Table of Contents

1.0 Purpose (8)

1.1 Scope (8)

2.0 References (8)

3.0 Definitions (9)

3.1 Support of Extended Character Sets (9)

4.0 Implementation Model (10)

4.1 Application Data Flow Diagrams (10)

4.1.1 Verification (10)

4.1.2 Storage (11)

4.1.3 Storage Commitment - Push Model (11)

4.1.4 Print (12)

4.1.5 Modality Worklist (12)

4.1.6 Modality Performed Procedure Step (13)

4.2 AE Functional Definitions (15)

4.2.1 Verification Real-World Activities (15)

4.2.2 Storage Real-World Activities (15)

4.2.3 Storage Commitment Real-World Activities (15)

4.2.4 Print Real-World Activities (15)

4.2.5 Worklist Real-World Activity (16)

4.2.6 Modality Performed Procedure Step Real-World Activity (17)

4.2.7 Sequencing of Real-World Activities (17)

5.0 AE Specifications (18)

5.1 Verification AE Specification (19)

5.1.1 Association Initiation by Real-World Activity (19)

5.1.1.1 Real-World Activity - Verification SCU/SCP (19)

5.1.1.1.1 Associated Real-World Activity - Verification SCU/SCP (19)

5.1.1.1.2 Proposed Presentation Contexts - Verification SCU/SCP (19)

5.1.1.1.3 SOP Specific Conformance Statement - Verification SCU/SCP (19)

5.2 Storage AE Specification (20)

5.2.1 Association Establishment Policies (20)

5.2.1.1 General (20)

5.2.1.2 Number of Associations (20)

5.2.1.3 Asynchronous Nature (20)

5.2.1.4 Implementation Identifying Information (20)

5.2.1.5 Maximum PDU Size offered (20)

5.2.1.6 DICOM Application Context (20)

5.2.2 Association Initiation by Real-World Activity (21)

5.2.2.1 Real World Activity - Storage SCU (21)

5.2.2.1.1 Associated Real World Activities - Storage SCU (21)

5.2.2.1.2 Proposed Presentation Context (21)

5.2.2.1.3 SOP Specific Conformance to Storage Service SOP Classes (22)

5.2.2.1.4 Error Handling (40)

5.3 Storage Commitment AE Specification (41)

5.3.1 Association Establishment Policies (41)

5.3.1.1 General (41)

5.3.1.2 Number of Associations (41)

5.3.1.3 Asynchronous Nature (41)

5.3.1.4 Implementation Identifying Information (41)

5.3.2 Association Initiation by Real-World Activity (41)

5.3.2.1 Real-World Activity - Send Storage Commitment Request (41)

5.3.2.1.1 Associated Real-World Activity - Send Storage Commitment Request (41)

5.3.2.1.2 Proposed Presentation Contexts - Send Storage Commitment Request (42)

5.3.2.1.3 SOP Specific Conformance Statement - Send Storage Commitment Request 42

5.3.3 Association Acceptance Policy (42)

5.3.3.1 Real World Activity - Receive Storage Commitment Response (42)

5.3.3.1.1 Associated Real World Activity - Receive Storage Commitment Response (42)

5.3.3.1.2 Accepted Presentation Contexts - Receive Storage Commitment Response ..43

5.3.3.1.3 SOP Specific Conformance Statement - Receive Storage Commitment Response

(43)

5.4 Print AE Specification (44)

5.4.1 Association Establishment Policies (45)

5.4.1.1 General (45)

5.4.1.2 Number of Associations (45)

5.4.1.3 Asynchronous Nature (45)

5.4.1.4 Implementation Identifying Information (45)

5.4.2 Association Initiation by Real-World Activity (45)

5.4.2.1.1 Associated Real World Activities - Print (46)

5.4.2.1.2 Proposed Presentation Contexts (47)

5.4.2.1.2.1 SOP Specific Conformance Statement (47)

5.4.2.1.2.1.1 SOP Specific Conformance to Basic Film Session SOP Class (48)

5.4.2.1.2.1.2 SOP Specific Conformance to Basic Film Box SOP Class (49)

5.4.2.1.2.1.3 SOP Specific Conformance to Basic Grayscale Image Box SOP

Class (51)

5.4.2.1.2.1.4 SOP Specific Conformance to Basic Color Image Box SOP Class52

5.4.2.1.2.1.5 SOP Specific Conformance to Printer SOP Class (53)

5.4.2.1.2.1.6 SOP Specific Conformance to Print Job SOP Class (54)

5.5 Modality Worklist AE Specification (56)

5.5.1 Association Establishment Policies (56)

5.5.1.1 General (56)

5.5.1.2 Number of Associations (56)

5.5.1.3 Asynchronous Nature (56)

5.5.1.4 Implementation Identifying Information (56)

5.5.2 Association Initiation by Real-World Activity (56)

5.5.2.1 Real World Activity—Worklist (56)

5.5.2.1.1 Proposed Presentation Context (57)

5.5.2.1.2 SOP Specific Conformance to Modality Worklist Service SOP Class (57)

5.5.2.1.3 Error Handling (60)

5.6 Modality Performed Procedure Step AE Specification (61)

5.6.1 Association Establishment Policies (61)

5.6.1.1 General (61)

5.6.1.2 Number of Associations (61)

5.6.1.3 Asynchronous Nature (61)

5.6.1.4 Implementation Identifying Information (61)

5.6.2 Association Initiation by Real-World Activity (61)

5.6.2.1 Real World Activity (62)

5.6.2.1.1 Associated Real-World Activity (62)

5.6.2.1.2 Proposed Presentation Contexts (62)

5.6.2.1.3 SOP Specific Conformance Statement (62)

6.0 Communication Profiles (67)

6.1 TCP/IP Stack Supported (67)

6.1.1 Physical Media Supported (67)

7.0 Extensions/Specializations/Privatizations (67)

7.1 Standard extended/specialized/private Syntaxes (67)

7.2 Private Transfer Syntaxes (67)

8.0 Configuration (68)

8.1 Basic System Configuration (68)

8.2 DICOM Network Configuration (69)

8.2.1 Local Host - TCP/IP and General (69)

8.3 DICOM Storage Configuration (69)

8.4 DICOM HIS/RIS Configuration (69)

8.5 DICOM Print Configuration (70)

8.6 “Print/Store” Key Configuration (71)

8.6.1 Multiple Destinations - Auto Transfer (71)

8.6.2 Overlay or Burned in graphics (71)

8.6.3 Secondary Capture Image Storage SOP Class (71)

8.6.4 Auto Retry (71)

8.6.5 Compression (71)

8.6.6 Print Routing - Auto Transfer (72)

9.0 Media Storage (73)

9.1 Implementation Model (73)

9.2 Application Data Flow Diagram (73)

10.0 Media Storage AE Functional Definition (74)

10.1 Real-World Activities for Media Storage (74)

10.1.1 Browse Directory Information—Real World Activity (74)

10.1.2 Import into Local Storage—Real-World Activity (74)

10.1.3 Export to Local Archive Media—Real-World Activity (74)

10.1.4 Sequencing of Real World Activities (74)

10.2 AE Specifications (75)

10.3 Antares Media Storage AE Specification (75)

10.3.1 File Meta Information Options (75)

10.4 Media Storage Application Profile (76)

10.4.1 DICOMDIR keys (76)

10.4.2 Compliance to STD-GEN-CDR (78)

10.4.3 Compliance to STD-US-SC-MF-CDR (78)

10.4.4 Compliance to STD-US-ID-MF-CDR (79)

10.5 Augmented and Private Profiles (80)

10.5.1 Augmented Application Profiles (80)

10.5.1.1 AUG-GEN-CDR, AUG-US-SC-MF-CDR, AUG-US-ID-MF-CDR (80)

10.5.2 Private Application Profiles (80)

10.6 Extensions, Specializations and Privatizations of SOP Classes and Transfer Syntaxes (80)

10.7 Configuration (81)

10.7.1 AE Title Mapping (81)

10.7.1.1 DICOM Media Storage AE Title (81)

10.8 Support of Extended Character Sets (81)

List of Tables

Table1: References (8)

Table2: Acronyms and Abbreviations (9)

Table3: Patient Registration fields populated from Worklist (16)

Table4: Supported SOP Classes (18)

Table5: Verification Presentation Context Table (19)

Table6: Supported Storage SOP Classes (20)

Table7: Store Presentation Context Table (21)

Table8: US Image IOD Attributes (23)

