Table 6-2: 840 Ventilator diagnostic codes
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AE0110Unable to
establish O2
flow EST Flow
sensors cross
check Test
Alert Flow controller unable
to establish and control
oxygen flow at
120 L/min.
1.Make sure oxygen supply is
connected and unrestricted.
2.Verify oxygen regulator pres-
sure is set between 9-12 psi.
3.Switch PSOLs to see if failure
transfers to the other gas side.
If yes, return the air PSOL to its
original position and replace
the oxygen PSOL.
4.Run the leak test in EST to
check for leaks/occlusions.
5.Perform a flow sensor calibra-
tion.
6.Replace Q1.
7.Replace the Inspiratory Elec-
tronics PCB.
8.Replace the AI PCB.
AE0111Unable to
establish air flow EST Flow
sensors cross
check Test
Alert Flow controller unable
to establish and control
air flow at 120 L/min.
1.Make sure air supply is con-
nected and unrestricted.
2.Verify air regulator pressure is
set between 9-12 psi.
3.Switch PSOL's to see if failure
transfers to the other gas side.
If yes, return Oxygen PSOL to
its original position and replace
the Air PSOL.
4.Run the leak test in EST to
check for leaks/occlusions.
5.Perform a flow sensor calibra-
tion.
6.Replace Q2.
7.Replace the Inspiratory Elec-
tronics PCB.
8.Replace the AI PCB.
AE0213Air PSOL leak EST Gas
Supply/SV Test Alert Air PSOL (PSOL2)
forward leak was
detected via excessive
pressure buildup
(>50 cmH2O but
<100cmH2O) in
blocked inspiratory
module.
1.Check for leaks around the Q2
O-rings.
2.Make sure air supply is con-
nected and unrestricted.
3.Reseat Air PSOL.
4.Replace Air PSOL.
5.Replace AI PCB.
AE0214O2 PSOL leak EST Gas
Supply/SV Test Alert Oxygen PSOL (PSOL1)
forward leak was
detected via excessive
pressure buildup
(>50 cmH2O but
<100cmH2O) in
blocked inspiratory
module.
1.Check for leaks around the Q1
O-rings.
2.Make sure oxygen supply is
connected and unrestricted.
3.Reseat oxygen PSOL.
4.Replace oxygen PSOL.
5.Replace AI PCB.
AE0306Test circuit not
connected EST Leak Test Alert Pressure not detected
on expiratory side
1.Make sure test circuit is prop-
erly connected.
2.Replace expiratory bacteria
filter.
3.Check for leaks around the Q3
flow sensor.
4.Check/replace the exhalation
valve.
5.Replace the expiratory pressure
transducer PCB.
AE0601GUI High Alarm
LED fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace the GUI LED PCB.
AE0602GUI Medium
Alarm LED fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace the GUI LED PCB.
AE0603GUI Low Alarm
LED fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace the GUI LED PCB.
AE0604GUI Normal LED
fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace the GUI LED PCB.
AE0605GUI Batt Backup
LED fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB
2.Replace the GUI LED PCB.
AE0606GUI On Batt Pwr
LED fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace the GUI LED PCB.
AE0607G UI Compressor
Ready LED fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace the GUI LED PCB.
AE0608G UI Compressor
Operating LED
fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace the GUI LED PCB.
AE0609GUI 100% O2
LED fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace keyboard.
AE0610GUI Alarm
Silence LED fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace the GUI LED PCB.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AE0611GUI Screen Lock
LED fails.EST GUI Lamp
Test
Alert CLEAR key pressed to
indicate LED not on.
1.Check/replace interconnect
cable between the GUI LED
PCB and the GUI CPU PCB.
2.Replace the GUI LED PCB.
AE0702Bad Vent inop
LED EST BD Lamp
Test
Alert CLEAR key pressed to
indicate one or both
ventilator inoperative
LEDs not on.
Replace BD LED PCB.
AE0703Bad SVO LED EST BD Lamp
Test Alert CLEAR key pressed to
indicate one or both
SVO LEDs not on.
Replace BDU LED PCB.
AE0704Bad Loss of GUI
LED EST BD Lamp
Test
Alert CLEAR key pressed to
indicate loss of GUI LED
not on.
Replace BDU LED PCB.
AE1001Air PSOL
loopback
current OOR EST PSOL
Loopback Test
Alert Air PSOL (PSOL2)
loopback current out of
range of drive current.
1.Verify that the air supply is
good.
2.Verify air regulator pressure set
to between 9 and 12 psi.
3.Switch PSOLs to see if failure
transfers to the other gas side.
If yes, return oxygen PSOL
(PSOL1) to its original position
and replace PSOL2.
4.Switch Q1 and Q2, run a flow
sensor calibration and rerun
test. If the problem transfers to
the other gas side, return Q1 to
its original position and replace
Q2.
