The Agilent Technologies 8902A measuring receiver delivers the accuracy and resolution of a high per-formance power meter at frequencies from 150 kHz to 1.3 GHz (50 MHz to 26.5 GHz with the Agilent 11793A microwave converter) and levels from

+30 dBm to –127 dBm. It accurately measures AM, FM, and f M, including residuals and incidentals, with a single keystroke. The 8902A measuring receiver, with the 11793A, counts RF signals to 26.5 GHz with 10 Hz resolution and excellent long-term frequency stability. The 8902A measuring receiver with Option 050 offers increased power measurement accuracy. This option specifies Tuned RF Level on the 8902A measuring receiver to an accuracy of ±(0.015 dB + 0.005 dB/10 dB step). Agilent 8902A

Measuring Receiver Technical Specifications

Agilent 11722A Sensor Module

Agilent 11792A Sensor Module

Agilent 11793A Microwave Converter

Agilent 11812A Verification Kit

2

AGILENT 8902A MEASURING RECEIVER*TECHNICAL SPECIFICATIONS

Specifications describe the test set’s warranted performance and are valid over the entire operation and environmental ranges unless otherwise noted. All specifications are valid after a 30-minute warm-up period of continuous opera-tion, and within the frequency ranges defined below.Supplemental characteristics are intended to provide additional information useful in applying the instru-ment by giving typical, but non-warranted performance parameters. These characteristics are shown in Italics and labeled as “nominal,” “typical,” or “supplemental.”*Shaded text signifies measurements made with the 8902A measuring receiver using the 11793A microwave converter and 11792A sensor module. With this config-uration, all standard 8902A specifications apply except where changes are shown as shaded text.

Frequency Modulation

RATES 1:

20 Hz to 10 kHz, 150 kHz ≤f c <10 MHz.20 Hz to 200 kHz, 10 MHz ≤f c ≤1300 MHz.20 Hz to 200 kHz, 10 MHz ≤f c ≤26.5 GHz.DEVIATIONS 1:

40 kHz peak maximum, 150 kHz ≤f c <10 MHz.400 kHz peak maximum, 10 MHz ≤f c ≤1300 MHz.400 kHz peak maximum, 10 MHz ≤f c ≤26.5 GHz.ACCURACY 1, 2, 3:

FM Accuracy Frequency Range Rates Deviations ±2% of reading 250 kHz – 10 MHz 20 Hz – 10 kHz ≤40 kHz peak ±1 digit ±1% of reading 10 MHz – 1300 MHz 50 Hz – 100 kHz ≤400 kHz peak ±1 digit ±5% of reading 10 MHz – 1300 MHz 20 Hz – 200 kHz ≤400 kHz peak ±1 digit

For rms detector add ±3% of reading.

DEMODULATED OUTPUT DISTORTION 1, 4:

THD Frequency Range Rates Deviations <0.1%400 kHz – 10 MHz 20 Hz – 10 kHz <10 kHz <0.1%10 MHz – 1300 MHz 20 Hz – 100 kHz <100 kHz AM REJECTION (50 Hz to 3 kHz BW)3:

AM Rejection

Frequency Range Rates AM Depths <20 Hz peak 150 kHz – 1300 MHz 400 Hz or 1 kHz ≤50%deviation RESIDUAL FM (50 Hz to 3 kHz BW):

<8 Hz rms at 1300 MHz, decreasing linearly with frequency to <1 Hz rms for 100 MHz and below.<17 Hz rms'1300 MHz

MAXIMUM FM DEVIATION, RESOLUTION, AND MAXIMUM DEMODULATED OUTPUT SENSITIVITY

ACROSS AN OPEN CIRCUIT (600 ?output impedance)5:

Maximum Maximum Demodulated Deviations Resolution Output Sensitivity

(D F)100 Hz 0.01 mV/Hz D F peak ≥40 kHz 10 Hz 0.1 mV/Hz 4.0 kHz ≤D F peak <40 kHz 1 Hz 1.0 mV/Hz D F peak < 4 kHz 0.1 Hz

1.0 mV/ Hz

D F rms < 0.3 kHz

(rms detector only)

Resolution is increased one digit with 750 μs de-empha-sis and pre-display on.

The demodulated output signal present at the

Modulation Out/Audio In connector is increased in amplitude by a factor of 10 with 750 μs de-emphasis.DEMODULATED OUTPUT DISTORTION 1, 4

:

THD Frequency Range Rates Deviations <0.3%150 kHz – 400 kHz 20 Hz – 10 kHz <10 kHz DETECTORS:+peak, – peak, ±peak/2, peak hold, average (rms sinewave calibrated), rms.

STEREO SEPARATION (50 Hz to 15 kHz): >47 dB.

1.But not to exceed: 20 kHz rates and 40 kHz peak deviations with 750 μs de-emphasis filter.

2.Not to exceed for stated accuracy: 50 Hz to 40 kHz rates with rms detector.

3.Peak residuals must be accounted for in peak readings.

4.With 750 μs de-emphasis and pre-display "off," distortion is not specified for modulation outputs >4V peak. This condition can occur near maximum deviation for a measurement range, at rates <2 kHz.

5.For optimum flatness, cables should be terminated with their characteristic impedance.

3

Amplitude Modulation

RATES:

20 Hz to 10 kHz, 150 kHz ≤f c <10 MHz.20 Hz to 100 kHz, 10 MHz ≤f c ≤1300 MHz.DEPTH:to 99%.ACCURACY 2, 3, 6:

AM Accuracy Frequency Range Rates Depths ±2% of reading 150 kHz – 10 MHz 50 Hz – 10 kHz 5% – 99%±1 digit ±3% of reading 150 kHz – 10 MHz 20 Hz – 10 kHz to 99%±1 digit ±1% of reading 10 MHz – 1300 MHz 50 Hz – 50 kHz 5% – 99%±1 digit ±3% of reading 10 MHz – 1300 MHz

20 Hz – 100 kHz to 99%±1 digit

FLATNESS 5, 7:

Flatness Frequency Range Rates

Depths

±0.3% of reading

10 MHz – 1300 MHz 90 Hz – 10 kHz 20% – 80%±1 digit DEMODULATED OUTPUT DISTORTION:<0.3% THD for ≤50% depth. <0.6% THD for ≤95% depth.For f c >1300 MHz add 0.4% THD.FM REJECTION (50 Hz to 3 kHz BW)3:

FM Rejection Frequency Range Rates Deviations <0.2% AM

250 kHz – 10 MHz 400 Hz or 1 kHz <5 kHz peak <0.2% AM 10 MHz – 1300 MHz 400 Hz or 1 kHz <50 kHz peak RESIDUAL AM (50 Hz to 3 kHz BW):<0.01%rms .Supplemental Characteristics:

DETECTORS:+peak, –peak, ±peak/2, peak hold, average (rms sinewave calibrated), rms.

MAXIMUM DEPTH, RESOLUTION, AND MAXIMUM DEMODU-LATED OUTPUT SENSITIVITY ACROSS AN OPEN CIRCUIT (600 ?output impedance)5:

Maximum Maximum Demodulated Depths

Resolution Output Sensitivity

0.1%0.01V / percent AM peak ≥40.0%0.01%0.1V / percent AM peak <40.0%0.001%

(rms detector only)

0.1V / percent

AM rms <3.0%

Phase Modulation

RATES:

200 Hz to 10 kHz, 150 kHz ≤f c <10 MHz.200 Hz to 20 kHz, 10 MHz ≤f c ≤1300 MHz.200 Hz to 20 kHz, 10 MHz ≤f c ≤26.5 GHz.ACCURACY 3:

±4% of reading ±1 digit, 150 kHz ≤f c <10 MHz.±3% of reading ±1 digit, 10 MHz ≤f c ≤1300 MHz.±3% of reading ±1 digit, 10 MHz ≤f c ≤26.5 GHz.For rms detector add ±3% of reading.DEMODULATED OUTPUT DISTORTION:<0.1% THD.AM REJECTION (for 50% AM at 1 kHz rate)3:<0.03 radians peak (50 Hz to 3 kHz BW).

MAXIMUM DEVIATION, RESOLUTION, AND MAXIMUM DEMODULATED OUTPUT SENSITIVITY ACROSS AN OPEN CIRCUIT (600 ?output impedance)5:

Supplemental Characteristics:

MODULATION RATES:usable from 20 Hz to 100 kHz with degraded performance.

DETECTORS:+peak, – peak, ±peak/2, peak hold, average (rms sinewave calibrated), rms.

2.Not to exceed for stated accuracy: 50 Hz to 40 kHz rates with rms detector.

3.Peak residuals must be accounted for in peak readings.

5.For optimum flatness, cables should be terminated with their characteristic impedance.

6.For peak measurements only: AM accuracy may be affected by distortion generat-ed by the measuring receiver. In the worst case this distortion can decrease accu-racy by 0.1% of reading for each 0.1% of distortion.

7.Flatness is the variation in indicated AM depth for constant depth on input signal.

Modulation Reference

AM CALIBRATOR DEPTH AND ACCURACY:

33.33% depth nominal, internally calibrated to an accuracy of ±0.1%. FM CALIBRATOR DEVIATION AND ACCURACY:

34 kHz peak deviation nominal, internally calibrated to an accuracy of ±0.1%.

Supplemental Characteristics:

CARRIER FREQUENCY:10.1 MHz.

MODULATION RATE: 10 kHz.

OUTPUT LEVEL:– 25 dBm.

Frequency Counter

RANGE:

150 kHz to 1300 MHz.

150 kHz to 26.5 GHz.

SENSITIVITY:

12 mV rms(–25 dBm), 150 kHz ≤f c≤650 MHz.

22 mV rms(–20 dBm), 650 MHz ≤f c≤1300 MHz.

40 mV rms(–15 dBm), 150 kHz ≤f c≤650 MHz.

71 mV rms(–10 dBm), 650 MHz

40 mV rms(–15 dBm), 1300 MHz

MAXIMUM RESOLUTION:

1 Hz.

10 Hz.

ACCURACY:

±reference accuracy ±3 counts of least-significant digit,

f c<100 MHz.