Table9: Secondary Capture IOD Attributes (29)

Table10: US-MF Image IOD Attributes - Clips (32)

Table11: US-MF Image IOD Attributes - 3D Volumetric Data (35)

Table12: C-STORE Status Responses (40)

Table13: Standard SOP Classes as Storage Commitment Push Model (41)

Table14: Initiation presentation context Storage Commitment Request (42)

Table15: Presentation context accepted for Storage Commitment (43)

Table16: Basic Gray Scale Print Management Meta SOP Classes (44)

Table17: Basic Color Print Management Meta SOP Classes (44)

Table18: Presentation Context Table (47)

Table19: Basic Film Session N_CREATE_RQ attributes (48)

Table20: Attributes of the N_DELETE_RQ - Basic Film Session SOP Class (48)

Table21: Basic Film Session SOP status (48)

Table22: Supported DIMSE Services for Basic Film Box SOP Class (49)

Table23: Used Basic Film Box N_CREATE_RQ attributes (49)

Table24: Attributes of the N_DELETE_RQ on the Basic Film Session SOP Class (50)

Table25: Basic Film Box SOP status (50)

Table26: Used Basic Grayscale Image Box N-Set attributes (51)

Table27: Basic Grayscale Image Box SOP status (52)

Table28: Used Basic Color Image Box N-Set attributes (52)

Table29: Basic Color Image Box SOP status (53)

Table30: Used Printer N-Event reports (53)

Table32: Used Print Job N-Event reports (54)

Table31: Mandatory Printer N_GET_RSP, N_EVENT_REPORT_RQ attributes (54)

Table33: SOP Classes as an SCU (56)

Table34: Worklist Presentation Context Table (57)

Table35: Modality Worklist Matching Key Attributes (57)

Table36: Modality Worklist C_FIND_RSP Return Key Attributes (57)

Table37: C-FIND Status Responses (60)

Table38: SOP Classes as an SCU (61)

Table39: Proposed presentation contexts (62)

Table40: Performed Procedure Step N-CREATE Attributes (62)

Table41: N-SET Response Status (64)

Table42: Performed Procedure Step N-SET Attributes (65)

Table43: N-SET Response Status (66)

Table44: User Configurable Printer parameters (70)

Table45: Application profiles, Activities, and Roles for DICOM Exchange Media (75)

Table46: DICOMDIR keys (76)

Table47: STD-GEN-CDR Supported SOP Classes (78)

Table48: STD-US-SC-MF-CDR Supported SOP Classes (78)

Table49: STD-US-ID-MF-CDR Supported SOP Classes (79)

Table50: Augmented Application profiles, Activities, and Roles for DICOM Exchange Media (80)

Table51: AUG-GEN-CDR, AUG-US-SC-MF-CDR and AUG-US-ID-MF-CDR Supported SOP Classes (80)

List of Figures

Figure1: Verification Model (10)

Figure2: Store Model (11)

Figure3: Storage Commitment Model (11)

Figure4: Print Model (12)

Figure5: Modality Worklist Model (12)

Figure6: Modality Performed Procedure Step Model (13)

Figure7: Implementation Model (14)

Figure8: Media Storage Application Data Flow Diagram (73)

1.0 Purpose

This document describes the conformance to the ACR-NEMA DICOM 3.0 Standard by the SONOLINE? Antares ultrasound system software version 3.0 from Siemens Medical Solutions, Inc. Ultrasound Division. It shall establish the conformance specifications for this system only, and does not apply to other products offered by Siemens Medical Solutions, or its affiliates.

The SONOLINE Antares system is a device that generates ultrasound images that can be sent using DICOM standard protocols and definitions to other DICOM compliant devices that support the SOP classes defined in Table4on Page18.

1.1 Scope

The DICOM standard provides a well-defined set of structures and protocols that allow inter-operability of a wide variety of medical imaging devices. The SONOLINE Antares system provides support for essential services related to ultrasound scanning and connectivity to DICOM compliant devices. SONOLINE Antares system products will not support all features supported by the DICOM standard. This document clearly states the DICOM services and data classes that are supported by the applications included with the SONOLINE Antares. The intent of this document is to allow users and other vendors who also conform to the DICOM standard to exchange information within the specific context of those elements of the DICOM standard that SONOLINE Antares system supports.

This document is written with respect to the adopted portions of the DICOM standard, Version 3. The following sections of this document follow the outline specified in the DICOM Standard NEMA publication PS3.2 [1].

2.0 References

Specifications of the DICOM 3.0 standard may be obtained from ACR-NEMA for customers who require detailed information.

Table1: References

Document Title Location

[1]Second part of the DICOM standard: NEMA Standards

Publication PS 3.2-2001, Digital Imaging and Communications in Medicine (DICOM), Part 2: Conformance https://www.wendangku.net/doc/a112956308.html,/ dicom/2001.html

3.0 Definitions

Table2: Acronyms and Abbreviations

Acronym or Abbreviation Definition

ACR-NEMA American College of Radiology - National Electrical Manufacturer's

Association

AE Application Entity

Conformance Statement A formal statement associated with a specific implementation of the DICOM

Standard. It specifies the Service Classes, Information Objects,

Communications Protocols and Media Storage Application Profiles supported

by the implementation

DICOM 3.0Digital Imaging and Communications in Medicine, Version 3.0.

DIMSE DICOM Message Service Element

EBE Explicit Big Endian transfer syntax

ELE Explicit Little Endian transfer syntax

ILE Implicit Little Endian transfer syntax

C-STORE Composite Store

Ethernet Network topology devised in 1976 by DIX (DEC/Intel/Xerox) which is the

most common in practice today.

IOD Information Object Definition

MWL Modality Worklist

OOG Object Oriented Graphics

PACS Picture Archiving and Communications Systems

PDU Protocol Data Unit

RWA Real-World Activity

SCP Service Class Provider

SCU Service Class User

SOP Service-Object Pair

syngo Siemens DICOM Application Framework - Common User SW

UID Unique identifier

3.1 Support of Extended Character Sets

The Antares platform supports the ISO 8859 Latin 1 (ISO-IR 100) character set family and the same family with code extensions (ISO 2022 IR 100 Latin-1).

4.2 AE Functional Definitions

4.2.1 Verification Real-World Activities

The Antares Verification application entity supports Verification Service Class as SCU thus allowing the operator to verify the ability of an application on a remote node to receive DICOM messages. (C-ECHO DIMSE)

4.2.2 Storage Real-World Activities

The Antares Storage application entity performs all of the functions necessary to transmit ultrasound images and associated data to DICOM network archive servers and/or workstations. The Antares Storage AE supports the Ultrasound Image Storage SOP class as SCU. The Antares Storage AE also supports the Secondary Capture SOP class, as SCU, for devices that do not support the US Image SOP. The Antares Storage AE initiates separate associations to the storage SCP servers, each time the user selects “Store/Print” or “Send/Archive To Destination”.

4.2.3 Storage Commitment Real-World Activities

The Storage Commitment Push Model SCU (SC-SCU) uses the Storage Commitment Service Class to request commitment for one or more composite objects from the Storage Commitment Push Model SCP (SC-SCP). SC-SCU issues an N-ACTION to SC-SCP containing a list of refer-ences to composite objects, requesting that the SC-SCP takes responsibility for storage commit-ment of the composite objects. If the SC-SCP has determined that all the composite objects exist and it has successfully completed storage commitment for the set of composite objects, it issues an N-EVENT-REPORT with status successful and the list of stored images. SC-SCU now knows that the composite objects have been committed by SC-SCP and can delete its copies of composite ob-jects. The information from the N_EVENT-REPORT is propagated back to the user interface.

If committed storage cannot be done for some reason for any of the list of composite objects the SC_SCP issues an N-EVENT-REPORT with a status of completed-failures exist and would in-clude both the successful and failed list. The Antares SCU can be configured to receive the

N_EVENT_REPORT on the same association on which N_ACTION was sent or on a different as-sociation.

4.2.4 Print Real-World Activities

The Antares Print AE provides all aspects of the Print Management SCU. The AE initiates separate associations to the print servers, verifying their on-line status when the user selects “Print/Store” or “Expose Film Job.” The Antares Print AE accommodates both grayscale and color print servers.

To invoke the Print “REAL WORLD ACTIVITY” the user selects a “Print/Store” key. Film sheets can be transferred immediately after being filled, at the end of exam, or queued to hard disk for transfer later.