5.Replace the AI PCB.
6.Replace the Inspiratory
Electronics PCB.
AE1002O2 PSOL
loopback
current OOR
EST PSOL
Loopback Test
Alert Oxygen PSOL (PSOL1)
loopback current out of
range of drive current.
1.Verify that the oxygen supply is
good.
2.Verify oxygen regulator pres-
sure set to between 9 and 12
psi.
3.Switch PSOLs to see if failure
transfers to the other gas side.
If yes, return the air PSOL to its
original position and replace
PSOL1.
4.Switch Q1 and Q2, run a flow
sensor calibration and rerun
test. If the problem transfers to
the other gas side, return Q2 to
its original position and replace
Q1.
5.Replace the AI PCB.
6.Replace the Inspiratory
Electronics PCB.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AE1104Insp check valve
test failed EST Safety
System Test
Alert It took too little time to
relieve excess pressure
through open safety
valve, indicating
inspiratory check valve
(CV3) may be damaged
or incorrectly mounted.
1.Make sure test circuit is con-
nected.
2.Make sure CV3 is not installed
backward.
3.Replace CV3.
AE1201Exp valve
loopback
current OOR EST Exp Valve
Loopback Test
Alert Exhalation valve
loopback current is out
of range of drive
current.
1.Verify that the system has no
leaks or occlusions by running
the leak test in EST.
2.Clean exhalation valve
diaphragm.
3.Calibrate the exhalation valve.
4.Clean/replace the exhalation
valve.
5.Replace the AI PCB.
AE1305Seal test failed EST Exp Valve
Seal Test Alert Seal test Δ pressure is
above alert level but
below failure level.
1.Verify that the system has no
leaks or occlusions by running
the leak test in EST.
2.Clean the exhalation valve.
3.Calibrate exhalation valve.
4.Replace the exhalation valve.
5.Replace the AI PCB.
AE1600Compressor
Test - Not
installed EST
Compressor
Test
Status Ventilator did not sense
a compressor attached
and skipped test.
No action required.
AE1601Wall air pressure
detected EST
Compressor
Test
Alert Wall air pressure switch
(PS2) detected air after
user was prompted to
disconnect air.
1.Make sure air supply is discon-
nected.
2.Disconnect PS2 and rerun test.
If test passes, replace PS2.
AE1602ac power not
connected EST
Compressor
Test
Alert System is still running
on battery power after
prompting user to
connect ac power.
Compressor can only
run on ac (facility)
power.
1.Plug in ventilator power cord
and check the cord connec-
tion at the ventilator.
2.Check ac.
3.Disconnect BPS to isolate prob-
lem.
4.Replace power supply.
AE1603Compressor
pressure not
detected
EST
Compressor
Test
Alert Compressor pressure
transducer (PC)
indicates that
compressor air is not
present, although
compressor motor is
on.
1.Run compressor leak test to
check for leaks.
http://www.wendangku.net/doc/a2627e20aaea998fcc220ec5.htmle leak detector to check
for leaks at the accumulator
fittings and other tubing
connections.
b.Verify no leaks at CV2 within
the ventilator by plugging
the air inlet fitting.
2.Replace compressor PCB.
3.Contact Puritan-Bennett Tech-
nical Support.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AE1604Run mode time
OOR EST
Compressor
Test
Alert Compressor timer is not
running while
compressor motor is
on.
Listen for motor. If motor is on,
replace compressor PCB.
Otherwise, replace compressor.
AE1606Compressor
pressure
detected EST
Compressor
Test
Alert Compressor pressure
transducer (PC)
indicates that
compressor air is
present after
compressor motor is
disabled and
accumulator drained.
Replace compressor PCB.
AE1607Disabled mode
time OOR EST
Compressor
Test
Alert Compressor timer is
running while
compressor motor is
off.
Replace compressor PCB.
AE1608Unable to test
standby mode EST
Compressor
Test
Alert Compressor standby
mode check cannot be
run, because of inability
to verify timer
functionality during
compressor run and
disabled tests.
1.Troubleshoot code AE1604 or
AE1607, if present.
2.Replace compressor PCB.
AE1609Standby mode
time OOR EST
Compressor
Test
Alert Compressor motor is
still running. During
standby mode test
phase, compressor
motor should
eventually turn off.
Replace compressor PCB.
AE1610Unable to
perform
compressor load
test EST
Compressor
Test
Alert Compressor load test
cannot be run, because
of inability to verify PC
during compressor run
and disabled tests.
Troubleshoot code AE1603 or
AE1606.
AE1611Compressor
load test failed EST
Compressor
Test
Alert Compressor unable to
maintain minimum
pressure under worst-
case breath delivery
waveform.
