Automotive Emission Analyzer MEXA-584L

Communication Software Instruction Manual

CODE:GZ0000106461

Preface

This manual describes the operation of the Automotive Emission Analyzer, MEXA-584L, Communication Software.

Be sure to read this manual before using the product to ensure proper and safe operation of the instrument. Also safely store the manual so it is readily available whenever necessary.

Product specifications and appearance, as well as the contents of this manual are subject to change without notice.

Warranty and Responsibility

HORIBA warrants that the Product shall be free from defects in material and workmanship and agrees to repair or replace free of charge, at HORIBA’s option, any malfunctioned or damaged Product attributable to HORIBA’s responsibility for a period of one (1) year from the delivery unless otherwise agreed with a written agreement. In any one of the following cases, none of the warranties set forth herein shall be extended;

z Any malfunction or damage attributable to improper operation

z Any malfunction attributable to repair or modification by any person not authorized by HORIBA

z Any malfunction or damage attributable to the use in an environment not specified in this manual

z Any malfunction or damage attributable to violation of the instructions in this manual or operations in the manner not specified in this manual

z Any malfunction or damage attributable to any cause or causes beyond the reasonable control of HORIBA such as natural disasters

z Any deterioration in appearance attributable to corrosion, rust, and so on

z Replacement of consumables

HORIBA SHALL NOT BE LIABLE FOR ANY DAMAGES RESULTING FROM ANY MALFUNCTIONS OF THE PRODUCT, ANY ERASURE OF DATA, OR ANY OTHER USES OF THE PRODUCT.

Trademarks

Generally, company names and brand names are either registered trademarks or trademarks of the respective companies.

October, 2007 ? 2007 HORIBA, Ltd.

Notes

MEXA-584L COMMUNICATION SOFTWARE is the optional program for communication with MEXA-584L, the Automotive Emission Analyzer.

HORIBA,Ltd. grants you a license to use the program and related materials (“Software”), provided that you agree the followings:

Q Copyright

HORIBA, Ltd. owns the copyright and any other rights on the Software.

Q Use of the Software

You may install and use the Software on a personal computer that you own.

The license is nonexclusive and nontransferable.

You may make a backup copy of the Software.

Q Restrictions

Do not sublicense any third party to use the Software and its reproduction.

Do not make reverse engineering of the Software.

Q No warranty

HORIBA, Ltd. does not and cannot warrant the performance or results you may obtain by using the Software.

Q Licensing period

This license is established when you receive the Software.

HORIBA, Ltd. may cancel the license with prior notification or when you act against items mentioned above.

Q Modification

The specifications of this product and the contents of this manual may be changed without

notice.

Description in This Manual

Note

This interprets the necessary points for correct operation and notifies the important points for handling the unit.

Reference

This indicates the part of where to refer the information.

Tip

This indicates reference information.

Contents

Outline . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .1 Conditions. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .1 Contents. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .1

Installation. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .2

Basic screen . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .3 Screen items. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .4 Operation buttons. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .5 Menu operation. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .7

HOST communication settings . . . . . . . . . . . . . . . . . . . . . . . . . . .9

Outline

1

Outline

The communication software for MEXA-584L “MEXA-584L COMMUNICATION SOFTWARE”allows you easy operation to import measurement data obtained with MEXA-584L into a computer as a CSV format file (logging).

Conditions

This software can run on Windows XP and/or Windows VISTA.

The recommended conditions of the personal computer that this software will be installed into are as follows; Note

Make sure to use a dedicated RS-232C cable (optional) for communication between the personal computer and MEXA-584L. A commercially manufactured cable may be useless for the communication.

Contents

CPU:

Pentium 1 GHz or faster (minimum: Pentium 600 MHz)

Memory: 512 MB or more Free area in hard disk: 110 MB or more

z Installation CD:

z

Instruction manual (this book): 1 in total 1

Installation

Installation

1.Exit all applications currently running on the personal computer.

