卫群*, 汪涓涓, 徐玲玲, 朱 慧 (杭州娃哈哈集团有限公司, 杭州 310018) 摘 要: 目的 建立微波消解-原子吸收法测定食品中钙含量的方法。方法 采用微波消解法对样品进行前处理, 在检测样品中加入氯化镧(8 g/L)屏蔽剂, 用火焰原子吸收法进行检测。比较经消解后样品中不同的硝酸浓度对钙含量测定结果的影响, 探究微波消解法测定食品中钙含量时结果偏低的原因。结果 经微波消解后, 样品中的硝酸含量大于0.5%时, 会导致钙含量的检测结果偏低。经湿法消解处理的样品, 其加标回收率在97.2%~106.0%之间, 经微波消解法处理的样品, 其加标回收率在96.8%~104.0%之间。将采用上述两种消解方法处理的样品的钙含量检测结果进行比较, 测定值间的相对误差为1.64%~3.08%, 在可接受范围内。结论 微波消解-原子吸收法可以用于食品中钙含量的检测。 关键词: 微波消解法; 原子吸收法; 钙含量 Determination of calcium content in food by microwave digestion-atomic absorption spectrometry LI Wei-Qun *, WANG Juan-Juan, XU Ling-Ling, ZHU Hui (Hangzhou Wahaha Group Co ., Ltd., Hangzhou 310018, China ) ABSTRACT: Objective To establish a method for determination of calcium content in foods by microwave digestion-atomic absorption spectrometry. Methods The samples were pretreated with microwave digestion and detected by flame atomic absorption spectrometry with lanthanum chloride (8 g/L) as screening agent. The effects of different concentrations of nitric acid in sample after digestion on the determination of calcium content were compared for exploring the causes of lower calcium content detected by atomic absorption spectrometry with microwave digestion. Results The detection results of calcium content were lower when the concentration of nitric acid residue in samples was larger than 0.5% after microwave digestion. The recoveries of samples treated by wet digestion were between 97.2%~106.0% and the samples treated by microwave digestion were between 96.8%~104.0%. The detection results of calcium content from above methods were compared and the relative errors (RE) were between 1.64%~3.08%, which were in the acceptable range. Conclusion Microwave digestion-atomic absorption spectrometry can be used for the detection of calcium content in foods. KEY WORDS: microwave digestion method; atomic absorption spectrometry; calcium content 1 引 言 钙是人体的必需元素之一, 是构成骨骼与牙齿的重 要成份, 在调节细胞代谢、维持肌肉收缩和保证神经传导 等方面都有重要作用。缺钙将可能导致严重的疾病, 但是过量补钙则会影响铁和锌的吸收[1]。因此, 准确测定食品