1.Run compressor leak test to
check for compressor leak.
http://www.wendangku.net/doc/a2627e20aaea998fcc220ec5.htmle leak detector to check
for leaks at the accumulator
fittings and other tubing
connections.
b.Verify no leaks at CV2 within
the ventilator by plugging
the air inlet fitting.
2.Replace compressor.
3.Replace compressor PCB.
AE1700Compressor
Leak Test - Not
installed
EST
Compressor
Leak Test
Status Ventilator did not sense
a compressor attached
and skipped test.
No action required. Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AE1701Unable to
perform
compressor leak
test EST
Compressor
Leak Test
Alert Compressor leak test
cannot be run, because
of inability to verify
timer functionality.
1.Troubleshoot accompanying
code AE1604 or AE1607.
2.Replace compressor PCB.
AE1702Wall air pressure
detected EST
Compressor
Leak Test
Alert Wall air pressure switch
(PS2) detected air
presence after user was
prompted to
disconnect air.
1.Make sure air supply is discon-
nected.
2.Disconnect PS2 and rerun test.
If test passes, replace PS2.
AE1703Wall air pressure
not detected EST
Compressor
Leak Test
Alert Wall air pressure switch
(PS2) did not detect air
presence after user was
prompted to connect
air.
1.Make sure air supply is con-
nected.
2.Run Gas Supply/SV Test.
AE1704Compressor
leak detected EST
Compressor
Leak Test
Alert Compressor leak was
detected using
compressor timer to
detect compressor
turning on momentarily
during test.
Troubleshoot compressor
compartment or inspiratory
module for leaks.
AE1901GUI touch: Error EST GUI Touch
Test Alert Touch screen error
occurred (e.g., blocked
beam or low-level
error), buffer cannot be
read, or report is
invalid.
1.Clean touch screen, removing
any obstructions.
2.Replace touch frame PCB.
3.Replace GUI CPU PCB.
AE2001Bad GUI serial
port EST GUI Serial
Port Test
Alert While in loopback
mode, failed to verify
received message was
identical to transmitted
message.
Replace GUI CPU PCB.
AE2101Battery not
charged EST Battery
Test
Alert BPS not fully charged at
start of test.
1.Allow BPS to fully charge, then
repeat test.
2.Replace BPS PCB.
3.Replace battery pack.
4.Replace BD CPU PCB.
AE2102Battery not
discharging
EST Battery
Test
Alert BPS not discharging
after ac power was
disconnected.
1.Verify ac power is disconnected
when prompted.
2.Replace BPS PCB.
3.Replace battery pack.
4.Replace BD CPU PCB.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AE2103Bad Backup
Power Supply EST Battery
Test
Alert While BPS was
discharging, BPS
voltage dropped below
accepted level or
dropped too quickly.
1.Ensure that the battery pack is
fully charged by checking that
the green LED on the BPS is lit
prior to initiating an EST. If the
amber LED is lit, allow the unit
to charge the batteries prior to
rerunning EST.
2.Replace the battery pack.
3.Replace the BPS PCB.
4.Replace the AI PCB.
5.Replace the power supply.
AE2104Battery not
charging EST Battery
Test
Alert BPS not charging after
ac power was
reconnected.
1.Verify that ac power is recon-
nected when prompted and
that ac is good.
2.Replace BPS PCB.
3.Replace the battery pack.
4.Replace the power supply.
5.Replace the BD CPU.
6.Replace the AI PCB.
AE2300GUI Nurse Call
Test - Not
installed
EST Test Status User pressed CLEAR to
indicate nurse’s call
device not installed.
Test was skipped.
No action required. Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AS0010Unable to
establish O2
flow
SST flow sensor
test
Alert Flow controller unable
to establish and control
oxygen flow at
120 L/min.
1.Make sure oxygen supply is
connected and unrestricted.
2.Verify that the oxygen supply
meets minimum pressure
requirements.
3.Verify that the patient circuit
system has no leaks or occlu-
sions. If not sure, run the first
four tests in EST to get to the
leak test. If the leak test passes
in EST, exit out of EST. Correct
the leak or occlusion in the
patient circuit used in SST and
rerun or try another patient cir-
cuit.
4.Check the oxygen inlet filter
assembly.
5.Verify that the oxygen regula-
tor pressure is set between 9-
12 psi.
6.Run a flow sensor calibration.
7.Switch Q1 and Q2. Rerun the
flow sensor calibration and
EST. If the failure transfers to
the air side, return Q2 back to
its original position and replace
Q1.
8.Switch PSOLs to see if the fail-
ure transfers to the air side. If
yes, return PSOL2 back to its
original position and replace
PSOL1.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AS0011Unable to
establish air flow SST flow sensor
test
Alert 1.Flow controller
unable to establish
and control air flow
at 120 L/min.