±reference accuracy ±3 counts of least-significant digit, or

30 Hz, whichever is larger, f c≥100 MHz.

Supplemental Characteristics:

MODES:Frequency and Frequency Error (displays the difference between the frequency entered via the keyboard and the actual RF input frequency).

SENSITIVITY IN MANUAL TUNING MODE:

Approximate frequency must be entered from keyboard.

0.22 mV rms(–60 dBm).

0.71 mV rms(–50 dBm).

Using the RF amplifier and the IF amplifiers, sensitivity can be increased to approximately:

–100 dBm.

–90 dBm, f c≤1300 MHz.

–75 dBm, 1300 MHz

8.After 30-day warm-up.

9.The 8902A fundamental RF power measurement units are watts. Further internal

processing is done on this number to display all other units.

10.When using a power sensor, the noise specification may mask the linearity speci-

fication and become the predominant error. When operating on the top RF power range, add the power sensor's linearity percentages found in the power sensor's specifications.Internal Time Base Reference

FREQUENCY:10 MHz.

AGING RATE:

<1 x 10–6/month.

<1 x 10–9/day (Option 002)8.

Supplemental Characteristics:

INTERNAL REFERENCE ACCURACY:

Overall accuracy is a function of timebase calibration, aging rate, temperature effects, line voltage effects, and short-term stability.

Standard Option 002 Aging Rate<1 x 10–6/mo.<1 x 10–9/day Temperature Effects<2 x 10–7/°C<2 x 10–10/°C

Line Voltage Effects

(+5%, –10% Line<1 x 10–6<6 x 10–10

Voltage Change)

Short-Term Stability—<1 x 10–9for

1 second average

RF Power

The Agilent 8902A measuring receiver, with

11722A sensor module, performs RF power meas-urements from –20 dBm (10 μW) to +30 dBm (1 W) at frequencies from 100 kHz to 2.6GHz.

The 8902A measuring receiver, with 11792A sensor module, performs RF power measurements from

–20 dBm (10μW) to +30 dBm (1 W) at frequencies from 50 MHz to 26.5 GHz.

RF POWER RESOLUTION9:

0.01% of full scale in watts or volts mode.

0.01 dB in dBm or dB relative mode.

LINEARITY (includes sensor non-linearity):

RF range linearity ±RF range-to-range change error.

RF RANGE LINEARITY (using recorder output)10:

±0.02 dB, RF ranges 2 through 5.

±0.03 dB, RF range 1.

Using front-panel display add ±1 count of least-significant digit. RF RANGE-TO-RANGE CHANGE ERROR (using recorder output): ±0.02 dB/RF range change from reference range.

Using front-panel display add ±1 count of least-significant digit. INPUT SWR:

Using 11722A sensor module: <1.15.

Using 11792A sensor module:

<1.15, 1300 MHz ≤f c.

<1.25, 1300 MHz

<1.40, 18.0 GHz

4

ZERO SET (digital settability of zero):

±0.07% of full scale on lowest range.

Decrease by a factor of 10 for each higher range. Supplemental Characteristics:

ZERO DRIFT OF METER:

±0.03% of full scale/°C on lowest range. Decrease by a

factor of 10 for each higher range.

NOISE (at constant temperature, peak change over any one-minute interval for the 11722A or 11792A sensor modules):

0.4% of full scale on range 1 (lowest range).

0.13% of full scale on range 2.

0.013% of full scale on range 3.

0.0013% of full scale on range 4.

0.00013% of full scale on range 5.

ZERO DRIFT OF SENSORS (1 hour, at constant temperature after 24-hour warm-up):

±0.1% of full scale on lowest range for 11722A

and 11792A sensor modules.

RF POWER RANGES OF AGILENT8902A MEASURING RECEIVER WITH 11722A AND 11792A SENSOR MODULES:

–20 dBm to –10 dBm (10 μW to 100 μW), range 1.

–10 dBm to 0 dBm (100 μW to 1 mW), range 2.

0 dBm to +10 dBm (1 mW to 10 mW), range 3.

+10 dBm to +20 dBm (10 mW to 100 mW), range 4.

+20 dBm to +30 dBm (100 mW to 1 W), range 5. RESPONSE TIME (0 to 99% of reading):

<10 seconds, range 1.

<1 second, range 2.

<100 milliseconds, ranges 3 through 5.

DISPLAYED UNITS:

Watts, dBm, dB relative, %relative, volts, mV, μV, dB V, dB mV, dB μV. INTERNAL NON-VOLATILE CAL-FACTOR TABLES

(user-modifiable using special functions):

Maximum number of cal factor/frequency entries:

Table #1 (primary):16 pairs plus Reference Cal Factor.

Table #2(frequency offset): 22 pairs plus Reference Cal Factor. Maximum Allowed Frequency Entry:42 GHz.

Frequency Entry Resolution: 50 kHz.

Cal Factor Range:40 to 120%.

Cal Factor Resolution: 0.1%.

Power Reference

POWER OUTPUT:

1.00 mW. Factory set to ±0.7%, traceable to the U.S. National

Bureau of Standards.

ACCURACY:±1.2% worst case (±0.9% rss) for one year (0 °C to 55 °C).

Supplemental Characteristics:

FREQUENCY:50 MHz nominal.

SWR:1.05 nominal.

FRONT PANEL CONNECTOR:N-type female.Tuned RF Level

POWER RANGE:–127 dBm to 0 dBm, using IF synchronous detector (200 Hz BW).

–100 dBm to 0 dBm, using IF average detector (30 kHz BW). POWER RANGE (Using 11792A Sensor Module):

For IF Synchronous Detector:

+10 dBm to –117 dBm, 2.5 MHz ≤f c≤1300 MHz.

+5 dBm to –105 dBm, 1300 MHz ≤f c≤12.4 GHz.

+5 dBm to –100 dBm, 12.4 GHz ≤f c≤18.0 GHz.

+5 dBm to –95 dBm, 18.0 GHz ≤f c≤26.5 GHz.

For IF Average Detector:

+10 dBm to –90 dBm, 2.5 MHz ≤f c≤1300 MHz.

+5 dBm to –80 dBm, 1300 MHz ≤f c≤12.4 GHz.

+5 dBm to –75 dBm, 12.4 GHz ≤f c≤18.0 GHz.

+5 dBm to –70 dBm, 18.0 GHz ≤f c≤26.5 GHz.

1.9 Special Function degrades Tuned RF Level minimum

sensitivity by 10 dB.

FREQUENCY RANGE:

2.5 MHz to 1300 MHz.

2.5 MHz to 26.5 GHz.

DISPLAYED RESOLUTION11:

4 digits in watts or volts mode.

0.01 dB or 0.001 dB in dBm or dB relative mode.

4 digits in watts or volts mode.

0.01 dB in dBm or dB relative mode.

RELATIVE MEASUREMENT ACCURACY (at constant temperature and after RF range calibration is completed)12:

Detector linearity + IF range-to-range error + RF range-to-range error + frequency drift error + noise error ±1 digit.

Detector linearity + mixer linearity + IF range-to-range error + RF range-to-range error + frequency drift error + noise error ±1 digit.

11.The 8902A fundamental Tuned RF Level measurement units are volts. Further

internal processing is done on this number to display all other units.

12.Tuned RF Level accuracy will be affected by residual FM of the source-under-test.

If the residual FM peak is >50 Hz measured over a 30 second period in a 3 kHz BW.

Tuned RF Level measurements should be made using the IF average detector (30 kHz BW) by using Special Function 4.4. The Tuned RF Level measurement sensi-tivity when using the IF average detector is –100 dBm.

5

6

Carrier Noise (Options 030-037)

FREQUENCY RANGE:10 MHz to 1300 MHz.

CARRIER POWER RANGE: +30 dBm to –20 dBm;

12.5 kHz, 25 kHz and 30 kHz filters.

+30 dBm to –10 dBm; carrier noise filter.

DYNAMIC RANGE:115 dB.

CARRIER REJECTION (temp. ≤35 °C):>90 dB; for offsets of at least 1 channel spacing or 5 kHz, whichever is greater. RELATIVE MEASUREMENT ACCURACY:

±0.5 dB; levels ≥–95 dBc; 12.5 kHz, 25 kHz and

30 kHz filters.

±0.5 dB; levels ≥–129 dBc/Hz; carrier noise filter.

CARRIER NOISE FILTER:

Filter Noise Bandwidth:2.5 kHz nominal.

Noise Bandwidth Correction Accuracy (stored in non-volatile memory): ±0.2 dB.

Supplemental Characteristics:

ADJACENT/ALTERNATE CHANNEL FILTERS:

6 dB Filter Bandwidth:

8.5 kHz, 12.5 kHz adjacent-channel filter.

16.0 kHz, 25 kHz adjacent-channel filter.

30.0 kHz, 30 kHz (cellular radio) alternate-channel filter. TYPICAL NOISE FLOOR:–150 dBc/Hz, 0 dBm carrier power level. For system noise performance add LO contribution.

14.With the low-pass and high-pass audio filters used to stabilize frequency readings.Audio Frequency Counter

FREQUENCY RANGE:

20 Hz to 250 kHz. (Usable to 600 kHz.)

MAXIMUM EXTERNAL INPUT VOLTAGE:3V rms.

ACCURACY(for demodulated signals)14:

Accuracy Frequency Modulation (Peak)±3 counts of least-significant digit>1 kHz AM ≥10%

±Internal Reference Accuracy FM ≥1.0 kHz

f M≥1.5 radians

±0.02 Hz≤1 kHz AM≥10%

±Internal Reference Accuracy FM≥1.0 kHz

f M≥1.5 radians

±0.2 Hz≤3 kHz 1.5%≤AM<10% ±Internal Reference Accuracy0.15 kHz≤FM (3 kHz low-pass filter inserted)<1.0 kHz

0.15 radian≤f M

<1.5 radians

ACCURACY (for external signals)14:

Accuracy Frequency Level

±3 counts of least-significant digit>1 kHz≥100 mV rms ±Internal Reference

±0.02 Hz≤1 kHz≥100 mV rms

±Internal Reference Accuracy

Supplemental Characteristics:

DISPLAYED RESOLUTION:6 digits.