After an imaging exam is complete, the user has the ability to Print images stored on the hard drive using the Filming UI. Invoking the “Expose Film Job” Real World activity invokes the DICOM Print activity for selected exams or individual images. “Expose Film Job” is available through the Filming UI function

4.2.5 Worklist Real-World Activity

Patient registration can be automated by using the ‘Worklist’ Real World Activity.

Double clicking the ‘Scheduler’ icon in the patient browser UI initiates a manual Worklist query. If no matches are found, a message will be presented to the operator indicating so. If one or more patients are found, the scheduler list will be updated with the scheduled procedures. The Worklist feature can be configured to query for either the procedures scheduled for the Local Site (Antares Worklist SCU AE Title) or for Ultrasound scheduled procedures (Modality = US). Queries can be made for today’s scheduled procedures, or a user specified range of scheduled procedure dates and times. The system can also be configured to perform automatic Worklist queries at user specified intervals.

Patient based queries are not supported in the Antares 2.0 release.

Selection of a patient from the list will cause all demographic information for that patient to be loaded in to the patient data fields in the Patient Registration screen. The following patient registration fields will be populated with matching Worklist attributes:

Table3: Patient Registration fields populated from Worklist

Data Field DICOM Attribute DICOM Tag

Name Patient's Name(0010,0010)

ID Patient ID(0010,0020)

Date of Birth Patient's Birth Date(0010,0030)

Sex Patient's Sex(0010,0040)

Height Patient's Size(0010,1020)

Weight Patient's Weight(0010,1030)

Institution name Institution name(0008,0080)

Performing Physician Performing Physicians’ Name(0008,1050)

Referring Physician Referring Physician's Name(0008,0090)

Operator Operators' Name(0008,1070)

Accession No Accession Number(0008,0050)

Indication Admitting Diagnoses Description(0008,1080)

Request ID Requested Procedure ID(0040,1001)

Requested Procedure Requested Procedure Description(0032,1060)

Scheduled Procedure Step Description(0040,0007)

Additional Info Additional Patient’s History(0010,21B0)

4.2.6 Modality Performed Procedure Step Real-World Activity

The Antares supports the DICOM Modality Performed Procedure Step Service as SCU. The modality performed procedure step SCU informs the performed procedure step SCP about the procedure performed at the modality in the N-CREATE and N-SET DIMSE service. Immediately after a new patient, study or scheduled procedure is registered (via Patient Registration) the Antares automatically performs an MPPS N-CREATE operation with a status of IN-PROGRESS for the newly created Performed Procedure Step. When the current patient procedure ends, (either with a End Exam, or new Patient / Study / Procedure) the Antares automatically performs an MPPS N-SET final operation with a status of COMPLETED. The user may also manually complete or discontinue the current Performed Procedure Step through the MPPS dialog. An MPPS N-SET final operation is performed with the appropriate status of COMPLETED or DISCONTINED.

The modality performed procedure step SCP responds to the N-CREATE and N-SET and confirms that it received the information from the modality.

4.2.7 Sequencing of Real-World Activities

In order for any of the remote processes to be able to provide the SCP services which the SONOLINE Antares system has requested, the appropriate associations must first be opened. This initiation occurs with the following activities:

1.When a “Send To Destination” operation occurs.

2.When one or more Storage SCP devices are configured for transfer “During the exam” and

the associated Print/Store key(s) are pressed. The current image displayed on the system screen in captured and transferred.

3.When one or more Storage SCP devices are configured for transfer at “End of exam” and

the current patient exam ends (e.g. New Patient, Study or End Exam button is pressed). All images in the current performed procedure are transferred.

4.When configured, Storage Commitment trigger is automatically derived from the

successful completion of a Send Job.

5.When an “Expose Film Job” print operation occurs.

6.When one or more Print SCP devices are configured for transfer “During the exam” and

the associated Print/Store key(s) are pressed. The current film sheet is transferred after

becoming filled.

7.When one or more Print SCP devices are configured for transfer at “End of exam” and the

patient procedure ends (e.g. New Patient, Study or End Exam button is pressed). All film sheets in the current performed procedure are transferred.

8.When user initiated or automatic Worklist query operations occur.

9.When configured, MPPS N-CREATE (status = IN PROGRESS) operation is

automatically performed when a patient is successfully registered. An MPPS N-SET

(status = COMPLETED) operation is automatically performed when the patient exam

ends.

5.0 AE Specifications

The following specifications apply to the AE as depicted in Figure7on page14. The Antares AE provides conformance to the following DICOM Service SOP Classes as an SCU.

Table4: Supported SOP Classes

Service SOP Class Name SOP Class UID Verification 1.2.840.10008.1.1

Ultrasound Image Storage 1.2.840.10008.5.1.4.1.1.6.1

Secondary Capture Image Storage 1.2.840.10008.5.1.4.1.1.7

Ultrasound Multi-frame Image Storage ? 1.2.840.10008.5.1.4.1.1.3.1

Basic Grayscale Print Management Meta SOP Class 1.2.840.10008.5.1.1.9

Basic Color Print Management Meta SOP Class 1.2.840.10008.5.1.1.18

Print Job SOP Class 1.2.840.10008.5.1.1.14

Modality Worklist Information Model - FIND 1.2.840.10008.5.1.4.31

Storage Commitment - Push Model 1.2.840.10008.1.20.1

Modality Performed Procedure Step 1.2.840.10008.3.1.2.3.3

? The Ultrasound Multi-frame Image Storage SOP is used to transfer Antares Ultrasound clips & 3D volumetric data sets. When used for 3D volumetric data sets, each frame represents a single slice from the 3D volume. These 3D images may be viewed as Cine images.

5.1 Verification AE Specification

5.1.1 Association Initiation by Real-World Activity

The Antares DICOM Service Tool application attempts to initiate a new association for DIMSE C-ECHO service related operations.

5.1.1.1 Real-World Activity - Verification SCU/SCP

5.1.1.1.1 Associated Real-World Activity - Verification SCU/SCP

The associated Real-World activity is a C-ECHO request initiated by the DICOM Service Tool ap-plication. If the process successfully establishes an association to a remote Application Entity, it will send the C-ECHO-Request via the open association to verify that the remote Application En-tity is responding to DICOM messages. If a remote process successfully establishes an association,the Antares responds to C-ECHO-Requests made from remote applications

5.1.1.1.2 Proposed Presentation Contexts - Verification SCU/SCP

The Antares DICOM application will propose Presentation Contexts as shown in the following table.

5.1.1.1.3 SOP Specific Conformance Statement - Verification SCU/SCP

The Antares conforms to the definition of a Verification SCU & SCP in accordance to the DICOM Standard.

Table 5: Verification Presentation Context Table

Presentation Context Table

Abstract Syntax Transfer Syntax

Role

Extended Negotiation

Name

UID

Name List

UID List

Verification

1.2.840.10008.1.1

Implicit VR Little Endian

1.2.840.10008.1.2SCU / SCP None

Explicit VR Little Endian 1.2.840.10008.1.2.1Explicit VR Big Endian

1.2.840.10008.1.2.2

5.2 Storage AE Specification

The Antares DICOM system provides Standard Conformance to the following DICOM V3.0 SOP Classes as an SCU.

Table6: Supported Storage SOP Classes

Service SOP Class Name SOP Class UID

Ultrasound Image Storage 1.2.840.10008.5.1.4.1.1.6.1

Secondary Capture Image Storage 1.2.840.10008.5.1.4.1.1.7

Ultrasound Multi-frame Image Storage 1.2.840.10008.5.1.4.1.1.3.1

This is accomplished using the DIMSE C-STORE Service to whom the SCU issues a service request with a SOP instance that meets the requirements of the desired ultrasound IOD.

5.2.1 Association Establishment Policies

5.2.1.1 General

The configuration of the Antares DICOM application defines the Application Entity Titles, the port numbers, the host names and IP addresses.

5.2.1.2 Number of Associations

The Antares DICOM application initiates several associations at a time, one for each destination to which a transfer request is being processed in the active job queue list.

5.2.1.3 Asynchronous Nature

The Antares DICOM software does not support asynchronous communication (multiple outstand-ing transactions over a single association).

5.2.1.4 Implementation Identifying Information

The Antares DICOM software provides a single Implementation Class UID of

? <“1.3.12.2.1107.5.9.20000101”>

and an Implementation Version Name of

? <“SIEMENS_SWFVC10E”>.

