Thermopile Sensor
OTP-538U
Revision Date: 2009/04/16
Introduction
The OTP-538U is a thermopile sensor in classic TO-46 housing. The sensor is composed of 116 elements of thermocouple in series on a floating micro-membrane having an active diameter of 545 μm and with blacken surface to absorb the incident thermal infrared radiation, which induces a voltage response at output terminals. The sensor chip is fabricated by a unique front-surface bulk micromachining technology, which result in smaller size and faster to response ambient temperature change.
The IR window is a bandpass filter having its 50%cut-on wavelength at 5 μm, and cut-down at 14μm. The sensor responses proportionally to the incident IR radiation and has a constant signal response up to its cut-off frequency, which is limited by the sensor thermal time constant of tens millisecond range.
The OTP-538U thermopile sensor provides nearly Johnson-noise-limited performance, which can be calculated by its ohmic series resistance. A thermistor element, with a lead connected to ground, is also provided inside the TO package for ambient temperature reference.
Features
Non-contact temperature detection
Voltage output, easy to take signal
Zero power consumption
Wide detection temperature range
Applications
Medical Application:Ear thermometers
Home Facility:Microwave oven, Hair dryer, Safety system, Home security & Air conditioner
Industry Application:Process monitor and controller, Infrared non-contact thermometers
Automobile Application:Thermal sensing system
Table of Contents
1 General Characteristics
1.1 Absolute Maximum Ratings
1.2 Handling Guidelines
2 Device Characteristics
2.1 Device Descriptions
2.2 Sensor Characteristics
2.3 Signal Output Characteristics
2.4 Frequency Response
2.5 Thermistor Characteristics
2.6 Optical Characteristics
2.7 Filter Characteristics
2.8 Mechanic Drawing and Pin Assignment
3 Liability Statement
List of Figures
Figure 1: Typical output voltage versus object temperature with sensor at 25° C Figure 2: Frequency response
Figure 3: Field of view curve
Figure 4: Typical spectral transmittance curve
Figure 5: Mechanical drawing and pin assignment
List of Tables
Table 1: Absolute maximum ratings
Table 2: Device characteristics
Table 3: Thermopile sensor characteristics
Table 4: Typical numerical data of Thermopile’s output voltage (sensor at 25° C) Table 5: Thermistor information
Table 6: Tabulated thermistor data
Table 7: Optical characteristics
Table 8: Filter parameters
1. General Characteristics
1.1 Absolute Maximum Rating
Table 1: Absolute maximum rating
Parameter Min. Typ. Max. Unit
Ambient Operating Temperature -20 100 ℃
Storage Temperature -40 100 ℃
1.2 Handling Guidelines
Stresses above the absolute maximum ratings may cause damages to the device.
Do not expose the sensor to aggressive detergents such as Freon, Trichloroethylene, etc.
Avoid touching or cleaning of the window. If necessary, do cleaning only very carefully with alcohol and cotton swab.
Hand soldering and wave soldering may be applied by a maximum temperature of 260° C for a dwell time less than 10 s. Reflow soldering is not recommended.
2. Device Characteristics
2.1 Device Descriptions
The materials of OTP-538U are lead-free and fully complied with the RoHS regulations.
Table 2: Device characteristics
Parameter Description
Cap Metal cap with integrated IR window
Header TO 46
Leads(3 isolated + 1 ground) pins with solderable gold coating
Filter Si-based IR filter
Insulation gas sealing The sensor is sealed hermetically in a nitrogen environment.
2.2 Sensor Characteristics
Table 3: Thermopile sensor characteristics
Parameter Min Typ Max Unit Conditions
Output Voltage 0.77 1.44 mV T amb.=25℃T obj. =50℃
Sensitivity 70 85 100 V/W * TC of sensitivity 0.10 0.11 0.12 %/K Typical Sensitivity area in
diameter
--- 545 --- μm
Resistance of thermopile 50 65 80 KΩ 25℃TC of resistance --- 0.09 --- %/K Typical
Time constant --- 16 --- ms *
Noise voltage 28 32 36 nV/Hz1/2
NEP 0.28 0.36 0.48 nW/Hz1/2
Normalized detectivity (D*) 1.0*108 1.3*108 1.7*108cm*Hz1/2/W* Notes: * All optical performances are measured at 1 Hz chopper frequency using a blackbody radiator of 500K temperature with filtered spectrum ranging from 5 to 14μm.
Distance between packaged thermopile and blackbody is 11cm.
2.3 Signal Output Characteristics
Table 4: Typical numerical data of Thermopile’s output voltage (sensor at 25° C) Temp. (℃) V_out (mV) Temp. (℃) V_out (mV) -20 -1.29 50 1.02
-10 -1.06 60 1.49
0 -0.80 70 1.99
10 -0.51 80 2.52
20 -0.18 90 3.09
25 0.00 100 3.69
30 0.19 110 4.33
40 0.59 120 5.00
Note: Blackbody source size: 250 x 250mm
Distance from sensor to source: 110mm
2.4 Frequency Response
Figure 2: Frequency response
2.5 Thermistor Characteristics
Table 5: Thermistor information
Parameter Min Typ Max Unit Conditions
Resistance 97 100 103 KΩ 25℃
β value 3944 3964 3984 K 0℃/50℃
Table 6: Tabulated thermistor data
(K?)
(K?) Resistance Resistance
Temp.(o C) Min Center Max.Temp.(o C)Min Center Max.
-20 882.3 919.7 957.94542.6844.18 45.69
-15 668.9 696.3 724.25035.2336.50 37.78
-10 511.7 532.0 552.55529.2330.31 31.41
-5 394.8 409.8 425.16024.3625.29 26.23
0 307.0 318.3 329.86520.4121.20 22.01
5 240.5 249.1 257.77017.1717.85 18.55
10 189.8 196.3 202.97514.5115.10 15.70
15 150.7 155.7 160.78012.3112.82 13.34
20 120.5 124.4 128.38510.4810.93 11.38
25 97.0 100.0 103.0908.9609.347 9.742
30 78.32 80.83 83.34957.6858.023 8.369
35 63.61 65.72 67.83100 6.615 6.911 7.214
40 51.96 53.73 55.51105 5.712 5.972 6.238
2.6 Optical Characteristics
Table 7: Optical characteristics
Parameter
Min
Typ Max Unit Conditions Field of view
66
degree
At 50% target signal
Figure 3: Field of view curve
2.7 Filter Characteristics
Table 8: Filter parameters
Parameter
Min Typ Max Unit Conditions Cut on wavelength
4.7
5.0
5.3
μm
At 25℃,
50% transmittance
Figure 4: Typical spectral transmittance curve
2.8 Mechanic Drawing and Pin Assignment
3. Liability Statement
The contents of this document are subject to change without notice and customers should consult with Oriental System Technology (OST) sales representatives before ordering. Customers considering the use of OST thermopile devices in applications where failure may cause personal injury or property damage, or where extremely high levels of reliability are demanded, are requested to discuss their concerns with OST sales representatives before such use. The Company’s responsibility for damages will be limited to the repair or replacement of defective product. As with any semiconductor device, thermopile sensors or modules have a certain inherent rate of failure. To protect against injury, damage or loss from such failures, customers are advised to incorporate appropriate safety design measures into their product.