2.Expand the installation CD, and double-click setup.exe to start installation.

3.According to the displayed messages, input the installation settings.

4.If https://www.wendangku.net/doc/fb4708513.html, Framework 2.0 has not been installed on your computer, an

installer program will start up. Follow the wizard to install https://www.wendangku.net/doc/fb4708513.html, Framework

2.0.

2

Basic screen

3

Basic screen

Click the [Start] button on the Windows task bar and select [Program] → [MEXA-584L].

MEXA-584L COMMUNICATION SOFTWARE will start up and the basic screen will be displayed.

Fig. 1 Basic screen

Basic screen

4

Screen items

①Menu bar

The following operations are available with using a mouse or short-cut keys (see “ Menu operation ” (page 7)).

[Setting (S)]: Check/change of communication port settings Check/change of HC full-scale [Help (H)]: Check of version information

②Measurement data

Measurement data by component (CO/HC/CO2/LAMBDA/AFR/O2/NO/RPM/TEMP) are displayed. (O2/NO/RPM/TEMP are optional.)The update interval is 500 ms.

When a measurement value is beyond the measurement range, the value is displayed in red.

("---" is displayed for items that contain no data.)

③P .E.F value

The converted value from propane to normal hexane, which is the unique conversion factor for the measuring instruments, is displayed.④Alarm

These indicators are shown in red when an alarm occurs. They are blank under normal conditions.

z Oil temperature error (TEMP): The value input from the oil temperature sensor is not correct.

z Leak check error (LEAK): Leak has occurred.

z Hang-up error (HANG): An HC hang-up has occurred.

z Engine speed error (RPM): The value input from the engine speed sensor is not correct.

z Warm-up error (WARMING UP): The unit is warming up.

z Probe alarm (PROBE): The probe has become misaligned/removed during measurement. (The CO 2 measurement value is 5 %vol or less.)

z FILTER alarm (FILTER): Filter clogging or measurement gas flow blocking.

z CAL ERROR (CAL ERROR): Incorrect calibration or bad sensitivity.Refer to MEXA-584L instruction manual for countermeasures for these conditions.

⑤STATUS display

The current status of the MEXA-584L instrument is shown.z STANDBY: Standby mode z MEAS: Measurement mode z WARMUP: Warm-up mode

⑥ LOGGING display The current logging status is shown (ON/OFF).⑦LOGGING SAMPLING RATE value

The sampling interval (s) during logging is displayed.

⑧SAMPLING TIME value The sampling set time (min) during logging is displayed. Blank is shown for no settings.

⑨PASS TIME value

The lapsed time (min) from logging start is displayed.

⑩Operation buttons

The following operations are available with using a mouse or

corresponding function keys shown on buttons (see “ Operation buttons ” (page 5)).

z [MEAS (F1)]: Measurement z [STANDBY (F2)]: Standby

z [LOGGING (F3)]: Logging (Data saving)

z [SET LOGGING PARAMETER (F4)]: Logging condition setting z [EXIT (F5)]: Exit from the software

Basic screen

5

Operation buttons

[MEAS (F1)]

This button makes the MEXA-584L instrument enter measurement mode.

This button is available only when the MEXA-584L instrument is in standby mode.

[STANDBY (F2)]

This button makes the MEXA-584L instrument enter standby mode.

This button is available only when the MEXA-584L instrument is in measurement mode.

[LOGGING (F3)]

This button starts logging, or saving sequential measurement data as a CSV file.

1.Click the [LOGGING (F3)] button, and the LOGGING dialog box is displayed.

Fig. 2 LOGGING

2.Specify “Save in” and “file name” and click the [Open (O)] button. Logging starts, and the basic screen returns to LOGGING ON display.

Sequential measurement data (lapsed time, value for measured component) are saved.

To cancel logging start and return to the basic screen, click the [Cancel] button on the LOGGING dialog box.