5.2.1.5 Maximum PDU Size offered

-The maximum PDU size offered by the Antares is configurable with default set to 28672 Bytes

5.2.1.6 DICOM Application Context

-DICOM Application context name: 1.2.840.10008.3.1.1.1

超级电容基本参数概念

超级电容基本参数概念 寿命Lifetime 超级电容器具有比二次电池更长的使用寿命,但它的使用寿命并不是无限的,超级电容器基本失效的形式是电容内阻的增加( ESR)与(或) 电容容量的降低.,电容实际的失效形式往往与用户的应用有关,长期过温(温度)过压(电压),或者频繁大电流放电都会导致电容内阻的增加或者容量的减小。在规定的参数范围内使用超级电容器可以有效的延长超级电容器的寿命。通常,超级电容器具有于普通电解电容类似的结构,都是在一个铝壳内密封了液体电解液,若干年以后,电解液会逐渐干涸,这一点与普通电解电容一样,这会导致电容内阻的增加,并使电容彻底失效。 电压Voltage 超级电容器具有一个推荐的工作电压或者最佳工作电压,这个值是根据电容在最高设定温度下最长工作时间来确定的。如果应用电压高于推荐电压,将缩短电容的寿命,如果过压比较长的时间,电容内部的电解液将会分解形成气体,当气体的压力逐渐增强时,电容的安全孔将会破裂或者冲破。短时间的过压对电容而言是可以容忍的。 极性Polarity 超级电容器采用对称电极设计,也就说,他们具有类似的结构。当电容首次装配时,每一个电极都可以被当成正极或者负极,一旦电容被第一次100%从满电时,电容就会变成有极性了,每一个超级电容器的外壳上都有一个负极的标志或者标识。虽然它们可以被短路以使电压降低到零伏,但电极依然保留很少一部分的电荷,此时变换极性是不推荐的。电容按照一个方向被充电的时间越长,它们的极性就变得越强,如果一个电容长时间按照一个方向充电后变换极性,那么电容的寿命将会被缩短。 温度Ambient Temperature 超级电容器的正常操作温度是-40 ℃~70℃,温度与电压的结合是影响超级电容器寿命的重要因素。通常情况下,超级电容器是温度每升高10℃,电容的寿命就将降低30%~50%,也就说,在可能的情况下,尽可以的降低超级电容器的使用温度,以降低电容的衰减与内阻的升高,如果不可能降低使用温度,那么可以降低电压以抵清高温对电容的负面影响。比如,如果电容的工作电压降低为1.8V,那么电容可以工作于65℃高温下。如果在低于室

GE彩超参数

1:高档实时四维彩色多普勒超声波诊断仪 主要要求及系统概述 1.设备要求 1.1整机为原装进口设备。 1.2*该设备有美国FDA和德国CE认证。 2.主机系统性能概括 2.1彩色多普勒超声波诊断仪包括: 2.1.1高分辨率彩色逐行液晶显示器≥19英寸 2.1.2全数字化彩色超声诊断系统主机,数字化二维灰阶成像单元,数字化彩色多普勒单元,数字化频谱多普勒显示和分析单元,数字化能量血流成像单元2.1.3超高细微分辨血流技术HD-Flow,双向PDI 编码显示血流方向和密度信息,对微小血管显示的高度灵敏度,减少彩色过溢,支持所有探头 2.1.4全数字波束形成器,实时二维扫描成像组件,实时三维扫描成像组件2.1.5编码激励技术,组织二次谐波成像,实时三同步能力 2.1.6凸型扩展技术,用于二维和彩色血流 2.1.7频率复合成像技术FFC,屏幕可显示 *2.1.8实时空间复合成像用于腹部和线阵探头(声束偏转线数均可调,均分成3,5,7,9,11),8级别可调,应用于2D, 3D,CFM、PD、HD-Flow和STIC模式(附图) *2.1.9智能化斑点噪声抑制技术,可调级别6级,应用于所有成像模式,与基础图像双侧对比,可实时或后处理实现

*2.1.10二维、胎儿面部三维成像,频谱多普勒模式自动图像优化调整 *2.1.11一体化实时立体成像技术, 容积成像速率达到40容积/秒 2.1.12容积探头扫查角度自动偏转技术βView,支持腔内容积探头 *2.1.13自动颈后透明层厚度测量SonoNT技术,在获取合适切面的前提下,可自动识别早孕期胎儿颈后透明层的边界,并自动测量颈后透明层厚度,帮助使用者通过超声检查结果有效地评估21三体及18三体综合症等染色体异常的风险率 2.1.14智能三维重建(SonoRender Start),可自动识别胎儿及周围组织结构并自动调整容积取样框位置及大小,帮助使用者能快速获得胎儿表面三维容积结构的工具,提高工作效率 2.1.15反转成像模式,显示低回声或液性暗区的立体结构,结合VOCAL可对低回声区域的不规则体积进行测量 2.1.162D/3D直方图技术,作用于2D/CFM/PD模式,可计算灰度直方图和彩色直方图 2.1.17 3D/4D 曲线取样成像技术,任意曲线或直线切割3D平面 2.1.18对3D/4D图像具有MagicCut“魔术剪”功能,可随意切除3D组织或伪像:可分别切除2D或CFM或者2D+CFM一起切除 *2.1.19 STIC空间时间成像相关技术,可应用于4D 胎儿心脏成像技术,可以适用于B/CFM /B Flow/PDI / HD Flow / CRI /SRI等多种模式或技术,无需心电导联线可计算心率;可应用于腹部、腔内容积探头。 2.1.20实时4D支持腹部,腔内等类型容积探头(提供白皮书证明) *2.1.21系统支持多语言操作界面(包括中文)

金风风力机技术参数

金风风力机技术参数

技术数据一览表 表1、GW121/2500kW机组主要技术数据编 号 项目技术参数与规格备注1 认证级别IEC IIIB及以下 2 最大风速37.5 m/s 50年一遇(10分钟平均) 3 极大风速52.5 50年一遇(3秒平均) 4 运行环境温度-30℃~+40℃ 5 生存环境温度-40℃~+50℃ 6 相对湿度≤95% 7 防护等级IP54 发电机 8 转子转速7~15.5 rpm 9 叶尖线速度85.8 m/s 10 切入风速 3 m/s 11 额定风速(静 态) 9.3 m/s 12 切出风速(10 分钟均值) 22 m/s

号 项目技术参数与规格备注 13 再启动风速风机切出后,机组持续检测风速小于切出风速(同时满足10分钟平均风速、10s持续风速、3s瞬时风速),机组即可并网。 14 再启动温度-18 ℃ 15 功率调整方式变桨变速调节 16 紧急刹车方式3个叶片顺桨实现气动刹车 17 风速-桨角关系高于额定风速变桨控制 18 风速-转速关系低于额定风速变速控制 19 噪音(整个风电 机组) <103 dB 20 轮毂中心高度90 m 21 叶轮直径121.5 m

号 项目技术参数与规格备注 22 叶片数目 3 23 扫风面积11595 m2 24 风向排列(取 向) 上风向 25 叶轮倾斜度 3 deg 26 叶片锥度-3 deg 27 叶片制造商见报价文件 28 叶片加工工艺真空吸注 29 叶片材料玻璃纤维增强树脂 30 叶片长度59.5 m 31 叶片根部连接 件 螺栓 32 叶片根部到轮 毂中心的距离 1.17 m 33 叶片纵剖面图 34 弦(根/尖)长最大弦长约3 m

产品需求规格书模板

XX项目 产品需求规格说明书模板

目录 1文档介绍 (2) 1.1文档目的 (2) 1.2文档范围 (2) 1.3读者对象 (2) 1.4参考文档 (3) 1.5术语与缩写解释 (3) 2综合描述 (3) 2.1产品介绍 (3) 2.2产品面向的用户群体(可选) (3) 2.3产品应当遵循的标准或规范 (4) 2.4产品范围 (4) 2.5产品涉众(涉及角色) (4) 2.6设计和实现的限制 (4) 2.7假设和约束(依赖) (5) 3产品需求 (5) 3.1需求分类 (5) 3.2用例图 (6) 3.3功能需求 (7) 3.3.1需求描述 (7) 3.3.2特殊需求 (8) 3.3.3数据规范 (8) 3.4非功能需求(包括但不限制于以下几项) (8) 3.4.1时间特性要求 (8) 3.4.2精度要求 (9) 3.4.3业务量估算 (9) 3.4.4灵活性 (9) 3.4.5可用性 (9) 3.4.6安全性 (10) 3.4.7兼容性 (10) 3.4.8易用性 (11) 3.4.9可维护性 (11) 3.5运行环境 (11) 3.5.1设备及分布 (11) 3.5.2支撑软件 (12) 3.6接口 (12) 3.6.1硬件接口 (12) 3.6.2软件接口 (12) 3.6.3通讯接口 (12) 3.6.4用户接口 (13) 4验收标准 (14) 4.1功能验收标准 (14) 4.2非功能性验收标准 (14) 附录A:需求建模与分析报告 (14) A.1需求模型1 (15) A.2需求模型N (15) 附录B:需求确认 (15)