1.Make sure air supply is con-
nected and unrestricted.
2.Verify that the air supply meets
minimum pressure require-
ments.
3.Verify that the patient circuit
system has no leaks or occlu-
sions. If not sure, run the first
four tests in EST to get to the
leak test. If the leak test passes
in EST, exit out of EST. Correct
the leak or occlusion in the
patient circuit used in SST and
rerun or try another patient cir-
cuit.
4.Check the air inlet filter (F2).
5.Verify that the air regulator
pressure is set between 9-12
psi.
6.Run a flow sensor calibration.
7.Switch Q1 and Q2. Rerun the
flow sensor calibration and
EST. If the failure transfers to
the oxygen side, return Q1
back to its original position and
replace Q2.
8.Switch PSOLs to see if the fail-
ure transfers to the oxygen
side. If yes, return PSOL1 back
to its original position and
replace PSOL2.
AS0012O2 pressure not
detected SST flow sensor
test
Alert Oxygen pressure not
detected via PS1. Only
air available for SST.
Make sure oxygen supply is
connected.
AS0013Wall air pressure
not detected SST flow sensor
test
Alert Air pressure not
detected via PS2 or PC.
Only oxygen available
for SST.
Make sure air supply is connected.
AS0202Excessive leak SST Circuit leak Alert Pressure drops to 85
cmH2O in 10 s,
then in 10 s more by
≥ 10 cmH2O.1.Make sure patient circuit is
connected and is not leaking.
2.Check exhalation valve opera-
tion.
3.Verify connections of bacteria
filters and humidifier are
secure.
AS0305Occluded
expiratory filter SST Expiratory
filter
Alert Pressure drop across
filter > 2cmH2O, but
<3 cmH2O.
Consider replacing expiratory
filter.
AS0308Occluded
expiratory
compartment
SST Expiratory
filter
Alert Exhalation
compartment pressure
>3 cmH2O but
< 4cmH2O.
1.Check compartment for
obstruction.
2.Verify exhalation valve opera-
tion by running EST.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AS0311Low expiratory
filter ΔP SST Expiratory
filter
Alert Pressure drop
across expiratory filter
< 0.4 cmH2O, but
> 0.1 cmH2O.
1.Repeat test, following direc-
tions more closely.
2.Replace filter.
AS0403Occluded
inspiratory limb SST Circuit
Resistance
Alert Inspiratory limb
pressure > 8.5 cmH2O
(adult) or 5.5cmH2O
(pediatric), but
< 12.5 cmH2O (adult)
or 7.5cmH2O
(pediatric), indicating
occlusion.
Check for occluded patient tubing.
AS0406Occluded
exhalation limb SST Circuit
Resistance
Alert Expiratory limb
pressure > 8.5cmH2O
(adult) or 5.5cmH2O
(pediatric) but > 12.5
cmH2O (adult) or
7.5cmH2O (pediatric),
indicating occlusion.
Check for occluded patient tubing.
AS0407Unable to reach
min peak flow SST Circuit
Resistance
Alert During characterization
of total circuit
resistance over a range
of flows, peak flow
< 80L/min but > 60 L/
min.
Check for kinked or occluded
patient tubing.
AS0411Unable to reach
min peak flow SST Circuit
Resistance
Alert During characterization
of total circuit
resistance over a range
of flows, peak flow was
less than alert
threshold.
Check for kinked or occluded
patient tubing.
AS0413Insp limb
resistance low SST Circuit
Resistance
Alert Inspiratory limb
pressure < 0.6 cmH2O
(adult) or 5.5cmH2O
(pediatric), but
> 0.2 cmH2O.
1.Make sure inspiratory filter is
installed.
2.Repeat test, following direc-
tions more closely.
3.Replace inspiratory filter.
4.Replace patient circuit.
AS0414Exp limb
resistance low SST Circuit
Resistance
Alert Expiratory limb
pressure < 0.6cmH2O
(adult) or 0.5 cmH2O
(pediatric), but
> 0.2cmH2O.
Replace patient circuit.
AS0505Excessive
compliance
SST
Compliance
calibration
Alert High compliance
>6 mL/cmH2O (adult)
or 4.5mL/cmH2O
(pediatric), but
< 12mL/cmH2O (adult)
or 9 mL/cmH2O
(pediatric).
1.Make sure correct patient tub-
ing type was specified in SST.
2.Replace patient circuit. Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
AS0507Compliance low SST
Compliance
calibration Alert High compliance falls
below 1.56 mL/cmH2O
(adult) or 1.34 mL/
cmH2O (pediatric) but
not lower than
1.05 mL/cmH2O.
1.Make sure correct patient tub-
ing type was specified in SST.
2.Replace patient circuit with a
known good circuit and filter
set.