3.When the LOGGING display is ON, Click the [LOGGING (F3)] button again, and logging terminates.

The following logging information is added to the header of the saved file.MEXA-584L Sampling Data, DATE, 2007/9/18, 12:34:56RATE:, 1, sec TIME:,1, min

Basic screen

6

[SET LOGGING PARAMETER (F4)]

This button allows to set logging conditions (sampling interval and sampling time).

The logging conditions can be set only when LOGGING is OFF.

1.Click the [SET LOGGING PARAMETER (F4)] button,

and the LOGGING PARAMETER SETTING dialog box is displayed.

Fig. 3 LOGGING PARAMETER SETTING

Note

Set a larger value for Sampling time than the value of Sampling rate.

Example:

The available Sampling time range for 61 s Sampling rate: 2 min to 10000 min

The available Sampling time range for 1800 s Sampling rate: 31 min to 10000 min

2.Set logging conditions and click the [OK (F4)] button. The logging conditions are

changed and the basic screen returns.

To cancel the logging condition change and return to the basic screen, click the

[CANCEL (F5)] button.

①Sampling rate

Set the data sampling frequency by the second (1 s to 1800 s).

One sampling will be executed every set time.

②Sampling time check box

Checking the box enables the upper limit for the sampling time that has

been set.

In this case, logging automatically ends at the set time.

Unchecking the box makes the set sampling time invalid.

If there is insufficient space available at the save destination, the data

will not be saved correctly, which will prevent the data from being

loaded. Make sure there is sufficient space available at the save

destination.

③Sampling time Set the upper limit of sampling time by the minute (1 min to 10000 min).

Basic screen

7

[EXIT (F5)]

Clicking this button during LOGGING OFF makes the software just terminate.

Clicking this button during LOGGING ON displays a confirmation message of saving the current logging file and exiting the software.

Fig. 4 EXIT

To save the current logging file and exit the software, click the [OK (F4)] button.

To continue logging, click the [CANCEL (F5)] button.

Menu operation

Q

Setting (S) Port (P)

This is used to specify the communication port to be connected with MEXA-584L on the computer that has two or more RS-232C communication ports.

1.Click the [Setting (S)] from the menu bar and Click the [Port (P)] command, the PORT SETTING dialog box is displayed.

Fig. 5 PORT SETTING

2.Click to select the option button of the port number to be used for communication with MEXA-584L.

3.Click the [OK (F4)] button, the port settings are changed and the basic screen returns.

To cancel the port settings change and return to the basic screen, click the [CANCEL (F5)] button.

Basic screen

8

Full scale (F)

You can check and change HC full scale.

1.From the menu bar, click [Setting (S)] and click the [Full Scale (F)] command. The

FULL SCALE SETTING dialog box appears.

Fig. 6 FULL SCALE SETTING

2.Select the HC full scale range for the MEXA-584L you are using.

3.Click the [OK (F4)] button to update the range changes and return to the basic

screen.

To cancel the port settings change and return to the basic screen, click the [CANCEL (F5)] button.

Note

Once a full scale range has been set, it is not necessary for it to be reset when the same full scale range unit is used.

Q Help (H)

Version (V)

This is used to display the version information.

1.Click [Help (H)] from the menu bar and Click the [Version (V)] command, the

VERSION dialog box is displayed.

Fig. 7 VERSION

2.Click the [OK (F5)] button to return to the basic screen.

HOST communication settings

9

HOST communication settings

The communication states for this software and the MEXA-584L unit are shown in the table below.

Communication method half-duplex operation

Synchronization method asynchronous communication method Baud rate 9600 bps Data length 8 bit Stop bit 1 bit Error control Check sum Data format Binary Flow control None

Plug type

D-Sub 9 pin socket Pin assignment

Pin 2 TXD Pin 3 RXD Pin 5 GND Pin 7 CTS Pin 8 RTS

2 Miyanohigashi, Kisshoin Minami-ku, Kyoto 601-8510 Japan