【对本文档的说明: 本文档中黑色斜字体为说明性文字,黑色正常字体为需求规格说明书实际写作时必需部分。蓝色字体为举例说明文字。】 1文档介绍 1.1 文档目的 提示: 软件需求规格说明主要描述系统的概貌、功能要求、性能分析、运行要求和将来可能提出的要求。阐述一个软件系统必须提供的功能和性能以及它所要考虑的限制条件,它应该尽可能完整地描述系统预期的外部行为和用户可视化行为。 举例说明: 示例:本文档的主要目的是描述XXX项目中XXX模块的功能需求和非功能需求,功能需求采用用例的方式描述。以使所有涉众能够达成共识。本需求说明书,在需求固化之前,会有相应的变更。在文档历史中会详细记录变更的具体内容。 1.2 文档范围 提示: 文档范围包括:产品介绍,产品面向的用户群体,产品应当遵守的标准与规范,产品范围,产品中的角色,产品的功能性需求,产品的非功能性需求。 1.3 读者对象 提示: 1)各种管理人员及开发人员:专案经理、系统工程师、软件开发人员、硬件开发人员、测试人员、型态管理人员、品质保证人员、作业员和技术出版人员。 2)软件使用客户。

便携彩超 迈瑞M 技术参数

全数字化便携式彩色多谱勒超声诊断系统技术规格 一、货物名称:全数字化便携式笔记本彩色多谱勒超声诊断系统。 二、数量:一台。 三、产品用途说明:用于临床病人腹部、妇科、产科、心脏、浅表组织与 小器官、外周血管、颅脑及介入性超声等全身应用。 四、设备配置:主机+3把探头(腹部凸阵+心脏相控阵+浅表线阵)+全身应 用软件包(四维成像功能及软件),台车及图像工作站。 五、主要功能及技术指标: 1. 彩色多谱勒超声波诊断仪功能及参数: 便携式笔记本彩超,整机重量(含电池)≤, 14英寸高清晰、医用 专业彩色液晶显示器。 * 组织多普勒(支持TDI速度图、TDI能量图、TDI频谱图)。 具有三条可以360度任意旋转角度的M型取样线,并能够在离线分析 过程中重构M型图像。 组织谐波成像技术。 凸阵探头扩展成像技术。 二维和彩色多谱勒双幅显示。 图像局部放大功能(实时和冻结放大,放大倍率>6倍),同时支持将 超声图像显示区域放大至全屏14英寸。 具有组织特征成像,能够独立选择液性、肌肉、常规多种组织模式 进行成像。 具有二维、彩色、频谱、TDI等一键智能优化图像功能。

主机内置锂电池组,电池操作时间:待机时间不小于2个小时,连 续扫查时间不小于1小时。 支持热插拔探头,不关机直接更换探头,并快速自动识别相应探头,进入相应检查模式。 有腹部、妇科、产科、神经、肌骨、肝脏、胰腺、外伤急诊科等专科测量预置功能便于临床出诊和记录。 在线剪贴板功能,存储的图像能与实时扫查图像同屏显示并可在线 浏览、删除、拷贝等操作。 具有中文操作界面及输入中文注释功能。 多幅图像同屏回放切换功能。 支持4D容积成像,支持表面、骨骼等多种成像模式,支持立体方位在线引导。 支持4D 曲面成像技术,最大限度地满足观察组织的生理曲线 2.测量和分析:(B型,M型,频谱多谱勒,彩色多谱勒)。 血管内中膜自动描记测量:可进行前壁、后壁描记,自动同步显 示最大值、最小值、质量指标等7种参数值。 外周血管血流测量分析报告功能。 3.图像存储与(电影)回放重现单元: 支持同步存储,即存储或导出图像数据的同时可以完成实时扫描 可对回放的容积图像进行各种成像模式的切换、调节亮度、对比度、阈值、平滑等处理。 对存储的容积图像,可以多个方向旋转。

蓄电池基本知识(参数含义及各型号优缺点)

电池基本参数说明 额定电压:电池正常工作的电压。 额定容量:例如:28Ah(20hr,1.75V/cell,25℃) 是指在25℃时,20小时放电(即2.8A)使单个电池电压降到1.75V所放出的容量,折算到1小时放电的安培值。 尺寸:长、宽、高、总高。 内阻:例如:4.0mΩ(25℃,充满电) CCA:冷启动电流值:在-17.8℃和-28.9℃条件下,充满电的12V蓄电池在30s 内,其端电压下降到7.2V时,蓄电池所能供给的最小电流。 储备容量(25℃):完全充足电的12V蓄电池,在25±2℃的条件下,以25A恒流放电至蓄电池端电压下降到10.5±0.05V时的放电时间。 环境温度:电池工作的温度,有的细分充电温度与放电温度。 DODxx%:电池用掉xx%的电。如:“DOD80%,700次”则说明电池每次都用去80%的电,可循环使用700次。 最大充电电流:例如:4.5C20。是指在以20小时放电为标准的电池容量数值乘以4.5即为最大充电电流。 最大放电电流:算法同上,即为最大的放电电流。 循环充电电压:也有叫浮充电压,是指将蓄电池组与电源线路并联连接到负载电路上,电源线路仅略高于蓄电池组的断路电压,由电源线路所供的少量电流来补偿蓄电池组局部作用的损耗,以使其能经常保持在充电满足状态而不致过充电。电极L或R:有正极、反极电池之分。区分方法: 1、在外包装或者电池上,反极电池一般会标注"L"字样。正极电池一般不标注。 2、面对电池极柱靠近自己一侧,正极电池‘+’极柱在电池左侧,反之在右侧。比能量: 体积能量密度:以wh/L为单位,体现单位体积下电池可以存储的能量大小。 重量能量密度:以wh/kg为单位,体现单位重量下电池可以存储的能量大小。比功率:以kw/kg为单位,体现单位重量下电池可以输出的功率。 电池三段式充电 一、恒流段:当电池电压较低时,为了避免充电电流过大损坏电池,应该限制充电电流不能过大,又为了缩短充电时间,应使用最大允许充电电流充电。恒流充电阶段为主充电阶段,电池已经充入约85~90%的电量。 二、恒压段:保持这个恒定的电压对电池充电,在恒压充电过程中,电池电压会越来越高,电流会越来越小,当充电电流下降到0.5C时,恒压充电结束。 三、浮充段:浮充电阶段实际上也是恒压充电,在这个阶段的充电电压一般控制在13.6~13.8V左右,充电电流较自放电电流略大,一般为0.01~0.03C左右。通过涓流充电,可以将电池电量充到接近100%。 铅蓄电池外壳文字说明 例如:6-QAW-100-D 6:代表串联的电池数,每个2V,即12V

vivid 7 pro BT05彩超技术参数

Vivid 7 pro BT05 08/2005 1 Description Vivid 7 pro Console (220V 50Hz PAL) Vivid 7彩色多普勒超声系统主机 心脏(成人及儿童)、外周血管、腹部、妇产科、小器官、经颅、经食道、腔内检查等全面应用能力 TruScan architecture TruScan 平台 T ruDate image 真实成像技术 New level Digital Raw Data 第二代数字化原始数据处理技术更真实地获取和保留超声图像信息,提高图像调节的灵活性,数据测量的正确性。 1024 Channel & 12 BIT A/D 转换 1024 通道处理, 12 BIT 模数转换 Coded Octave Imaging 编码宽带二次谐波成像 1) 编码宽带二次谐波具有更宽的频带,较常规二次谐波有更高轴向分辨率。 2) 编码宽带 二次谐波较常规二次谐波有更高帧频。 3) 多组谐波频率连续可调,适合于不同透声条件的个体。 Code Coversion 编码反相成像技术 high frame rate 2D & color image 高帧频应用在2D 和彩色成像模式 Coronary Flow Imaging 专门设计的冠状动脉成像套件可以有效地减少心腔内高速血流对冠脉及心肌内相对低流速血流的干扰,清晰显示并测量冠脉和心肌内血流并作其功能储备分析。 Multiple Angle Compound Imaging 实时多角度复合成像 1.增加了线阵探头的扫描视野。 2.实现了超声束从不同方向同步扫描而成像,提高了超声扫描线交错覆盖区内的组织的分辨率。 3.消除强回声界面下的盲区 Speckle Reduce Imaging 斑点噪声去除技术 Automatic Optimization 自动优化功能