消解用酸种类和用量

消解试样使用最广泛的酸是HN03、HCl、HF、HCl04、H202等都是良好的微波吸收体,它们在微波炉中的稳定性、沸点和蒸汽压以及与试样的反应。 (1)HNO3(比重1.42,70%水溶液,w/w) 在常压下的沸点为120℃。在0.5MPa下,温度可达176℃,它的氧化电位显著增大,氧化性增强。能对无机物及有机物进行氧化作用。金属和合金可用硝酸氧化为相应的硝酸盐,这些硝酸盐通常易溶于水。部分金属元素,如Au、Pt、Nb、Ta、Zr不被溶解。Al 和Cr不易被溶解。硝酸可溶解大部分的硫化物。 硝酸消化的主要有机样品有:饮料、植物、废水、聚合物等。 (2)HCl(1.19,37%),沸点110℃ HCl不属于氧化剂,通常不用来消解有机物。HCl在高压与较高温度下可与许多硅酸盐及一些难溶氧化物、硫酸盐、氟化物作用,生成可溶性盐。许多碳酸盐、氢氧化物、磷酸盐、硼酸盐和各种硫化物都能被盐酸溶解。 (3)HCl04(72%),沸点130℃ 是一种强氧化剂,能彻底分解有机物。但高氯酸直接与有机物接触会发生爆炸,因此,通常都与硝酸组合使用。或先加入HN03反应一段时间后再加入HC104。HC104大都在常压下的预处理时使用,较少用于密闭消解中,要慎重使用。 (4)HF(38.3%),沸点112℃ 在密闭容器中达180℃,会产生约0.8Mpa 的分压,能有效地使硅酸盐变成可挥发的SiF4,而留下其他要测量的元素。少量HF与其他酸相结合使用,可有效地防止样品中待测元素形成硅酸盐。 (5)H2O2(30%),沸点107℃ 过氧化氢的氧化能力随介质的酸度增加而增加。H202 分解产生的高能态活性氧对有机物质的破坏特别有利。使用时通常先加HN03预处理后再加入H2O2 。 (6)H2S04(1.84,98.3%),沸点338℃ 硫酸是许多有机组织、无机氧化物、合金、金属及矿石等的有效溶剂。它几乎可以破坏所有的有机物。但在密闭消解时要严格监控反应温度,因为浓H2S04在达到沸点温度时可能熔化聚四氟乙烯内罐,浓H2S04的沸点是338℃,而聚四氟乙烯的使用温度不能超过240 ℃。所以,一般不单独用H2S04,而是与HN03一起组合使用。 (7)HP04(1.71,85%),沸点158℃ 磷酸有较低的蒸气压,在0.8MPa时温度可达240 ℃。热HP04 适用于消解那些用HCl消解时会使某些特定痕量组分挥发损失的铁基合金,磷酸还可溶解铬矿、氧化铁矿、铝炉渣等。在微波消解时,常常采用两种或两种以上的酸组合(混合)使用,消解效果更好。常使用的混合酸有以下几种。 (1)水,HCl:HNO3=3:1 v/v 王水需现配现用。王水可用来溶解许多金属和合金,其中包括钢、高温合金钢、铝合金、锑、铬和铂族金属等。植物体与废水也常使用它来进行消化。王水可从硅酸盐基质中酸洗出 部分金属,但无法有效的加以完全溶解。除王水外,硝酸和盐酸还常以另外的比例混合在一起使用, 所谓的勒福特(Lefort)王水,也叫逆王水,是三份硝酸与一份盐酸的混合物。 可用来溶解氧化硫和黄铁矿。 (2)HN03:H2S04, 常用的比例为1:1(v/) 这种混酸的最高温度仅比单纯HN03时的最高温度高10℃左右。高温条件下,易于形成硫酸盐络合物,还具有脱水和氧化的性质。通常在完成最初的消化后,可加入H2O2 以完成消化。但是,只有当液量减少且冒S02气体后才能添加H202 。用它消解的样品有:聚合物、脂肪及有机物质。

微波消解原子荧光法测定食品中的砷

微波消解原子荧光法测定食品中的砷【摘要】目的建立微波消解原子荧光法测定食品中砷的方法。方法采用微波消解处理样品,原子荧光法测定,进行了消解条件、仪器条件、精密度、回收率等实验。结果方法的检出限为1.68ng/ml,在0.0028-0.05μg/ml范围内,相关系数r=0.9993,回收率在92.4-101.4之间,相对标准偏差4.03%。结论微波密闭消解样品,消化过程节约试剂,防止试样中待测元素的损失,干扰少,适用于食品中砷的测定。 【关键词】微波消解原子荧光法食品砷 砷的化合物在自然环境中广泛存在。人体长期摄入被砷污染的食物后,可以引起严重中毒,因此砷在食品卫生检测中被列为常规检测的有害元素。对砷的检测技术要求较高。在食品分析中常用的消解方法为湿法消解,该方法消解时间长,酸和其他溶液用量大,基体干扰严重[1]。本文用微波消化食品,方法简单,消化速度快,大大缩短了检验周期,取得满意的结果。 1试验部分 1.1原理样品经过预处理,在全封闭的消化罐中经微波消解后,加入硫脲使五价砷还原为三价砷,在酸性条件下,以硼氰化钾作还原剂,将样品中待测的砷还原成挥发性共价氢化物,借助载气将其带入原子化器中进行原子化。在砷特制空心阴极灯照射下,发射特征波长的荧光,其荧光强渡与砷含量成正比,与标准系列比较定量。 1.2主要仪器与试剂