3.Run EST to check the operation
of PI and PE during the Circuit
Pressure Test. Replace applica-
ble pressure transducer.
4.Run Atmospheric Pressure
Transducer calibration.
AS0509Excessive
compliance SST
Compliance
calibration
Alert Low compliance >
6 mL/cmH2O (adult) or
4.5mL/cmH2O
(pediatric), but <
12mL/cmH2O (adult)
or 9 mL/cmH2O
(pediatric).
1.Make sure correct patient tub-
ing type was specified in SST.
2.Replace patient circuit with a
known good circuit and filter
set.
3.Run EST to check the operation
of PI and PE during the Circuit
Pressure Test. Replace applica-
ble pressure transducer.
4.Run Atmospheric Pressure
Transducer calibration.
AS0511Compliance low SST
Compliance
calibration Alert Low compliance falls
below 1.56 mL/cmH2O
(adult) or 1.34 mL/
cmH2O (pediatric), but
not below
1.05 mL/cmH2O.
1.Make sure correct patient tub-
ing type was specified in SST.
2.Replace patient circuit with a
known good circuit and filter
set.
3.Run EST to check the operation
of PI and PE during the Circuit
Pressure Test. Replace applica-
ble pressure transducer.
4.Run Atmospheric Pressure
Transducer calibration.
Dxxxxx Assertion Background
Checks (BD)Failure that
results in a
POST or a
reset
System generated a
reset to correct a
boundary check or
possible data
corruption of control
variables.
Replace the BDCPU PCB.
DT0002Bus error /
Access fault Background
Checks (BD)
Failure that
results in a
POST or a
reset
Hardware trap from an
access fault due to a
bad memory chip, bad
control line, or a
hardware timing issue.
Replace the BD CPU PCB.
Exxxxx Varies Background
Checks (GUI)Status A status message
indicating an event that
was intentionally
caused.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
NOTE:
For all "FE" codes (failures reported during EST), diagnose the problem and perform the repair.
Run a complete EST to reset the EST test failure in memory. If there is a failed EST test result in
memory, normal operation is prevented until the failed EST test is rerun and passes.
FE0001Inspiratory
autozero out of
range EST Circuit
Pressure Test
Failure Inspiratory pressure
transducer ADC count
at 0cmH2O is out of
range.
1.Replace SOL1.
2.Replace Inspiratory Electronics
PCB.
3.Replace AI PCB.
FE0002Expiratory
autozero out of
range EST Circuit
Pressure Test
Failure Exhalation pressure
sensor ADC count at
0cmH2O is out of
range.
1.Replace SOL
2.
2.Replace exhalation transducer
PCB.
3.Replace AI PCB.
FE0003Failed to reach
test pressure EST Circuit
Pressure Test
Failure Unable to build
pressure (air) to
10cmH2O at
5 L/min within time-out
period.
1.Make sure that test circuit is
installed and air supply or com-
pressor is available and good.
2.Verify that the air regulator
pressure is set between 9-12
psi.
3.Verify that the patient circuit
system has no leaks or occlu-
sions by running the first four
tests in EST to get to the leak
test.
FE0004Cross-check
failed EST Circuit
Pressure Test
Failure Inspiratory/expiratory
pressure transducer
readings at 10cmH2O
test pressure are too far
apart.
1.Verify that the system has no
leaks or occlusions. If not sure,
run the first four tests in EST to
run the leak test
2.If the leak test passes in EST,
exit out of EST. Review the
data for the Circuit Pressure
Test to determine which pres-
sure transducer is out of range:
PI or PE. Replace applicable
pressure transducer.
3.Replace AI PCB.
FE0005Bad insp
autozero
solenoid EST Circuit
Pressure Test
Failure Inspiratory pressure
reading (taken after
inspiratory pressure
transducer autozero
solenoid (SOL1)
actuated) out of range
(-0.60 to 0.60 cmH2O).
1.Replace SOL1.
2.Replace Inspiratory Electronics
PCB.
3.Replace AI PCB.
FE0006Bad exp
autozero sol
EST Circuit
Pressure Test
Failure Expiratory pressure
reading (taken after
expiratory pressure
transducer autozero
solenoid (SOL2)
actuated) out of range
(-0.60 to 0.60 cmH2O).
1.Replace SOL
2.
2.Replace exhalation transducer
PCB.
3.Replace AI PCB.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
FE0007Cross-check
failed EST Circuit
Pressure Test
Failure Inspiratory/expiratory
pressure transducer
measurements at 50 or
100cmH2O test
pressure are too far
apart.
1.Replace the expiratory bacteria
filter.
2.Verify that the system has no
leaks or occlusions. If not sure,
run the first four tests in EST to
run the leak test.
3.If the leak test passes in EST,
exit out of EST. Review the data
for the Circuit Pressure Test to
determine which pressure
transducer is out of range: PI or
PE. Replace applicable pres-
sure transducer.