SPD385-40A-MH产品规格书-V3.3

审核: Reviewed by 陈伟东 日期: Date 2011-06-29 批准: Granted by 李叶来 日期: Date 2011-06-29 深圳市海鹏信电子股份有限公司 SHENZHEN HAIPENGXIN ELECTRONICS CO.,LTD. 版权所有 侵权必究 All rights reserved

2009-02-01 未经许可不得扩散 第2页,共8页Page 2 , Total8 修订记录 日期 修订版本 描述 编写 2009-02-01 V1.0 初版制订 韦耀峰 2010-08-12 V1.1 技术参数更新,添加认证 李韵晴 2010-12-10 V3.0 产品说明书版本同步 李韵晴 2011-06-29 V3.2 海拔高度和安规版本更改 曾维霞 2012-01-09 V3.3 尺寸图加标公差、更改大气压力 曾维霞

2009-02-01 未经许可不得扩散 第3页,共8页Page 3 , Total8 目 录 1 适用范围..........................................................................................................................5 2 规范性引用文件...............................................................................................................5 3 功能性能..........................................................................................................................5 3.1 简述.................................................................................................................................5 3.2 性能指标..........................................................................................................................5 4 防护原理(仅供参考).....................................................................................................6 5 结构和外观.......................................................................................................................6 5.1 端口类型..........................................................................................................................6 5.2 结构.................................................................................................................................6 5.2.1 结构尺寸....................................................................................................................6 5.2.2 附件............................................................................................................................7 5.2.3 安装............................................................................................................................7 6 适用环境与安规...............................................................................................................7 7 环保.................................................................................................................................8 8 产品标签..........................................................................................................................8 9 产品照片.. (8)

超级电容规格书

WRITTEN CHECKED APPROVED To. : DATE : 200 . . . SPECIFICATION PRODUCT : STARCAP MODEL : DMS series KORCHIP CORP. KORCHIP B/D, 817-38, Anyang 2-dong, Manan-gu, Anyang-si, Gyeonggi-do, KOREA TEL : 82 - 31 - 361 - 8000 FAX : 82 - 31 - 361 - 8080

Page No.ITEM etc. 1Cover Page 2Index 3 1. Scope 2. Part Number System 3. Product Model Name 4. Photo 5. Nominal Specifications 4 6. Cell Structure 7. Product Construction And Dimension 5 8. Packing Specifications 6 9. Specifications And Test Method 7 10. Measuring Method Of Characteristics 8 11. Mounting And Soldering 9 12. Cautions For Use 10 13. Environmental Management Index

Items DMS 3R3 204 R DMS 3R3 224 R Cell Size ?6.8 × 1.4mm ?6.8 × 1.4mm OPERATING TEMPERATURE -10 ~ +60 ℃-10 ~ +60 ℃RATED VOLTAGE 3.3 VDC 3.3 VDC ELECTROSTATIC CAPACITANCE (F) 0.20 F 0.22 F CAPACITANCE TOLERANCE -20 ~ 80 %-20 ~ 80 %EQUIVALENT SERIES RESISTANCE (ESR)LESS THAN 200?LESS THAN 200?LEAKAGE CURRENT (LC) LESS THAN 150? LESS THAN 150? 1. Scope This specification applies to STARCAP(Electric Double Layer Capacitor), submitted to specified customer in cover page. 2. Part Number System DMS 3R3 204 R (Example) ① ② ③ ④ ① Series Name ② Rated Voltage : 3.3VDC ③ Capacitance : 0.20 F (204 = 20 × 10+4 uF) ④ Terminal Type : R-type 3. Product Model Name 1) Product : Electric Double Layer Capacitor 2) Model name : DMS3R3204R, DMS3R3224R 4. Photo 5. Nominal Specifications

彩超参数

高档彩色多普勒超声诊断仪 技术规格及要求 1.货物名称: 全数字高档彩色多普勒超声诊断仪 2.用途说明: 2.1.腹部、产科、妇科、心脏、小器官、泌尿、血管、儿科、急诊、麻醉、其它 2.2.要求为2012年最新版本及最新出产机型,具有用户现场升级能力,可满足将来临床 应用扩展需求 3.货物数量:壹套 4.交货期限:合同签订后30日 5.系统技术规格及概述: 5.1.全数字化彩色多普勒超声诊断系统主机 5.2.≥19寸高分辨率彩色液晶显示器 5.3.≥10.4寸高灵敏度防反光彩色触摸屏 5.4.控制面板可独立旋转和升降 5.5.数字波束增强器 5.6.多倍波束合成 5.7.探头接口≥4个 5.8.二维灰阶模式 5.9.谐波成像模式 5.10.M型模式 5.11.彩色M型模式 *5.12可配置解剖M型模式(≥2条取样线)* 5.12.可配置曲线M型模式 5.13.彩色多普勒成像(包括彩色、能量、方向能量多普勒模式) 5.14.频谱多普勒成像(包括脉冲多普勒、高脉冲重复频率、连续波多普勒) 5.15.组织多普勒成像 5.1 6.负荷成像 5.17.造影成像 5.18.自由臂三维成像 5.19.四维成像 *5.21宽景成像(要求所有探头可用,支持彩色宽景,扫描速度提示)* 5.20.空间复合成像 5.21.斑点抑制成像 5.22.频率复合成像 5.23.独立角度偏转 *5.26扩展成像(要求凸阵、线阵、容积探头可用)* 5.24.实时双幅对比成像 5.25.组织特异性成像 5.2 6.高分辨率血流成像 5.27.一键自动优化(包括应用于二维、彩色及频谱模式) 5.28.全屏放大

高档彩超参数

彩色多普勒超声诊断系统 主要技术要求和规格 一.设备名称及用途:高档全身应用彩色多普勒超声诊断仪1台 用于腹部、妇产、胎儿;血管(外周、脑血管);浅表、肌骨、造影及 介入等超声诊断和相关科研。 二.交货日期:合同签订后90天 三.投标时要求提供原厂家的Data Sheet,评标以此为准 四.具有“国家进口商品安全质量许可证书”和“中华人民共和国医疗器械注册证” 五.主要技术规格及系统概述: 1.主机系统性能概括 1.1显示器≥20英寸数字纯平高分辨率彩色超薄液晶监视器,无闪烁、不间断逐行扫描,高 度可调,可旋转,倾斜,主机具备彩色触摸屏 1.2 全数字化彩色超声诊断系统主机 1.3 数字化二维灰阶成像(部件)单元 1.4 数字化频谱多普勒显示及分析系统 1.5 数字化彩色血流成像(部件)单元 1.6 二次谐波成像(自然组织谐波成像)可用于全部2D探头和4D探头,具有明确谐波频率 显示,可观可调 1.7全数字式波束形成器,探头通道数≥57000通道 1.8 对比造影剂谐波成像,支持腹部、血管、小器官、经颅、腔内及术中的造影 1.8肥胖病人组织差异校正成像技术,可自动校正超声声速完成对肥胖、困难病人条件的超 声扫查。 1.9乳腺组织差异校正成像技术,可根据乳腺不同类型,如腺体为主或脂肪为主型乳腺自动 校正超声声速,提高组织细节及边界显示。 1.10三维成像,具备三维、四维实时成像功能三维容积定量,支持腹部、妇产科、小器官、 腔内检查成像,具备三维容积造影功能 1.11血管内中膜自动测量,可以分析血管前壁和后壁的测量数据,自动给出分析报告,自动 优化测量曲线,可以和血管造影相结合 ﹡1.12 弹性成像:色棒和曲线指导医生操作,可以结合融合成像、GPS定位