MEASUREMENT RATE: 2 readings per second.

COUNTING TECHNIQUE: Reciprocal with internal 10 MHz timebase. AUDIO INPUT IMPEDANCE: 100 k?nominal.

Audio RMS Level

FREQUENCY RANGE:50 Hz to 40 kHz.

VOLTAGE RANGE:100 mV to 3 V.

ACCURACY:±4.0% of reading.

Supplemental Characteristics:

FULL RANGE DISPLAY: 0.3000 V, 4.000 V.

AC CONVERTER: True-rms responding for signals with crest factor of ≤3.

MEASUREMENT RATE: 2 readings per second.

AUDIO INPUT IMPEDANCE:100 k?nominal.

7

Audio Distortion

FUNDAMENTAL FREQUENCIES:400 Hz ±5% and 1 kHz ±5%. MAXIMUM EXTERNAL INPUT VOLTAGE: 3 V.

DISPLAY RANGE:0.01% to 100.0% (–80.00 dB to 0.00 dB). DISPLAYED RESOLUTION: 0.01% or 0.01 dB.

ACCURACY: ±1 dB of reading.

SENSITIVITY:

Modulation:0.15 kHz peak FM, 1.5% peak AM or

0.6 radian peak f M.

External:100 mV rms.

RESIDUAL NOISE AND DISTORTION15:

0.3% ( –50 dB), temperature<40 °C.

Supplemental Characteristics:

MEASUREMENT 3 dB BANDWIDTH:20 Hz to 50 kHz. DETECTION: True rms.

MEASUREMENT RATE: 1 reading per second.

AUDIO INPUT IMPEDANCE: 100 k?nominal.

Audio Filters

DE-EMPHASIS FILTERS:25 m s, 50 m s, 75 m s, and 750 m s. De-emphasis filters are single-pole, low-pass filters with 3 dB frequen-cies of: 6366 Hz for 25 m s, 3183 Hz for 50 m s, 2122 Hz for 75 m s, and 212 Hz for 750 m s.

50 Hz HIGH-PASS FILTER (2 pole):

Flatness:<1% at rates ≥200 Hz.

300 Hz HIGH-PASS FILTER (2 pole):

Flatness:<1% at rates ≥1 kHz.

3 kHz LOW-PASS FILTER (5 pole):

Flatness:<1% at rates ≤1 kHz.

15 kHz LOW-PASS FILTER (5 pole):

Flatness:<1% at rates ≤10 kHz.

>20 kHz LOW-PASS FILTER (9 pole bessel)16:

Flatness:<1% at rates ≤10 kHz.

Supplemental Characteristics:

DE-EMPHASIS FILTER TIME CONSTANT ACCURACY:±3%. HIGH PASS AND LOW PASS FILTER 3 dB CUTOFF FREQUENCY ACCURACY: ±3%.

>20 kHz LOW PASS FILTER 3 dB CUTOFF FREQUENCY: 100 kHz nominal.

OVERSHOOT ON SQUARE WAVE MODULATION16: <1%.RF Input

FREQUENCY RANGE:150 kHz to 1300 MHz.

150 kHz to 26.5 GHz when using the 11793A sensor module.

OPERATING LEVEL:

Minimum Maximum Frequency Range Operating Level Operating Level

12 mV rms(–25 dBm)7 V rms(1 W peak)150 kHz – 650 MHz

Source SWR <4

22 mV rms(–20 dBm)7 V rms(1 W peak)650 MHz – 1300 MHz

Source SWR <4

Supplemental Characteristics:

TUNING:

Normal Mode: Automatic and manual frequency entry.

Track Mode: Automatic and manual frequency entry, f c≥10 MHz. Normal and Track Mode: Manual entry of approximate frequency. Acquisition Time (automatic operation): ~1.5 seconds.

INPUT IMPEDANCE:50 ?nominal.

MAXIMUM SAFE DC INPUT LEVEL:5 V dc.

General Specifications

TEMPERATURE:

Operating: 0 °C to 55 °C.

Storage:– 55 °C to 75 °C.

REMOTE OPERATION:GPIB; all functions except the line switch are remotely controllable.

DEFINED IN IEEE-488.2 GPIB COMPATIBILITY: SH1, AH1, T5, TE0, L3, LE0, SR1, RL1, PP0, DC1, DT1, C0, E1.

EMI:Conducted and radiated interference is within the require-ments of VDE 0871 (Level B), and CISPR publication 11.

POWER: 100, 120, 220, or 240V (+5%, –10%); 48 to 66 Hz; 200 VA maximum.

WEIGHT: Net 23.4 kg (52 lb); Shipping 31.4 kg (69 lb).

DIMENSIONS:190 mm H x 426 mm W x 551 mm D

(7.5" x 16.8" x 21.7").

15.For demodulated signals, the residual noise generated by the 8902A must be

accounted for in distortion measurements (that is residual AM, FM or f M).

16.The >20 kHz low-pass filter is intended for minimum overshoot with squarewave

modulation.

8

OPTION 050 SPECIFICATIONS

FREQUENCY RANGE:2.5 MHz to 26.5 GHz.

TUNED RF LEVEL DYNAMIC RANGE:

–120 dBm to 0 dBm.

–110 dBm to –15 dBm.

POWER ACCURACY:

Using an Agilent 8902A Option 050 with 11722A sensor module (10 to 1300 MHz):

Relative accuracy:

±0.005 dB/10 dB step (0 to –100 dBm).

±0.050 dB/10 dB step (–100 to –120 dBm).

±0.015 dB ±1 digit.

Absolute accuracy:

±0.005 dB/10 dB step (0 to –100 dBm).

±0.050 dB/10 dB step (–100 to –120 dBm).

±0.120 dB ±1 digit.

Using an Agilent 8902A Option 050 with 11722A sensor module and 11793A microwave converter

(1300 to 2600 MHz, –15 to –110 dBm):

Relative accuracy, 85 dB dynamic range:

±0.005 dB/10 dB step (0 to 60 dB).

±0.050 dB/10 dB step (60 to 85 dB).

±0.015 dB ±1 digit.

Absolute accuracy:

±0.005 dB/10 dB step (–15 to –100 dBm).

±0.050 dB/10 dB step (–100 to –110 dBm).

±0.120 dB ±1 digit.

Using an Agilent 8902A Option 050 with 11792A sensor module and 11793A microwave converter

(1300 MHz to 26.5 GHz, –15 to –100 dBm):

Relative accuracy, 85 dB dynamic range:

±0.005 dB/10 dB step (0 to 60 dB).

±0.050 dB/10 dB step (60 to 85 dB).

±0.015 dB ±1 digit.

Absolute accuracy:

±0.005 dB/10 dB step (–15 to –100 dBm).

±0.120 dB ±1 digit.

INPUT SWR:

<1.18, RF range 1 and 2.

<1.40, RF range 3.

TEMPERATURE:

Operating:15 °C to 30 °C.

Storage:–55 °C to 74 °C.

Supplemental Characteristics:

MEASUREMENT TIME:

10 to 30 seconds.AGILENT 11793A MICROWAVE CONVERTER SPECIFICATIONS

LO AMPLITUDE RANGE:

+8 dBm to +13 dBm, 2 GHz to 18 GHz.

+7 dBm to +13 dBm, 18 GHz to 26.5 GHz.

0 dBm to + 5 dBm, 18 GHz to 26.5 GHz with Option 001,

011, or 021.

TEMPERATURE:

Operation:0 °C to 55 °C.

Storage:–55 °C to 75 °C.

–25 °C to 75 °C (Options 001, 011, and 021).

POWER:100, 120, 220, or 240 (+5%, –10%); 48 to 66 Hz;

20 VA maximum.

WEIGHT: Net 7.5 kg (16.5 lb); shipping 10.9 kg (24 lb). DIMENSIONS:88 mm H x 425 mm W x 528 mm D. Supplemental Characteristics:

RF INPUT CONNECTOR:3.5 mm male.

LO INPUT CONNECTOR:3.5 mm male.

IF OUTPUT CONNECTOR:N-type female.

REAR PANEL CONTROL CONNECTOR:BNC female. INCLUDED ACCESSORIES:

Control Cable: 11170A BNC cable.

LO Output to 11793A LO Input Cable:3.5 mm female to 3.5 mm female flexible cable and 3.5 mm male to N-type male adapter; Options 001, 011, and 021 delete the 3.5 mm to N-type adapter. 8902A RF input to 11793A IF output cable: N-type male to N-type male flexible cable.

9

AGILENT 11722A SENSOR MODULE SPECIFICATIONS

FREQUENCY RANGE:100 kHz to 2.6 GHz.

POWER RANGE:+30 dBm (1 watt) to – 20 dBm (10 m W). INPUT SWR (connected to an 8902A):

<1.15, for RF Power measurements.

<1.33, for Tuned RF Level measurements, RF range 1 and 2.

<1.5, for Tuned RF Level measurements, RF range 3.

<1.33, for Tuned RF Level measurements, RF range 3 with

Special Function 1.9.

POWER SENSOR LINEARITY:

+2%, – 4%; +30 dBm to +20 dBm.

Negligible deviation, levels <+20 dBm.

CALIBRATION FACTORS:

Each 11722A sensor module is individually calibrated. The calibra-tion factors are printed on the 11722A sensor module for easy ref-erence.

CAL FACTOR UNCERTAINTY:

Frequency RSS Uncertainty Worst Case (MHz)Uncertainty

0.10.7 % 1.6%

0.30.7% 1.6%

1.00.8% 1.7%

3.00.8% 1.7%

10.00.9% 2.0%

30.00.9% 2.0%

50.00.0% (ref)0.0% (ref)

100.0 1.1% 2.2%

300.0 1.1% 2.2%

1000.0 1.1% 2.2%

2600.0 1.2% 2.3%

Supplemental Characteristics:

MAXIMUM PEAK POWER:100 Wpeak or 300 W ms per pulse. INPUT IMPEDANCE:50 ?nominal.

INPUT CONNECTOR:N-type male.