4.Replace AI PCB.
FE0008Failed to reach
test pressure EST Circuit
Pressure Test
Failure Unable to build
pressure (air) to 50 or
100 cmH2O at 5 L/min
within time-out period.
1.Make sure air supply or com-
pressor is available.
2.Check for system leak at the
expiratory bacteria filter or O2
sensor. If not sure, run the first
four tests in EST to run the leak
test.
FE0010ac power not
connected EST Circuit
Pressure Test
Failure System is still running
on battery power after
prompting user to
connect ac power. EST
can only run on ac
(facility) power.
1.Plug in ventilator power cord.
2.Disconnect BPS to isolate prob-
lem.
3.Replace power supply.
FE0101O2 flow sensor
cross check
failed
EST Flow
sensors cross
check Test
Failure Oxygen flow sensor
(Q1) cross-check failed.
1.Make sure oxygen supply is
connected and unrestricted.
2.Verify that the oxygen supply is
good.
3.Verify that the system has no
leaks or occlusions. If not sure,
run the first four tests in EST to
run the leak test.
4.Run a flow sensor calibration.6.
Switch Q1 and Q2. Rerun the
flow sensor calibration and
EST. If the failure transfers to
the air side, return Q2 back to
its original position and replace
Q1.
5.Switch PSOLs to see if the fail-
ure transfers to the air side. If
yes, return PSOL2 back to its
original position and replace
PSOL1.
6.Replace exhalation flow sensor
(Q3)
7.Replace AI PCB.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
FE0102O2 PSOL current
out of range EST Flow
sensors cross
check Test
Failure Oxygen PSOL (PSOL1)
current is out of range
with respect to flow
sensor (Q1).
1.Make sure oxygen supply is
connected and unrestricted.
2.Verify that the oxygen supply is
good.
3.Verify that the system has no
leaks or occlusions. If not sure,
run the first four tests in EST to
run the leak test.
4.Switch PSOLs to see if the fail-
ure transfers to the air side. If
yes, return PSOL2 back to its
original position and replace
PSOL1.
5.Run a flow sensor calibration.
6.Switch Q1 and Q2. Rerun the
flow sensor calibration and
EST. If the failure transfers to
the air side, return Q2 back to
its original position and replace
Q1.
FE0103Air flow sensor
cross check
failed
EST Flow
sensors cross
check Test
Failure Inspiratory module air
flow sensor (Q2) cross-
check failed.
1.Make sure air supply is con-
nected and unrestricted.
2.Verify that the air supply is
good.
3.Verify that the system has no
leaks or occlusions. If not sure,
run the first four tests in EST to
run the leak test.
4.Run a flow sensor calibration.
5.Switch Q1 and Q2. Rerun the
flow sensor calibration and
EST. If the failure transfers to
the oxygen side, return Q1
back to its original position and
replace Q2.
6.Switch PSOLs to see if the fail-
ure transfers to the air side. If
yes, return PSOL back to its
original position and replace
PSOL2
7.Replace exhalation flow sensor
(Q3).
8.Replace AI PCB.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
FE0104Air PSOL current
out of range EST Flow
sensors cross
check Test
Failure Air PSOL (PSOL2)
current is out of range
with respect to air flow
sensor (Q2).
1.Make sure air supply is con-
nected and unrestricted.
2.Verify that the air supply is
good.
3.Verify that the system has no
leaks or occlusions. If not sure,
run the first four tests in EST to
run the leak test.
4.Switch PSOLs to see if the fail-
ure transfers to the air side. If
yes, return PSOL back to its
original position and replace
PSOL2
5.Run a flow sensor calibration.
6.Switch Q1 and Q2. Rerun the
flow sensor calibration and
EST. If the failure transfers to
the oxygen side, return Q1
back to its original position and
replace Q2.
FE0106Unable to
establish O2
flow EST Flow
sensors cross
check Test
Failure Flow controller unable
to establish and control
oxygen flow at 60, 5,
and 1 L/min.
1.Make sure oxygen supply is
connected.
2.Check regulated oxygen pres-
sure.
3.Replace PSOL1 or Q1.
4.Replace AI PCB.
FE0107Unable to
establish air flow EST Flow
sensors cross
check Test
Failure Flow controller unable
to establish and control
air flow at 60, 5, and 1
L/min.
1.Make sure air supply is con-
nected.
2.Check regulated air pressure.
3.Perform flow sensor calibra-
tion.
4.Replace PSOL2 or Q2.
5.Replace AI PCB.
FE0108O2 zero flow
check failed EST Flow
sensors cross
check Test
Failure Inspiratory flow >
0.153 L/min with
oxygen PSOL (PSOL1)
commanded to 0
(closed).