便携彩超技术参数及要求-0401资料

便携式彩色多普勒超声诊断系统 主要技术要求和规格 一.设备名称:便携式彩色多普勒超声诊断系统 二.数量:1套 三.设备用途: 主要用于心脏、腹部、妇产科、胎儿、小器官、外周血管等常规诊断 四.交货日期:合同签订后30天 五.主要技术要求和规格: 1.彩色多谱勒超声波诊断仪包括: 1.1 15寸高清晰、医用专业彩色液晶显示器 1.2 二维灰阶成像部件 1.3 频谱多谱勒显示及分析系统 1.4 彩色多谱勒超声波诊断部件 1.5 多谱勒能量图 1.6 Steer 角度独立偏转技术(提供图片证明) 1.7 组织谐波成像 1.8 凸阵扩展成像技术 1.9 回声信号离线分析及处理(要求支持动态范围、频谱基线、图像效果等处理) 1.10 具有空间复合成像技术 1.11 二维和彩色多谱勒双幅显示 1.12 图像局部放大功能(实时和冻结放大) 1.13 智能宽景成像(要求支持线阵探头及凸阵探头,具有实时宽景成像速度提示、宽景长度 达到90cm、有多种伪彩显示) 1.14 具有组织特征成像能够独立选择肌肉、常规、脂肪、液性成像模式 2.测量和分析:(B型,M型,频谱多谱勒,彩色多谱勒) 2.1 一般测量 2.2 产科测量 2.4 心脏功能测量 2.5 自动左室定量分析 2.6 外周血管血流测量分析报告功能 3.图像存储与(电影)回放重现单元: 支持同步存储(支持单帧图像文件包含: DCM、TIFF单帧,电影文件包括:CIN、AVI、DCM),即存储或导出图像数据的同时可以完成实时扫描 4.参考信号: 心电,心音,脉冲波,心电触发 5.输入/输出信号: 5.1 输入: VCR,外部视频, RGB彩色视频

vivid-7-pro-BT05彩超技术参数

Description Vivid 7 pro Console (220V 50Hz PAL) Vivid 7彩色多普勒超声系统主机 心脏(成人及儿童)、外周血管、腹部、妇产科、小器官、经颅、经食道、腔内检 查等全面应用能力 TruScan architecture TruScan 平台 T ruDate image 真实成像技术 New level Digital Raw Data 第二代数字化原始数据处理技术更真实地获取和保留超声图像信息,提高图像调 节的灵活性,数据测量的正确性。 1024 Channel & 12 BIT A/D 转换 1024 通道处理, 12 BIT模数转换 Coded Octave Imaging 编码宽带二次谐波成像 1)编码宽带二次谐波具有更宽的频带,较常规二次谐波有更高轴向分辨率。 2)编码宽带二次谐波较常规二次谐波有更高帧频。 3)多组谐波频率连续可调,适合于不同透声条件的个体。 Code Coversion 编码反相成像技术 high frame rate 2D & color image 高帧频应用在2D和彩色成像模式 Coronary Flow Imaging 专门设计的冠状动脉成像套件可以有效地减少心腔内高速血流对冠脉及心肌内相 对低流速血流的干扰,清晰显示并测量冠脉和心肌内血流并作其功能储备分析。 Multiple Angle Compound Imaging 实时多角度复合成像 1.增加了线阵探头的扫描视野。 2.实现了超声束从不同方向同步扫描而成像,提 高了超声扫描线交错覆盖区内的组织的分辨率。3.消除强回声界面下的盲区 Speckle Reduce Imaging 斑点噪声去除技术 Automatic Optimization 自动优化功能 5 Vivid 7 pro BT0508/2005

揭开四维彩超的真相

如果查不出潜在危机,价格再低,路程再近,你,还会选择吗? 揭开四维彩超的真相 随着人们对下一代健康的重视,越来越多的人开始运用四维彩超来排查胎宝宝畸形。但是,你知道吗?并不是所有四维彩超都能查出异常情况。 今天,我们就特地请到从事妇产科超声影像工作近30年,曾多次应邀参加国际超声学术会议,并获得无数赞誉的李井成医师,为大家揭开一下四维彩超的真相: 不是所有“四维”都一个样 工欲善其事,必先利其器。要令检查无误,首先需要一台高分辨率的设备。仅就品牌而言,四维彩超就有美国GE,日本菲力浦、韩国三星,甚至还有国产的。其中,尤以美国GE 系列最被业内人士认可。“如果设备分辨率不够的话,像心脏小失缺之类的情况就无法显现。”杭州艾玛妇产医院李井成主任介绍说。而GE系列中,又包含了GE730专家版、E6、E8等不同型号,每种型号可对应使用于不同的准妈妈。 GE Voluson E8 Expert:该设备是目前世界上先进的彩色超声设备,被誉为"神奇彩超",是筛查胎儿畸形的尖端设备。具有艺术性的容积显像工具使图像采集和分析更快捷和方便,通过自动获取胎儿心脏的容积图像,有助于胎儿心脏检查图像的规范化采集。 GE Voluson E6:它的容积超声技术已经得到了世界范围的广泛认可,并一直保持着其在该领域的旗舰产品的地位。具有成像速度快,图像质量好,操作灵活简便及心血管追踪技术等优点,减少了检查时间。其断层成像功能,更是能更精确定位,明确诊断神经系统疾病。 GE Voluson 730 Expert专家版:超越普通四维彩超的局限性,引领超声领域的先驱,是世界上第一个具有真正容积超声显像技术的产品。功能强大,可以对三维数据进行实时的采集、浏览和分析处理,从面获得高品质的容积图像,有助于医生准确做出临床诊断。 准确率70%取决于医生技术 可能很多人都有这样的认知,最好的设备出最准确的结果。但在B超领域,可不是这样。李井成主任表示,B超检查结果,30%来自于仪器设备,70%取决于医生的诊断技术和经验。尤其像早期胼胝体缺失、小脑蚓部疾病等神经系统疾病,往往要依靠于医生丰富的经验和高超的技术才能检测出来。“国家规定10年以上高年资职业医师,才有资格做这类超声检查,但实际上,光有资历远远不够,只有在胎儿排畸方面亲手经历过数万个案例,才能精准筛查出一些细微病变。”艾玛妇产医院李井成主任对此表示。 据悉,艾玛妇产医院每个月都会发现50例左右的畸形胎儿,这些畸形儿情况各异,有骨骼系统畸形、泌尿系统畸形、消化系统畸形、心脑血管系统畸形和面部器官畸形,而且有的畸形儿还几种类型兼而有之。 “今年10月,在筛查出的58例畸形胎儿中,唇裂的就占了13例,其它诸如无脑儿、右侧桡骨缺如畸形、拇指发育不良及软骨发育不全等各种畸形也不在少数。最不可思议的是有一天,连续四个接受检查的孕妇都怀了畸形胎儿。” 畸形儿的产生是家庭的不幸,社会的不幸。在短短一个月时间里发现这么多畸形胎儿,是不是代表杭州的畸形儿正越来越多呢? “由于环境污染、工作压力等原因,现在畸形胎儿确实有愈来愈多的趋势,但像艾玛妇医院这种情况还是有一定特殊性的。”李井成主任解释说:“其实来我们医院检查出畸形儿的,有些都曾在其它各大医院进行过同类产检,但由于技术、设备等原因不能确诊,于是来这里进行复查确认。可以说,有部分畸形儿是集中到我们医院了,因此产生这种‘井喷’现象。”

便携式彩超技术参数

便携式彩色多普勒超声诊断仪技术参数 一、总体要求: 1. 设备用途说明:主要用于心脏、腹部、泌尿科、妇产科、儿科、新生儿、经颅、 腔内、小器官及外周血管。 2.系统标准配置包括: 1)彩色多普勒血流成像 2)二维灰阶成像 *3)频谱多普勒显示分析系统,PW频谱多普勒,CW连续多普勒 4)能量多普勒成像 5)方向能量多普勒成像 6)组织谐波成像 *7)腹部大凸探头具备空间复合成像工作模式 *8)配全方位360度M型成像(提供≥3条取样线图片证明) *9)配实时宽景成像(提供扫描过程速度提示功能证明文件) 3. 测量和分析 1)一般测量:包括距离、面积、周长、容积、时间、斜率、心率等 2)产科测量软件:生长曲线显示,胎儿超声心动图计测量,妇产科报告 3)心脏功能测量与分析,心脏报告可编辑 *4)IMT(血管内中膜)测量与分析报告,要求能同时自动测量前壁及后壁内中膜厚度(附图证明) 5)血管血流测量与分析 6) 自定义注释:包括插入、删除、编辑、保存等 4. 输入/输出信号:输入:具备USB接口。输出:复合视频、RGB彩色视频 5.连通性:医学数字图像 6. 图像管理与记录装置:硬盘、U盘存储 7. 超声图像存档与病案管理功能:在主机中完成病人静态图像和动态图像的存储、 管理及回放 8. 产品安全性能: 1)电气安全:符合CE要求 2)声输出安全:系统具备声学输出功率、机械指数 二、技术要求: 1.系统通用功能 *1)彩色监视器:≥15吋高分辨率彩色LCD监视器,无闪烁,不间断逐行扫描 *2)主机笔记本式设计,显示屏可前后倾斜角度150度以上,主机厚度9cm以内 *3)整机重量6.5公斤以下(含电池),便于出诊应用 2. 探头规格 1)可选探头群工作频率范围(2.0-12.0MHz) 2)谐波成像:能应用于配置的腹部探头、线阵探头及心脏探头 3)所配的每种探头的谐波频率均≥2组 4)腹部凸阵探头:128阵元 工作频率:3.0-6.0MHz,最大扫描深度指标≥370mm 5)线阵探头):128阵元