SWITCH LIFE: >1,000,000 switchings.

SWITCH ISOLATION:>90 dB.

WEIGHT:Net 0.8 kg (1.75 lb); Shipping 1.2 kg (2.6 lb). DIMENSIONS:51.2 mm H x 62.4 mm W x 1935 mm D

(2" x 2.5" x 76.2").AGILENT 11792A SENSOR MODULE SPECIFICATIONS

FREQUENCY RANGE:

RF Power measurements:

50 MHz to 26.5 GHz.

50 MHz to 18.0 GHz, Option 001.

POWER RANGE:+30 dBm (1 watt) to – 20 dBm (10 m W). INPUT SWR (connected to an Agilent 11793A):

<1.15, 1300 MHz ≤f c.

<1.25, 1300 MHz

<1.40, 18.0 GHz

POWER SENSOR LINEARITY:

+2%, – 4%; +30 dBm to +20 dBm.

Negligible deviation, levels <+20 dBm.

CALIBRATION FACTORS:

Each 11792A sensor module is individually calibrated. The calibra-tion factors are printed on the 11792A sensor module for easy ref-erence.

CAL FACTOR UNCERTAINTY:

Frequency RSS Uncertainty Worst Case

Uncertainty

2.0 GHz 2.3 4.6%

6.0 GHz 2.5 5.0%

10.0 GHz 2.9 5.7%

14.0 GHz 3.4 6.6%

18.0 GHz 3.7 6.9%

22.0 GHz 3.87.8%

26.5 GHz 4.18.3%

Supplemental Characteristics:

INPUT CONNECTOR: 3.5 mm male (N-type male, Option 001). INPUT IMPEDANCE: 50 ?nominal.

SWITCH LIFE: >1,000,000 switchings.

WEIGHT: Net 0.8 kg (1.75 lb); Shipping 1.2 kg (2.6 lb). DIMENSIONS: 51.2 mm H x 62.4 mm W x 1935 mm D

(2" x 2.5" x 76.2").

10

AGILENT 11812A VERIFICATION KIT

SPECIFICATIONS

FREQUENCY:30 MHz.

11812A ACCURACY:±(0.003 dB + 0.003 dB/10 dB step).

OPTION 050 WORST CASE CUMULATIVE TUNED RF LEVEL

ACCURACY VERIFIED WITH 11812A:

±0.010 dB/10 dB step (0 to –100 dBm).

±0.050 dB/10 dB step (–100 to –120 dBm).

±0.015 dB ±1 digit.

TEMPERATURE:

Operation:15 °C to 30 °C.

Storage:–55 °C to 74 °C.

AGILENT 8902A REAR PANEL

INPUTS/OUTPUTS

Supplemental Characteristics:

FM OUTPUT: 10 k?impedance, –9 V to 6 V into an open circuit,

~6 V/MHz, dc coupled, 16 kHz bandwidth (one pole).

AM OUTPUT: 10 k?impedance, –4 V to 0 V into an open circuit,

~8 mV/%, dc coupled, 16 kHz bandwidth (one pole).

RECORDER OUTPUT: DC voltage proportional to the measured

results, 1 k?impedance, 0 V to 4 V for each resolution range into

an open circuit.

IF OUTPUT: 50 ?impedance, 150 kHz to 2.5 MHz, –27 dBm to –3

dBm.

10 MHz REFERENCE OUTPUT: 50 ?impedance, TTL levels (0 V to

>2.2 V into an open circuit). Available only with Option 002

1x10–9/day internal reference.

10 MHz REFERENCE INPUT 17:>500 ?impedance, 0.5 V peak-to-peak

minimum input level.

LO INPUT (Option 003): 50 ?impedance, ~1.27 MHz to 1301.5

MHz, 0 dBm nominal.

RF SWITCH REMOTE CONTROL OUTPUT: Provides output signals

necessary to remotely control either an Agilent 33311B,C Option

011 or an 8761A RF switch.

FREQUENCY OFFSET MODE REMOTE CONTROL OUTPUT:TTL

high output if in frequency offset mode (Special Function 27.1 or

27.3) with an external LO frequency >0, TTL low output for all

other cases.

17.External reference accuracy affects accuracy of all measurements.

11

Agilent Technologies’ Test and Measurement

Support, Services, and Assistance

Agilent Technologies aims to maximize the value you receive, while minimizing your risk and problems. We strive to ensure that you get the test and measurement capabilities you paid

for and obtain the support you need. Our extensive support resources and services can help you choose the right Agilent products for your applications and apply them successfully. Every instrument and system we sell has a global warranty. Support is available for at least five years beyond the produc-tion life of the product. Two concepts underlie Agilent’s overall support policy: “Our Promise” and “Your Advantage.”Our Promise

“Our Promise” means your Agilent test and measurement equip-ment will meet its advertised performance and functionality. When you are choosing new equipment, we will help you with product information, including realistic performance specifica-tions and practical recommendations from experienced test engineers. When you use Agilent equipment, we can verify that it works properly, help with product operation, and provide basic measurement assistance for the use of specified capabili-ties, at no extra cost upon request. Many self-help tools are available.

Your Advantage

“Your Advantage” means that Agilent offers a wide range of additional expert test and measurement services, which you can purchase according to your unique technical and business needs. Solve problems efficiently and gain a competitive edge by contracting with us for calibration,extra-cost upgrades, out-of-warranty repairs, and on-site education and training, as well as design, system integration, project management, and other professional services. Experienced Agilent engineers and tech-nicians worldwide can help you maximize your productivity, optimize the return on investment of your Agilent instruments and systems, and obtain dependable measurement accuracy

for the life of those products.By internet, phone, or fax, get assistance with all your test and measurement needs.

Online Assistance

https://www.wendangku.net/doc/cb11330568.html,/find/assist

Phone or Fax

United States:

(tel) 1 800 452 4844

Canada:

(tel) 1 877 894 4414

(fax) (905) 206 4120

Europe:

(tel) (31 20) 547 2323

(fax) (31 20) 547 2390

Japan:

(tel) (81) 426 56 7832

(fax) (81) 426 56 7840

Latin America:

(tel) (305) 269 7500

(fax) (305) 269 7599

Australia:

(tel) 1 800 629 485

(fax) (61 3) 9210 5947

New Zealand:

(tel) 0 800 738 378

(fax) (64 4) 495 8950

Asia Pacific:

(tel) (852) 3197 7777

(fax) (852) 2506 9284

Product specifications and descriptions in this

document subject to change without notice.

Copyright ? 1985, 2000 Agilent Technologies

Printed in U.S.A. 10/00

5968-5312E

LWHB汽车外廓尺寸自动测量仪说明书

L W H-25B型汽车外廓尺寸自动测量仪 随机技术文件 使用说明书

目次 使用说明书 一、用途及适用范围............................................ 二、主要技术参数.............................................. 三、测量仪组成和测试原理...................................... 四、测量仪特点 (2) 五、安装与标定 (3) 六、使用方法 (12) 七、维护与保养 (23) 八、常见故障及排除方法 (23) 九、电气接口参数 (24) 测量仪基础图 (25) 正文 (26)

一、用途及适用范围 LWH-25B型汽车外廓尺寸自动测量仪(以下简称测量仪)用于自动测量汽车长宽高外廓尺寸、货车栏板高度。 二、主要技术参数 三、测量仪组成和测量原理 (一)测量仪组成 测量仪由计算机(工控机)、激光雷达、测控软件、高速光电(轴距测量仪需要)、高速信号采集卡(轴距测量仪需要)等安装附件等组成。 1、计算机(工控机) 工控机硬件配置:Intel新一代酷睿I7处理器,DDR3 2G内存,32G固态硬盘,2个千兆以太网口等。 2、激光雷达 德国原装进口,检测距离50米,扫描频率50HZ,支持270度扫描。 3、测控软件 自主研发的测控软件具有高度的灵活性、实时性、可视化。

4、高速信息采集卡 自主研发的高速采集卡,配备16通道的模数转换器、8通道光电等,配置两个串口,波特率高达38400bit/s。 (二)测量原理 由左,右两个扫描仪把车辆切割成许多切片,由前扫描仪确定切片的厚度,通过软件系统中的算法,对切片数据进行合成,运算,得到车辆轮廓的长,宽,高数据、并结合高速信号采集卡对高速光电信号的分析计算出所有轴距数据 图1 测试过程示意图 四、测量仪特点 (一)满足新修订的国标GB21861的要求; (二)结构简单。可以迅速完成安装调试工作。除开始检测和检测结束时由操作员电脑操作外,实现了无人参与自动测量,汽车不停车,低速通过(0~5 km/h); (三)全自动测量车辆外廓长、宽、高尺寸,包括普通客货车辆以及目前公安交管部门重点监察的危化品运输车,各种特种工程车辆、挂车货箱长度栏板高度,测量过程无需人工干预; (四)测量过程耗时短,在车辆通过15秒内输出测量结果; (五)电脑屏幕上动态显示出车辆的三维外廓图形及尺寸数据; (六)采用SQL 对测试结果进行保存,方便查阅和审核; (七)车辆外廓尺寸检测后,依据国家标准GB1589-2004《道路车辆外廓尺寸、轴荷及质量限值》要求进行数据处理、判定,便于对车辆的监督管理; (八)系统软件基于WINDOWS操作系统; (九)检测系统具有与各级管理部门的联网接口,可方便的接入上级系统; (十)环境适应性强,对光线变化不敏感,适用于夜间和室外的工作环境,除恶劣气候条件(如大风、大雾、雨天等)外,可实现全天候、24小时和-30℃~50℃条件下正常作业(用于严寒条件下,采用自带加热的测量雷达); (十一)设备维护简单、迅速。设备使用一段时间后,如果因各种原因造成雷达定位偏差,可在软件帮助下迅速完成雷达的重新定位,使整套系统迅速重新投入工作;