AFS-820双光道原子荧光分光光度计(北京吉天仪器有限公司);MK型光纤压力自控密闭微波消解仪,附聚四氟乙烯样杯;DKP—型电子控温加热板(上海新仪微波化学科技有限公司)。砷标准溶液:国家标准物质研究中心(GBW08611),ρ(As)=1000μg/ml。用超纯水稀释成ρ(As)=1μg/ml标准使用液(临用现配);硝酸:优级纯(ρ20=1.42g/ml);30%过氧化氢,分析纯;硫脲—抗坏血酸(50g/L),称取硫脲和抗坏血酸各5g,溶于100ml水中。硼氰化钾溶液(10g/L):称取2.5g硼氰化钾并使其溶于250ml5g/L的氢氧化钾溶液中;载液:盐酸溶液(2%)。 1.3样品处理 称取约1.00g食品样品于溶样杯内,加硝酸6ml,放在180℃的电热板上,加热到硝酸黄烟帽尽(约15-20min),取下,放置室温;再加硝酸5ml,过氧化氢1.5ml,置入微波炉内,一档3min,二档5min,三档3min,加压消解;取出后加5ml水,置于180℃的电热板上加热,待水挥发尽干,再加入1.25ml硫酸,继续放在180℃的电热板上驱赶硝酸,至水挥发完。冷却后,用水将内容物定量转入25ml比色管中,其间加入2.5ml50g/L硫脲—抗坏血酸,补水至刻度,混匀备测。 1.4仪器条件[2] 灯电流:50mA,负高压:270V;载气流量:400ml/min,屏蔽气流量800ml/min;延时时间0s,读数时间:10s;原子化器高度8mm。 1.5标准曲线 分别吸取砷标准应用液0.50、1.00、2.00、4.00、8.00、10.00ml

微波消解法测定-COD

实验六微波消解法测定COD 1. 实验目的 掌握COD测定仪装置测量污水中COD(Cr)的办法。 2.原理 微波消解COD测定仪,采用硫酸一重铬酸钾消解体系,利用微波作用于反应内部引起分子间产生高摩擦作用所产生的热量来消解产品。 3.仪器的主要技术性能及结构 本仪器由主机、密封消解罐组成。 密封法测量范围:COD(Cr): 10~800mg/L, COD (Cr)>800mg/L(稀释测定); 消解时间:能同时消解数个水样(3~9个任意),耗时不超过30分钟。 4.测试方法与步骤 4.1 试剂的选用与配置 (1 )重铬酸钾溶液 重铬酸钾消解液(用于密封消解法):0.25N,称取经过120度烘干2小时的分析纯重铬酸钾12.259 克,溶于约500ml 蒸馏水中,边搅拌边缓慢加入浓硫酸100ml,冷却后移入1000ml容量瓶中,加入30克固体硫酸汞(用于测定低氯离子或无氯离子时可不加),并用蒸馏水稀释至刻度。 重铬酸钾标准液,0.25N (用于非密封微回流法和标定):称取经过120度烘干2小时的基准纯或分析纯12.259克,置于1000ml容量瓶中,并用蒸馏水稀释至刻度。 (2)试亚铁灵指示溶液:分别称取1.485 克邻菲罗玲和0.695 克硫酸亚铁溶于水中,稀释至100ml,贮于棕色瓶内。 (3)硫酸亚铁铵标准溶液,约0.1N,准确浓度应在使用的当天用重铬酸钾标准液标定。 配制方法:取39.5 克分析纯的六水硫酸亚铁铵溶于蒸馏水中,边搅拌边缓慢加入20ml浓硫酸,冷却后移入1000ml容量瓶中,并稀释至刻度。 标定方法:量取5.00ml 重铬酸钾标准溶液,稀释至大约45 毫升。加入10 毫升浓硫酸,冷却后,加入2 滴试亚铁灵指示剂,用硫酸亚铁铵滴定,溶液的颜色由黄色经蓝绿色至红褐色即为终点。 C= ( 0.250X5.00 ) /V