1.Verify no leaks at the Q1
O-rings.
2.Run EST Gas Supply Test to
check for PSOL1 leak. Remove
and then reseat PSOL1
3.Replace PSOL1.
4.Perform a flow sensor calibra-
tion.
5.Replace Q1.
FE0109Air zero flow
check failed
EST Flow
sensors cross
check Test
Failure Inspiratory flow >0.153
L/min with air PSOL
(PSOL2) commanded
to 0 (closed).
1.Verify no leaks at the Q2
O-rings.
2.Run EST Gas Supply Test to
check for PSOL2 leak. Remove
and then reseat PSOL2
3.Replace PSOL2.
4.Perform a flow sensor calibra-
tion.
5.Replace Q2.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
FE0204Wall air pressure
not detected EST Gas
Supply/SV Test
Failure Wall air not detected
initially and wall air still
not detected after user
was prompted to
connect wall air.
1.Make sure air supply is con-
nected.
2.Check PS2: Remove the wiring
connectors from PS2 and
jumper the connectors to each
other. If the system now rec-
ognizes the air supply, replace
PS2.
FE0205O2 pressure not
detected EST Gas
Supply/SV Test
Failure Oxygen pressure not
detected initially and
oxygen pressure still
not detected after user
was prompted to
connect oxygen.
1.Make sure oxygen supply is
connected.
2.Check PS1: Remove the wiring
connectors from PS1 and
jumper the connectors to each
other. If the system now rec-
ognizes the air supply, replace
PS1.
FE0206O2 pressure
detected (O2
not
disconnected)EST Gas
Supply/SV Test
Failure Oxygen pressure switch
detected presence of
oxygen after user was
prompted to
disconnect oxygen.
1.Make sure oxygen supply is dis-
connected.
2.Replace Inspiratory Electronics
PCB.
3.Replace PS1.
FE0207Air PSOL leak EST Gas
Supply/SV Test Failure Air PSOL (PSOL2)
forward leak was
detected via excessive
pressure buildup (>100
cmH2O) in blocked
inspiratory module.
1.Check for leaks around the Q2
O-rings.
2.Verify that the safety valve
relieves pressures above 100
cmH2O. Listen for the pressure
relief while watching the
numeric digital display of sys-
tem pressure in the upper
screen. If the safety valve does
not crack open, replace the
safety valve.
3.Replace PSOL2.
FE0208Wall air pressure
detected EST Gas
Supply/SV Test
Failure Wall air pressure switch
(PS2) detected wall air
after user was
prompted to
disconnect it.
1.Make sure air supply is discon-
nected.
2.Replace Inspiratory Electronics
PCB.
3.Replace PS2.
FE0209O2 PSOL leak EST Gas
Supply/SV Test Failure Oxygen PSOL (PSOL1)
forward leak was
detected via excessive
pressure buildup (>100
cmH2O) in blocked
inspiratory module.
1.Check for leaks around the Q1
O-rings.
2.Verify that the safety valve
relieves pressures above 100
cmH2O. Listen for the pressure
relief while watching the
numeric digital display of sys-
tem pressure in the upper
screen. If the safety valve does
not crack open, replace the
safety valve.
3.Replace PSOL1.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
FE0210SV pressure
relief failed EST Gas
Supply/SV Test
Failure Safety valve cracking
pressure and/or peak
steady-state pressure is
out of range.
1.Make sure To patient port is
blocked.
2.Check for leaks at the O2 sen-
sor, SOL1 or PI.
3.Replace the safety valve.
4.Replace PI.
FE0211O2 pressure not
detected EST Gas
Supply/SV Test
Failure PS1 did not detect
oxygen after user was
prompted to connect it.
1.Make sure oxygen supply is
connected.
2.Replace Inspiratory Electronics
PCB.
FE0212Compressor
pressure
detected EST Gas
Supply/SV Test
Failure Compressor pressure
transducer (PC)
detected that
compressor was
pressurized.
Replace compressor PCB.
FE0215Air zero flow
check failed EST Gas
Supply/SV Test
Failure During zero-flow check,
air flow sensor (Q2)
reads >0.05 L/min.
1.Make sure gas supplies are dis-
connected.
2.Make sure compressor is not
running.
3.Verify no leaks at the Q2
O-rings.
4.Check for PSOL2 leak.
5.Remove and then reseat
PSOL2.
6.Perform a flow sensor calibra-
tion.
7.Replace PSOL2.
8.Replace Q2.
FE0216O2 zero flow
check failed EST Gas
Supply/SV Test
Failure During zero-flow check,
oxygen flow sensor
(Q1) reads >0.05 L/
min.
1.Make sure gas supplies are dis-
connected.
2.Make sure compressor is not
running.
3.Verify no leaks at the Q2
O-rings.
4.Check for PSOL2 leak.