测量仪器说明书

目录 一、GeoPluse浅地层剖面仪操作规程 (1) 1、仪器简介 (1) 1)功能简介 (1) 2)系统配置 (1) 2、GeoPluse浅地层剖面仪系统配置连接 (1) 1)换能器安装 (1) 2)5430A收发机与5210A接收机连接 (2) 3)接通电源 (4) 3、5210A与5430A收发机功能键简介 (4) 1)5430A收发机功能键简介 (5) 2)5210A接收机功能键简介 (5) 4、数据采集后处理 (7) 二、Knudsen 320Ms双频测深仪操作规程 (14) 1、仪器简介 (14) 1)工作原理 (14) 2)功能简介 (14) 2、系统配置连接 (15) 1)换能器连接 (15) 2)Knudsen 320Ms主机与电脑的连接 (15) 3)接通电源 (16) 3、Knudsen 320Ms菜单结构 (16) 4、数据采集后处理 (21) 三、TideMaster型潮位仪操作规程 (29) 1、仪器硬件设置 (29) 1)主要设备仪器 (29) 2)操作及安装使用 (31) 2、临时验潮站站址选择原则 (31) 3、仪器的软件设置 (31) 四、GPS操作规程 (41) 1、工作原理 (41) 2、基准站操作 (41) 1)仪器架设 (41) 2)用手簿启动基准站 (44) 3、Trimble SPS461 GPS罗经设置及使用说明 (46) 1)网络连接方法设置461 (46)

2)SPS461 信标机定位定向仪液晶屏设置说明 (51) 五、海底管线铺设导航、定位技术 (64) 1、GPS定位原理 (64) 2、海洋定位技术 (65) 1)差分GPS技术 (65) 2)信标差分技术 (65) 3、GPS 控制网及基准站的设立解算 (66) 1)基准站的选定和设立 (66) 2)GPS控制网的布设、施测和解算 (67) 3)测区的坐标七参数的解算 (68) 4)利用转化参数转换坐标 (69) 4、海底管道施工导航定位技术 (69) 1)海底管线临时定位桩施工 (69) 2)铺管船法海底管线铺设导航定位 (71) 六、海底管线预、后调查方案 (75) 1、概述 (75) 1)项目概述 (75) 2)海底管线状态简介 (75) 2、使用检测仪器进行海底管线铺设后调查内容 (76) 1)海底管线外观检查 (76) 2)经处理后的完工调查 (77) 3、后调查作业设备及选用原则 (77) 1)测深设备 (77) 2)旁侧声纳 (78) 3)浅地层剖面仪 (78) 4、调查作业施工 (79) 1)作业准备 (79) 2)计划线布设 (79) 3)作业方法和步骤 (80) 4)作业注意事项 (82) 5、数据采集、记录以及报告 (83) 1)数据采集 (83) 2)数据处理 (84) 6、组织机构与职责 (85)

教你各类温度测试仪的正确使用方法!

教你各类温度测试仪的正确使用方法! 热电偶、无线炉温测试仪都是用来测量温度的仪器。使用它们,能够给我们的工作带来很大的便利。这种温度测试仪功率高、但是功耗低,使用寿命长;而且产品体积小,存储容量大,任何意外都不会丢掉数据。这就很好地解决了安全隐患问题。 虽然说这两种温度测试仪具有相同的用途,但是工作的原理是不一样的,产品结构不同,使用方法也是不同的。 那么,我们该如何正确使用它们呢?下面就分别来了解下吧。 一、热电偶的正确使用方法 众所周知,热电偶可以直接测量温度并把温度信号转换成热电动势信号,再转换成被测介质的温度。常作为测温元件用于跟踪仪中,所以跟踪仪热电偶的正确使用是非常重要,事关着温度曲线的变化情况。 在使用时最容易出现问题的地方就是热电偶了,热电偶是易耗品,但是可以进行维修的。掌握正确的使用方法势在必行,具体如下 一、跟踪仪热电偶在粘贴工件时一定得记住紧密贴合产品工件,不能让热电偶的焊点在里面晃动,特别是热电偶焊点当跟金属碰到一起的时候,容易产生瞬间电压,而导致测出来的温度曲线某个地方温度突然升得很高,这样又会引起分析软件Y轴坐标也跟着变得很高,那么整个曲线就会看起来很小了。 二、热电偶正确的接线方法是:红线接负极,黄线接正极,接的时候两股线一定要往上顶到公插头的三角处,防止裸露的部份短路 三、针对玻璃纤维的热电偶,在插拔时千万要记得用手捏住公插头往外拔,不要去拉线,这样容易造成保护套被拉松或拉出公插头,造成线芯裸露,而导致短路的现象。 四、使用时千万不要打结或折成90度以上,这样很容易造成内部的两根线芯断裂,断了之后你都不知道在哪里断的,那么这根热电偶就没用了,使用时要细心一点,不能野蛮施工。 在操作上要特别注意不要损坏热电偶,不然就会导致跟踪仪测量出来的温度曲线出现异常,影响产品的质量。 二、无线炉温测试仪的正确使用方法 温度测试仪可能你知道,但是大家对温度测试仪的操作方法知道多少,是不是只是照着说明书上面的看一看就开始操作起来了,而对于温度测试仪的操作要点仍然不了解多少。那么小编在这里就给大家详细介绍一下温度测试仪的具体操作方法。 1、首先将温度测试仪接通电源:断定“电压调理”旋钮已置“0”位,然后翻开电源开关。 2、设定“漏电流”值:按下开关“15”,调理“漏电流预置”电位器“14”将“漏电流”预置在所需值。 3、衔接被测件:依据被测件的需求,将测验线和被测件衔接好。 4、“守时测验”:将守时开关“17”置在“守时”方位,调理守时拨盘开关,设定所需的守时时刻,然后按下“发动”开关,并调理“电压调理”旋钮使输出电压至所需值。

二次元影像测量仪(兆丰VIP系列)使用说明书

VIP系列 影像座标测量仪用户手册

目录 前言 (2) 1.仪器规格及技术参数 (2) 1.1影像测量仪具体规格及参数 (3) 1.2仪器所需电脑推荐配置 (3) 2.仪器工作原理及结构 (4) 2.1工作原理 (4) 2.2仪器总体结构 (4) 3.仪器安装 (6) 3.1仪器使用环境 (7) 3.2仪器的安装方法 (7) 4.仪器的使用方法 (8) 5.仪器的维护和保养 (8) 6.仪器成套性 (8) 7.售后服务 (8) 8.常见问题 (9)

前言 VIP系列影像座标测量仪是集光学、精密机械、电子、计算机于一体的精密高效测量仪器。它是一种由高分辨率CCD彩色摄像机、连续变倍物镜、电脑、精密光学尺、高精度工作台与测量软件等组成的高精度、高效率的视频测绘系统。以二维测量为主,也可作为三维视频测量系统,可轻易实现测量、检验、校准、逆向工程等目的,被广泛应用于各种行业。如:手机配件、液晶面板、触摸屏、饮料瓶坏瓶盖、灯饰制品、电子元件、精密模具、刀具、弹簧、导电橡胶、油封止阀、照相机零件、脚踏车零件、汽车零件、PCB加工等,也可用于教学、科研、产品研发等领域。

1.仪器的规格及技术参数 1.1 VIP系列影像坐标测量仪技术参数 附件:随机配送机台工作桌一张、防尘罩一个、自动机标配工业电脑 1.2 VIP系列影像坐标测量仪电脑标准配置(推荐)

2. 仪器工作原理及结构 2.1 工作原理: 影像座标测量仪是通过连续变倍物镜、彩色CCD,通过轮廓透射光或表面光照明将被测工件放大后成像在显示器上的影像放大测量系统。利用专业测量软件对精密光学尺传输的数据和实时影像画面进行图像数据处理,由操作者使用鼠标或键盘在电脑上进行快速描边、标注测量。 2.2VIP300 400 500 仪器总体结构(如图一) 图1 VIP300 400 500影像式测量仪 2.2.1影像式测量仪主要有支撑部分、视频部分、工作台部分及照明部分。 2.2.2 支撑部分包括大理石底座(3)、大理石立柱(10),机台机架(2); 2.2.3 视频部分包括Z轴滑轨座组,镜头,CCD,程控分区照明灯(6)通过软件灯光控制区,可调节灯光的开关亮暗与八方向开关或亮暗,通过Z轴传动组升降,可实现对不同高度工件的测量; 2.2.4 工作台部分包括上层工作台(4)、线性导轨、X轴光栅尺、X轴传动组(11)中层工作台(5)Y 轴传动组、工作台玻璃、V型导轨、Y轴光栅尺、(13、14)通过鼠标或键盘进行各种任务操作,(15)通过摇杆可以快速的移动定位工作台。

红外线体温测温仪使用说明

2020年疫情一直在反反复复,日常出门都必须戴着口罩,进入商城或者小区、办公楼都需要经过体温检测,不少人对红外体温枪是否会产生产生交叉感染有疑惑,下面将详细介绍一下。 一、工作原理 英国物理学家F. W. 赫胥尔在1800 年作各色光研究时发现了红外线,当时称作「不可见之光」,赫胥尔用三棱镜将太阳光分解,并在各色光位置上放上温度计,结果发现位于红外线位置的温度计升温最快,红外线热作用强。之后人们花了一百多年的时间认识红外辐射的电磁本质,了解探索热辐射的基本规律,随着光学技术、电子技术等不断发展,红外技术也日趋完善,其中红外测温技术目前广泛应用于各个领域,其原理是利用物体表面的红外辐射来求得被测温度的。 任何物体只要它的温度高于绝对零度(-273 度),就有热能转变的热辐射向外部发射,物体温度不同,其辐射出的能量不同,且辐射波的波长也不同,但总是包含着红外辐射在内,当物体的温度在千摄氏度以下时,其热辐射中最强的电磁波是红外波。 依据此原理,红外测温枪基本的测温过程是这样的:由人体发射出的能量经光学系统汇聚到红外探测器上,探测器将入射的辐射转换成为电压信号,电压信号送入接收系统后,经过数据处理及曲线自动拟合,最后准确推算出被测人体温度,以数字方式显示输出。 那测到红外辐射能量是怎么计算出物体温度,它们之间一个什么样的关系呢? 19 世纪科学家斯特藩和玻耳兹曼通过实验和计算得出了黑体辐射定律:MB(T) = σT4 (σ为常数),这个定律告诉我们,单位时间从黑体单位面积上辐射出的总辐射能和其本身的热力学温度的4 次方成正比。 当然实际物体(非黑体)的辐射定律一般比较复杂,需借助于黑体的辐射定律来研究,主要是受物体的发射率影响,不同物体的发射率不同,可通过查表或实验得到,红外测温枪可以因物体材质、结构、厚度等等所导致的红外幅射力误差作出校正,比较准确地测出该物体的表面温度。 二、注意事项:

全自动生化仪使用说明书.doc

便携式生化检测仪 340 使用说明书便携式生化检测仪

【产品名称】便携式生化检测仪 【型号】340 【产品性能】 便携式生化检测仪(以下简称POC)。 POC专用于检测本公司体外诊断试剂盒“同型半胱氨酸检测试剂盒”,用于定量检测临床血清或血浆样本中同型半胱氨酸(HCY)。 POC是集样本处理、检测及分析报告一体化的便携式生化检测仪,无需外置电脑和安装软件。一次检测一份样本,约15分钟内完成检测并报告定量检测结果,具有机体小巧、携带及安装简便,操作简单快捷的特点。 POC控制过程:将含有检测试剂及样本的专用检测管放入测试盒内后,通过触摸屏控制,读取RFID 卡上的参数,自动完成搅拌、孵育、检测;自动计算样品中被检物的浓度并报告检测结果。 产品主要性能参数如下: 重量:3.5kg 外形尺寸:260×145×140cm 检测波长:340nm 自动控温:37℃ 电源:由电源适配器将电网电源AC100-240V,50/60Hz转换为DC12V电流4.0 A。 额定功率:30VA 工作温度:15℃~30℃ 相对湿度:40%~85% 大气压力:86.0 kPa~106.0 kPa 储存:经包装后的POC应存储在0℃~40℃,相对湿度不超过85%,无腐蚀性气体和通风良好的环境内。 运输:运输过程中应防止受到剧烈冲击、雨淋和曝晒。 【适用范围】 本仪器仅与本公司生化检测试剂盒“同型半胱氨酸检测试剂盒”配套使用,用于定量检测临床血清或血浆样本中生化成分检测。 【禁忌症】 无。 【主要结构】 由主机和电源适配器组成,仪器外观见图1,接口见图2。

图2仪器背面接口 【注意事项、警示以及提示性内容】 1.严禁非授权维修人员自行拆开机体。 2.禁止使用非专用管,以免损坏仪器。 3.检测操作时,放入检测管以前,确认管盖盖严,拭净管体外残留液体。 4.当系统工作时,切勿接触系统上的运动部件。 5.不可手动开检测盖。 6.使用触摸屏,只能用手指接触,禁止使用笔或尖锐物体接触。 7.必须使用专用的试剂盒,使用前确认试剂盒的适用性。 8.必须使用专用的试剂盒专用RFID卡,否则无法检测。 9.必须在有效期内使用试剂盒和RFID卡。 10.使用试剂、样本应严格按照相关管理规范执行。 11.剩余试剂、样本及废弃物的处理严格执行国家有关医疗废弃物处理规范执行。 12使用过的仪器进行运输、维修或储存前,应用75%的酒精对检测盒及仪器表面仔细清洁消毒,以防止污染及可能的生物风险。 【图形、符号、缩写的解释】 图形、符号、缩写名称解释 警告指本部位存在一定的危险,操作时应小心。 参考说明书参考说明书 怕晒表明运输包装件不能直接照晒 怕雨表明包装件怕雨淋 禁止翻滚表明不能翻滚运输包装

点着温度测定仪 使用说明书

GLMDW-02型 点着温度测定仪 使用说明书 武汉格莱莫检测设备有限公司

DW-02型点着温度测定仪是根据国家标准 GB4610-84《塑料燃烧性能试验方法—点着温度的测定》而设计的测试设备。点着温度可以相对比较各种材料在特定条件下的燃烧特性,本方法测试简单方便,可以为设计应用选材提供参考数据。另有DW-04型武汉格莱莫检测设备公司。一、试验步骤 打开电源,把锭炉加热到预定温度,并使之恒温,允许误差±2℃。 将装有1克试样的容器放入铜锭炉的孔中,盖上盖子(盖子预先放在锭炉上加热),并打开秒表。 将点火火焰置于盖的喷嘴上方2毫米处晃动,火焰长度10~15毫米左右。如果在开始5分钟内,喷嘴上没有(或有)连续5秒钟的火焰,则每次将炉温升高(或降低)10℃,用新的试样重新试验,直到测得喷嘴上出现连续5秒钟以上火焰时的最低温度为止,并记录此温度。 在每个预定的温度做三个试样,若有两个没有5秒钟以上的火焰,则将炉温度升高10℃,再做三个试样,如有两个出现5秒钟以上火焰的最低温度,将其修约到十位数,即为材料的点着温度。

在预热性塑料的测定中有发泡溢出现象时,可以将试样减少到0.5克,如果仍有溢出,则不能用本方法试验。 二、报告 在试验报告中,注明试验方法和参考标准,材料的鉴别特征,试样的来源、粒度和试 样量,试验的结果,观察到的现象(烟气、火焰颜色等)应详细记录。 三、试样粒度:制备成0.5-1.0毫米。 四、试样量:1克。 五、炉温:150℃~450℃之间任意点着温度恒 定不大于±2℃,在测试过程中,如果设定 温度与水银温度计指示不统一时,以水银 温度计为准,因为铂电阻的设定误差是全 量程的±1℃%。 六、试验结束时,应拨动开关气手柄,置于开 气,进行降温,使炉温降到常温。 七、仪器装箱清单 1.加热炉1台 2.控制器1台 3.4孔盖板(硅钙板)1个

精密影像测量仪使用说明书

二次元精密影像量测仪使用说明书 1.1简介 本套系统综合了机器视觉、精密光机、精密测量算法等技术。广泛应用于电子、军工、模具、橡胶、塑胶制品、五金、刀具、弹簧、汽车零件 及教学、科研、产品研发等领域。 软件特点: ◆ 拥有超强及完善的2D几何测量功能和CNC自动控制功能。 ◆ 适合用于解决批量工件或复杂尺寸之检测效率;强调快速、精 准;品质保证。 ◆ 提供影像截取、对比、校正、补偿、自动寻边、自动取点、自动 对焦、智能学习等功能。 ◆ 针对重复工件的测量具有记忆、学习、自动编程的功能,快速准 确的现场实时测量。 1.2测控系统型号及含义 软件版本测量方式控制方式驱动器适用仪器 VMD 系列VMM 系列手动 自动 手动 手动 无 无 金相显微镜 手动型影像测量仪 VMA 系列自动自动伺服电机二次元影像测量仪 1.3测控系统的基本组成 精准测控系统并不是仅仅一个软件而已,它包括一定的硬件系统才能发挥软件的功能,一套完善的测控系统应包括如下组成部分:

1.4计算机的配置要求 1.4.1硬件要求 最低配置: CPU:Celeron 2.0GHz 内存:256 MB 独立显卡:显存64 MB 硬盘:800MB 空间 显示器:支持1024*768 平面直角显示 推荐配置: CPU:Pentium4 2.4GHz 内存:512 MB 独立显卡:显存128MB 硬盘:800MB 空间 显示器:支持1024*768 平面直角显示 1.4.2操作系统要求 精准测控系统支持Windows 2000/XP 简体中文操作系统,需要安装 DirectX8.0 及以上版本的软件;如果要在英文、繁体等其它语言版本的Windows操作系统中安装精准测控系统,则需要选择软件的专业版本。 软件测量结果可以支持以Microsoft Word 、Excel 输出并加注公司名 称及背景资料,且可与AutoCAD同步动态传输。基于以上的功能,所以要求 安装系统的计算机系统要预装Microsoft Office和AutoCAD 以支持软件的 功能。

全自动馏程测定仪操作使用说明书

S Y Q-6536D 石油产品蒸馏测 定仪 (低温单管式) 使用说明书 扬州市菲柯特电气有限公司 一、概述 自动馏程测定仪按照国标GB/T2282。GB/T6536设计生产。

型自动馏程测定仪采用了模块化设计,检测部分采用了先进的传感器和高精度AD转换电路,主控部分采用了多个工业应用、高可靠性16位RISC结构、超低功耗微处理器,良好可靠的通讯将各模块组成一个统一的、可靠的测控平台。 自动馏程测定仪所有运转活动部分全部采用步进电机带动,具有精度高、低噪音、运行可靠、维护量小、使用时间长的特点。 自动馏程测定仪的运行程序,采用高质量、最简捷的模块化程序设计,并与硬件有机的结合,使得馏程测定过程的升温和冷却、液位跟踪、记录温度、打印等全部工作自动完成,达到了一键出结果的操作方式。 自动馏程测定仪自动检测所在环境的大气压和测定仪内的工作温度,并对测定结果进行了在标准大气压下的自动补偿,使测试结果增加了可比性。 自动馏程测定仪预设了16组测定参数,供检测不同试样时选用,便于检测操作。同时预设参数具有可修改性,来满足测定特殊试样的要求。 二、特点 ·良好人机界面,方便操作 ·一键完成馏程蒸馏测定,简化操作 ·八组预置参数,供选用 ·可修改预置参数,适应特殊要求 ·红外液位检测不受室内光线、灯光干扰 ·液位跟踪,灵活自如 ·全部模块化设计稳定、可靠性高 ·自动储存100个检测结果,并随时查看或打印 ·检测过程遵守标准规定,数据可靠 ·检测方法可靠,重复性好 ·可长期连续工作,故障率极低 三、安全指导 为确保****型自动馏程测定仪安全运行,必须遵守以下指导: 1. 在安装使用前,请仔细阅读本使用说明书。 2. 请注意包装上的警告标志。