女士泌尿系统彩超

仁记说体检(七)女士妇科彩超 近年来越来越多女性为妇科炎症疾病所困扰,妇科肿瘤成为女性健康的主要威胁。妇科肿瘤主要指女性生殖系统发生的肿瘤,如子宫、卵巢、输卵管、外阴肿瘤等。大概1/3的成年女性都有子宫肌瘤,是发病率最高的妇科肿瘤。专家指出,妇科肿瘤是可以预防的。早期发现、早期治疗,大部分是可以完全治愈的,且能节省很大的经济费用。妇科彩超检查就是其中一项常规检查,但是具体检查什么,相信很多女性却不了解,今天就为大家做一介绍。 妇科彩超能检查什么 1、妇科彩超是可以检查子宫妇科炎症。彩超检查近距离观察盆腔里脏器:子宫、卵巢、盆腔肿块以及是否积液等。 2、妇科彩超还能提供血流信号,提高阳性病人的诊断率,特别是滋养细胞疾病(绒癌葡萄胎的诊断),子宫肌瘤与腺肌瘤的鉴别(子宫肌瘤可见环状或半环状血流信号);卵巢黄体与囊肿的鉴别;轻度输卵管积水和盆腔血管扩张的鉴别;以及盆腔恶性疾病的原发灶和转移灶的诊断,是否有周边组织浸润等。 妇科彩超检查时应该注意什么 在检查前半小时至1小时需要饮水1000ml左右,并且要憋尿憋到最大的限度。因为只有膀胱充盈到一定程度,才能将子宫从盆腔深处挤到下腹部,用彩超观察到子宫及卵巢。盆腔彩超在检查的时候需要露出下腹部,所以妇科彩超注意事项中,最好穿裤子或者半裙,以方便检查。 高危人群 肥胖、不育、未产、延迟绝经(52岁以后绝经)、糖尿病、高血压、多囊卵巢综合症、卵巢颗粒细胞瘤、子宫内膜增生或不典型增生史和子宫肌瘤有不规则出血者、有使用外源性雌激素史者、有癌家族史、多发癌及重复癌倾向者等。 高龄、未孕、促排卵药物的使用,以及乳腺癌、结肠癌、或子宫内膜癌病史,卵巢癌家族史均被视为是卵巢癌发病的高危因素。绝境后激素替代治疗可轻度增加卵巢癌的发病率。

彩超招标参数

彩超招标参数 序号招标要求 *1 设备为全数字化彩色多普勒超声诊断仪:为原装进口机 2 用途:腹部、心脏、妇产科、泌尿科、浅表组织与小器官、外周血管 3. 探头配置:配腹部、心脏、腔内、 4. 提供设备为全数字化彩色多普勒超声诊断仪包括: *4.1 彩色监视器:17寸高分辨率彩色监视器,无闪烁,不间断逐行扫描 4.2 数字化高分辨率二维灰阶成像部件 4.3 数字化实时动态频谱多普勒显示及多参数分析系统 4.4 全数字化彩色多普勒血流显示,血流成像清晰 4.5 数字化多普勒能量图(CDE/CPA)单元 组织谐波成像(THI, 谐波成像基波频率3组, 至少一组为脉冲反相谐波4.6 成像) 4.7 脉冲反相谐波成像功能 4.8 动态范围:170DB 4.9 全数字化超宽频带波束形成器 4.10 14336数字化通道 4.11 灰阶256 4.13 高级编码组织多谱勒成像(ColorTDI/PowerTDI/TDI-M/TDI-PW)功能 4.14 合成成像技术( 作用于发射和接收) 4.15 一键式图像优化技术 第四代空间复合成像技术用于二维和彩色模式( 可以用于腹部及浅表小*4.16 器官,可以多角度调节,可以同幅双侧实时对比 , 可以和其他技术结合)附

图 智能化斑点噪音抑制技术( 可以支持所有探头,可以多级调节,可以同屏*4.17 双幅实时对比 , 可以支持3D , 可以和其他技术结合) 4.17.1 拥有原始数据的超声平台 4.17.2 全中文操作界面 4.17.3 具有实时解剖M型功能,扫描线可任意角度调整 B-Flow二维灰阶血流成象 *4.17.4 4.18 探头接口:3个 4.19 安全性能:符合进口商品安全质量要求 4.20 投标机型为用户认可的机型与配套 5 测量和分析:(B型、M型、频谱多普勒、彩色多普勒)功能 一般测量(距离、面积、周长、容积、角度、时间、斜率、心率、流速、5.1 压力、流速比等)功能 5.2 产科测量,具有产科应用测量软件及胎儿超声心动图测量软件。 5.3 心血管多普勒频谱自动勾画与测量及分析功能 5.4 心脏功能测量(B型、M型、D型、B/CFI/M型) 5.5 外周血管血流测量与分析软件 5.6 全数字化硬盘,磁光盘图像存储,电影回放重现单元1000帧以上 5.7 参考信号:同步心电显示、心电触发装置 5.8 输入/输出信号包括: 5.9 输入:VCR、外部视频、RGB彩色视频,USB接口 5.10 输出:复合视频、RGB彩色视频/S-视频,USB接口 6 图像存储与电影回放单元 6.1 超声图像静态、动态存储、原始数据回放重现

技术规格书

技术规格书 项目名称:轨顶摩擦控制装置采购 《技术规格书》中标注有“★”号的条款必须实质性响应,不允许负偏离(不满足要求)。 ★一、技术参数 轨顶摩擦控制装置包含以下几个主要组成部分:车轮探测单元(车轮传感器)、控制单元(控制箱)、分配泵和电机、摩擦调节剂储存单元(储液箱)、摩擦调节剂涂覆组件(输液管/涂覆板)。 1、摩擦调节剂储存单元(储液箱):独立的调节剂和电气隔箱,配有单独的可锁盖设计和一体式斜面板设计。防水密封盖,配有可调节下拉锁扣。箱体尺寸(长宽高):(860~870) mm *(450~460)mm *(680~690)mm 2、电气箱:符合NEMA12标准的固态控制板。 泵周期调整范围:0.05s——1.95s 泵轴数调整范围:1轴——256轴 熔断保护器:20A 3、车轮探测单元(车轮传感器):非接触式敏感车轮传感器(SMS),钢轨安装支架和J型螺栓。支架尺寸(长宽高):(1365~1375)mm *(510~530)mm *(805~810)mm。支架材质:碳钢喷漆。J型螺栓尺寸(英制):10-1/2,3/4,镀锌。 4、摩擦调节剂涂覆组件(输液管/涂覆板):每台设备含有2个涂敷板,涂敷板上均有内置分配器,以保证均匀喷涂到钢轨。涂覆板尺寸(长宽高):(912~916)mm *(62~66)mm *(24~26)mm 涂覆板材质:合金钢。内置分配器材质:不锈钢、铜。 5、必备条件: A 噪音控制距离300米以上,波磨控制距离800米以上。 B 轨顶涂覆,对横向力、波磨、噪音、轨顶剥离掉块、压溃控制效果明显。其中横向力平均降低值27%或以上,噪音降低25%或以上。 C 调节剂是非油或者脂类产品,不含挥发性溶剂,必须是水基固化干性薄膜产品。 此产品是一水基态摩擦调节剂,包含由无机固体,聚合物,以及其他添加剂组成的工程

相关文档