食品中铅测定的试验设计

食品中铅测定的试验设计 铅是一种毒性很强的重金属,不是人体必需的微量元素,食品中铅主要来源于原料污染和生产工艺、容器、包装、储存和运输等环节的污染,被世界卫生组织列为食品污染物加以控制。人体摄入0.04g的铅就会引起急性中毒,铅中毒具有蓄积性、持久性和不可逆性。因此,加强食品检测防止铅中毒非常重要,而使用快速、灵敏、准确的测定方法测定食品中铅显得十分必要。 1.铅的理化性质及对人体健康的影响 1.1铅的理化性质 铅是一种重金属元素,化学符号为Pb,原子序数为82,熔点327.502℃,沸 点1740℃,密度11.3437g/cm 3 ,莫氏硬度1.5,很柔软,金属铅有良好的展性, 能压成薄片,但没有延性,不能拉成丝。不与水作用,与盐酸反应时,生成溶解度小的氯化铅覆盖在铅的表面,使反应终止。与硫酸的作用和盐酸相似。能溶于 浓热的硫酸中,生成可溶性的硫酸氢铅;溶于稀硝酸,生成硝酸铅[1] ,故测定 铅含量时常配制成硝酸铅溶液。铅为重金属,可导致蛋白质性,对人体有毒。 1.2铅在人体内分布及对人体健康的影响 人体吸收的铅大部分来自食物,少部分来自污染的空气,铅通过肠道和呼吸道吸收入人体后,随血流分布到全身各器官和组织,血液中的铅部分通过肾脏由尿液排出体外,部分从大便排出,部分储存在骨骼里。人体内的铅95%以上都以不溶性磷酸盐形式沉积在骨骼中,而且很难出来再回到血液,骨骼中的铅的半衰期约为20~30年,这部分铅对人体来说相对安全。少部分储存在肝、肾、肌肉和 中枢神经系统[2] 。急性铅中毒比较少见,其毒性主要是由于铅在人体蓄积所造 成的神经性和血液性中毒[2] 。铅的毒性机理主要是对蛋白质及酶中的半胱氨酸 残基的反应。铅慢性中毒对人体危害分为三个阶段:(1)低血色素贫血导致溶血和红细胞寿命缩短,还会出现无相关的行为异常或组织功能障碍(包括消化、免疫等);(2)中枢神经系统失调,并诱发多发性神经炎。表现为机能亢进,冲动行为、知觉紊乱和学习能力下降。在许多严重的病例中,症状包括坐立不安、易怒、头痛、肌肉震颤、运动失调和记忆力丧失;(3)肾衰竭、痉挛、昏迷甚 至死亡。对婴幼儿、儿童及孕妇的伤害尤为明显[3] 。

微波消解试题(含答案)

单位: 科室: 姓名: 微波消解培训测试 一、列举5种不能进行微波消解的物质。(10分) 答:炸药(TNT,硝化纤维等)、推进剂(肼,高氯酸胺等)、高氯酸盐、二元醇(乙二醇,丙二醇等)、航空燃料(JP-1等)、引火化学品、醚(熔纤剂-乙二醇苯基醚等)、丙烯醛、酮(丙酮,甲基乙基酮等)、烷烃(丁烷,己烷等)、乙炔化合物、双组分混合物(硝酸和苯酚,硝酸和三乙胺,硝酸和丙酮等)、硝酸甘油酯,硝化甘油或其它有机硝化物、丙三醇(甘油),以上物质任意五种即可。(每种正确得分2分) 二、简述MARS方法编辑中设置功率平台的意义和选择依据。(10分) 答:功率平台三个1600W、800W、400W,代表功率控制精度,控制为0-100%。1600W的控制精度是16W,400W的控制精度是4W。当样品数量很少时,控制精度要求高,所以功率设置要低。选择的依据为1-3个消解罐使用400W。3-6个消解罐使用800W,6个以上使用1600W。(两个要点,各5分) 三、在微波消解仪中,样品为什么要尽量分散对称摆放?(10分) 答:以保证微波加热的均匀性以及测温的准确性。微波吸收体过于集中比列于加热和散热的效果。 四、微波消解后上机分析元素回收率偏低的原因有哪些?(10分) 答:结果偏低可能原因是多方面的。主要有以下几个:1、操作问题(2分),比如称量错误,样品预处理不好等。2、样品损失(2分),来源多方面的,如加样到消解罐中时损失等。3、消解效果不好(3分),需要改进消解方法或者调整酸体系配置。4、后期分析问题(3分),调整标准曲线或者提高信噪比,或者使用更高精度的分析仪器。 五、为什么要保持罐体的清洁与干燥?(10分) 答:由于微波是分子内加热,只要有溶液态物质存在,既可以吸收微波。所以罐体上如果有污染或者溶液存在,就会被加热,从而导致罐体变形甚至泄露的发生。带来不必要的损失。 六、微波消解使用的酸试剂体系有什么要求?(10分) 答:HNO3、HCl、HF无限制使用(2分),H2O2可以适当添加,以促进消化效果,H2SO4、H3PO4限制使用(2分),H3ClO4禁止使用(2分)。消解试剂体积在8-30ml(2分)。试剂体系要求8-30ml(2分)。 七、高压消解罐存在金属元素污染如何清洗?(10分)