检测仪使用说明书

检测仪使用说明书 一.概述 核酸蛋白检测仪、紫外检测仪是液湘色谱仪中的一种紫外检测装置,核酸蛋白检测仪、紫外检测仪是根据生命科学的发展对于现代色谱仪器的要求而改进设计的一种新型紫外检测仪。该仪器在创新方面的主要特点为: 1.该仪器除配有输出10mV记录仪信号外,还配有输出适合计算机积分仪的输口,这 样很方便构成色谱工作站系统。(可同时进行计算机和记录仪信号输出,亦可省去记录仪) 2.该仪器的数字显示设计为固定光吸收,A显示计算机用和可变量程光吸收A显示记 录仪用两种可选模式,这样可方便于规范化读数(特别是可应用于药品生产的GMP 工艺规范化管理),同时亦可根据科研需要进行可变量程的高灵敏度读数,这样可方便于对低浓度样品检测。 3.该仪器采用新型进口IP28光电倍增管和改进型电路结构,使仪器工作更为稳定可 靠。 该仪器配有上层析柱、恒流泵、部分收集器等等,即组成一套完整的液色湘色谱分离分析系统。它可应用于现代生物学研究,药物测定、农业科研、化工、食品及医疗单位对具有紫外吸收的样品作定量分析。本仪器主要元器件均采用进口,仪器全部采用LED数字显示,使用方便。 二.主要技术性能 (1)核酸蛋白检测仪提供波长:254nm、280nm。 (2)紫外检测仪提供波长:220nm、254nm、280nm、340nm。 (3)量程范围:0~100%T、0~2A、0~1A、0~0.5A、0~0.2A、0~0.1A、0~0.05A。 (4)流式样品池:容积100微升、光程3毫米。 微量样品池:容积30微升、光程10毫米。 (5)记录仪输出:10mV (6)积分仪输出:0.1A/mV (7)数显模式:固定A量程读数(0~2.0A);可变A量程读数(0~2.0A、0~1.0A、0~0.5A、0~0.2A、0~0.1A、0~0.05A)。 (8)量程在0.05A档时:噪音≦0.002A。 (9)工作环境温度:0℃~35℃。 (10)仪器可连续工作。 (11)电源:220VAC±10%50HZ。 (12)单体外形尺寸:280×180×158(mm)。 (13)主机重量:5㎏。 三.工作原理 从光源发出的光经狭缝,滤色器聚焦到样品池上,此单色光通过样品池射到光电倍增管的光阴极面上,使光束由于样品浓度不同所引起透光强度的变化转换成光电流变化,此光电流经放大器放大,并输入到对数转换器、使透光率T转换成光吸收A输出即A=lgT/1=ε·CL式中ε为待测样品的摩尔消光系数,C为样品浓度,采用摩尔/升单位,L为光程,用厘米作单位。根据上式只要测出了A、L和ε就可求出样品浓度C。若从放大器直接输入到记录仪,则在记录仪上绘出的是样品透光率T变化的图谱,若从对数转换器输入到记录仪上,在记录仪上绘出的是样品光吸收变化的图谱。 四.仪器结构 核酸蛋白检测仪、紫外检测仪是单光路结构,由紫外检测器、和记录仪部分组成现将其构造分别说明如下: 1.紫外检测仪: 它由光源、干涉滤色片、样品池、光电倍增管、放大和对数板、低压板和高压板等组成。面板上有四氟塑料管的进样口和出样口,A调零以及调节“光量”大小旋(光

TR-10数字温度计使用说明书

TR-10便携式记录型测温仪使用说明 一,概述: TR-10是一款具备数据记录功能的温度测量仪表,仪表可记录100个温度点和时间,摄氏华氏转换,超温报警等功能。广泛应用于暖通制冷维修、食品、宠物等行业。 二,技术参数: 1、温度传感器:NTC K=103,B=3435 2、测温范围:-40℃~+110℃, 3、测温精度:±1℃(-20℃~+80℃),±2℃(-40℃~-20℃,+80℃~+110℃) 4、记录点数:100个, 5、采样周期:记录状态下为间隔时间,非记录状态下为10S 6、显示未定要求—电磁兼容测试:(1)EFT干扰测试>2级 (2)ESD测试>2级 7、时间:2009年1月1日—2099年12月31日 产品出厂参数值:日期为09 01 01,时间为12:00 00 间隔时间为001,(1分钟) 上限温度值都为:000.0度 下限温度值都为:000.0度 三,产品示意图: 正面图片:要求有液晶屏全部显示,以及能看清按键上的字。

背面图片:要求说明有背面各个部分的功能,及按键的图片,必要时增加局部放大的图片 液晶屏显示的说明: 说明液晶屏各部分显示代表的参数 四,按键操作说明: 按键使用模式说明:按一下按键立即抬起为“时间按”,按住按键查过五秒后抬起为“长时间按”前置按键的使用说明: Record: 功能一:开启和关闭记录功能 功能二:在记录过程中或记录完成后,按此键可以查看温度记录点的参数。 ▲▼: 功能一:增加和减小所要设置的数值;长按可以连续增加或减小参数值。 功能二:查看记录的温度点; Clear:清除所有已经记录的温度点值。 后置按键使用说明: Set: 功能一:长时间按此键五秒为进入或退出参数设置模式; 功能二:短时间按可退出温度查看状态。 Time: 功能一:短时间按为切换年月日和时分秒, 功能二:长时间按此键五秒进入或退出时间或日期设置状态。 Switch: 功能一:短时间按为摄氏华氏转换。 功能二:长按五秒为12/24小时转换功能。

源兴-影像测量仪.

YVM-T系列 影像测量仪使用说明书 东莞市源兴光学仪器有限公司

目录 一、仪器用途 (1) 二、仪器规格参数 (1) 三、仪器结构与工作原理 (1) 四、仪器开箱与安装 (3) 五、仪器测量方法 (4) 六、仪器维修与保养 (4) 七、仪器成套性 (4) 八、产品装箱清单 (5)

一、 仪器用途 非接触影像测量仪是一种由高解析度CCD 彩色摄像机、连续变倍物镜、PC 电脑显示器、转接盒、精密光学尺、2D 资料测量软体与高精度工作台等精密机械结构组成的高精度、高效率光电测量仪器,以二维测量为主,也能作三维测量。它被广泛应用在各种不同的精密产业中,如电子元件、精密模具、精密刀具、弹簧、螺丝加工、塑胶、橡胶、油封止阀、照相机零件、脚踏车零件、汽车零件、导电橡胶、PCB 加工等各种精密加工业,是机械、电子、仪表、钟表、轻工、塑胶等行业,院校、研究所和计量检定部门的计量室、试验室以及生产车间不可缺少的计量检测设备之一。 二、 仪器规格参数 -1- 型号 参数 YVM1575T YVM2010T YVM3020T YVM4030T YVM5040T 工 作台 X,Y 轴行程 150X75mm 200X100mm 300X200mm 400X300mm 500X400mm Z 轴行程(mm) 220调焦及辅助测量 工作 台尺寸 305X205mm 356X256mm 456X306mm 600X400mm 706X606mm 传动方式 X 、Y 轴光杆传动,Z 轴齿轮传动 数位测量系统 光学尺解析度:X 、Y 、Z 轴0.001mm 测量软件一套 脚踏开关采样接口,RS232打印机接口 影像描准系统 高解析度CCD 彩色摄像机,PC 电脑一套,影像卡 0.7-4.5X 变焦镜头,总视频放大倍率为34-220X 配2X 物镜时,总视频放大倍率可到440X 数据转接卡 照明系统 可调试LED 环形表面光及轮廓光 电气参数 输入电压AC220V/AC110V 、50-60HZ 外形尺寸(mm) 500X550X680 500X550X680 600X900X680 960X820X940 1100X750X940 仪器重量 127Kg 140Kg 190Kg 350Kg 430Kg

直径测量仪DMS-600说明书

烟台双合汇升测控设备有限公司

目录 1.概述----------------------------------------------------------------3 2.仪器的结构-------------------------------------------------------- 3 3.主要性能参数与及使用条件--------------------------------- 4 4.操作指南--------------------------------------------------------- 4 5.仪器常见故障及注意事项------------------------------------ 10

一、概述 DMS-600智能测微仪是利用电感测量原理:电感传感器的把微小位移变化转换为电压信号并放大,再把放大后的信号量进行数字化并采集,送入嵌入式微机进行处理。可直接将采集来的数据经过运算显示出被测工件直径尺寸的绝对值。根据预先设定的工件尺寸公差界限,用模拟光柱及颜色指示出被测工件合格与否,操作简便直观。 ●本产品采用工业级彩色TFT触摸屏,亮度高、寿命长,特别适合各种工业生产环 境的应用,制作了中文菜单操作界面,大大方便了操作和人机交互。 ●本产品可以进行多达30种规格被测参数进行设置,也具有分组功能,还具有测 量数据的存储和查询功能等。 ●数据通讯功能可使现场多台量仪组成网络或者直接将数据存入USB移动盘里, 通过专用软件形成SPC统计图表以实现现场在线质量管理。也可以根据用户的特殊要求,修改和补充产品的功能,以满足不同应用场和的不同要求。 ●本产品为用户提供极大方便,半自动校对零位,数据存储,统计分析。 二、仪器结构 1.触摸屏; 2. 电源开关;

红外测温仪使用说明书

红外测温仪及二次表现场使用 说明书

双波长红外测温仪 为了解决温度的测量问题,温度的自由选择问题,以及长期稳定的校准需要等,威廉姆森设计了双波长高温计,这使得威廉姆森温度的测量上远远超过了业界的其它测温产品,显示出威廉姆森显著的优势 传感器概述: 相对与单波长温度传感器,双波长红外测温仪的主要优点在于: ●对于难测量的物体(如灰色金属表面),红外测温仪采用自动 补偿的方法从而增加准确度。 ●目标大小小于传感器目标直径,如电线,或移动的目标等,它 也可以准确无误的测量。 ●目标在部分受到阻挡镜头模糊时,或干预媒体,如烟雾,灰尘, 和/或水喷雾,双波长红外测温仪仍然可以准确和可靠的测量

williamson 有两种类型的高温计的设计。双波长及双色彩设计。这两种温度测量技术是基于相同的物理原理主要涉及测量红外能量 在两个相邻的波长之间计算的比例通过这两项测量,确定温度。两者的设计不同点在于:双色彩设计采用了两个层次的红外探测器被称为“夹心探测器” ,而双波长技术采用“单一探测器”的设计(见图) 。 基于其独特的技术测量红外能量,双波长红外测温仪设计提供了一些优势。 一, 在恶劣的环境下更高的稀释信号因子。提高了传感器的控制能力,使它可以穿过脏的窗口或水喷淋,喷雾油,烟,和尘埃等。从而也提高了测量精度这使得它对被测物体表面的氧化物,熔融金属,有光泽的金属(低辐射)等都不会受到影响 ,包括应用目标大小小于传感器目标直径,如电线,或移动的目标等,它也可以准确无误的测量。 双波长 双色彩