5.Remove and then reseat
PSOL2.
6.Perform a flow sensor calibra-
tion.
7.Replace PSOL2.
8.Replace Q2.
FE0217Exp zero flow
check failed
EST Gas
Supply/SV Test
Failure During zero-flow check,
exhalation flow sensor
(Q3) reads >0.1 L/min.
1.Make sure gas supplies are dis-
connected.
2.Make sure compressor is not
running.
3.Verify no leaks at the Q3
O-rings.
4.Perform a flow sensor calibra-
tion.
5.Replace Q3.
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
FE0301Excessive leak EST Leak Test Failure System pressure
dropped below failure
pressure level.1.Make sure test circuit is con-
nected and is not leaking.
2.Check for system leaks or
occlusions especially at the O2
sensor and expiratory filter. 3.Verify secure connection of
expiratory filter.
4.Check exhalation valve opera-
tion.
FE0305Unable to
establish
pressure EST Leak Test Failure System cannot attain
leak test starting
pressure using oxygen
or air within time-out
period.
1.Make sure test circuit is con-
nected and is not leaking.
2.Check exhalation valve opera-
tion, and verify secure connec-
tion of expiratory filter.
FE0401Accept key fails.EST GUI
Keyboard Test Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0402Clear key fails.EST GUI
Keyboard Test Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0403Insp. Pause key
fails.EST GUI
Keyboard Test
Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0404Exp. Pause key
fails.EST GUI
Keyboard Test
Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0405Man Insp fails.EST GUI
Keyboard Test Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0406100% O2/CAL
key fails.EST GUI
Keyboard Test
Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0407Info key fails.EST GUI
Keyboard Test Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0408Alarm Reset key
fails.EST GUI
Keyboard Test
Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0409Alarm Silence
key fails.EST GUI
Keyboard Test
Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0410Alarm Volume
key fails.EST GUI
Keyboard Test
Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0411Screen
brightness fails.
EST GUI
Keyboard Test
Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard. Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action
FE0412Screen contrast
key fails.EST GUI
Keyboard Test
Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0413Screen lock key
fails.EST GUI
Keyboard Test
Failure Wrong key pressed or
key not pressed within
15 s.
1.Repeat test.
2.Replace keyboard.
FE0501Bad knob EST GUI Knob
Test Failure Knob was not turned in
direction as prompted
within 15 s.
1.Repeat test.
2.Replace keyboard.
FE0801SAAS
(Safety Audible
Alarm System)
test failed EST GUI Audio
Test
Failure CLEAR key pressed to
indicate GUI audio
diagnostic failed.
1.Verify the GUI alarm cable con-
nection to the GUI CPU PCB.
2.Replace GUI alarm assembly.
3.Replace GUI CPU PCB.
FE0901Bad alarm cable EST BDU Audio
Test Failure Alarm cable voltage is
out of range (<3.5 or
>5.05 V).
1.Make sure BD alarm cable is
connected.
2.Replace BD alarm.
3.Replace AI PCB.
4.Replace BD alarm cable.
FE0902Bad power fail
cap EST BDU Audio
Test
Failure Power failure capacitor
initial voltage is out of
range (<4.5 or >5.05
V).
1.Make sure BD alarm cable is
connected.
2.Replace BD alarm.
3.Replace AI PCB.
4.Replace BD alarm cable.
5.Replace Motherboard PCB.
FE0903Bad power fail
cap EST BDU Audio
Test
Failure Power failure capacitor
final voltage is out of
range or RC constant
<60 s.
1.Make sure BD alarm cable is
connected.
2.Replace BD alarm.
3.Replace AI PCB.
FE0904Bad BD audio EST BDU Audio
Test Failure CLEAR key pressed to
indicate user did not
hear alarm, although
alarm was active.
1.Make sure BD alarm cable is
connected.
2.Replace BD alarm.
3.Replace AI PCB.
4.Replace BD alarm cable.
FE1101Safety valve
occluded EST Safety
System Test
Failure Excessive safety valve
back pressure when
safety valve is open.
Replace safety valve.
FE1102Bad safety valve
driver or
loopback EST Safety
System Test
Failure Safety valve loopback
current is out of range
during one or more of
timed test points.
1.Replace safety valve.
2.Replace AI PCB.
3.Replace Inspiratory Electronics
PCB.
FE1103Insp check valve
test failed EST Safety
System Test
Failure It took too long to
relieve excess pressure
through open safety
valve, indicating
inspiratory check valve
(CV3) is occluded or
test circuit is too large.
1.Make sure proper test circuit is
used.
2.Make sure CV3 is not installed
backward.
3.Replace CV3.
4.Replace Safety Valve (SV).
Table 6-2: 840 Ventilator diagnostic codes (continued)
Code Message
displayed
Test Response
Information / Possible
cause
Corrective action