食品中铅镉砷的测定国标

食品中铅的测定: 第一法石墨炉原子吸收光谱法 3 原理 试样经灰化或酸消解后,注入原子吸收分光光度计石墨炉中,电热原子化后吸收nm 共振线, 在一定浓度范围,其吸收值与铅含量成正比,与标准系列比较定量。 4 试剂和材料 硝酸:优级纯。 过硫酸铵。 过氧化氢(30%)。 高氯酸:优级纯。 硝酸(1+1):取50 mL 硝酸慢慢加入50 mL 水中。 硝酸(mol/L):取mL 硝酸加入50 mL 水中,稀释至100 mL。 硝酸(l mo1/L):取mL 硝酸加入50 mL 水中,稀释至100 mL。 磷酸二氢铵溶液(20 g/L):称取g 磷酸二氢铵,以水溶解稀释至100 mL。 混合酸:硝酸十高氯酸(9+1)。取9 份硝酸与1 份高氯酸混合。 铅标准储备液:准确称取g 金属铅(%),分次加少量硝酸(),加热溶解,总量不超过37 mL,移入1000 mL 容量瓶,加水至刻度。混匀。此溶液每毫升含mg 铅。 铅标准使用液:每次吸取铅标准储备液mL 于100 mL 容量瓶中,加硝酸()至刻度。如此经多次稀释成每毫升含ng,ng,ng,ng,ng 铅的标准使用液。 5 仪器和设备 原子吸收光谱仪,附石墨炉及铅空心阴极灯。 马弗炉。 天平:感量为1 mg。 干燥恒温箱。 瓷坩埚。 压力消解器、压力消解罐或压力溶弹。 可调式电热板、可调式电炉。 6 分析步骤 试样消解(可根据实验室条件选用以下任何一种方法消解) 湿式消解法:称取试样1 g~5 g(精确到g)于锥形瓶或高脚烧杯中,放数粒玻璃珠,加10 mL 混合酸(),加盖浸泡过夜,加一小漏斗于电炉上消解,若变棕黑色,再加混合酸,直至冒白烟,消化液呈无色透明或略带黄色,放冷,用滴管将试样消化液洗入或过滤入(视消化后试样的盐分而定)10 mL~25 mL 容量瓶中,用水少量多次洗涤锥形瓶或高脚烧杯,洗液合并于容量瓶中并定容至刻度,混匀备用;同时作试剂空白。 测定 仪器条件:根据各自仪器性能调至最佳状态。参考条件为波长nm,狭缝nm~nm,灯电流5 mA~7 mA,干燥温度120 ℃,20 s;灰化温度450 ℃,持续15 s~20 s,原子化温度:1700 ℃~2300 ℃,持续4 s~5 s,背景校正为氘灯或塞曼效应。 标准曲线绘制:吸取上面配制的铅标准使用液ng/mL(或μg/L),ng/mL(或μg/L),ng/mL(或μg/L),ng/mL(或μg/L),ng/mL(或μg/L)各10 μL,注入石墨炉,测得其吸光值并求得吸光值与浓度关系的一元线性回归方程。 试样测定:分别吸取样液和试剂空白液各10 μL,注入石墨炉,测得其吸光值,代入标准系列的一元线性回归方程中求得样液中铅含量。