二、可根据需要定制温度范围,测量目标的温度可以低至300 C 以 下 三、长期稳定的校准过程监测与控制等方面的应用,使得测量结果准 确无误。 红外测温仪现场连接方式按现场接线图连接 工作正常时LCD上应显示LO TEMP 红外测温仪工作基本原理

二次元影像测量仪(MVP系列)使用说明书

MVP系列 影像座标测量仪用户手册

目录 前言 (2) 1.仪器规格及技术参数 (3) 1.1影像测量仪具体规格及参数 (3) 1.2仪器所需电脑推荐配置 (3) 2.仪器工作原理及结构 (4) 2.1工作原理 (4) 2.2仪器总体结构 (4) 3.仪器安装 (7) 3.1仪器使用环境 (7) 3.2仪器的安装方法 (7) 4.仪器的使用方法 (8) 5.仪器的维护和保养 (8) 6.仪器成套性 (9) 7.售后服务 (9) 8.常见问题 (10)

前言 MVP系列影像座标测量仪是集光学、精密机械、电子、计算机于一体的精密高效测量仪器。它是一种由高分辨率CCD彩色摄像机、连续变倍物镜、电脑、精密光学尺、高精度工作台与测量软件等组成的高精度、高效率的视频测绘系统。以二维测量为主,也可作为三维视频测量系统,可轻易实现测量、检验、校准、逆向工程等目的,被广泛应用于各种行业。如:电子元件、精密模具、刀具、弹簧、导电橡胶、油封止阀、照相机零件、脚踏车零件、汽车零件、PCB加工等,也可用于教学、科研、产品研发等领域。

1.仪器的规格及技术参数 1.1 MVP系列影像坐标测量仪技术参数 附件:随机配送机台工作桌一张、防尘罩一个、自动机标配工业电脑(手动机电脑选配)备注:M-手动 A-自动

2. 仪器工作原理及结构 2.1 工作原理: 影像座标测量仪是通过连续变倍物镜、彩色CCD,通过轮廓透射光或表面光照明将被测工件放大后成像在显示器上的影像放大测量系统。利用专业测量软件对精密光学尺传输的数据和实时影像画面进行图像数据处理,由操作者使用鼠标或键盘在电脑上进行快速描边、标注测量。 2.2MVP系列仪器总体结构(如图一所示) 图1 MVP系列影像式测量仪 2.2.1影像式测量仪主要有支撑部分、视频部分、工作台部分及照明部分。 2.2.2 支撑部分包括大理石底座(3)、大理石立柱(13),机台机架(2); 2.2.3 视频部分包括Z轴升降组(9),镜头(6),CCD(7),上灯(5)通过软件灯光控制区,手动机台旋转Z轴手轮组(12)电动机台点击软件运动区,可实现对不同高度工件的测量; 2.2.4工作台部分包括大理石上层工作台(1),工作台玻璃(2),底灯(3),工作台中层(6),V型导轨(5),X轴传动组(8),Y轴传动组(9),,电动机台通过软件控制运动区或手动机台通过旋转X/Y轴开合手柄(4)可以快速的移动定位工作台。(如图三)

手持测距仪使用说明书

1.先要给激光测距仪装上电池,对于那些可以直接充电的激光测距仪,我们在使用前要先把电充满。 2.每一个激光测距仪上都会有一个开关电源,有的是通过轻按“发射键”,测距仪内部电源就可以打开,通过目镜可看见测距仪处于待机状态。 3.打开电源后,在测量前,我们还要选择好单位,操作方法是长按“模式键”,就可以直接选择你要选择的单位了。 4.一切准备工作都做好之后,我们可以通过测距仪目镜中的“内部液晶显示屏”瞄准被测物体,注意手不要抖动,这样可以减小误差,测量结果会更准确。 5.确定描准之后,轻按“发射键”,这时测量的距离就会显示在“内部液晶显示屏”上,我们可以记下这个数值,如果担心测量不准确,可以多测几次。 6.在瞄准被测物体时,如果感觉被测物体不是很清晰,我们可以通过“+/-2屈光度调节器”来调节被测物体远近的清晰度,可以通过顺转或逆转来调节远近,以达到最理想的清晰度。 注:各种品牌各种型号可能会有所差异,但基本使用方法都是大同小异,看看说明书应该操作都不会有问题。 扩展资料: 手持式测距仪,是根据利用电磁波学、光学、声学等原理且具有

小巧机身,用于距离测量的仪器。 原理:手持式测距仪在工作时向目标射出一束很细的激光,由光电元件接收目标反射的激光束,计时器测定激光束从发射到接收的时间,计算出从观测者到目标的距离。 一般采用两种方式来测量距离:脉冲法和相位法。脉冲法测距的过程是这样的:测距仪发射出的激光经被测量物体的反射后又被测距仪接收,测距仪同时记录激光往返的时间。光速和往返时间的乘积的一半,就是测距仪和被测量物体之间的距离。脉冲法测量距离的精度是一般是在+/- 1米左右。另外,此类测距仪的测量盲区一般是15米左右。

WTZK-02温度指示控制器说明书

WTZK-02(BWY-02)型 温度控制器 (电力变压器温度测量与控制的专用仪表) 使用说明书 (我厂选用的是杭州富阳仪表有限公司) 杭州华立仪表有限公司

最大插入深度 260

一、用途及原理 WTZK-02型温度控制器(BWY-02型温度控制器),适用于测量变压器油温或其他的液体、气体和蒸气的温度,并能在被测温度达到和超过设定值时发出接点信号。本仪表主要用于500KV A以上的变压器油温的测量和控制。控制器的主要结构是由温包(感温探头)毛细管和弹性元件组成。这三个部件构成的密封系统内充灌了感温介质、当被测温度发生变化时,温包内的液体压力随着发生变化,通过毛细管的传递使表头的弹性元件产生一个相对应的位移量,这个位移经机械放大后便可指示出被测温度,并带动微动开关动作输出信号。 二、主要技术参数 1.测量范围:0~100℃ 2.指示精确度:1.5级 3.开关特性 (1)开关设定精度:±3℃ (2)开关差:6±2℃ (3)开关额定负荷:AC220V、1.5A 4.温包尺寸:ф14×150 5.温包安装尺寸:外螺纹M27×2 6.表头安装尺寸:3孔ф6 144×210 7.开关在仪表出厂时标准设定值:第一上限55℃ 第二上限80℃ 如果用户需调整开关设定值,可打开表盖,转动表盘上旋钮设定针即可,然后按原样合上表盖。 三、安装和使用 1.温度设定:打开表盖,转动旋钮设定针即可(一般不用再设定)。 2.接线:打开表盖,将电缆线从引线接点穿入,按图2要求接线。电缆线推荐使用KW4×1.5控制电缆。 接线后拧紧引线接头处螺母,不得松动。表盖按原样合好,均匀旋紧四只螺钉。 温控器表头安装前应先完成安装接线和开关设定两项工作。

测距仪使用说明

Forestry pro measurement procedure guide To start :开始 1,press POWER button to turn on按电源键POWER打开 2,confirm your mode with the internal display (default is “last used” setting)确认你的模式与内部显示(默认是“最后的使用“设置) 3,Set your desired mode and start measuring设置你想要的模式,开始测量 Mode setting:模式设置 Target priority modes目标优先模式 1、press and hold MODE button,then press and hold POWER button within 0. 5 second.按下并保持住模式按钮,然后同时按住电源按钮在0.5秒之内 2、Continue to press and hold both buttons(more than 2 seconds)until first target priority mode and distant target priority mode are switched.继续同时按住两个按钮(超过2秒),直到第一个目标优先模式和遥远的目标优先模式切换。 3、if the button is not pressed within 0.5 second ,the display unitwill be switched.如果按钮不是压在0.5秒之内,显示单元<米/码/英尺。>将切换。Dst:Distant priority mode (factory setting)遥远的优先模式(工厂设置) 1st:first priority mode第一优先模式 Measurement modes测量模式 Press and release MODE button.按下和释放模式按钮 Measurement mode changes in the order of 1 (Act)to 6 (Hor)as shown below.测量模式变化的顺序1(Act)到6(Hor)如下所示。 Holding down MODE button longer switches the display unit.按住模式MODE按钮开关的显示装置 1.Act:linear distance直线距离 2.Hor(solid):horizontal distance水平距离 3.Hgt(solid):height from horizontal高度 4.Ang: angle (degrees)from horizontal角度 5.Hgt(blinking):vertical separation(height between two points) Hgt闪动为垂直分离(两点之间高度)测试模式 6.Hor(blinking):three-point measurement (height between two points)三分测量One-point(Act/Hor/Hgt/Ang)measurement单点测量 (measurement order:1-2-3-4-2,then repeat)测量顺序从1-2-3-4-2,然后重复Target point目标点 Internal display(“linear distance mode”)内部展示,直线距离模式 (display will be different depending on the setting mode.)显示的不同取决模式 1.set your desired mode.设置你想要的模式 2.Aim at the target.瞄准 3.Press POWER button to measure,keeping the button pressed allows continuous measurement of up to 20 seconds. 按电源按钮保持20秒来衡量 4.Results display example of each mode结果显示每个模式的例子 (1)Act :Linear distance mode直线距离模式 (2)Hor: horizontal distance mode水平距离模式

相关文档
相关文档 最新文档