微波消解说明书

特别警示! 对于从未使用过本设备的人员,请认真阅读本说明书 重要提示 1、禁止使用高氯酸消解样品。 2、禁止使用浓硫酸消解样品。 3、禁止消解芳香类化合物。 4、消解时必须在微波炉腔内装入消解罐,仪器不可以空载。 5、消解罐使用期限正常为3年,超过期限请更换全新消解罐。 一、微波简介 微波是一种频率在300MHz~300GHz,即波长在100cm~0.1 cm 范围内的电磁波。它位于电磁波谱的红外光谱和无线电波之间。通常用来加热的频率是2450±50 MHz,其性能近似太阳光,波速与光速相同,波长为12.24cm,振荡频率为每秒24.5亿次。二、微波加热的机理 微波加热主要是通过分子极化和离子导电两个效应对物质直接加热。物质吸收的能量迅速使其在分子和均匀加热介质间的重新分配,因此微波加热是一种“内加热”。诸如含水或酸的物质分子都是有极性的,这些极性分子在微波电场的作用下,以每秒24.5亿次的速率不断改变其正负方向,使分子产生高速的碰撞和摩擦,而形成高热;同时一些无机酸类物质溶于水后,分子电离成离子,在微波电场作用下,离子定向流动,形成离子电流,离子在流动过程中与周围的分子和离子发生高速摩擦和碰撞,使微波能转化成热能。 三、密闭增压消解的原理 密闭增压是指试样和消解溶液在密闭容器里通过微波的快速加热,使试样在高温高压下,表面层搅动、破裂,不断产生新的试样表面与溶液接触,直至试样消解完毕。

四、仪器安装与放置: 1、 建议不要把主机整体放入通风橱内,建议把排风管引入通风橱内,保证良好排风,若没有通风橱,也可以通过通 风管道把气体排到室外。 2 220V ,负载大于2.5KW 3 顶部安装排风机(后面上部寻找风机电源插孔) 五、面板及使用操作 1、基本功能介绍 1.1、控制按键页面 1.2、开机液晶屏3秒后显示画面 门锁 (使用时务必锁上) 温度传感器安装位 压力导气管安装位 控制面板 消解罐放置板 电源开关 炉门

微波消解技术在测定食品微量元素中的应用

微波消解技术在测定食品微量元素中的应用 发表时间:2010-05-04T11:11:37.437Z 来源:《中国医药卫生》杂志作者:杨福军[导读] 杨福军 洮南市疾病预防控制中心(吉林洮南137100)【摘要】目的探讨微波消解技术在测定食品微量元素中的优越性。方法依据GB/T5009-2006 ,以铜为例,微波消解与干、湿法消解对测定食品微量元素结果的影响。结果微波消解与干、湿法消解在测定结果上无显著性差异。结论微波消解技术在测定食品微量元素中具有广泛的应用。【关键词】微波消解微量元素应用[中图分类号]R151.1 [文献标识码]A [文章编号]1810-5734(2009)12-0086-02食品是人类生存的基本要素,但是食品中可能含有被污染有危害人体健康的物质。准确测定食品中的微量或痕量元素的关键是样品的前处理,样品的前处理直接影响分析结果的精密度和准确度。在先进的分析仪器被广泛应用时,再用陈旧、落后的传统消化加热来处理样品,就会严重制约现代分析仪器的先进性。传统的干法灰化和湿法灰化操作时间长,挥发元素易损失、易污染环境。而微波消解可以克服易挥发元素的损失,微波消解具有操作简便、溶剂用量少以及样品消化完全的特点。因此微波消解作为一种全新的样品处理方法已在食品、环境、化妆品等行业得到了广泛应用。 1 微波消解原理 微波是一种频率范围在300-300000兆HZ的电磁波,具有内加热及吸收极化作用等传统加热方式不具备的独特优点。微波产生的电场作用于极性分子,分子即以每秒24.5亿次的速度不断改变正负方向,分子间高速碰撞和摩擦,产生高热。样品因微波作用表面层不断搅动破裂,产生新鲜表面与酸反应,促使样品迅速消解。 [1] 2 材料与方法 [2] 2.1 仪器与设备: TAS-986原子吸收分光光度计(北京普析仪器有限责任公司);MDS-2002A微波密闭消解炉(上海新仪微波化学科技有限公司)。 2.2 试剂 2.2.1 硝酸(优级纯) 2.2.2 过氧化氢(分析纯) 2.2.3 标准铜溶液:1.0g/mL(国家标准物质研究中心) 2.3 样品消化:准确称取0.3-0.5g样品于消化罐内,加入一定量的硝酸和过氧化氢,晃动几次,使样品与消化液混合均匀,放置过夜或水浴上加热0.5h,之后将内胆放入消化罐中,旋好盖子。根据不同样品设置不同的控制压力与时间,消化后待冷却至常压,开罐取出消化液定容,待测。 2.4 测定:以铜为指标,依据GB/T5009.13-2003,用原子吸收法测定。 3 结果与讨论